Mr Poore 

While you are correct in the intent and meaning of the standard; my point was,
that if the customer specs product compliance with standard xyz, than the
requirements of xyz becomes your design and test target. I have lost count of
the number of times that a customer has spec'd 601-1 or whatever, when either
x950 or x1010-1 is applicable for the intended end use. What the customer
wants, the customer gets (if he pays)...

Brian 



There are no requirements for using an isolation transformer as long as you
can meet the requirements for "touch" current (leakage current) without one.
These requirements are similiar to 950, and not as stringent as the medical
requirement.

 Chris Poore 
  
-----Original Message----- 
From: owner-emc-p...@majordomo.ieee.org 
Sent: Wednesday, February 19, 2003 11:42 AM 
To: emc-p...@majordomo.ieee.org 
Subject: RE: IEC 61010 requirements 

If end-use wiil never see contact w/patient or test subject, and equipment can
be classified as Electrical Test and/or Measurement equipment, and does NOT
fall under the scope of the Med Directive, and can be scoped under the LVD,
then 61010-1 is applicable standard.

Of course, the customer spec can require compliance to the Med Dev Dir. 

Brian 

-----Original Message----- 
From: Han, Delphina 
Sent: Wednesday, February 19, 2003 9:10 AM 
To: emc-p...@majordomo.ieee.org 
Subject: IEC 61010 requirements 

Hi 

I am trying to find out safety requirements for devices that control and 
monitor equipment in a hospital lab (used for pathology). Does it fall under 
the IEC 61010 standard? If so, are there any requirements for use of 
isolation transformers in that standard? 
  