Peter,

Please let me have the name of the surgeon with the cellphone, so I'll be
careful to avoid him!!


Best Wishes,

Jon


On 21 Feb 2003 at 15:17, peter merguerian wrote:

> 
> Jon,
> I must say that I somewhat disagree with you. A medical equipment as defined
in IEC60601-1 is one 
> that is for "professional use. The surgeon's e-mails PC can also be used by
his son and I do not 
> consider it "professional" in the sense of the standard. 
> Let's not exagerate here! A person with too many medical devices should stay
in bed.
> The other day, I called a surgeon and he happened to be in the operating
room with his cellphone 
> performing an operation. Does that make his cellular comply with IEC 601-1?
> Peter
> 
> Jon Griver <jgri...@601help.com> wrote: 
>     
>     Delphina,
>     
>     It's not only an issue of intended use, but also a question of location.
With regard to the example 
>     you give, if a PC for the surgeon's urgent emails is located within the
'patient vicinity', as 
>     defined in IEC 60601-1-1 (within about 6 feet of the patient's bed),
then I believe that it should 
>     comply with IEC 60601-1, even though it's not a medical device.
>     
>     The reasoning behind this is that the patient could possibly touch the
PC (that's the logic behind 
>     the 'patient vicinity' concept), and close a leakage current loop,
together with other medical 
>     instruments with which he is in contact. This could be particularly
nasty if there is a fault in 
>     the PC's earth wire.
>     
>     Regards,
>     
>     Jon Griver
>     http://www.601help.com
>     The Medical Device Developers Guide to IEC 60601-1
>     
>     
>     On 20 Feb 2003 at 12:38, Han, Delphina wrote:
>     
>     &! gt; 
>     > Hi all,
>     > 
>     > Thanks for your responses. When I originally asked the question, I was
sure
>     > that the particular equipment I was asking about does not fall under
the
>     > scope IEC60601, but I wanted to confirm if IEC61010 is the relevant
>     > standard. As you may notice, I have a copy of the IEC60601 standard
but not
>     > IEC61010! I also needed to find out if there were special requirements
for
>     > isolation transformers in IEC61010. 
>     > 
>     > >From the discussion, it seems that which standard to comply with is
>     > dependent on the intended use, and not where the equipment is placed.
So, if
>     > we put a PC in the hospital operating room, but it is just for the
surgeon
>     > to let's say, check email (not practical...but this is just an
example),
>     > would 60950 be the relevant standard? I doubt it'd have to meet 60601
>     > requirements.
>     > 
>     > How about EMC standards? (maybe I should start a ne! w thread...).
Will this
>     > PC have to meet IEC 60601-1-2! for medical? Or EN 55022 for ITE? Bear
in mind
>     > this is in the hospital operating room. I am familiar with medical
equipment
>     > requirements but not so much with ITE, so forgive me if I'm making
mistakes
>     > here.
>     > 
>     > My thoughts are all from the regulatory point of view. Of course if the
>     > customer wants compliance with a more stringent requirement, that is
what he
>     > gets...
>     > 
>     > I appreciate all your thoughts/comments!
>     > 
>     > -Delphina
>     > 
>     > 
>     > 
>     > 
>     > >
>     > > At 09:10 AM 2/19/2003 -0800, Han, Delphina wrote:
>     > >
>     > > >Hi
>     > > >
>     > > >I am trying to find out safety requirements for devices that
control and
>     > > >monitor equipment in a hospital lab (used for pathology). Does it
fall
>     > under
>     > > >the IEC 61010 standard? If so, are there any requirements for use of
>     > > >isolation transformers in! that standard?
>     > > >
>     > > >Thanks in advance for your response!
>     > > >
>     > > >-Delphina
>     > > >
>     > > >-------------------------------------------
>     > > >This message is from the IEEE EMC Society Product Safety
>     > > >Technical Committee emc-pstc discussion list.
>     > > >
>     > > >Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/
>     > > >
>     > > >To cancel your subscription, send mail to:
>     > > > majord...@ieee.org
>     > > >with the single line:
>     > > > unsubscribe emc-pstc
>     > > >
>     > > >For help, send mail to the list administrators:
>     > > > Ron Pickard: emc-p...@hypercom.com
>     > > > Dave Heald: davehe...@attbi.com
>     > > >
>     > > >For policy questions, send mail to:
>     > > > Richard Nute: ri...@ieee.org
>     > > > Jim Bacher: j.bac...@ieee.or! g
>     > > >
>     > > >Archive is being moved, ! we will announce when it is back on-line.
>     > > >All emc-pstc postings are archived and searchable on the web at:
>     > > > http://www.ieeecommunities.org/emc-pstc
>     > 
>     > 
>     > -------------------------------------------
>     > This message is from the IEEE EMC Society Product Safety
>     > Technical Committee emc-pstc discussion list.
>     > 
>     > Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/
>     > 
>     > To cancel your subscription, send mail to:
>     > majord...@ieee.org
>     > with the single line:
>     > unsubscribe emc-pstc
>     > 
>     > For help, send mail to the list administrators:
>     > Ron Pickard: emc-p...@hypercom.com
>     > Dave Heald: davehe...@attbi.com
>     > 
>     > For policy questions, send mail to:
>     > Richard Nute: ri...@ieee.org
>     > Jim Bacher: j.bac...@ieee.org
>     > 
>     > Archive is being moved, we will announce when it is back on-line.
>     > All emc-pstc postings are ! archived and searchable on the web at:
>     > http://www.ieeecommunities.org/emc-pstc
>     > 
>     
>     
>     
>     -------------------------------------------
>     This message is from the IEEE EMC Society Product Safety
>     Technical Committee emc-pstc discussion list.
>     
>     Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/
>     
>     To cancel your subscription, send mail to:
>     majord...@ieee.org
>     with the single line:
>     unsubscribe emc-pstc
>     
>     For help, send mail to the list administrators:
>     Ron Pickard: emc-p...@hypercom.com
>     Dave Heald: davehe...@attbi.com
>     
>     For policy questions, send mail to:
>     Richard Nute: ri...@ieee.org
>     Jim Bacher: j.bac...@ieee.org
>     
>     Archive is being moved, we will announce when it is back on-line.
>     All emc-pstc postings are archived and searchable on the web at:
>     http://www.ieeecommunities.org/emc-pstc
> 
> 
> Do you Yahoo!?
> Yahoo! Tax Center - forms, calculators, tips, and more 




This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     majord...@ieee.org
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              emc-p...@hypercom.com
     Dave Heald:               davehe...@attbi.com

For policy questions, send mail to:
     Richard Nute:           ri...@ieee.org
     Jim Bacher:             j.bac...@ieee.org

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to