All very true Randy,

But nevertheless every hkl has an FOM assigned to it, and that is used to calculate the map.  Statistical distribution or not, the trend is that hkls with big amplitude differences get smaller FOMs, so that means large model-to-data discrepancies are down-weighted. I wonder sometimes at what point this becomes a self-fulfilling prophecy?  If you look in detail and the Fo-Fc differences in pretty much any refined structure in the PDB you will find huge outliers. Some are hundreds of sigmas, and they can go in either direction.

Take for example reflection -5,2,2 in the highest-resolution lysozyme structure in the PDB: 2vb1.  Iobs(-5,2,2) was recorded as 145.83 ± 3.62 (at 5.4 Ang) with Fcalc^2(-5,2,2) = 7264.  A 2000-sigma outlier!  What are the odds?   On the other hand, Iobs(4,-6,2) = 1611.21 ± 30.67 vs Fcalc^2(4,-6,2) = 73, which is in the opposite direction.  One can always suppose "experimental errors", but ZD sent me these images and I have looked at all the spots involved in these hkls.  I don't see anything wrong with any of them.  The average multiplicity of this data set was 7.1 and involved 3 different kappa angles, so I don't think these are "zingers" or other weird measurement problems.

I'm not just picking on 2vb1 here.  EVERY PDB entry has this problem.  Not sure where it comes from, but the FOM assigned to these huge differences is always small, so whatever is causing them won't show up in an FOM-weighted map.

Is there a way to "change up" the statistical distribution that assigns FOMs to hkls?  Or are we stuck with this systematic error?

-James Holton
MAD Scientist

On 10/4/2019 9:31 AM, Randy Read wrote:
Hi James,

I'm sure you realise this, but it's important for other readers to remember that the FOM is a statistical quantity: we have a probability distribution for the true phase, we pick one phase (the "centroid" phase that should minimise the RMS error in the density map), and then the FOM is the expected value of the phase error, obtained by taking the cosines of all possible phase differences and weighting by the probability of that phase difference.  Because it's a statistical quantity from a random distribution, you really can't expect this to agree reflection by reflection!  It's a good start to see that the overall values are good, but if you want to look more closely you have to look a groups of reflections, e.g. bins of resolution, bins of observed amplitude, bins of calculated amplitude.  However, each bin has to have enough members that the average will generally be close to the expected value.

Best wishes,

Randy Read

On 4 Oct 2019, at 16:38, James Holton <jmhol...@lbl.gov <mailto:jmhol...@lbl.gov>> wrote:

I've done a few little experiments over the years using simulated data where I know the "correct" phase, trying to see just how accurate FOMs are.  What I have found in general is that overall FOM values are fairly well correlated to overall phase error, but if you go reflection-by-reflection they are terrible.  I suppose this is because FOM estimates are rooted in amplitudes. Good agreement in amplitude gives you more confidence in the model (and therefore the phases), but if your R factor is 55% then your phases probably aren't very good either.  However, if you look at any given h,k,l those assumptions become less and less applicable.  Still, it's the only thing we've got.

2qwAt the end of the day, the phase you get out of a refinement program is the phase of the model.  All those fancy "FWT" coefficients with "m" and "D" or "FOM" weights are modifications to the amplitudes, not the phases.  The phases in your 2mFo-DFc map are identical to those of just an Fc map.  Seriously, have a look! Sometimes you will get a 180 flip to keep the sign of the amplitude positive, but that's it.  Nevertheless, the electron density of a 2mFo-DFc map is closer to the "correct" electron density than any other map.  This is quite remarkable considering that the "phase error" is the same.

This realization is what led my colleagues and I to forget about "phase error" and start looking at the error in the electron density itself (10.1073/pnas.1302823110).  We did this rather pedagogically.  Basically, pretend you did the whole experiment again, but "change up" the source of error of interest.  For example if you want to see the effect of sigma(F) then you add random noise with the same magnitude as sigma(F) to the Fs, and then re-refine the structure.  This gives you your new phases, and a new map. Do this 50 or so times and you get a pretty good idea of how any  source of error of interest propagates into your map.  There is even a little feature in coot for animating these maps, which gives a much more intuitive view of the "noise".  You can also look at variation of model parameters like the refined occupancy of a ligand, which is a good way to put an "error bar" on it.  The trick is finding the right source of error to propagate.

-James Holton
MAD Scientist


On 10/2/2019 2:47 PM, Andre LB Ambrosio wrote:
Dear all,

How is the phase error estimated for any given reflection, specifically in the context of model refinement? In terms of math I mean.

How useful is FOM in assessing the phase quality, when not for initial experimental phases?

Many thank in advance,

Andre.

------------------------------------------------------------------------

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1



------------------------------------------------------------------------

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1


------
Randy J. Read
Department of Haematology, University of Cambridge
Cambridge Institute for Medical Research     Tel: + 44 1223 336500
The Keith Peters Building                 Fax: + 44 1223 336827
Hills Road E-mail: rj...@cam.ac.uk <mailto:rj...@cam.ac.uk>
Cambridge CB2 0XY, U.K. www-structmed.cimr.cam.ac.uk <http://www-structmed.cimr.cam.ac.uk>



########################################################################

To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1

Reply via email to