Dear Graeme,

at the ECM last year Arwen Pearson suggested an even more sophisticated
method than the sum of runs. It was based on a set of random sums based
on Hadamard-matrices and the subsequent 'deconvolution'.

It sounded very promising to improve signal-to-noise and to turn your
sentence "you may get better data" into "you will get better data". I
think this would be worth implementing at beamlines - do you know if
anything in this direction is on its way?

Best,
Tim

On 05/01/2014 09:25 AM, Graeme Winter wrote:
> Hi All,
> 
> A major opportunity with Pilatus detectors is the chance to redistribute
> the dose in reciprocal space i.e. measure a lot more data, with less dose /
> frame, then decide in hindsight where you probably should have cut off the
> data set.
> 
> It is certainly true that "strategies" such as 0.2 s/0.2 degree (I would
> call this a tactic myself ;oD) seem to work well, and that it often seems
> that you need a reasonable dose to be able to process the data properly
> (see below). I would however agree strongly that unless you are not
> vulnerable to radiation damage the use of a strategy program such as EDNA
> is critical as continuous readout of a fast detector can let you kill your
> sample really quickly... and it would be a shame to measure the wrong part
> of reciprocal space.
> 
> Also the 0.2s / 0.2 degree rate is very beamline dependent. Here at Diamond
> it is certainly routine to measure data with 0.05 s / 0.1 degree exposure
> times with Pilatus2 and end up with very good data, and the latest Pilatus3
> machines can run with 0.01s exposure times. As Nukri said earlier, once you
> start running at these very high rates you become much more sensitive to
> beamline and source characteristics, so your mileage may vary and so on.
> It's certainly worth spending some time exploring the capability and what
> works well for *your* samples. I would however strongly agree with the
> recommendations for fine slicing, and avoid e.g. 1 degree images.
> 
> In terms of "a reasonable dose to process the data properly" there are some
> major challenges when dealing with exceedingly weak data in measuring the
> reflections at high resolution well: the statistics start to become poorly
> behaved with current analysis software. One tactic I have been playing with
> is to record the same wedge of data (for example from an EDNA strategy)
> with exceedingly low dose perhaps 20 times, then to process this and look
> for signs of radiation damage. After arbitrarily deciding which "pass"
> radiation damage kicked in at then *sum* the *raw images* from each pass up
> to this point e.g.
> 
> pass_1_0001.cbf + pass_2_0001.cbf + .... pass_N_0001.cbf => sum_0001.cbf
> 
> Then process these summed images as if this was the original data. Funnily
> enough you may get better data than processing pass_1 to pass_N separately
> and then scaling and merging all of the measurements, which leads me to
> pointing the pointy finger of blame at the behaviour of the statistics, and
> that statistics and things like background subtraction become hard when you
> have very sparse data.
> 
> This summing process may seem like manipulating your raw data (naughty!!)
> but in essence it is really just performing the same process as when you
> recorded multiple exposures / passes on a single CCD image. It also has the
> happy side effect of averaging out any random / high frequency effects
> induced from source / beamline effects, but will also average in any
> radiation damage effects as well! This by the way is what I was getting at
> with redistributing your dose in reciprocal space...
> 
> Cheerio, Graeme
> 
> 
> 
> 
> On 30 April 2014 17:41, Harry Powell <ha...@mrc-lmb.cam.ac.uk> wrote:
> 
>> Hi
>>
>> Marcus Mueller (from Dectris, who develop and manufacture the Pilatus) did
>> some work on this a couple of years ago and determined that an oscillation
>> angle ~ 0.5x the mosaicity of the crystal (using the XDS value of
>> mosaicity, which is not the same as Mosflm's); the abstract says -
>>
>>  The results show that fine ’-slicing can substantially improve scaling
>>> statistics and anomalous signal provided that the rotation angle is
>>> comparable to half the crystal mosaicity.
>>>
>>>
>>> Acta Cryst. (2012). D68, 42-56    [ doi:10.1107/S0907444911049833 ]
>>> Optimal fine
>>
>>
>> -slicing for single-photon-counting pixel detectors
>>>
>>> M. Mueller, M. Wang and C. Schulze-Briese
>>>
>>>
>> My reading of this is that there is still a place for strategy
>> calculations.
>>
>>
>>
>> On 30 Apr 2014, at Wed30 Apr 15:06, Sanishvili, Ruslan wrote:
>>
>>  Hi Jacob,
>>>
>>> I'll take a first crack as I am sure many will follow.
>>> It is true that with CCD detectors one has to be careful how small an
>>> oscillation range to use for a frame before read noise starts to eat into
>>> the data quality.
>>> Pilatus offers two major new features - is fast and is photon counting as
>>> opposed to integrating detector.
>>> The speed allows to collect data without a shutter and it is very
>>> important as it can dramatically improve data quality. Now there are fast
>>> CCD detectors as well on the market.
>>> Being a photon counter, Pilatus has no "read" noise which, as you have
>>> pointed out, allows you to collect as thin a frame as you want. However, it
>>> is if you consider the detector only. In reality, if you go very thin and
>>> very fast, you may not have enough flux to record the data. Also, even once
>>> we get rid of the shutter, there are still other sources of instabilities
>>> and they do affect the fast data collection adversely. One could try going
>>> (very) thin sliced and somewhat slow but there is another gotcha there.
>>> Most rotation stages used for rotating the sample crystal, do not like
>>> going extremely slow which would be the case with thin frames and long
>>> exposure times. In this case the speed may not remain as constant as we
>>> would like during data collection.
>>> I think there was a publication from Diamond Synchrotron discussing
>>> strategies of data collection with Pilatus.
>>> We've done a little bit of systematic studies as well and while things
>>> may well be sample- and facility-dependent, ~0.2 degree frames with ~0.2
>>> sec exposure time seemed to make good compromise between above-mentioned
>>> issues. Here I would like to emphasize again - there certainly will be
>>> samples which will benefit from somewhat different parameters.
>>> Hope it helps,
>>> Cheers,
>>> N.
>>>
>>> Ruslan Sanishvili (Nukri)
>>> Macromolecular Crystallographer
>>> GM/CA@APS
>>> X-ray Science Division, ANL
>>> 9700 S. Cass Ave.
>>> Lemont, IL 60439
>>>
>>> Tel: (630)252-0665
>>> Fax: (630)252-0667
>>> rsanishv...@anl.gov
>>>
>>>
>>> ________________________________________
>>> From: CCP4 bulletin board [CCP4BB@jiscmail.ac.uk] on behalf of Keller,
>>> Jacob [kell...@janelia.hhmi.org]
>>> Sent: Wednesday, April 30, 2014 7:49 AM
>>> To: CCP4BB@jiscmail.ac.uk
>>> Subject: [ccp4bb] Pilatus and Strategy wrt Radiation Damage
>>>
>>> Dear Pilatus/Radiation Damage Cognoscenti,
>>>
>>> I read a few years ago, before the advent of Pilatus detectors, that the
>>> best strategy was a sort of compromise between number of images and
>>> detector readout noise "overhead." I have heard that Pilatus detectors,
>>> however, have essentially no readout noise, so I am wondering whether
>>> strategies have changed in light of this, i.e., is the best practice now to
>>> collect as many images as possible at lowest exposure possible?
>>>
>>> JPK
>>>
>>> *******************************************
>>> Jacob Pearson Keller, PhD
>>> Looger Lab/HHMI Janelia Farms Research Campus
>>> 19700 Helix Dr, Ashburn, VA 20147
>>> email: kell...@janelia.hhmi.org
>>> *******************************************
>>>
>>
>> Harry
>> --
>> ** note change of address **
>> Dr Harry Powell, MRC Laboratory of Molecular Biology, Francis Crick
>> Avenue, Cambridge Biomedical Campus, Cambridge, CB2 0QH
>> Chairman of European Crystallographic Association SIG9 (Crystallographic
>> Computing)
>>
>>
>>
>>
>>
>>
> 

-- 
Dr Tim Gruene
Institut fuer anorganische Chemie
Tammannstr. 4
D-37077 Goettingen

GPG Key ID = A46BEE1A

Attachment: signature.asc