I have been following this thread, and read 
some more after Googling around. A copy of the
Nancy Leveson IEEE report is available on the net.

http://courses.cs.vt.edu/~cs3604/lib/Therac_25/Therac_1.html

It is very complete, and written in terms of
Understatement......

A few remarks:

The single fact that even one patient complained
About a burning pain during radiation exposure should,
no MUST have been enough to ring all available
alarm bells. If one reads to what extend
safety measures were ignored, one can say only
one thing: this company has acted amoral.
(even in 1985) Hey this was not prehistoric !!!

Even from a medical/functional point of view:
Giving hazardous treatments to patient without 
a operator/treatment log available on paper
(excuse: no memory in a PDP-11 ???? )
is asking for errors.  The absence
of this log made it impossible to trace down
what exactly went wrong. 


The type of software bugs (that are discussed in detail
in this report) are those made by programming beginners.
Even In 1985. I have been programming in 1984 on
modem technology and clearly remember how
magazines like Byte discussed programming 
technologies on microprocessors, and discussions
about testing software were already common stuff.
I personally made use of a Byte published text editor
In 1982 (assembler), and of a multitasking kernel for
the 8080 series. The article discussed the problems
that go with parallel processing  and offered several 
mechanisms for the problems like race conditions
that deal with that, such as semaphores.

The technique of incrementing a byte as a flag
for setting it, and relying on not being zero 
for a fault condition goes wrong every 256
settings. How stupid can one be !!!!
It's a beginners fault one makes once only!

I think that article is still very nice to
the manufacturer of this Therac 25.







Gert Gremmen
Ce-test, qualified testing bv



Van: emc-p...@ieee.org [mailto:emc-p...@ieee.org] Namens John Woodgate
Verzonden: zaterdag 10 mei 2008 7:46
Aan: emc-p...@ieee.org
Onderwerp: Re: Ethics grad work Therac-25

In message 
<4bea4e50d3d4344da9d84367ae31706402195...@dcexvs02.tennant.tco.corp>, 
dated Fri, 9 May 2008, "Bender, Curtis" <curtis.ben...@tennantco.com> 
writes:

>What EXACTLY are the manufacturer's root ethical issue(s) here? I have 
>the following so far:
>
>0.) doing a software safety analysis AFTER the machine has been 
>commercially available - what kind of motivation would the manufacturer

>have to pass!!

Even from the 'amoral' business viewpoint, the motivation is that the 
consequences of doing it might be more favourable than those of not 
doing it - as proved to be so.

>1.) Releasing software that that had an insufficient safety analysis - 
>was fault tree bad or was that accepted practice in 1983?

Not enough was generally known about software safety at that time. 
Probably even less about how to check for safety.

>   if 1.) is not the root, then:

>2.) Inadequate response to 6 injuries/overdoses in 9 months. this was a

>million dollar machine but hey, what is the value of a human life?

It's not clear 'who knew what, when?'. Was there enough evidence that at

least some of the accidents were not due to operator error, for example.

>3.) Knowingly leaving an unsafe product on the market for 10 months 
>until units were required to be shutdown by the FDA  & CRPB. same 
>comment as above.

'Knowingly' is a big word. Did anybody actually 'know'? Or even 
'suspect'?
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Murphy's Law has now been officially re-named The Certainty Principle
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           emcp...@ptcnh.net
     Mike Cantwell           mcantw...@ieee.org

For policy questions, send mail to:

     Jim Bacher:             j.bac...@ieee.org
     David Heald:            emc-p...@daveheald.com

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           emcp...@ptcnh.net