I suggest you reference IEC 1010, Safety Requirements for Electrical 
Equipment for Measurement, Control, and Laboratory Use.  Clause 12.5.1 
indicates that equipment shall not produce a sound pressure level above 85 
dbA when measured both at the operator's position and at 1 meter from the 
position with the highest sound pressure.

I have been away from IEC 601.1 (Medical Electrical Equipment) for awhile, 
but my last recollection is that requirements were under consideration only 
and currently not included in the standard.

Hope this helps.