hi,
as a comment to zhijie's question, the biggest problem with marion bradford's 
"protein" assay is that it measures the SDS concentration too! therefore, where 
it would be most useful – that is in quantifying the total protein in an 
SDS-treated gel sample – it cannot be used! the amido-black method is far 
better, at least as far as i know. does anyone know why it's seldom used? our 
protocol is below if anyone's interested.
best
jon

Amido-Black assay.
Dilute 10µl of the clarified protein extract (also in SDS sample buffer) to 
200µl with water, add 600µl Amido-Black reagent, mix and incubate at RT for 
5min.  Centrifuge for 5min at 13krcf.  Discard the supernatant and wash the 
pellet twice in 500µl Acid-MeOH [10%(v/v) HAc in MeOH].  The final supernatant 
should be colourless.  Dissolve the pellet in 1ml 1M NaOH (warm if necessary) 
and assay at 615nm.

Amido-Black reagent
Stock = 0.13g Amido-Black 10B (Merck) + 1ml HAc + 9ml MeOH.  Stir well then 
filter.  Stable at 4°C indefinitely.  To use, dilute 1ml of this stock in 50ml 
Acid-MeOH. Stable upto 1 week at 4°C. Calibrate with a dilution series of a 
"standard" protein such as BSA.


Von: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] Im Auftrag von Alex Lee
Gesendet: Freitag, 5. Oktober 2018 05:37
An: CCP4BB@JISCMAIL.AC.UK
Betreff: Re: [ccp4bb] [Offtopic] Protein stain in Coomassie Blue dyes

Hi All,

Thanks for all of your inputs!

Alex

On Thu, Oct 4, 2018 at 7:31 PM Zhijie Li 
<zhijie...@utoronto.ca<mailto:zhijie...@utoronto.ca>> wrote:
Hi,

At high concentration (1-2%) the published saturating SDS:protein binding ratio 
is about 1.4:1 by weight, that is roughly one SDS molecule per two aa on 
average. It is dense but  not that dense to prevent any further interaction.  
More importantly, as a quite hydrophilic small molecule SDS should have no 
trouble dissociating from the peptide when its in-solution concentration drops 
(therefore you can use SDS gel bands for MS). With common procedure, during 
staining the SDS should partially fall off( especially if the gel is heated), 
and partially remain with the protein in the gel, depending on: how hydrophobic 
the protein is, how low the environmental SDS concentration becomes, how much 
organic solvent there is in the solution, etc.. The coomassie should simply 
find whatever hydrophobic/positively charged patch to bind and aggregate. 
Besides, coomassie-R is probably slightly more hydrophobic than SDS so it is 
capable of competing SDS off if necessary. (The even less hydrophobic Coomassie 
G250 definitely binds protein in the presence of detergents - that how Blue 
Native gel works for membrane proteins) Finally, since the only thing you are 
looking for is some deeper blue to indicate the presence of protein, even if 
SDS did prevent dye binding to some extent, your gel still will work. This is 
different from when you want to use the dye to do some quantitative work such 
as the Bradford assay. It would be interesting to know the effect of detergents 
on Bradford.

Zhijie



On Oct 4, 2018, at 9:26 PM, Alex Lee 
<alexlee198...@gmail.com<mailto:alexlee198...@gmail.com>> wrote:
Dear All,

I am thinking that in an SDS-PAGE experiment, if protein samples are boiled in 
SDS containing loading dye, and supposedly SDS interacts with proteins, why the 
Coomassie Blue dyes could still interact with and stain the proteins?      I am 
thinking SDS is covering the proteins, making no room for the Coomassie Blue 
dyes interaction.  I'd appreciate it if any input from this forum.

Alex

________________________________

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1

________________________________

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1

########################################################################

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1

Reply via email to