Jessica

There are a couple of things that we will try here when we have problems with 
tissue adherence

1.  We utilize a particular brand of plus slides  - trubond 380 - these have 
worked very well in our hands, we purchase them from Newcomer Supply
2.  Post fix in 10% NBF for 10 minutes after you run the slides down to water - 
this does not impact the H&E staining but can impact IHC staining we have seen 
that is may decrease sensitivity so you would need to potentially rework your 
IHC protocol
3.  Our Ram-11 IHC protocol requires HIER we have two different retrieval 
protocols that we have validated for this particular antibody - our routine 95C 
for 30 minutes and then a 70C for 2 hours - we use the longer lower temp 
retrieval when we have problems with tissue adherence, again you may need to 
potentially rework your IHC protocol for the lower temp retrieval.

Hope this helps, good luck

Liz

Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC
Premier Laboratory, LLC
PO Box 18592
Boulder, CO 80308
(303) 682-3949 office
(303) 682-9060 fax
(303) 881-0763 cell
l...@premierlab.com
www.premierlab.com

Ship to Address:

Premier Laboratory, LLC
1567 Skyway Drive, Unit E
Longmont, CO 80504

-----Original Message-----
From: Jessica Riggleman via Histonet [mailto:histonet@lists.utsouthwestern.edu] 
Sent: Monday, September 12, 2016 8:05 AM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Egg Whites and IHC Staining

Hello,

We are cutting large Rabbit PLF formalin fixed paraffin embedded sections.

We have had issues staining H&E and Goldner’s Trichrome. Egg whites is working 
as a great slide adhesive for these two, but not so much for IHC Staining (RAM 
11 specifically).

Any ideas/suggestions on how to get these sections to adhere? We have tried 
many methods, such as positively charged slides, chrom-alum gel, and so on.

Thank you for all input,
Jessica


_____________________________________________________________________

Jessica Riggleman | Research Associate

Globus Medical, Inc.
Valley Forge Business Center
2560 General Armistead Avenue | Audubon, PA 19403
Ph: (610) 930-1800 ext. 2583 | Fax:

Confidentiality Note:  This email is confidential and intended solely for the 
use of the individual to whom it is addressed. If you are not the intended 
recipient, be advised that you have received this email in error and that any 
use, dissemination, forwarding, printing, or copying of this email is strictly 
prohibited. If you have received this email in error please contact the sender. 
Any views or opinions presented are solely those of the author and do not 
necessarily represent those of Globus Medical, Inc. Although this email and any 
attachments are believed to be free of any virus or other defects which might 
affect any computer or IT system into which they are received, no 
responsibility is accepted by Globus Medical, Inc. for any loss or damage 
arising in any way from the receipt or use thereof.


_______________________________________________
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
_______________________________________________
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

Reply via email to