We have never experienced any issue like that here we will use the HIER 
solutions in the PT links for up to three times within a period of 2 weeks.   
We have had our PT link units validated and they are calibrated yearly.  Do you 
review the Target Retrieval Run Reports, we print and keep all of ours.  Our 
units are primarily programed to heat to 80C we put the slides in, warm up to 
95C retrieve for 20 minutes cool down to 80C remove slides right after that.  I 
would think if you left slides in the retrieval solution for different times 
after they are completed you might see some changes in staining intensity, that 
why we try to be consistent and remove the slides as soon as they are 
completed.  We do use other retrieval times and temps on some occasions but 
what I stated is our standard protocol.   I hope this helps.  

There is one other thing, it think it is extremely important to clean the 
instrument as it is required, we have in our SOP's after 200 slides I don't 
know what Dako recommends but we actually clean more often typically after 
around 80 to 120 slides or so.  That will keep your probe nice and clean and 
decrease your variability, if you don't clean you can get build up in the probe 
and that is just going to cause inconsistent staining, its easy to set up a 
cleaning cycle at the end of the day.

There are also a lot of other factors that can affect staining consistency - 
tissue placement on the slide in one thing, we place our tissue in the same 
area on the slide, same number of drop zones, appropriate amount of reagents, 
cutting corners on amount of reagents will not lead to consistent high quality 
staining.    Section thickness can also lead to variation in staining 
intensity.  

Liz

Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC
Premier Laboratory, LLC
PO Box 18592
Boulder, CO 80308
(303) 682-3949 office
(303) 682-9060 fax
(303) 881-0763 cell
l...@premierlab.com
www.premierlab.com

March 10, 2014 is Histotechnology Professionals Day

Ship to Address:

Premier Laboratory, LLC
1567 Skyway Drive, Unit E
Longmont, CO 80504


-----Original Message-----
From: Tiana Baskin via Histonet [mailto:histonet@lists.utsouthwestern.edu] 
Sent: Tuesday, July 14, 2015 8:42 AM
To: 'histonet@lists.utsouthwestern.edu'
Subject: [Histonet] DAKO LINK PTs

Hi Histonetters,

I have encountered a problem with some of my staining and I am struggling to 
pinpoint the root cause. I was wondering if any other DAKO Autostainer Link/PT 
users are experiencing the same oddities.

It seems like the first runs with fresh HIER PT solution is very typical of 
what we have optimized and the second batch of slides has more background and 
in some cases nuclear staining (especially in Actin (1A4) and pan CK (AE1/AE3)) 
when the target is cytoplasmic. We do not use the PT solution more than twice. 
What do others do? Have you seen similar things?

Tiana