Bob,

I have added my view of what the requirements are of IEC 950, 2nd ed, and A3.

Brought to you by Johnson's Vax wrote:
> 
> I worked out what looks like the standards requirements in IEC 950 for
> telecom spacings. Do you see any obvious errors in what I did?
> The result is spacings which seem excessive to me.
> 
> Bob Johnson
> 
>               DETERMINATION OF SPACINGS IN TELECOM CIRCUITS
> 
> 7 from 6.2.1.1b, the maximum signal voltages may be up to 70.7 Vac or 120 Vdc.
> 7 from M.3.1.2, the maximum ringing voltage may be 300 Vpp or 200 Vp-to-earth
>   (M.2 for European ringing specifies only currents and cannot be used to
>   derive operating voltages.)
> 7 2.2.7.1 requires the maximum voltage difference be considered.
>   (2.2.7.2 allows SELV voltages to be consided zero with respect to primary,
>   but not TNV)
> 7 2.2.7.4 states ringing signals can be ignored for creepage distances
>   (but not for clearances)
> 
> Assuming pollution degree 2, overvoltage category II, material group III,
> without the routine testing option.
> 
> CLEARANCES
> ----------

Based on your assumptions I see the following being required:
 
CIRCUIT                 TRANS.  CLEARANCE       OPERAT. BASIC   REINF.INS.
mains 120 (= 150) Vac   1,5 kV  Tables 3 + 4    0,7 mm  1,7 mm  2,4 mm
mains 230 (= 300) Vac   2,5 kV  Table 3         1,4 mm  2,0 mm  4,0 mm
TNV network (150)       1,5 kV  Table 5         0,7 mm  1,0 mm  2,0 mm
TNV network (300)       1,5 kV  Table 5         1,6 mm  1,9 mm  3,8 mm

A TNV circuit is a secondary circuit (1.2.8.8), Table 3 applies for primary to
secondary, and Table 5 for TNV to TNV.
Note: 2.2.6 requires only operational insulation between TNV and another TNV
circuit.
The 300 Vpp ringing signal is above the 120 V mains peak voltage, therefore,
additional 0,2 mm need to be considered here, but is below the 230 V mains peak
voltage.

> Mains to TNV:
>         peak difference
>         120Vac = 170Vp to earth, 230Vac = 325Vp to earth
>         ringing signal is 200Vp to earth
>                          170 Vp + 200 Vp = 370 Vp
>                          325 Vp + 200 Vp = 525 Vp
>         these are normal repetitive peak voltages (requiring table 4)
> 
>                 mains           120 Vac 230 Vac
>                 Table 3         2.0 mm  3.2 mm
>                 Table 4         0.2 mm  0.2 mm
>         Basic insulation        2.2 mm  3.4 mm
>         Reinforced insulation   4.4 mm  6.8 mm
> TNV to TNV
>         peak difference between two (independent) ringing signals
>         (respect to earth) is +200 Vp to -200 Vp = 400 Vp;
>         from table 5:
> 
>         Basic insulation        1.9 mm
>         Reinforced insulation   3.8 mm
> 
> CREEPAGE
> --------
> Mains to TNV:
> 
>         rms difference  120 Vrms + 120 Vdc = 170 Vrms
>                         230 Vrms + 120 Vdc = 260 Vrms
>         from table 6:
> 
>                 mains           120 Vac 230 Vac
>         Basic insulation        2.0 mm  3.2 mm
>         Reinforced insulation   4.0 mm  6.4 mm
> 
> TNV to TNV
>         dc difference   +120 Vdc to -120 Vdc = 240 V dc
> 
>         Basic insulation        2.5 mm
>         Reinforced insulation   5.0 mm
> 

When using the interpolation option (Condition 7 to Table 6), the spacings may 
be
slightly reduced as follows:

 Mains to TNV:
                mains           120 Vac 230 Vac
        Basic insulation        1,8 mm  2,7 mm
        Reinforced insulation   3,6 mm  5,4 mm

 TNV to TNV:
        Basic insulation        2,4 mm
        Reinforced insulation   4,8 mm

-- 
Kind regards, Horst Dierich, Germany
EMAIL: [email protected]

Reply via email to