I do have one story from the trenches, involving a laptop, ESD and
"non-standard testing".  

...back in the day, at a previous company...  

we had a customer complaint about a mysterious lockup.  There is a
side-story about sales, management, pressure, schedule, cost and all
that noise.  But sparing that, the lockup could only be duplicated by
floating the laptop and discharging to the MODEM connector several
times.  No bleed-off between zaps.  Eventually a secondary discharge was
heard inside the laptop.  

Apparently the MODEM circuit has some floating circuits due to TELCO
hi-pot requirements.  The circuit did fine under TELCO testing, which I
recall went to 3kV.  But if the stored charge in the MODEM circuit got
to about 7 kV, then there was a secondary arc to another circuit and
havoc ensued.

It was an interesting lesson for me about the (negative)virtues of
isolated circuits inside an apparatus.  As far as I know, this type of
test was never adopted by any standards body, but it became a regular
part of my internal engineering testing.  


//
Patrick 


-----Original Message-----
From: Brian Oconnell [mailto:oconne...@tamuracorp.com] 
Sent: Tuesday, April 24, 2012 6:05 PM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] Safety cost and ESD

For automotive stuff (ISO10605), I have to be careful about bleeding
charge
after each iteration because test level = 25kV. I use a 470k Resistor
attached in series with a short braid that is screwed into the ref
plane,
and touch the UUT. The strap is used for discharge only and is not
attached
during testing.  I have seen the commercial CISPR25 labs do similar.

I have not seen the discharge create additional problems - perhaps I
have
just been lucky. Opinions whether this is a 'respectable' test
technique?

Brian

-----Original Message-----
From: emc-p...@ieee.org [mailto:emc-p...@ieee.org]On Behalf Of Conway,
Patrick
Sent: Tuesday, April 24, 2012 3:53 PM
To: McInturff, Gary; EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: [PSES] Safety cost and ESD

This is a classic test problem.  I've seen this several times.  For this
setup, there is no place for the charge to dissipate between zaps.  So,
how
to discharge between zaps?  The easy answer  is to briefly connect a
strap
from apparatus to ground.  But that casues lots of problems, including
false
"failures".
 
The reason?  The discharge created when using a zero Ohm strap is
uncontrolled.  If you remove the "bleed" resistor, then the discharge is
not
bleeding slowly, it becomes another form of an ESD discharge.  The
problem
is that the waveform will not be representative of the human-body-model.

 
Designers of ESD equipment go to a lof trouble designing those ESD guns.
They must conform to an exact waveshape.  But the zero-Ohm ground strap
has
none of the circuit elements to shape the curve.  So there is a
likelyhood
of a faster rise time, more ringing, etc.  All things that no longer
represent the HBM.    If ths waveshape causes upset in the appratus it
cannot be considered a failure since the waveform is not HBM.
 
Check with your customer on how they are testing.  If the appratus
survives
the zap from the gun, but is upset when they discharge with the strap,
then
you have the asnwer.  The easy solution for "Test" is to place the bleed
resistor back into the discharge strap and see if the appratus survives.

 
The zero Ohm ground strap is not a real-world HBM scenario, and
certainly
not in conformance with EN61000-4-2 or any of the HBM standards.  On the
other hand, if your product is not subject to HBM, or your product needs
testing to a user-scenario that includes a strap discharge, then that is
a
perfectly good test.
 
//
Patrick
 
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
 
Different question about ESD.
 
I have a component we tested on the normal 55024 directed ESD table for
a
table top mounted device. Worked fine, problem is that the customer
places
this on a large metallic roll around pedestal on rubber wheels. When
they
send it to a lab and test it this way there is a problem. I don't have
the
pedestal so I'm trying to simulate with my table and removing the 1 mohm
bleeder resistor. Between discharges to the table, I ground a braided
strap
attached to the table top to my reference plane below. I get very
similar
results then. What should the braid really look like - should it just be
a
short, should it have some bleed resistance in it. I chose none since
the
discharge is going to be people touching the pedestal or other furniture
that is grounded. 
 
What does the standard say about  the VCP and HCP?
 
Gary McInturff
Reliability/Compliance Engineer  

 
//
Patrick 
 
From: McInturff, Gary [mailto:gary.mcintu...@esterline.com] 
Sent: Monday, April 23, 2012 12:37 PM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: [PSES] Safety cost and ESD
 
Given the implementation differences between US/Canada(?) and EU on the
medical 60601-1 standards. EU the June, US June 2013. How are folks
handling
new products being introduced now or in the very near future? Just got a
quote back and the US certifier wants to charge me twice once for the
2nd
edition and then to transition to the 3rd edition. I anyone else running
aground on this. Seems like this should be happening on both sides of
the
pond - since a CB report to 3rd edition would run into the second
edition
enforcement in the US (and maybe Canada - I don't know their
implementation
date).
 
To be fair - I get good service once I get past the sticker shock and
don't
have any complaints from that standpoint. In fact I enjoy the
engineering
staff I work with.
 
Different question about ESD.
 
I have a component we tested on the normal 55024 directed ESD table for
a
table top mounted device. Worked fine, problem is that the customer
places
this on a large metallic roll around pedestal on rubber wheels. When
they
send it to a lab and test it this way there is a problem. I don't have
the
pedestal so I'm trying to simulate with my table and removing the 1 mohm
bleeder resistor. Between discharges to the table, I ground a braided
strap
attached to the table top to my reference plane below. I get very
similar
results then. What should the braid really look like - should it just be
a
short, should it have some bleed resistance in it. I chose none since
the
discharge is going to be people touching the pedestal or other furniture
that is grounded. 
 
What does the standard say about  the VCP and HCP?
 
Gary McInturff
Reliability/Compliance Engineer

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list. To post a message to the list, send your
e-mail to <emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site
at http://product-compliance.oc.ieee.org/ can be used for graphics (in
well-used formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <emcp...@radiusnorth.net>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>



This message and any enclosures are intended only for the addressee.  Please 

notify the sender by email if you are not the intended recipient.  If you are 

not the intended recipient, you may not use, copy, disclose, or distribute this 

message or its contents or enclosures to any other person and any such actions 

may be unlawful.  Ball reserves the right to monitor and review all messages 

and enclosures sent to or from this email address.

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <emcp...@radiusnorth.net>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

Reply via email to