RE: [Histonet] Eosin Leaching

2014-06-24 Thread Michael LaFriniere
I have seen this happen with high levels of Isopropanol in alcohols after the 
eosins, I switch to regular reagent alcohols  and seemed to eliminate the 
bleeding of the eosin.

Michael
Michael R. LaFriniere, HT (ASCP) 
Executive Director
 

Capital Choice Pathology Laboratory
12041 Bournefield Way, Suite A * Silver Spring, MD 20904  
P: 240.471.3427 * F: 240.471.3401 * Cell 410-940-8844
michael.lafrini...@ccplab.com
 
-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Adrienne 
Anderson
Sent: Thursday, June 19, 2014 5:24 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Eosin Leaching

Hello all,

We're having a problem with our eosin bleeding out of the sections. I've read 
that this can be caused by not dehydrating adequately after eosin, but we're 
still having an issue when using fresh alcohols. Does anyone have any other 
ideas as to the cause of this problem?

Thanks in advance for your help!

Adrienne



Adrienne Anderson, BS, HTL(ASCP)
Histotechnologist
Phylogeny, Inc. 
1476 Manning Pkwy, Powell, Ohio 43605
Phone: (614) 846-6161 
Fax:  (877) 591-1815

This message, including any attachments, is confidential and may be privileged 
or may contain health information protected by state and federal law.
Information and opinions expressed in this message and/or attachments are those 
of the author and are not necessarily those of the company. 
If you are not the intended recipient, please notify the sender and delete this 
message from your system. 
Any use of this information by individuals other than the intended recipient is 
strictly prohibited. ___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Eosin leaching

2014-06-23 Thread Debra Siena
Hi Histonet,


Eosin bleaching is usually due to the alcohols that are used after the eosin, 
if you are in a humid area, the alcohols can actually absorb water from the 
humidity in the air. This may mean a change of how often you rotate or change 
the alcohols, possibly more often than you did during the winter months when 
usually less humid.  Also if you use a graded alcohol after the eosin, maybe 
worth looking into going straight to 100%, I usually always recommend at least 
3 changes 100% alcohol for 1 minute and 3 changes of xylene for 1 minute.  If 
the last alcohol in your staining line is pink, it means that there is eosin 
and that means that there is water.  Eosin loves water and loves to come out of 
solution into the one containing the water.  I have even seen pink xylene in 
some staining lines due to water in xylene, however, not enough that it has 
turned cloudy yet but still have a bit of water.  Xylene will tolerate about 
.5% water before becoming too milky or cloudy to use.  Also when you rotate 
your solutions make sure that you dry the containers well before  refilling and 
by graduating the levels of the alcohols and xylenes to make sure that your 
last container completely covers the slides. Lastly, adding a 95% alcohol 
before your eosin, if you don't have one may help to keep excess water out of 
your eosin which will cause the eosin to break down faster, again eosin loves 
water.  Just spouting off the top of my head but hope something helps out.

Debbie Siena
800.442.3573 ext. 229 | www.statlab.com

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Eosin Leaching

2014-06-22 Thread Tony Reilly
Hi Adrienne

Also make sure the wash after your bluing agent is higher than the blueing 
solution or alkalinity of the solution will leach out your eosin.

Regards
Tony


Tony Reilly B.App.Sc,  M.Sc
Chief Scientist
Anatomical Pathology
Pathology Queensland PAH
_
Health Services Support Agency| Department of Health

Building 15, Level 1, 
199 Ipswich Road 
WOOLLOONGABBA  Queensland 4102
Ph: 07 3176 2412
Mob: 0402139411
Fax: 07 3176 2930
Email: tony.reil...@health.qld.gov.au | www.health.qld.gov.au 
     





-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Rathborne, Toni
Sent: Friday, 20 June 2014 10:11 PM
To: Sanders, Jeanine (CDC/OID/NCEZID); 'Adrienne Anderson'; 
'histo...@listsutsouthwestern.edu'
Subject: RE: [Histonet] Eosin Leaching

Is your alcohol level as high as or higher than the level of eosin in the 
containers? If there is residual eosin left at the upper portion of the slide, 
that could be causing the problem.

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Sanders, 
Jeanine (CDC/OID/NCEZID)
Sent: Thursday, June 19, 2014 5:26 PM
To: 'Adrienne Anderson'; 'histonet@lists.utsouthwestern.edu'
Subject: RE: [Histonet] Eosin Leaching

Have you changed your coverslipping mountant?

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Adrienne 
Anderson
Sent: Thursday, June 19, 2014 5:24 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Eosin Leaching

Hello all,

We're having a problem with our eosin bleeding out of the sections. I've read 
that this can be caused by not dehydrating adequately after eosin, but we're 
still having an issue when using fresh alcohols. Does anyone have any other 
ideas as to the cause of this problem?

Thanks in advance for your help!

Adrienne



Adrienne Anderson, BS, HTL(ASCP)
Histotechnologist
Phylogeny, Inc. 
1476 Manning Pkwy, Powell, Ohio 43605
Phone: (614) 846-6161 
Fax:  (877) 591-1815

This message, including any attachments, is confidential and may be privileged 
or may contain health information protected by state and federal law.
Information and opinions expressed in this message and/or attachments are those 
of the author and are not necessarily those of the company. 
If you are not the intended recipient, please notify the sender and delete this 
message from your system. 
Any use of this information by individuals other than the intended recipient is 
strictly prohibited. ___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


This email, including any attachments sent with it, is confidential and for the 
sole use of the intended recipient(s). This confidentiality is not waived or 
lost, if you receive it and you are not the intended recipient(s), or if it is 
transmitted/received in error.
Any unauthorised use, alteration, disclosure, distribution or review of this 
email is strictly prohibited.  The information contained in this email, 
including any attachment sent with it, may be subject to a statutory duty of 
confidentiality if it relates to health service matters.
If you are not the intended recipient(s), or if you have received this email in 
error, you are asked to immediately notify the sender by telephone collect on 
Australia +61 1800 198 175 or by return email.  You should also delete this 
email, and any copies, from your computer system network and destroy any hard 
copies produced.
If not an intended recipient of this email, you must not copy, distribute or 
take any action(s) that relies on it; any form of disclosure, modification, 
distribution and/or publication of this email is also prohibited.
Although Queensland Health takes all reasonable steps to ensure this email does 
not contain malicious software, Queensland Health does not accept 
responsibility for the consequences if any person's computer inadvertently 
suffers any disruption to services, loss of information, harm or is infected 
with a virus, other malicious computer programme or code that may occur as a 
consequence of receiving this email.
Unless stated otherwise, this email represents only the views o

RE: [Histonet] Eosin Leaching

2014-06-20 Thread Vanessa Avalos
We had the same problem. Increase your Alcohol/Reagents (following Eosin) times 
as well as using fresh ones. I use 80% for about 20 sec and (2)100% for a 
minute each and (3) Xylene Subs for 2 min each. I also use a fresh 80% and 
rotate the 100% each day. If using a Xylene Sub it could also be a possibility 
that your mounting medium is not compatible. 

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Adrienne 
Anderson
Sent: Thursday, June 19, 2014 2:24 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Eosin Leaching

Hello all,

We're having a problem with our eosin bleeding out of the sections. I've read 
that this can be caused by not dehydrating adequately after eosin, but we're 
still having an issue when using fresh alcohols. Does anyone have any other 
ideas as to the cause of this problem?

Thanks in advance for your help!

Adrienne



Adrienne Anderson, BS, HTL(ASCP)
Histotechnologist
Phylogeny, Inc. 
1476 Manning Pkwy, Powell, Ohio 43605
Phone: (614) 846-6161 
Fax:  (877) 591-1815

This message, including any attachments, is confidential and may be privileged 
or may contain health information protected by state and federal law.
Information and opinions expressed in this message and/or attachments are those 
of the author and are not necessarily those of the company. 
If you are not the intended recipient, please notify the sender and delete this 
message from your system. 
Any use of this information by individuals other than the intended recipient is 
strictly prohibited. ___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Eosin Leaching

2014-06-20 Thread Anita Buchiane
I'd advise checking your dehydration alcohols with a hydrometer.  This 
"bleeding" happened to me once and it turned out the alcohols after the Eosin 
were not graded.  My lab aide accidently used 100% only instead of 70%, 95% and 
then 100%.  
Anita

Anita

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Adrienne 
Anderson
Sent: Friday, June 20, 2014 8:15 AM
To: Rathborne, Toni
Cc: histonet@lists.utsouthwestern.edu
Subject: Re: [Histonet] Eosin Leaching

We haven't changed our mounting medium, but I'll have to check on the alcohol 
levels. When I stain (we hand stain only - very small lab), I always make sure 
the alcohol is higher than the eosin, but I have a lab assistant now, and it's 
very possible I didn't make that a point. I'll try that.

Any more tips are welcome still - thank you all!

Adrienne


On Jun 20, 2014, at 8:11 AM, Rathborne, Toni  wrote:

> Is your alcohol level as high as or higher than the level of eosin in the 
> containers? If there is residual eosin left at the upper portion of the 
> slide, that could be causing the problem.
> 
> -Original Message-
> From: histonet-boun...@lists.utsouthwestern.edu 
> [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Sanders, 
> Jeanine (CDC/OID/NCEZID)
> Sent: Thursday, June 19, 2014 5:26 PM
> To: 'Adrienne Anderson'; 'histonet@lists.utsouthwestern.edu'
> Subject: RE: [Histonet] Eosin Leaching
> 
> Have you changed your coverslipping mountant?
> 
> -Original Message-
> From: histonet-boun...@lists.utsouthwestern.edu 
> [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Adrienne 
> Anderson
> Sent: Thursday, June 19, 2014 5:24 PM
> To: histonet@lists.utsouthwestern.edu
> Subject: [Histonet] Eosin Leaching
> 
> Hello all,
> 
> We're having a problem with our eosin bleeding out of the sections. I've read 
> that this can be caused by not dehydrating adequately after eosin, but we're 
> still having an issue when using fresh alcohols. Does anyone have any other 
> ideas as to the cause of this problem?
> 
> Thanks in advance for your help!
> 
> Adrienne
> 
> 
> 
> Adrienne Anderson, BS, HTL(ASCP)
> Histotechnologist
> Phylogeny, Inc. 
> 1476 Manning Pkwy, Powell, Ohio 43605
> Phone: (614) 846-6161 
> Fax:  (877) 591-1815
> 
> This message, including any attachments, is confidential and may be 
> privileged or may contain health information protected by state and federal 
> law.
> Information and opinions expressed in this message and/or attachments are 
> those of the author and are not necessarily those of the company. 
> If you are not the intended recipient, please notify the sender and delete 
> this message from your system. 
> Any use of this information by individuals other than the intended recipient 
> is strictly prohibited. ___
> Histonet mailing list
> Histonet@lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> 
> ___
> Histonet mailing list
> Histonet@lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> 


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___

The information contained in, or attached to, this e-mail, may contain 
confidential information and is intended solely for the use of the individual 
or entity to whom it is addressed and may be subject to legal privilege. If you 
have received this e-mail in error you should notify the sender immediately by 
reply e-mail, delete the message from your system and notify your system 
manager. Please do not copy it for any purpose, or disclose its contents to any 
other person. The views or opinions presented in this e-mail are solely those 
of the author and do not necessarily represent those of the company. The 
recipient should check this e-mail and any attachments for the presence of 
viruses. The company accepts no liability for any damage caused, directly or 
indirectly, by any virus transmitted in this email.
___

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Eosin Leaching

2014-06-20 Thread Adrienne Anderson
We haven’t changed our mounting medium, but I’ll have to check on the alcohol 
levels. When I stain (we hand stain only - very small lab), I always make sure 
the alcohol is higher than the eosin, but I have a lab assistant now, and it’s 
very possible I didn’t make that a point. I’ll try that.

Any more tips are welcome still - thank you all!

Adrienne


On Jun 20, 2014, at 8:11 AM, Rathborne, Toni  wrote:

> Is your alcohol level as high as or higher than the level of eosin in the 
> containers? If there is residual eosin left at the upper portion of the 
> slide, that could be causing the problem.
> 
> -Original Message-
> From: histonet-boun...@lists.utsouthwestern.edu 
> [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Sanders, 
> Jeanine (CDC/OID/NCEZID)
> Sent: Thursday, June 19, 2014 5:26 PM
> To: 'Adrienne Anderson'; 'histonet@lists.utsouthwestern.edu'
> Subject: RE: [Histonet] Eosin Leaching
> 
> Have you changed your coverslipping mountant?
> 
> -Original Message-
> From: histonet-boun...@lists.utsouthwestern.edu 
> [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Adrienne 
> Anderson
> Sent: Thursday, June 19, 2014 5:24 PM
> To: histonet@lists.utsouthwestern.edu
> Subject: [Histonet] Eosin Leaching
> 
> Hello all,
> 
> We're having a problem with our eosin bleeding out of the sections. I've read 
> that this can be caused by not dehydrating adequately after eosin, but we're 
> still having an issue when using fresh alcohols. Does anyone have any other 
> ideas as to the cause of this problem?
> 
> Thanks in advance for your help!
> 
> Adrienne
> 
> 
> 
> Adrienne Anderson, BS, HTL(ASCP)
> Histotechnologist
> Phylogeny, Inc. 
> 1476 Manning Pkwy, Powell, Ohio 43605
> Phone: (614) 846-6161 
> Fax:  (877) 591-1815
> 
> This message, including any attachments, is confidential and may be 
> privileged or may contain health information protected by state and federal 
> law.
> Information and opinions expressed in this message and/or attachments are 
> those of the author and are not necessarily those of the company. 
> If you are not the intended recipient, please notify the sender and delete 
> this message from your system. 
> Any use of this information by individuals other than the intended recipient 
> is strictly prohibited. ___
> Histonet mailing list
> Histonet@lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> 
> ___
> Histonet mailing list
> Histonet@lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> 


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Eosin Leaching

2014-06-20 Thread Rathborne, Toni
Is your alcohol level as high as or higher than the level of eosin in the 
containers? If there is residual eosin left at the upper portion of the slide, 
that could be causing the problem.

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Sanders, 
Jeanine (CDC/OID/NCEZID)
Sent: Thursday, June 19, 2014 5:26 PM
To: 'Adrienne Anderson'; 'histonet@lists.utsouthwestern.edu'
Subject: RE: [Histonet] Eosin Leaching

Have you changed your coverslipping mountant?

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Adrienne 
Anderson
Sent: Thursday, June 19, 2014 5:24 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Eosin Leaching

Hello all,

We're having a problem with our eosin bleeding out of the sections. I've read 
that this can be caused by not dehydrating adequately after eosin, but we're 
still having an issue when using fresh alcohols. Does anyone have any other 
ideas as to the cause of this problem?

Thanks in advance for your help!

Adrienne



Adrienne Anderson, BS, HTL(ASCP)
Histotechnologist
Phylogeny, Inc. 
1476 Manning Pkwy, Powell, Ohio 43605
Phone: (614) 846-6161 
Fax:  (877) 591-1815

This message, including any attachments, is confidential and may be privileged 
or may contain health information protected by state and federal law.
Information and opinions expressed in this message and/or attachments are those 
of the author and are not necessarily those of the company. 
If you are not the intended recipient, please notify the sender and delete this 
message from your system. 
Any use of this information by individuals other than the intended recipient is 
strictly prohibited. ___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Eosin Leaching

2014-06-19 Thread Adrienne Anderson
Hello all,

We’re having a problem with our eosin bleeding out of the sections. I’ve read 
that this can be caused by not dehydrating adequately after eosin, but we’re 
still having an issue when using fresh alcohols. Does anyone have any other 
ideas as to the cause of this problem?

Thanks in advance for your help!

Adrienne



Adrienne Anderson, BS, HTL(ASCP)
Histotechnologist
Phylogeny, Inc. 
1476 Manning Pkwy, Powell, Ohio 43605
Phone: (614) 846-6161 
Fax:  (877) 591-1815

This message, including any attachments, is confidential and may be privileged 
or may contain health information protected by state and federal law.
Information and opinions expressed in this message and/or attachments are those 
of the author and are not necessarily those of the company. 
If you are not the intended recipient, please notify the sender and delete this 
message from your system. 
Any use of this information by individuals other than the intended recipient is 
strictly prohibited. ___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Eosin Leaching

2014-06-19 Thread Sanders, Jeanine (CDC/OID/NCEZID)
Have you changed your coverslipping mountant?

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Adrienne 
Anderson
Sent: Thursday, June 19, 2014 5:24 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Eosin Leaching

Hello all,

We're having a problem with our eosin bleeding out of the sections. I've read 
that this can be caused by not dehydrating adequately after eosin, but we're 
still having an issue when using fresh alcohols. Does anyone have any other 
ideas as to the cause of this problem?

Thanks in advance for your help!

Adrienne



Adrienne Anderson, BS, HTL(ASCP)
Histotechnologist
Phylogeny, Inc. 
1476 Manning Pkwy, Powell, Ohio 43605
Phone: (614) 846-6161 
Fax:  (877) 591-1815

This message, including any attachments, is confidential and may be privileged 
or may contain health information protected by state and federal law.
Information and opinions expressed in this message and/or attachments are those 
of the author and are not necessarily those of the company. 
If you are not the intended recipient, please notify the sender and delete this 
message from your system. 
Any use of this information by individuals other than the intended recipient is 
strictly prohibited. ___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Eosin leaching out of sections

2010-02-03 Thread Rathborne, Toni
Where do you get the beads from?

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu]on Behalf Of Joyce
Cline
Sent: Wednesday, February 03, 2010 2:02 PM
To: histonet@lists.utsouthwestern.edu
Subject: RE: [Histonet] Eosin leaching out of sections 


Humidity has played a large part in our slides leaching Eosin. We have a 
dehumidifier in our staining room and we also use H2Blue Beads in our Xylene 
substitute. The beads absorb water from the substitute and help prevent 
leaching of the stain.

The beads are ordered from Amercian Mastertech.

Joyce Cline, Technical Specialist
Hagerstown Medical Laboratory
301-665-4980
fax 301-665-4941

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Scott
Hendricksen
Sent: 03 February 2010 04:16
To: Histonet@lists.utsouthwestern.edu
Subject: [Histonet] Eosin leaching out of sections

Hi again,

Does anyone have an issue with eosin leaching out of sections a few days
after they have been coverslipped on an H&E stained slide?

Thanks,

Scott Hendricksen HT(ASCP)
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

* CONFIDENTIALITY NOTICE * This message contains confidential 
information and is intended only for the individual named. If you are not the 
named addressee you should not disseminate, distribute or copy this e-mail. 
Please notify the sender immediately by e-mail if you have received this e-mail 
by mistake and delete this e-mail from your system.

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


CONFIDENTIALITY NOTICE
This message and any included attachments are from Somerset Medical Center
and are intended only for the addressee.  The information contained in this
message is confidential and may contain privileged, confidential,
proprietary and/or trade secret information entitled to protection and/or
exemption from disclosure under applicable law.  Unauthorized forwarding,
printing, copying, distribution, or use of such information is strictly
prohibited and may be unlawful.  If you are not the addressee, please
promptly delete this message and notify the sender of the delivery error
by e-mail or you may call Somerset Medical Center's computer Help Desk
at 908-685-2200, ext. 4050.

Be sure to visit Somerset Medical Center's Web site - 
www.somersetmedicalcenter.com - for the most up-to-date news, 
event listings, health information and more.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Eosin leaching out of sections

2010-02-03 Thread Joyce Cline
Humidity has played a large part in our slides leaching Eosin. We have a 
dehumidifier in our staining room and we also use H2Blue Beads in our Xylene 
substitute. The beads absorb water from the substitute and help prevent 
leaching of the stain.

The beads are ordered from Amercian Mastertech.

Joyce Cline, Technical Specialist
Hagerstown Medical Laboratory
301-665-4980
fax 301-665-4941

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Scott
Hendricksen
Sent: 03 February 2010 04:16
To: Histonet@lists.utsouthwestern.edu
Subject: [Histonet] Eosin leaching out of sections

Hi again,

Does anyone have an issue with eosin leaching out of sections a few days
after they have been coverslipped on an H&E stained slide?

Thanks,

Scott Hendricksen HT(ASCP)
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

* CONFIDENTIALITY NOTICE * This message contains confidential 
information and is intended only for the individual named. If you are not the 
named addressee you should not disseminate, distribute or copy this e-mail. 
Please notify the sender immediately by e-mail if you have received this e-mail 
by mistake and delete this e-mail from your system.

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


FW: [Histonet] Eosin leaching out of sections

2010-02-03 Thread Ian Montgomery
Scott,
Did you mount from water or a clearing agent?

Dr. Ian Montgomery,
Histotechnology,
I.B.L.S. Support Unit,
Thomson Building,
University of Glasgow,
Glasgow,
G12 8QQ.

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Scott
Hendricksen
Sent: 03 February 2010 04:16
To: Histonet@lists.utsouthwestern.edu
Subject: [Histonet] Eosin leaching out of sections 

Hi again,

Does anyone have an issue with eosin leaching out of sections a few days
after they have been coverslipped on an H&E stained slide? 

Thanks,

Scott Hendricksen HT(ASCP)
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Eosin leaching out of sections

2010-02-03 Thread Bonner, Janet
Your xylene and/or alcohols may need changing.
 
Janet 



From: histonet-boun...@lists.utsouthwestern.edu on behalf of Scott Hendricksen
Sent: Wed 2/3/2010 9:02 AM
To: Histonet@lists.utsouthwestern.edu
Subject: [Histonet] Eosin leaching out of sections 



> Hi again,
>
> Does anyone have an issue with eosin leaching out of sections a few days 
> after they have been coverslipped on an H&E stained slide?
>
> Thanks,
>
> Scott Hendricksen HT(ASCP)


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



===
The information contained in this message may be privileged and/or confidential
and protected from disclosure.  If the reader of this message is not the 
intended 
recipient or an employee or agent responsible for delivering this message to 
the 
intended recipient, you are hereby notified that any dissemination, 
distribution 
or copying of this communication is strictly prohibited.  If you have received 
this
communication in error, please notify the sender immediately by replying to 
this 
message and deleting the material from any computer.
===
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Eosin leaching out of sections

2010-02-03 Thread Scott Hendricksen
> Hi again,
> 
> Does anyone have an issue with eosin leaching out of sections a few days 
> after they have been coverslipped on an H&E stained slide? 
> 
> Thanks,
> 
> Scott Hendricksen HT(ASCP)


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Eosin leaching out of sections

2010-02-03 Thread Scott Hendricksen
Hi again,

Does anyone have an issue with eosin leaching out of sections a few days after 
they have been coverslipped on an H&E stained slide? 

Thanks,

Scott Hendricksen HT(ASCP)
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Eosin leaching out of sections

2010-02-02 Thread Scott Hendricksen
Hi again,

Does anyone have an issue with eosin leaching out of sections a few days after 
they have been coverslipped on an H&E stained slide? 

Thanks,

Scott Hendricksen HT(ASCP)
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet