Pretty much what is in place here for existing/established instruments. 



Joelle Weaver MAOM, HTL (ASCP) QIHC
 
> From: mill...@emhs.org
> To: joellewea...@hotmail.com; l...@premierlab.com; mw...@wakehealth.edu; 
> robin...@mercyhealth.com; tbr...@holyredeemer.com; 
> histonet@lists.utsouthwestern.edu
> Subject: RE: [Histonet] RE: Instrument Verification
> Date: Fri, 4 Apr 2014 10:08:54 +0000
> 
> I must say, the first time I read ANP.23045 I had the same reaction you all 
> did. After reading it over a couple of times, we started looking at it 
> differently. The question is not about validation--- it is about verifying 
> that the equipment will function and perform as intended. It doesn't have 
> anything to do with protocols and procedures for staining, cutting, 
> processing, etc. If you look at any service report for service performed on 
> your instrument, it is likely that the service engineer has some sort of 
> notation on the report that states the unit was run and monitored for 
> performance and meets all specs.
> To comply with the CAP question we merely wrote a procedure for 
> Instrument-Equipment Performance Verification stating that 1)The verification 
> of instrument/equipment function will be performed by a qualified service 
> engineer upon installation, after scheduled preventative maintenance, major 
> instrument repairs, or relocation. 2) Use of the instrument/equipment will 
> resume after the verification of its performance has been successfully 
> completed.3) The service engineer will provide the Histology Laboratory with 
> a report indicating that performance function was tested and satisfactory.
> Of course, we also have all service reports (which include the notation that 
> the function checks were performed and acceptable) available for each 
> instrument.
> 
> Suzie Miller, MLT ASCP
> Senior Histotech
> 
> Mercy Health System of Maine Laboratory
> 
> -----Original Message-----
> From: histonet-boun...@lists.utsouthwestern.edu 
> [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of joelle weaver
> Sent: Thursday, April 03, 2014 7:52 PM
> To: Elizabeth Chlipala; Martha Ward-Pathology; Cynthia Robinson; Terri Braud; 
> histonet@lists.utsouthwestern.edu
> Subject: RE: [Histonet] RE: Instrument Verification
> 
> I do understand and sympathize with the situation in many clinical labs with 
> staff , sometimes barely enough to do the work and it is challenging to keep 
> up with expanding documentation also.  I would like to meet the GLP, but do 
> struggle to be as extensive in my documentation. I do try to get as close to 
> the ISO standards as possible, just to cover myself. I agree with Elizabeth's 
> post that this seems to be the direction CAP has been heading over the years.
> I think that if you get new instruments, methodology, technology they will 
> certainly want to see the more robust documentation. For example ( see the 
> current CAP today on IHC validation), this will surely be the "guideline" of 
> tomorrow....
> But for those older, in long use instruments and technology,  my opinion is 
> that if you have documentation in line with what the checklist stipulated 
> when it went into use, and also all PM, maintenance, and QC- and have 
> documented any corrective actions, this will probably "fly" for now? What 
> does everyone else think?
> 
> 
> 
> 
> Joelle Weaver MAOM, HTL (ASCP) QIHC
> 
> > From: l...@premierlab.com
> > To: mw...@wakehealth.edu; robin...@mercyhealth.com;
> > tbr...@holyredeemer.com; histonet@lists.utsouthwestern.edu
> > Date: Thu, 3 Apr 2014 14:16:58 -0600
> > CC:
> > Subject: [Histonet] RE: Instrument Verification
> >
> > Hello All
> >
> > Coming from a GLP environment this type of equipment validation is standard 
> > in our setting.  This is just my opinion but I think the CAP checklist is 
> > moving towards the type of equipment documentation that is already required 
> > in a GxP or ISO environment.  I always thought that instrument 
> > qualification (IQ) - operational qualification (OQ) and process 
> > qualification (PQ) or simply stated IQ/OQ/PQ were used only in GxP settings 
> > but you now see some of the larger clinical labs running these types of 
> > validations on their equipment and processes.  To me it does make sense 
> > that some type of equipment validation should be required whether it  is a 
> > two page document on the microtomes, waterbaths, etc. or complete 
> > IQ/OQ/PQ's  on major pieces of equipment such as tissue processors, 
> > immunostainers and IHC retrieval units.  I believe that all of these are 
> > important processes that should be completed in histology laboratories 
> > today.    We are a GLP compliant lab and every single piece of equipment is 
> > calibrated and validated as designated in our Master Validation Plan.  IHC 
> > stainers and retrieval units should be validated, even our refrigerators 
> > and freezers are calibrated and validated.  Our pipettors are calibrated 
> > quarterly, and any piece of equipment that generates a weight or 
> > temperature is calibrated yearly.
> >
> > For example if you do not validate your IHC retrieval units how can you 
> > really tell if they reach the temperature that they are programmed to 
> > reach, does the temperature stay consistent through the retrieval process, 
> > did it retrieve for the time programmed?  The only way to determine this is 
> > to perform a validation.  How do you troubleshoot problems if you do not 
> > know if your instruments are performing to their specification without 
> > testing those specifications - that's what equipment validation is and 
> > that's why in my opinion its important.
> >
> > Histology laboratories are now responsible for running IHC that directly 
> > effects a patients treatment - meaning the numerous therapeutic and 
> > prognostic markers we routinely run now.  Validation is an important 
> > process especially if you are using image analysis for these markers.  I 
> > hate to say it but we better get used to it, because this is not going away.
> >
> > And now the shameless plug -  I will be giving a 90 minute lecture at the 
> > Florida State Meeting  
> > https://classic.regonline.com/custImages/240000/241449/FSH2014OnlineProgram.pdf
> >  on this exact topic, so if you want to learn how to create a Master 
> > Validation Plan and learn how to perform a basic validation or a more 
> > detailed IQ/OQ and PQ and to what extent you need to validate a particular 
> > piece of equipment -  sign up for the meeting plus there are lots of other 
> > great topics being presented too.
> >
> > Liz
> >
> > Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC Premier Laboratory, LLC PO
> > Box 18592 Boulder, CO 80308
> > (303) 682-3949 office
> > (303) 682-9060 fax
> > (303) 881-0763 cell
> > l...@premierlab.com
> > www.premierlab.com
> >
> > March 10, 2014 is Histotechnology Professionals Day
> >
> > Ship to Address:
> >
> > Premier Laboratory, LLC
> > 1567 Skyway Drive, Unit E
> > Longmont, CO 80504
> >
> >
> > -----Original Message-----
> > From: histonet-boun...@lists.utsouthwestern.edu
> > [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Martha
> > Ward-Pathology
> > Sent: Thursday, April 03, 2014 1:22 PM
> > To: Cynthia Robinson; Terri Braud; histonet@lists.utsouthwestern.edu
> > Subject: [Histonet] RE: Instrument Verification
> >
> > I'm with you.   There really appears to be no value to this particular 
> > requirement.    I would only be concerned with it if I had just purchased 
> > it, or moved it into our lab from another location.
> >
> >
> >
> > Martha Ward, MT (ASCP) QIHC
> > Manager
> >
> > Molecular Diagnostics Lab
> > Medical Center Boulevard  \  Winston-Salem, NC 27157 p 336.716.2109  \
> > f 336.716.5890 mw...@wakehealth.edu
> >
> >
> >
> >
> >
> > -----Original Message-----
> > From: histonet-boun...@lists.utsouthwestern.edu
> > [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of
> > Cynthia Robinson
> > Sent: Thursday, April 03, 2014 3:03 PM
> > To: Terri Braud; histonet@lists.utsouthwestern.edu
> > Subject: [Histonet] RE: Instrument Verification
> >
> > I agree with you in that CAP is just looking for things to change and 
> > doesn't seem to be considering the change and decrease in staffing seen in 
> > clinical settings. Cryostat validation? Really....cut a slide after you 
> > have cleaned and pm'd the thing and go on. Good grief...I don't need any 
> > more paper and documentation on routine processes. As for tissue 
> > processors, I have 20 year old VIP's that have been running and producing 
> > specimens acceptably. I did validate them prior to being put in use but we 
> > didn't document like we do now. And I don't see the need to do it at this 
> > stage of use. We did do a very extensive validation on the Peloris we put 
> > into use last year and will going forward on new equipment.  To me the 
> > daily QC of stain should provide our 'validation' of the process and 
> > include the processor. I am interested in others thoughts as well.
> >
> > Thanks for allowing me to rant.
> >
> > Cindi Robinson, HT(ASCP)
> > Mercy Medical Center-Sioux City
> > Dunes Medical Laboratories
> > 350 W Anchor Drive
> > Dakota Dunes SD 57049
> >
> >
> >
> >
> > -----Original Message-----
> > From: histonet-boun...@lists.utsouthwestern.edu
> > [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Terri
> > Braud
> > Sent: Thursday, April 03, 2014 1:00 PM
> > To: histonet@lists.utsouthwestern.edu
> > Subject: [Histonet] RE: Instrument Verification
> >
> > I just received my midcycle CAP and for cryostat validation, we are 
> > planning to cut and stain a piece of frozen tonsil and have the path sign 
> > off on it. For the tissue processors, we will run a one minute test 
> > program. I hope this will fly.  Is it just me, or is CAP insanely out of 
> > control with new or modified regulations and policies for AP?
> >
> > Terri L. Braud, HT(ASCP)
> > Anatomic Pathology Supervisor
> > Holy Redeemer Hospital Laboratory
> > 1648 Huntingdon Pike
> > Meadowbrook, PA 19046
> > Ph: 215-938-3676
> > Fax: 215-938-3874
> >
> > 6. Validation of cryostat (Gloria Tharp)
> >
> > Message: 6
> > Date: Thu, 3 Apr 2014 09:59:26 -0500
> > From: "Gloria Tharp" <gth...@pcasoutheast.com> Could anyone tell me how you 
> > are handling the new CAP ANP.23045 question on function and verification of 
> > equipment regarding a cryostat.
> > Gloria Tharp, BA, HTL(ASCP)
> > ------------------------------
> >
> > Message: 7
> > Date: Thu, 3 Apr 2014 15:26:17 +0000
> > From: "Leann M. Murphy" <lmurp...@aultman.com> How is everyone validating 
> > the tissue processor for new CAP ANP.23045 question on function and 
> > verification of equipment?
> > LeAnn Murphy
> > Aultman Hospital
> > Canton, Ohio
> > ************************
> > ----------------------------------------------------------------------
> > -----------
> >
> >
> >
> > CONFIDENTIALITY NOTICE:
> >
> > This E-Mail is intended only for the use of the individual or entity to 
> > which it was sent. It may contain information that is privileged and/or 
> > confidential, and the use or disclosure of such information may also be 
> > restricted under applicable federal and state law. If you received this 
> > communication in error, please do not distribute any part of it or retain 
> > any copies, and delete the original E-Mail.
> > Please notify the sender of any error by E-Mail.
> >
> > Thank you for your cooperation.
> >
> >
> > _______________________________________________
> > Histonet mailing list
> > Histonet@lists.utsouthwestern.edu
> > http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> >
> > _______________________________________________
> > Histonet mailing list
> > Histonet@lists.utsouthwestern.edu
> > http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> >
> >
> > _______________________________________________
> > Histonet mailing list
> > Histonet@lists.utsouthwestern.edu
> > http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> >
> > _______________________________________________
> > Histonet mailing list
> > Histonet@lists.utsouthwestern.edu
> > http://lists.utsouthwestern.edu/mailman/listinfo/histonet
>                                           
> _______________________________________________
> Histonet mailing list
> Histonet@lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> -------------------------------------------------------------------------------------------------------------------
> This email message, including any associated files, is for the sole use of 
> the intended recipient(s) and may contain information that is confidential, 
> privileged, or subject to copyright, trade secret or other protection. This 
> message also may contain information protected by state and federal privacy 
> laws that are enforced through serious civil and criminal sanctions. Any 
> unauthorized review, use, disclosure, or distribution is prohibited. If you 
> are not an intended recipient of this message, please notify the sender 
> immediately by replying to this e-mail, and delete the original and all 
> copies of this message from your computer or other device.
                                          
_______________________________________________
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

Reply via email to