Martin Hi!

Competitor's product is Class I.
> 
> 
> From:  Martin Ansdell@MITEL on 03/19/99 09:20 AM
> Hello,
> 
> This is the first time that I've replied to TREG, so I hope all goes well.
> 
> UL requires two levels of protection.
> 
> Class I equipment uses protective earth as one level, so the protective
> earthing terminal has to be permanent and hence the "stud on the chasis"
> method.
> 
> Class II equipment satisfies two levels of protection without reliance on
> protective earth.  Earth is only present for operational requirements.
> 
> Could you confirm whether your competitor's equipment is Class I or Class
> II.  The class makes a significant difference in the earthing requirements.
> 
> Regards,
> Martin.
> 
> 
> 
> 
> From: pe...@itl.co.il   on   03/18/99 11:47 PM ZE2
> 
> Please respond to t...@world.std.com
> 
> To:   emc-p...@majordomo.ieee.org, t...@world.std.com
> cc:    (bcc: Martin Ansdell/Kan/Mitel)
> 
> Subject:  Protective Earthing Terminal
> 
> 
> 


PETER S. MERGUERIAN