OK... so NONE of that worked...  So, I reinstalled everything and checked it 3 times...

Here is something....

What should the voltage be on Pin #1 of P95 (XV12DC) be with respect to chassis?

I am reading 13.75 VDC

Confirming what the picture APPEARS to be in the manual for the connection of the 1-wire, 3-pin connector to the 3-pin 12V power IN on the 2 meter module:

As you are looking at the open top of the K3s from the front, over the front panel....

Does the connector plug in such that the 1-wire is on "Pin #3" on the 2 meter module (IAW, the pin closest to the PA shield/rear of the chassis?

I took the PCB out of the 2 meter module enclosure and inspected the pins and traces on the PCB....

Pin #1 trace appears to be a (+) line which is bypassed to ground with a n SMD device...  believe it was stenciled as a cap (input filtering of some sort)?  (Pin #1 to Pin #3 is approx. 0.2 Ohms)

Pin #2 solders to the ground plane on the bottom of the PCB. (0.1 Ohms or so between Pin #2 and enclosure; Open from Pin#2 to Pin #3)

Pin #3 appears to be the unfiltered (+) as it leads to a rather large trace on the bottom of the board leading from the connector to the interior of the board.

So the power wire from P95/Pin #1 goes eother to Pin #1 or Pin #3 on the 2 meter module 12V IN.  The picture in the manual APPEARS to be Pin #3 and it follows, as the connector has Pin #1 marked and Pin #1 is traditionally on the left side of a PC Board header.

I'm running out of ideas...

73,

______________________
Clay Autery, KY5G
(318) 518-1389

On 01/11/20 23:22, Clay Autery wrote:

Update...

1) Strictly speaking, removing the RefLock option board and associated SINGLE TMP cable did not "fix" anything... Still have the "HIGH CUR" error. 2) Incidentally, IF you fire the K3s up on ANY other band (after disco and reconnecting power to clear error) WITH the KXUSB cable connected to the computer AND the K144XV "RS-232" port, and THEN switch from any other band TO 2 meters, the status light flashes once as expected (ready for firmware). BUT, the K144XV Utility will NOT connect to the K144XV.  Then, of course, if you shut the K3s down while on 2M band and remove the KXUSB cable from its port, then power back up....

a) Powers up, status light flashes once or twice as expected.
b) the K3s shows ANT3 briefly, and then starts trying to display the power-on banner, but is interrupted repeatedly by the HIGH CUR error.... which remains after the banner scroll completes.
c) You have to power down, and disco/reconnect the power to reset.

I pulled the left side of the case off ans inspected the IF IN and IF out cables to verify that they are connected correctly AND still fully plugged, and had NOT been pinched off.

I think I will try again, with the ANT 3 cable disconnected to remove the BNC cable assembly as a possible cause....  I'm reaching here now.

_*Question:*_  Is there any chance that having the "grip" portion of ANY of the TMP connectors making contact with the K144XV enclosure a "problem"?  The ANT3 TNP board connector is set at a slightly OFF angle, and the TMP outer (coax shiield) might be touching the enclosure.  Just trying t figure out what is "wrong", and if it anything I have control over...

73,

______________________
Clay Autery, KY5G
(318) 518-1389
On 01/11/20 22:01, Clay Autery wrote:
Just what the title says....  I installed a new K144XV and Reference Oscillator Phase Lock Option tonight.

When I hit the band switch to go up to the 2 meter band, it thru a HIGH CURR error.

I followed the directions in both install manuals precisely (as appropriate for a K3s as enhanced), and double checked BOTH sets of steps after getting the error. IF I disco the power, wait, reconnect, and then power up (provided I switch bands out of 2 meters), the error goes away. But if I switch back to 2 meters, it throws the error again.

I am going to go in and remove the REFLock option and try to just get the K144XV to "work". The status light flashes just as described in the manual for "normal operation".  (once or twice on power up, then OFF) I've verified the supply voltage and current are "normal"  14.2 VDC at 1.47 amps (higher because the subRX was installed a couple days ago.

Any ideas as to what is causing this "issue"? I'm missing all my weekend nets.  <wink>

73,

_._,_._,_
------------------------------------------------------------------------
Groups.io Links:

You receive all messages sent to this group.

View/Reply Online (#31816) <https://groups.io/g/Elecraft-K3/message/31816> | Reply To Group <mailto:[email protected]?subject=Re:%20Re%3A%20%5BElecraft-K3%5D%20%5BElecraft%5D%20K3s%20throwing%20a%20%22HIGH%20CURR%22%20error%20after%20installing%20K144XV%20and%20Reference%20Phase%20Lock%20Option> | Reply To Sender <mailto:[email protected]?subject=Private:%20Re:%20Re%3A%20%5BElecraft-K3%5D%20%5BElecraft%5D%20K3s%20throwing%20a%20%22HIGH%20CURR%22%20error%20after%20installing%20K144XV%20and%20Reference%20Phase%20Lock%20Option> | Mute This Topic <https://groups.io/mt/69637695/1171784> | New Topic <https://groups.io/g/Elecraft-K3/post>

Your Subscription <https://groups.io/g/Elecraft-K3/editsub/1171784> | Contact Group Owner <mailto:[email protected]> | Unsubscribe <https://groups.io/g/Elecraft-K3/leave/4370919/558102286/xyzzy> [[email protected]]

_._,_._,_
______________________________________________________________
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:[email protected]

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html
Message delivered to [email protected] 

Reply via email to