Dear Nan, All,

Certainly this topic has come up several times and we never seem to quite get 
to a solution that suits everybody.

I don’t know why 1e-3 was originally chosen for use in the standard name table, 
but even if you go back to version 1 it is in there, which means it was agreed 
prior to 2006 when the CF website at PCMDI was set up.

The last time the question of salinity units was aired in detail was during the 
TEOS-10 discussions in 2011. Unfortunately, the mailing list archive seems to 
be unavailable at the moment, but I can vouch for the fact that the current 
definitions of the salinity names came from the very detailed discussions that 
we had at that time. In particular, we added the following wording to the 
definition of sea_water_practical_salinity: ‘Practical Salinity, S_P, is 
defined on the Practical Salinity Scale of 1978 (PSS-78) and is calculated from 
the electrical conductivity of sea water (as well as temperature and pressure). 
Officially S_P is dimensionless so that, while convenient, and while it is 
common practice, it is not officially sanctioned to say S_P = 35 psu. Often 
authors use PSS-78, as in S_P = 35 PSS-78. If salinity was measured using 
remote sensing techniques and not conductivity, then it is recommended that 
additional metadata (calibration/validation information) be described in the 
variable comment attribute.’

Once upon a time (back in 2009) there was a discussion about allowing CF to use 
‘psu’ as a unit in its own right, but I think the TEOS-10 discussion made it 
clear that ‘psu’ is not really a unit at all, so that idea was dropped and we 
continued to use 1e-3.

We should remember that the canonical unit of ‘1e-3’ doesn’t prevent anyone 
using ‘1’ in their files if they prefer it, and vice versa. As Jim has already 
pointed out, UDunits can certainly cope with that. So in one sense, it doesn’t 
really matter to CF which we choose as the canonical unit as long as we can 
agree and, most importantly, make the definition really really clear so that 
consumers of the files know how to interpret the data.

In previous discussions there has never been unanimous agreement about whether 
it is better to use ‘1e-3’ or ‘1’. My niggling concern about changing the unit 
after all these years is whether it will lead to misinterpretation of existing 
data files. Is that going to be a problem?  We have in the past changed the 
canonical units of standard names, but only to correct outright errors, rather 
than to change the interpretation of a name. How big a problem is it for the 
oceanographic community if we don’t change the unit?

If we do decide to go with ‘1’ as the canonical unit, is there a reference, 
such as TEOS-10, which we can use to support our decision? It would be useful 
to include it in the definition and hopefully reduce the need to keep 
revisiting this same question.

Either way, I think we can improve further on the definition to help people 
better understand the data.

Best wishes,
Alison

------
Alison Pamment                                 Tel: +44 1235 778065
NCAS/Centre for Environmental Data Archival    Email: 
alison.pamm...@stfc.ac.uk<mailto:j.a.pamm...@rl.ac.uk>
STFC Rutherford Appleton Laboratory
R25, 2.22
Harwell Oxford, Didcot, OX11 0QX, U.K.


From: CF-metadata [mailto:cf-metadata-boun...@cgd.ucar.edu] On Behalf Of Nan 
Galbraith
Sent: 27 May 2015 15:45
To: Rich Pawlowicz
Cc: cf-metadata@cgd.ucar.edu
Subject: Re: [CF-metadata] Salinity units

Hi all -

The '.001' units for P.S. doesn't mean that stored values of practical
salinity differs from A.S. by 'a factor of around a 1000', as far as I
know. If that's the logical inference, then this unit is really a problem,
and maybe we should do something about it.

I wish my CF email archive went back a little further, because there's
nothing (since 2004) that I can find that explains the rationale for
this unit. It certainly *looks* like a compromise between a unit for a
non-dimensional variable and PPT ... When this was originally under
discussion, way back when, I'll bet someone argued that it would eventually
be a big problem.  I'd really love to see that email thread!

Cheers -
Nan


On 5/26/15 11:52 AM, Rich Pawlowicz wrote:

I’m not sure what the best answer is either, but I think the “correct” way is
to have people deal with Practical Salinity in some special fashion in
their workflow, because it *is* defined in a weird way that is generally
incompatible with the general idea of ‘quantities with units’) - getting
a salinity definition that is aligned with the way all other quantities
in the world are defined was one of the motivating factors behind TEOS-10!

So, essentially people would have to make their own choice about
what to do with ‘practical salinity’ for whatever they are doing.

I will point out, though, that having two kinds of data that differ
numerically by a factor of around a 1000 is a good way of getting
them to realize that they really are not exactly compatible - you
wouldn’t *want* Practical Salinity and Absolute salinity on the same
plot (“look - salinity increased by 0.16 g/kg everywhere in 2010!”)

But I understand that one might want to make this as painless as
possible.



On May 26, 2015, at 8:48 AM, Signell, Richard 
<rsign...@usgs.gov<mailto:rsign...@usgs.gov>> wrote:


Rich,
Thanks for this.   Yes, I guess my concern is that folks will do a
catalog search for *salinity* variables, and with a few spot checks,
see that they are have data values in the range of 29-36 or so, and
then go ahead and run a workflow that converts all units using the
units attribute.   And if "practical salinity" has units of "1" and
"absolute salinity" has units of "g/kg" = "0.001", then the data might
not appear on that fixed y-axis plot with [29 36].     But I don't
have a good alternative.   I guess we have to rely that people will
realize from the standard_names that for comparison, you need to
estimate absolute salinity from practical salinity using tools like
GSW toolbox.

-Rich

On Fri, May 22, 2015 at 3:59 PM, Rich Pawlowicz 
<rpawlow...@eos.ubc.ca<mailto:rpawlow...@eos.ubc.ca>> wrote:


Ummm…I’m not entirely what you are asking, but

a) PSS-78 Practical Salinity is a dimensionless number.  It was defined
such that "the numerical values of practical salinity would be similar to the
numerical values of previous salinity data, when expressed in ‰”, but
it isn’t in fact ppt or anything, and you shouldn’t be multiplying it up or
down by factors of 1000.

b) "Previous salinity data”, (Cox or Knudsen salinity) which
was obtained from titrations, does in fact represent a
mass fraction of something (because you are titrating
with a mass of silver). This was denoted by the ppt ‘unit'.

c) TEOS-10 Absolute Salinity is also a mass fraction (of dissolved solute
on the Reference Composition Salinity Scale). However, nowadays the
SI brochure suggests that different quantities should be distinguished
by their symbols, not their units.  So, there isn’t actually a recommended
unit for Absolute Salinity. You can write

S_A = 35 g/kg = 0.035 kg/kg = 35000 mg/kg

or, again using SI rules and treating the units as a ‘thing’:

S_A/(g/kg) = 35

and any of these are valid - the same way lengths can be in
meters or km or mm or whatever is handy (this is also
true for preformed salinity).

‘ppt’ is discouraged as a unit of mass fraction because (for example) it
could be confused with ‘part per trillion’


Now, the gsw toolbox assumes ‘g/kg’ for its TEOS-10 salinity inputs
and outputs, but YOU don’t have to do that if you don’t want to.

I admit it is a little magic how we can ESTIMATE Absolute Salinity (with
units) from Practical Salinity (without units), but keep in mind that this
is only ONE possible way of estimating Absolute Salinity, and in fact it is
a method that is metrologically somewhat suspect because of the
definition of PSS-78. S_A could also be obtained from density
measurements, for example - and then there is some other
conversion factor involving different units.

Rich.


On May 22, 2015, at 1:01 PM, Signell, Richard 
<rsign...@usgs.gov<mailto:rsign...@usgs.gov>> wrote:


Roy,

For sure dimensionless.  But "1.0", "0.001" or "g/kg"?

The latest version (27) of the CF Standard Name list
(http://cfconventions.org/Data/cf-standard-names/27/build/cf-standard-name-table.html)
states:

sea_water_salinity: "0.001"
sea_water_absolute_salinity: "g kg-1"
sea_water_practical_salinity:    "0.001"
sea_water_preformed_salinity:  "g kg-1"
sea_water_cox_salinity: "0.001"

and units packages, of course, would treat "g kg-1" the same as "0.001".

Yet in the IOC manual on equation of seawater:
http://www.teos-10.org/pubs/TEOS-10_Manual.pdf
it states (PDF page 176, printed page 166) that Practical Salinity
should have units of "1", while "Absolute Salinity" (the argument used
in the toolbox functions) and "Preformed Salinity" (used in numerical
ocean models) should have units "g kg-1".

So it appears that TEOS agrees with CF on units for Absolute Salinity
and Preformed Salinity, but not on Practical Salinity.

And OceanSites (as least here:
http://www.oceansites.org/docs/OS_PAP-3_201205_P_deepTS.txt)
is using "sea_water_practical_salinity" with units of "1", so they are
consistent with the TEOS publication, but not the current CF
convention (v27).

On the TEOS site, there is software to calculate Absolute Salinity
from Practical Salinity.   So it would seem that the technically
correct thing to do would be to use the "gsw_SA_from_SP" routine to
convert OceanSites Practical Salinity (in units of "1") to Absolute
Salinity (in units of "g/kg") before comparing with the "Preformed
Salinity" output "g/kg" from ocean models.

I'm pretty confused though, so I'm cc'ing Rich Pawlowicz on this,
hoping for his input.

Thanks,
-Rich



On Fri, May 22, 2015 at 1:49 PM, Lowry, Roy K. 
<r...@bodc.ac.uk><mailto:r...@bodc.ac.uk> wrote:

Dimensionless. Please????

This is the view of physical oceanographers for whom I have the greatest 
respect.

Cheers, Roy.
________________________________________
From: Reyna Jenkyns [re...@uvic.ca<mailto:re...@uvic.ca>]
Sent: 22 May 2015 18:06
To: cf-metadata@cgd.ucar.edu<mailto:cf-metadata@cgd.ucar.edu>; OceanSITES Data 
Management Team; Nan Galbraith
Subject: Re: [CF-metadata] Salinity units

I'm interested in this topic since I didn't realize what had been discussed 
previously, and now I think we must be non-compliant as well.  Is this 
documented formally in the CF documentation?

Reyna Jenkyns | Data Stewardship Team Lead - Digital Infrastructure
Ocean Networks Canada | T 250 853 3908 | oceannetworks.ca
University of Victoria PO Box 1700 STN CSC 2300 McKenzie Avenue Victoria, BC 
V8W 2Y2

________________________________________
From: CF-metadata 
<cf-metadata-boun...@cgd.ucar.edu><mailto:cf-metadata-boun...@cgd.ucar.edu> on 
behalf of Nan Galbraith <ngalbra...@whoi.edu><mailto:ngalbra...@whoi.edu>
Sent: Friday, May 22, 2015 10:03 AM
To: cf-metadata@cgd.ucar.edu<mailto:cf-metadata@cgd.ucar.edu>; OceanSITES Data 
Management Team
Subject: Re: [CF-metadata] Salinity units

Hello all -

It's been a long time, but is anyone interested in re-visiting the subject
of units for practical salinity in CF?

I was recently notified that my salinity data was likely to be
overlooked by
some users, because I'd used '1' as the units, not '.001'. Somehow, I'd
forgotten the (long-ago) discussion on the CF list about salinity units.

Some members of  the OceanSITES project are interested in revising our
format spec to encourage the use of '1' as an indication that salinity does
not have units - but, of course, we'd mostly rather remain CF-compliant.

Thanks for any feedback on this.

Cheers - Nan


On 6/17/09 2:48 AM, Lowry, Roy K wrote:


Dear All,

During an exercise with Alison mapping the CF Standard Names to a
units vocabulary in the BODC vocabulary server I noticed that the
units for salinity were '1.00E-03', i.e. parts per thousand. My
understanding in that since the introduction of the Practical
Salinity Scale that salinity is dimensionless with units of '1'.  Is
there agreement for our changing the units in the Standard Name
table?

Cheers, Roy.




--

*******************************************************

* Nan Galbraith        Information Systems Specialist *

* Upper Ocean Processes Group            Mail Stop 29 *

* Woods Hole Oceanographic Institution                *

* Woods Hole, MA 02543                 (508) 289-2444 *

*******************************************************




_______________________________________________
CF-metadata mailing list
CF-metadata@cgd.ucar.edu
http://mailman.cgd.ucar.edu/mailman/listinfo/cf-metadata

Reply via email to