Pat,

Thank you for that 'heads up'.
400 A/m is still around 1/2 mT, not the 2 mT I remember.  

Interesting reference will now appear in the 'design'
documentation ...reference spec: 'internet' 

I'm just trying to make certain that when the documents are
all here and we review my design versus specs, I won't have
to start over.  So far all is ok.

Thanks again.  

Regards,
Robert

On Tue, 2 Dec 2008 08:57:54 -0800
 pat.law...@slpower.com wrote:
> Hi Robert,
> 
> 3 A/m is indeed the test level called out in IEC
> 60601-1-2, the standard 
> used for general medical EMC testing.
> 
> However, there are many other medical safety standards in
> the IEC 
> 60601-2-* series which may have different test levels.
>  My favorite is IEC 
> 60601-2-24:1998, which calls out a test level of 400 A/m
> !   (Note -- I 
> haven't performed that test, and have no equipment or
> setup 
> recommendations.)
> 
> So long as you put in your product spec 'Reference
> specification for 
> power-frequency magnetic field testing: The Internet', I
> think you're 
> covered.   ; )
> 
> Pat Lawler
> EMC Engineer
> SL Power Electronics Corp.
> 
> emc-p...@ieee.org wrote on 12/01/2008 11:14:27 PM:
> > Brodie,
> > Thank you very much for your informative reply.
> > 3 A/m translates to 3.8 microTeslas, correct?
> > 10 A/m would then be 12.6 uT
> >
> > That is indeed benign.  For some reason 2 mT had stuck
> in
> > my mind and that would have been hard to deal with,
> > requiring a block diagram change during prototype
> design
> > phase.
> 
> > Thank you again.  Your answer has allowed me to jump
> ahead
> > on the Project as we wait for Purchasing to complete
> the
> > spec purchases.
> 
> > Regards,
> > Robert
> 
> > 
> > On Mon, 1 Dec 2008 14:59:16 -0600
> > "Brodie Pedersen" <brod...@nonin.com> wrote:
> > > EN 60601-1 points to IEC 60601-1-2 for EMC test
> > > requirements.  If you
> > > are making a device for Europe the frequency required
> at
> > > 3 A/m is 50 Hz.
> > > If you want to sell the device in the US it is at
> 60Hz as
> > > well following
> > > the method of IEC 61000-4-8, power frequency magnetic
> > > fields.  If the
> > > device is life supporting, then the test level is 10
> A/m.
> > >  These are
> > > very benign levels to pass unless you have a hall
> effect
> > > sensor built
> > > into your device.
> > >
> > > Brodie Pedersen Quality Manager Product Development
> > > Nonin Medical Inc.
> > >
> > >
> > > -----Original Message-----
> > > From: m...@california.com
> [mailto:m...@california.com]
> > > Sent: Monday, December 01, 2008 2:53 PM
> > > To: John Woodgate; emc-p...@ieee.org
> > > Subject: Re: EN60601-1 Susceptibility to magnetic
> fields
> > > from AC mains
> > >
> > > Going to, but need to design NOW, or at least make
> > > certain
> > > I'm not painted into a corner later.
> > >
> > >
> > > On Mon, 1 Dec 2008 19:35:37 +0000
> > >  John Woodgate <j...@jmwa.demon.co.uk> wrote:
> > > > In message <web-197812...@california.com>, dated
> Mon, 1
> > > > Dec 2008, m...@california.com writes:
> > > >
> > > >
> > > > >What is the requirement for susceptibility to
> magnetic
> > > > fields from the AC mains as listed in EN60601-1?
> > > >
> > > > It's really dangerous to rely on answers to such
> > > > questions, even from this erudite community. We
> can't
> > > > tell you everything that the standard says, and we
> > > don't
> > > > know what is relevant for your product. You DO need
> to
> > > > buy the standard and study it.
> > > > --
> > > > OOO - Own Opinions Only. Try www.jmwa.demon.co.uk
> and
> > > > www.isce.org.uk
> > > > Either we are causing global warming, in which case
> we
> > > > may be able to stop it,
> > > > or natural variation is causing it, and we probably
> > > can't
> > > > stop it. You choose!
> > > > John Woodgate, J M Woodgate and Associates,
> Rayleigh,
> > > > Essex UK
> > > >
> > > > -
> > > >
> > >
> >

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <emcp...@ptcnh.net>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>


Reply via email to