Dear Antony,
Have you considered MAIN - www-bmb.ijs.si ?
best regards,
dusan


On Jun 19, 2014, at 1:06 AM, CCP4BB automatic digest system 
<lists...@jiscmail.ac.uk> wrote:

> There are 6 messages totaling 843 lines in this issue.
> 
> Topics of the day:
> 
>  1. Rotamer Selection in Low Resolution Data (4)
>  2. Refinement
>  3. [phenixbb] Calculation of RSR for a fragment of PDB
> 
> ----------------------------------------------------------------------
> 
> Date:    Wed, 18 Jun 2014 15:39:02 +0000
> From:    Antony Oliver <antony.oli...@sussex.ac.uk>
> Subject: Rotamer Selection in Low Resolution Data
> 
> Dear Crystallographic Community,
> 
> Apologies for the cross-posting, but I *do* routinely use programs from all 
> three software packages.
> 
> I find myself refining a relatively low resolution structure (3.5 Angstrom) - 
> with 8 molecules in the asymmetric unit.
> Is there a *simple* automated way to place “optimal-fit to electron density" 
> side-chain rotamers into my model?
> Preferably in an NCS-independant manner?
> 
> With thanks,
> Antony.
> 
> - - - - - - - - - - - - - - - - - -
> Dr Antony W Oliver
> Senior Research Fellow
> CR-UK DNA Repair Enzymes Group
> Genome Damage and Stability Centre
> Science Park Road
> University of Sussex
> Falmer, Brighton, BN1 9RQ
> - - - - - - - - - - - - - - - - - -
> email: antony.oli...@sussex.ac.uk<mailto:antony.oli...@sussex.ac.uk>
> 
> tel (office): +44 (0)1273 678349
> tel (lab): +44 (0)1273 677512
> 
> http://www.sussex.ac.uk/lifesci/oliverlab
> http://tinyurl.com/aw-oliver
> - - - - - - - - - - - - - - - - - -
> 
> ------------------------------
> 
> Date:    Wed, 18 Jun 2014 09:13:57 -0700
> From:    Pavel Afonine <pafon...@gmail.com>
> Subject: Re: Rotamer Selection in Low Resolution Data
> 
> Hi Antony,
> 
> Apologies for the cross-posting, but I *do* routinely use programs from all
>> three software packages.
>> 
>> I find myself refining a relatively low resolution structure (3.5
>> Angstrom) - with 8 molecules in the asymmetric unit.
>> Is there a *simple* automated way to place “optimal-fit to electron
>> density" side-chain rotamers into my model?
>> Preferably in an NCS-independant manner?
>> 
> 
> naively assuming that one of the "three software packages" that you did not
> mention by name is Phenix:
> yes, you can do it in a number of different ways. Let me know if interested
> and I will list all options.
> 
> Pavel
> 
> ------------------------------
> 
> Date:    Wed, 18 Jun 2014 16:21:20 +0000
> From:    Antony Oliver <antony.oli...@sussex.ac.uk>
> Subject: Re: Rotamer Selection in Low Resolution Data
> 
> Hi Pavel,
> 
> Sorry… the current ‘triumvirate’ is, in no particular order:
> 
> CCP4, Phenix and Buster (Global Phasing).
> 
> Any suggestions would indeed be useful.
> 
> Many thanks,
> 
> Antony.
> 
> - - - - - - - - - - - - - - - - - -
> Dr Antony W Oliver
> Senior Research Fellow
> CR-UK DNA Repair Enzymes Group
> Genome Damage and Stability Centre
> Science Park Road
> University of Sussex
> Falmer, Brighton, BN1 9RQ
> - - - - - - - - - - - - - - - - - -
> email: antony.oli...@sussex.ac.uk<mailto:antony.oli...@sussex.ac.uk>
> 
> tel (office): +44 (0)1273 678349
> tel (lab): +44 (0)1273 677512
> 
> http://www.sussex.ac.uk/lifesci/oliverlab
> http://tinyurl.com/aw-oliver
> - - - - - - - - - - - - - - - - - -
> 
> On 18 Jun 2014, at 17:13, Pavel Afonine 
> <pafon...@gmail.com<mailto:pafon...@gmail.com>> wrote:
> 
> Hi Antony,
> 
> Apologies for the cross-posting, but I *do* routinely use programs from all 
> three software packages.
> 
> I find myself refining a relatively low resolution structure (3.5 Angstrom) - 
> with 8 molecules in the asymmetric unit.
> Is there a *simple* automated way to place “optimal-fit to electron density" 
> side-chain rotamers into my model?
> Preferably in an NCS-independant manner?
> 
> naively assuming that one of the "three software packages" that you did not 
> mention by name is Phenix:
> yes, you can do it in a number of different ways. Let me know if interested 
> and I will list all options.
> 
> Pavel
> 
> ------------------------------
> 
> Date:    Wed, 18 Jun 2014 13:25:24 -0400
> From:    Remie Fawaz-Touma <remiefa...@gmail.com>
> Subject: Refinement
> 
> Hi all, 
> 
> I need to refine a structure keeping the ligands or some of the ligands 
> intact (in same positions as before refinement). 
> Please give me specific instructions if you are familiar with this procedure. 
> I am using Refmac 5 in CCP4.
> 
> Thanks so much for your help.
> 
> Remie
> 
> ------------------------------
> 
> Date:    Wed, 18 Jun 2014 14:28:54 -0400
> From:    "Edward A. Berry" <ber...@upstate.edu>
> Subject: Re: [phenixbb] Calculation of RSR for a fragment of PDB
> 
> On 06/16/2014 04:26 PM, George Devaniranjan wrote:
>> Hi everyone,
>> 
>> I apologize at the beginning itself as I am new to using PHENIX/X-ray 
>> crystallography so this question might not make the most sense but I will 
>> try and explain it well.
>> 
>> I cut out a part of a map for a PDB fragment using phenix map cut out 
>> density resulting in a MTZ file.
>> 
>> Now I changed the fragment (changed its phi and psi externally) and want to 
>> calculate the RSR [real-space R -factor] for the model based on the old 
>> density (to check if I have made things worse or not).
>> 
>> how can I go about this? I tried using phenix.refine couldn't get it to work.
>> 
>> I have attached the MTZ fragment, the original PDB fragment and the modified 
>> PDB fragment.
>> native: nativeFragment.pdb
>> model: modelledFragment.pdb
>> MTZ: cut_out_density.MTZ
>> 
>> (I can see mine is worse but I want to quantify that "badness")
>> 
> 
> I'm sure phenix has tools to do what you want, but if you still have 
> difficulty
> and want to go back to G. Kleywegt's suggestions of using mapman, step by step
> instructions are below.
> 
> First you need to understand that an mtz file is not a map-
> a map is a list of the value of density at points, like pixels in a digital 
> picture
> but in 3D (voxels) instead of 2D.
> The mtz file contains structure factors which are the Fourier transform 
> coefficients
> of the map. With modern computers the map can be calculated rapidly from the 
> structure
> factors, and many applications work directly with mtz files and never save a 
> map file.
> Still the distinction is important- for examples you cannot calculate a CC 
> between
> mtz files in a defined area, because of the all-on-all relation between
> reflections and realspace. You can calculate the CC between two maps in a 
> defined
> area, which is what the suggested mapman command does.
> 
> As I understand it, mapman calculates the RSCC and RSR between two maps- the
> pdb file is only used to define the area in which the values are calculated.
> Since it uses the pdb to define the area, there is no need to cut out density
> around the fragment- I recommend using the original mtz to calculate the 
> "obs" map,
> that way you can use the same one for all fragments.
> 
> So you need to calculate
> a map from the original mtz covering all the fragments
> a map (Fc-Phic) from your fragment
> (They presumably must have the same cell parameters and grid spacing)
> and give them together with the fragment.pdb to mapman.
> 
> Do this once for nativefragment.pdb
> and again for modelledfragment.pdb
> and compare the results
> Remember phases in the original mtz wil be biased in favor of the original 
> model,
> so if the modified ragment is equally good it will still get a worse score.
> 
> step by step:
> 
> 1. Use fft to calculate a map from the original mtz:
> From the ccp4i gui, on the left side, select program list, then under that 
> fft.
> fill in according to the screen shot at  http://sb20.lbl.gov/ccp4/RunFFT-1.png
> (this is working with your cut-out-density.mtz but I recommend using the
> original mtz). see notes below.
> 
> 2. Use sfall to calculate SF's from nativefragment.pdb making 
> nativefragment.mtz
>       (http://sb20.lbl.gov/ccp4/RunSFALL.png)
>       (with a script sfall can make the map directly, don't see that in the 
> gui)
> 
> 3. use fft to calculate a map from nativefragment.mtz
>       (http://sb20.lbl.gov/ccp4/RunFFT-2.png)
> 
> 4. run mapman in the directory where these files were created:
> 
> MAPMAN > read m1 cut_out_density.map ccp4
> MAPMAN > read m2 nativefragment.map ccp4
> MAPMAN > RS_fit
>  Which obs_map ? (m2) m1
> Which calc_map ? (NATIVEFRAGMENT.MAP) m2
>  Model PDB file ? (m1.pdb) nativeFragment.pdb
>  RS-fit list output file ? (rs_fit.list) fragment27CC.txt
> 
> . .
>  Non-water residues with RSCC : (         28)
>  ... Average RSCC for these   : (   0.775)
> 
>  Non-water residues with RSR  : (         28)
>  ... Average RSR for these    : (   0.247)
> 
> 6. now repeat 2 - 4 using modified fragment instead of native fragment
> 
> 
> Notes-
> FFT-1:
>  Uncheck the box that says "append to an existing . . ."
> FFT will probably guess which columns to use, at least if coot was able to 
> auto-open.
> if this is blank and you don't know, send a list of the column labels.
> set the resolution say 100 to 2 A.
> click the bar that opens up "infrequently used options". You need to
> set the grid spacing so it will be the same in both maps, and fine
> enough that it doesn't distort either map. Should have no large prime
> factors- power of 2 like 64, 128 etc is fine.
> (Or let sfall choose grid based on resolution, note what it is and use the 
> same
> in step 3 when you make the other map)
> If your original mtz is not P1 and all of the fragments are not in the 
> asymmetric unit,
> you may need to use the option to extend map to cover (whole) pdb
> You only need to do this once in that case, for all the fragments.
> 
> sfall:  make sure the cell param are the same as in step 1, which will be 
> determined by orig mtz.
> 
> fft-2: use the same grid as in step 1. cell param taken from the mtz should 
> be right
> 
> ------------------------------
> 
> Date:    Wed, 18 Jun 2014 15:11:25 -0700
> From:    Pavel Afonine <pafon...@gmail.com>
> Subject: Re: Rotamer Selection in Low Resolution Data
> 
> Hi Antony,
> 
> (Sorry for replying with Phenix specific suggestions in non-Phenix forum.
> Since the whole conversation started here and seems to move on I thought
> it's best to keep it here rather than move off-list or to a more tailored
> list. Also several people expressed the interest asking directly.)
> 
> ok, here is the list. Some options are not exactly what you want but
> relevant and (very) close, some are supposed to do just what you want.
> Also, I assume you have the latest version of Phenix.
> 
> *1) Use phenix.real_space_refine:*
> 
> http://phenix-online.org/documentation/reference/real_space_refine.html
> For this you need a model (PDB file) and a map in CCP4 format or Fourier
> map coefficients.
> Since I'm not certain which one is going to work best in your specific
> case, I would run it in two ways (sorry, not GUI yet):
> 
> phenix.real_space_refine model.pdb map_coefficients.mtz
> phenix.real_space_refine model.pdb map_coefficients.mtz
> run=minimization_global+local_grid_search
> 
> Both should tidy up geometry while keeping best fit to the map, and also
> result in models with NO rotamer outliers or just a few (if you get
> outliers I would be interested to know, as this is not expected). The
> second option may run considerably longer. Resulting model should be a good
> starting point for usual reciprocal-space refinement in phenix.refine or
> other refinement program of your choice.
> 
> map_coefficients.mtz can be MTZ file out of a phenix.refine (it will
> contain 2mFo-DFc map coefficients) or any other MTZ file containing an
> equivalent map coefficients. You may be prompted to specify them when
> running phenix.real_space_refine. Also, you can calculate the map like this
> 
> phenix.maps model.pdb data.mtz
> 
> or using the GUI.
> 
> *2) Idealize model geometry minimally moving the model from starting
> (initial) position.*
> 
> This is done using phenix.geometry_minimization (available in the GUI; some
> command line defaults may be different in the GUI):
> http://phenix-online.org/documentation/reference/geometry_minimization.html
> 
> Using as simple as
> 
> phenix.geometry_minimization model.pdb
> 
> or drop in your file into the GUI, make sure you check "fix rotamers" box,
> and run.
> 
> By default it will idealize all usual targets: bond, angle, planarity,
> dihedral, chirality, as well as it will use C-beta deviations restraints
> and for each residue side chain rotamer outlier it will switch it to the
> nearest valid rotamer. On top of that, optionally, you can add
> secondary-structure and Ramachandran plot restrains. Ramachandran plot
> restrains come in two different flavors, which is a topic of a separate
> discussion - let me know if you get to this!
> 
> Important: since diffraction data is not used here, the resulting model can
> (potentially) deviate quite a lot from the starting one. Most likely you
> don't want this if you want to use this model to continue refinement. To
> make sure the model does not deviate much from initial state you can
> restrain it to starting point (I don't remember if these parameters are
> exposed in the GUI):
> 
> phenix.geometry_minimization model.pdb
> reference_restraints.restrain_starting_coord_selection=all
> reference_restraints.coordinate_sigma=0.7
> 
> You can be more specific, and restrain only selected atoms to initial
> position:
> 
> phenix.geometry_minimization model.pdb
> reference_restraints.restrain_starting_coord_selection="chain A and resseq
> 1:123" reference_restraints.coordinate_sigma=0.7
> 
> *3) Use phenix.refine.*
> 
> In theory a default phenix.refine run includes step #1 above. In reality
> I'm yet to optimize how real- and reciprocal-space refinements play
> together in most optimal way. This means that while real-space refinement
> part of phenix.refine will fix all rotamer outliers, its reciprocal-space
> refinement part my not keep them all. Also, in case of large molecule
> real-space refinement in phenix.refine may take quite some time.
> 
> *In summary*, my suggestion is to try all three options with all plausible
> to your case variants each one offers, and see which one works best.
> 
> Good luck!
> Pavel
> 
> 
> On Wed, Jun 18, 2014 at 9:21 AM, Antony Oliver <antony.oli...@sussex.ac.uk>
> wrote:
> 
>> Hi Pavel,
>> 
>> Sorry… the current ‘triumvirate’ is, in no particular order:
>> 
>> CCP4, Phenix and Buster (Global Phasing).
>> 
>> Any suggestions would indeed be useful.
>> 
>> Many thanks,
>> 
>> Antony.
>> 
>> - - - - - - - - - - - - - - - - - -
>> Dr Antony W Oliver
>> Senior Research Fellow
>> CR-UK DNA Repair Enzymes Group
>> Genome Damage and Stability Centre
>> Science Park Road
>> University of Sussex
>> Falmer, Brighton, BN1 9RQ
>> - - - - - - - - - - - - - - - - - -
>> email: antony.oli...@sussex.ac.uk
>> 
>> tel (office): +44 (0)1273 678349
>> tel (lab): +44 (0)1273 677512
>> 
>> http://www.sussex.ac.uk/lifesci/oliverlab
>> http://tinyurl.com/aw-oliver
>> - - - - - - - - - - - - - - - - - -
>> 
>> On 18 Jun 2014, at 17:13, Pavel Afonine <pafon...@gmail.com> wrote:
>> 
>> Hi Antony,
>> 
>>  Apologies for the cross-posting, but I *do* routinely use programs from
>>> all three software packages.
>>> 
>>> I find myself refining a relatively low resolution structure (3.5
>>> Angstrom) - with 8 molecules in the asymmetric unit.
>>> Is there a *simple* automated way to place “optimal-fit to electron
>>> density" side-chain rotamers into my model?
>>> Preferably in an NCS-independant manner?
>>> 
>> 
>> naively assuming that one of the "three software packages" that you did
>> not mention by name is Phenix:
>> yes, you can do it in a number of different ways. Let me know if
>> interested and I will list all options.
>> 
>> Pavel
>> 
>> 
>> 
> 
> ------------------------------
> 
> End of CCP4BB Digest - 17 Jun 2014 to 18 Jun 2014 (#2014-166)
> *************************************************************

Dr. Dusan Turk, Prof.
Head of Structural Biology Group http://bio.ijs.si/sbl/ 
Head of Centre for Protein  and Structure Production
Centre of excellence for Integrated Approaches in Chemistry and Biology of 
Proteins, Scientific Director
http://www.cipkebip.org/
Professor of Structural Biology at IPS "Jozef Stefan"
e-mail: dusan.t...@ijs.si    
phone: +386 1 477 3857       Dept. of Biochem.& Mol.& Struct. Biol.
fax:   +386 1 477 3984       Jozef Stefan Institute
                            Jamova 39, 1 000 Ljubljana,Slovenia
Skype: dusan.turk (voice over internet: www.skype.com