John,

I think you are getting your tests mixed up.  The Ohm Meter is a Ground
Bond Test but at a very low current. As I said in my earlier email, Annex F
doesn't specify a current value for this test so technically an ohm meter
will do the job and meet the requirement of the 61010-1.

The HiPot test puts a high voltage potential between Line (or neutral) to
chassis ground to test the Dielectric Strength of your insulation. This
test is typically performed at 1500Vac and/or 2100Vdc for electrical
equipment operating up to the 230-volt range, but I have seen lower voltage
levels used also.   This is a different test than the Ground Bond test and
is also required for 100% of production.

I hope this was helpful.

The Other Brian

On Wed, May 12, 2021 at 2:09 PM John Cochran <jcoch...@strongarm.com> wrote:

> I’m interested in these answers, but UL and Intertek do not require us to
> do more than Ground Continuity testing on 100% of the products.  The UL/cUL
> certified product is an Industrial LCD Monitor (NWGQ, NWGQ7) evaluated to
> the ITE standard IEC 60950-1 & CAN/CSA C22.2 No. 60950-1-07, with a
> detachable power cord.  The Intertek certified product is a mobile cart
> with a permanently attached 3/c cord with plug.  It is evaluated to UL
> 61010-1 & CSA C22.2#61010-1-12.  Both have an input range of 100-240VAC @
> 15A.  We use a Hi-Pot tester with a Ground Continuity, Open/Short and
> Dielectric tests.  Neither NRTL requires more than a buzzer/ohm-meter test
> of all exposed metal surfaces, which the Hi-Pot tester does.  The responses
> to this email challenge these requirements, but if not required, it is hard
> to convince management that we need to invest in more test equipment.  I
> feel more assured of the grounding on a system when we use our 25A Ground
> Bond Tester, but we only have one for engineering purposes.  We have never
> been required to do Ground Bond testing, only Ground Continuity.
>
>
>
> *John Cochran*
>
> *jcoch...@strongarm.com <jcoch...@strongarm.com>*
>
> 215-443-3400 x219
>
>
>
> *From:* MIKE SHERMAN <msherma...@comcast.net>
> *Sent:* Wednesday, May 12, 2021 11:08 AM
> *To:* EMC-PSTC@LISTSERV.IEEE.ORG
> *Subject:* Re: [PSES] Ground Bond Tests
>
>
>
> Rich Nute has 3 Technically Speaking columns on this that can be found at
> the In Compliance magazine website. The first is here
>
> https://incompliancemag.com/article/derivation-of-ground-impedance/
>
> and I find it quite informative.
>
> Mike Sherman
>
> Graco Inc.
>
>
>
> On 05/12/2021 8:01 AM Wiseman, Joshua <
> joshua.wise...@orthoclinicaldiagnostics.com> wrote:
>
>
>
>
>
> I agree with Brian. I’ve been involved with product safety over 20 yrs and
> worked for an NRTL for about 6 of those.
>
>
>
> The 40 A requirement came from the Canadian electrical code and all the
> NRTLs.
>
>
>
> The only exemptions I’ve seen to the 100 % ground bond requirement has
> been for Class II and Class III devices or devices using an external power
> supply (brick or wall-wart styles.)
>
>
>
> Josh
>
>
>
> *From:* Brian Kunde <bkundew...@gmail.com>
> *Sent:* Wednesday, May 12, 2021 8:42 AM
> *To:* EMC-PSTC@LISTSERV.IEEE.ORG
> *Subject:* Re: [PSES] Ground Bond Tests
>
>
>
> *EXTERNAL SENDER: Verify links, attachments and sender before taking
> action*
>
>
>
> 61010-1 § 4.1 says,  "Tests in this standard are TYPE TESTS to be carried
> out on samples of equipment or parts. Their only purpose is to check that
> the design and construction ensure conformity with this standard. In
> addition, *manufacturers shall perform the ROUTINE TESTS of Annex F on
> 100 % of equipment produced* which has both HAZARDOUS LIVE parts and
> ACCESSIBLE conductive parts."
>
>
>
> 61010-1 Annex F, § F2 calls out the "Protective earth" test, also known as
> the Ground Bond Test, as a required test to be performed on 100% of
> production.  Note the "NOTE" that says, "No value is specified for the test
> current.".   So some NRTL inspectors allow the ground bond test in
> production to be performed at any reasonable current value, but most will
> follow the requirements of 6.5.2.
>
>
>
>
>
> I have been doing Product Safety for over 30 years and have worked with
> UL, CSA, TUV, NEMKO, EOLAS, etc., and all of them require 100% testing of
> production units to both the Highpot test and the Ground Bond test.
>
>
>
> Regarding the 100mΩ verses 200mΩ question, all the NRTL inspectors I have
> worked with require the measured impedance of the Power Cord to be no more
> than 100mΩ.  Note the UL/ANSI/CSA Deviation to 6.5.2.4 that instead of the
> 100mΩ requirement has a "shall not cause a potential drop of more than 4
> V".  At a minimum 40 A ground bond test, a 100mΩ impedance would give you a
> 4V drop.  It doesn't say anything about an 8V drop for non-detachable power
> cords.  So NRTLs are going to stick with the 100mΩ across the board, I
> believe (this is opinion).
>
>
>
> Hope this information is helpful.
>
>
>
> The Other Brian
>
>
>
>
>
> On Tue, May 11, 2021 at 11:27 PM Steve Brody <sgbr...@comcast.net> wrote:
>
> So here is my question, or actually two of them:
>
>
>
> First, one of my clients has been told by their 3rd party NRTL that a
> ground bond test is required as part of factory/production routine tests,
> even though it is not required by 61010-1.  I have never run into this
> requirement in my work history and I would be interested in hearing if any
> of you have.
>
>
>
> Second, just to be sure my interpretation is correct, 61010-1, section
> 6.5.2.4, Impedance of protective bonding of plug-connected equipment, in
> the fist sentence, it has a limit of 100 mOhms impedance and my read is
> that this is for equipment with a power cord that has a plug on one end and
> a receptacle on the other.
>
>
>
> Then, in the same section, still under the title of plug-connected
> equipment, it says for equipment that has a non-detachable power cord  the
> limit is 200 mOhms.  My interpretation is that plug-connected equipment
> with a non-detachable power cord is when the power cord is hard wired to
> the unit under test, but still has a plug for mains connection.
>
>
>
> Thoughts?
>
>
>
> Thanks in advance and you can reply here or privately to
> stev...@productehsconsulting.com
>
>
>
>
>
> Steve Brody
>
> sgbr...@comcast.net
>
> C - 603 617 9116
>
> -
> ----------------------------------------------------------------
>
> This message is from the IEEE Product Safety Engineering Society emc-pstc
> discussion list. To post a message to the list, send your e-mail to &LT;
> emc-p...@ieee.org&GT;
>
> All emc-pstc postings are archived and searchable on the web at:
> http://www.ieee-pses.org/emc-pstc.html
> <https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Femc-pstc.html&data=04%7C01%7Cjoshua.wiseman%40orthoclinicaldiagnostics.com%7C5cb8d9181834401d3f7008d9154371a1%7C6e63ffc0c2fc4cc4b6c4666b2ce89d92%7C0%7C0%7C637564201689622581%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=HvQW7ZqoO5YZS7CRFVERci8wTOlw06x5rnPosmWGT2I%3D&reserved=0>
>
> Attachments are not permitted but the IEEE PSES Online Communities site at
> http://product-compliance.oc.ieee.org/
> <https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fproduct-compliance.oc.ieee.org%2F&data=04%7C01%7Cjoshua.wiseman%40orthoclinicaldiagnostics.com%7C5cb8d9181834401d3f7008d9154371a1%7C6e63ffc0c2fc4cc4b6c4666b2ce89d92%7C0%7C0%7C637564201689632540%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=DFlEIGK%2FXubJKcLlm3CjW9O9pz4s3g6%2FBIzQKjVbY64%3D&reserved=0>
> can be used for graphics (in well-used formats), large files, etc.
>
> Website: http://www.ieee-pses.org/
> <https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2F&data=04%7C01%7Cjoshua.wiseman%40orthoclinicaldiagnostics.com%7C5cb8d9181834401d3f7008d9154371a1%7C6e63ffc0c2fc4cc4b6c4666b2ce89d92%7C0%7C0%7C637564201689632540%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=FOkw7IT9ocKxHdHR5GEp5M10Z%2FwVFclMo2YCM0YaoME%3D&reserved=0>
> Instructions: http://www.ieee-pses.org/list.html (including how to
> unsubscribe)
> <https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flist.html&data=04%7C01%7Cjoshua.wiseman%40orthoclinicaldiagnostics.com%7C5cb8d9181834401d3f7008d9154371a1%7C6e63ffc0c2fc4cc4b6c4666b2ce89d92%7C0%7C0%7C637564201689642497%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=PXH%2FHIsFp6InxFWXx4Q%2FZWmSQIT0I8JTCP7srcPGGPA%3D&reserved=0>
> List rules: http://www.ieee-pses.org/listrules.html
> <https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flistrules.html&data=04%7C01%7Cjoshua.wiseman%40orthoclinicaldiagnostics.com%7C5cb8d9181834401d3f7008d9154371a1%7C6e63ffc0c2fc4cc4b6c4666b2ce89d92%7C0%7C0%7C637564201689642497%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=NbeSM5%2BTNQTjlhrl7diWuUlDYex0u0NATVhPOE%2Bhc40%3D&reserved=0>
>
> For help, send mail to the list administrators:
> Scott Douglas &LT;sdoug...@ieee.org&GT;
> Mike Cantwell &LT;mcantw...@ieee.org&GT;
>
> For policy questions, send mail to:
> Jim Bacher &LT;j.bac...@ieee.org&GT;
> David Heald &LT;dhe...@gmail.com&GT;
>
> -
> ----------------------------------------------------------------
>
> This message is from the IEEE Product Safety Engineering Society emc-pstc
> discussion list. To post a message to the list, send your e-mail to <
> emc-p...@ieee.org>
>
> All emc-pstc postings are archived and searchable on the web at:
> http://www.ieee-pses.org/emc-pstc.html
> <https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Femc-pstc.html&data=04%7C01%7Cjoshua.wiseman%40orthoclinicaldiagnostics.com%7C5cb8d9181834401d3f7008d9154371a1%7C6e63ffc0c2fc4cc4b6c4666b2ce89d92%7C0%7C0%7C637564201689652452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=e9sc0WamLt00Mphy%2B%2Fs6H1ba%2Bas%2FZ9HHcCIdHYKP4d8%3D&reserved=0>
>
> Attachments are not permitted but the IEEE PSES Online Communities site at
> http://product-compliance.oc.ieee.org/
> <https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fproduct-compliance.oc.ieee.org%2F&data=04%7C01%7Cjoshua.wiseman%40orthoclinicaldiagnostics.com%7C5cb8d9181834401d3f7008d9154371a1%7C6e63ffc0c2fc4cc4b6c4666b2ce89d92%7C0%7C0%7C637564201689652452%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=ATze1ijC%2B9lZbXDIBRi8PuDBkmlo29PxQ9nenxX4FEg%3D&reserved=0>
> can be used for graphics (in well-used formats), large files, etc.
>
> Website: http://www.ieee-pses.org/
> <https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2F&data=04%7C01%7Cjoshua.wiseman%40orthoclinicaldiagnostics.com%7C5cb8d9181834401d3f7008d9154371a1%7C6e63ffc0c2fc4cc4b6c4666b2ce89d92%7C0%7C0%7C637564201689662410%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=e%2BB%2FjWp%2Bj191nwQ6ICYHNAUX9FTFj5AnEyxv94PSbgo%3D&reserved=0>
> Instructions: http://www.ieee-pses.org/list.html (including how to
> unsubscribe)
> <https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flist.html&data=04%7C01%7Cjoshua.wiseman%40orthoclinicaldiagnostics.com%7C5cb8d9181834401d3f7008d9154371a1%7C6e63ffc0c2fc4cc4b6c4666b2ce89d92%7C0%7C0%7C637564201689662410%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=pOoTqThzM%2FBRTRQUJJ1wM1Lb4z5J291HZMXe8oiWvdE%3D&reserved=0>
> List rules: http://www.ieee-pses.org/listrules.html
> <https://nam11.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flistrules.html&data=04%7C01%7Cjoshua.wiseman%40orthoclinicaldiagnostics.com%7C5cb8d9181834401d3f7008d9154371a1%7C6e63ffc0c2fc4cc4b6c4666b2ce89d92%7C0%7C0%7C637564201689662410%7CUnknown%7CTWFpbGZsb3d8eyJWIjoiMC4wLjAwMDAiLCJQIjoiV2luMzIiLCJBTiI6Ik1haWwiLCJXVCI6Mn0%3D%7C1000&sdata=Fr0RoZL%2FNCIqgErKzX1r0nTMfXS8jHpfF6CUvzNO7rM%3D&reserved=0>
>
> For help, send mail to the list administrators:
> Scott Douglas <sdoug...@ieee.org>
> Mike Cantwell <mcantw...@ieee.org>
>
> For policy questions, send mail to:
> Jim Bacher <j.bac...@ieee.org>
> David Heald <dhe...@gmail.com>
>
> -
> ----------------------------------------------------------------
>
> This message is from the IEEE Product Safety Engineering Society emc-pstc
> discussion list. To post a message to the list, send your e-mail to <
> emc-p...@ieee.org>
>
> All emc-pstc postings are archived and searchable on the web at:
> http://www.ieee-pses.org/emc-pstc.html
>
> Attachments are not permitted but the IEEE PSES Online Communities site at
> http://product-compliance.oc.ieee.org/ can be used for graphics (in
> well-used formats), large files, etc.
>
> Website: http://www.ieee-pses.org/
> Instructions: http://www.ieee-pses.org/list.html (including how to
> unsubscribe) <http://www.ieee-pses.org/list.html>
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
> Scott Douglas <sdoug...@ieee.org>
> Mike Cantwell <mcantw...@ieee.org>
>
> For policy questions, send mail to:
> Jim Bacher <j.bac...@ieee.org>
> David Heald <dhe...@gmail.com>
>
> -
> ----------------------------------------------------------------
>
> This message is from the IEEE Product Safety Engineering Society emc-pstc
> discussion list. To post a message to the list, send your e-mail to <
> emc-p...@ieee.org>
>
> All emc-pstc postings are archived and searchable on the web at:
> http://www.ieee-pses.org/emc-pstc.html
>
> Attachments are not permitted but the IEEE PSES Online Communities site at
> http://product-compliance.oc.ieee.org/ can be used for graphics (in
> well-used formats), large files, etc.
>
> Website: http://www.ieee-pses.org/
> Instructions: http://www.ieee-pses.org/list.html (including how to
> unsubscribe) <http://www.ieee-pses.org/list.html>
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
> Scott Douglas <sdoug...@ieee.org>
> Mike Cantwell <mcantw...@ieee.org>
>
> For policy questions, send mail to:
> Jim Bacher <j.bac...@ieee.org>
> David Heald <dhe...@gmail.com>
> -
> ----------------------------------------------------------------
>
> This message is from the IEEE Product Safety Engineering Society emc-pstc
> discussion list. To post a message to the list, send your e-mail to &LT;
> emc-p...@ieee.org&GT;
>
> All emc-pstc postings are archived and searchable on the web at:
> http://www.ieee-pses.org/emc-pstc.html
>
> Attachments are not permitted but the IEEE PSES Online Communities site at
> http://product-compliance.oc.ieee.org/ can be used for graphics (in
> well-used formats), large files, etc.
>
> Website: http://www.ieee-pses.org/
> Instructions: http://www.ieee-pses.org/list.html (including how to
> unsubscribe) <http://www.ieee-pses.org/list.html>
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
> Scott Douglas &LT;sdoug...@ieee.org&GT;
> Mike Cantwell &LT;mcantw...@ieee.org&GT;
>
> For policy questions, send mail to:
> Jim Bacher &LT;j.bac...@ieee.org&GT;
> David Heald &LT;dhe...@gmail.com&GT;
>

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

Reply via email to