The operating/instruction manual is part of the safety submittal/approval
process and shall have correct ratings per the Type Test conditions. But
sales-infected spec sheets on websites or on advertising brochures are not
worthy of notice. Common problem in power industry.

If label and manual have different ratings - then the supplier's compliance
guy should be directly interrogated.

Brian

-----Original Message-----
From: emc-p...@ieee.org [mailto:emc-p...@ieee.org]On Behalf Of
ralph.mcdiar...@schneider-electric.com
Sent: Monday, December 03, 2012 10:36 AM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] IEC 60601-1-2 autoranging supplies

It certainly depends on how you rate the product, and what "ratings" really
means.  Typically the voltage rating would be marked on the nameplate of the
product, but some agencies will go look in the often embellished
specifications in the operator manual for the voltage range.

Some auto-ranging power supplies mark the a.c. input as a list of nominal
voltages, as in "100/120/220/240V".
____________________________________________________________________________
___

Ralph McDiarmid  |   Schneider Electric   |  Solar Business  |   CANADA  |
Regulatory Compliance Engineering



From: Brian Oconnell <oconne...@tamuracorp.com>
To: EMC-PSTC@LISTSERV.IEEE.ORG,
Date: 12/03/2012 08:55 AM
Subject: Re: [PSES] IEC 60601-1-2 autoranging supplies

Rated by test certificate and CB report, or by the sales dept-infected spec
sheet?

That is, what is the rating on the unit's label?

You can configure most variacs to provide well above the input voltage, but
this may do strange stuff to the source impendence. Use a really really
really big variac, or use an electronic source (Ametek, Chroma, etc).

Brian

-----Original Message-----
From: emc-p...@ieee.org [mailto:emc-p...@ieee.org]On Behalf Of David
Sent: Monday, December 03, 2012 8:21 AM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: IEC 60601-1-2 autoranging supplies

All,

IEC 60601-1-2 requires surge, burst, and dropouts be done at the maximum and
minimum rated voltages for autoranging supplies.  We typically test at 90
Vac and 240 Vac.  Many supplies I see now are rated up to 264 Vac though.
Is anyone testing at 264V?  Are there variacs out there to bump the input
voltage, since our generator can't go much above 240?

Thanks,

David

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <emcp...@radiusnorth.net>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

Reply via email to