Alex:
I agree with most of what you outlined with a few additional comments:

- Regarding your fourth point (and to some extent, your third point), there is 
an interpretation on this topic (which I requested back in 1996, which an OSHA 
response in 1997):               
https://www.osha.gov/laws-regs/standardinterpretations/1997-05-12-0
- Regarding your fifth point, I would clarify that there is no government 
agency mandated requirement for an NRTL “certification” for the manufacturer.  
The employer (hospital, medical provider, etc.) is regulated under OSHA, 
subject to the above interpretation.  Also, the some insurance companies may 
require the use of NRTL listed equipment as a condition of coverage for the 
facility where the equipment is used (I would include this as part of the 
market driven requirements which you mentioned).

Hope that helps.
Ken

    On Wednesday, July 23, 2025 at 12:18:21 PM EDT, Alex Grob 
<[email protected]> wrote:  
 
 
Hello Boštjan,
 
  
 
There are a couple of overlapping questions in your post below.
 
  
    
   - Classification is a commonly used type of “certification” of electrical 
medial devices by NRTLs.  Each NRTL can determine their own words to use to 
describe their “certification”.  Generally, Classification means that only some 
aspects of the equipment are covered by the certification.  Some NRTLs use this 
because they do not evaluate the medical applications and associated hazards as 
part of the certification.  For example, a Listed toaster oven likely has all 
the relevant hazards covered by the product standard.  For an X-Ray system the 
product standards don’t deal with the fact that exposure to radiation is a 
hazard…even when used correctly an X-Ray system can still cause harm (including 
death) to the patient.  Some NRTLs differentiate between these types of 
certifications by using Classified instead of Listed or Certified. Usually, the 
NRTL Mark has some words associated with the Classification such as “for 
electrical, mechanical and thermal hazards only”.
 
  
    
   
   - In summary, a Classification by an NTRL is equivalent to a Listing from 
the perspective of meeting an OSHA requirement.
 
  
    
   - The FDA is the regulatory body for medical devices (electrical or not) in 
the US.  An NRTL “certification” is not required for an FDA Approval.
 
  
    
   - The need for an NRTL “certification” for any medical device has been 
debated with varying interpretations for at least the last 25 years I have been 
in the medical device industry.  I work for a testing laboratory that tests 
mainly to IEC 60601-1 for medical electrical equipment; we have many customers 
that do not get a US NRTL certification for their electrical medical devices.  
These customers are able to sell their medical devices in the US usually 
without any issues.
 
  
    
   
   - For large medical devices (e.g., X-Ray, MRI) where they are permanently 
connected to the building wiring having an NRTL certification is likely 
mandated by the installation site or local AHJ (building / electrical 
inspector).
 
  
    
   - There is a general OSHA requirement regarding electrical / electronic 
products operating above 50 V needing an NRTL “approval”, but some read this 
section of the regulation as also allowing other options or not applying 
specifically to the device, but to its external enclosure and/or power cord.  I 
try to stay out of these discussions because they are all interpretation / 
opinion, I have never found an official (OSHA) interpretation on this topic.
 
  
    
   - Even if there is no mandated requirement for an NRTL “certification” from 
a legal / market access perspective, many hospitals, hospital chains and GPOs 
require a certification mark on devices they purchase. Having an NTRL 
certification may be a market-based requirement rather than a regulatory one.
 
  
 
I hope this helps.
 
  
 
Regards,
 

Alex
 
  
 
| 
Alex Grob
  |
| 
Chief Biomedical Engineer & Quality Manager
  |
| 
Direct: 262-672-6022
  |
| 

  |
|  |
|  |
|  |
| 
140 East Rawson Ave, Suite 301, Oak Creek, WI 53154 USA 
 
  
  |
|  |
| 
Experts in Product Safety     60601-1.com
  |
| 
  
 
MECA is Accredited by the FDA under the ASCA Program:
 
Click here forMore Info
  |


  
 
  
 
From: Boštjan Glavič <[email protected]>
Sent: Tuesday, July 22, 2025 5:03 AM
To: [email protected]
Subject: [PSES] CLASSIFIED MARK
 
  
 
Dear all,
 
  
 
customer has been informed by an NRTL that they are no longer able to issue 
NRTL mark for medical product. They can issue only CLASSIFIED MARK.
 
  
 
Is there any new rule about it?
 
  
 
Probably NRTL mark for medical product is not required - only FDA approval, 
correct?
 
  
 
Thank you for your thoughts.
 
  
 
Best regards,
 
Boštjan
 
  
 
  
 
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail 
[email protected] 
 
All emc-pstc postings are archived and searchable on the web at:
https://www.mail-archive.com/[email protected]/
 
Website: https://ewh.ieee.org/soc/pses/
Instructions: https://ewh.ieee.org/soc/pses/list.html (including how to 
unsubscribe)
List rules: 
https://pses.ieee.org/wp-content/uploads/2024/12/EM-PSTC-List-Rules.pdf 
 
For help, send mail to the list administrators:
Mike Sherman at: [email protected]
Rick Linford at: [email protected] 
 
For policy questions, send mail to:
Jim Bacher at: [email protected] 
 
To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1 

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
[email protected]
All emc-pstc postings are archived and searchable on the web at: 
https://www.mail-archive.com/[email protected]/
Website:  https://ewh.ieee.org/soc/pses/ 
Instructions: https://ewh.ieee.org/soc/pses/list.html (including how to 
unsubscribe)
List rules: 
https://pses.ieee.org/wp-content/uploads/2024/12/EM-PSTC-List-Rules.pdf

For help, send mail to the list administrators:
Mike Sherman at: [email protected]
Rick Linford at: [email protected]

For policy questions, send mail to:
Jim Bacher at:  [email protected]

 To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1 
  

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
[email protected]

All emc-pstc postings are archived and searchable on the web at:
https://www.mail-archive.com/[email protected]/

Website:  https://ewh.ieee.org/soc/pses/
Instructions:  https://ewh.ieee.org/soc/pses/list.html (including how to 
unsubscribe)
List rules: 
https://pses.ieee.org/wp-content/uploads/2024/12/EM-PSTC-List-Rules.pdf

For help, send mail to the list administrators:
Mike Sherman at: [email protected]
Rick Linford at: [email protected]

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
_________________________________________________
To unsubscribe from the EMC-PSTC list, click the following link: 
https://listserv.ieee.org/cgi-bin/wa?SUBED1=EMC-PSTC&A=1

Reply via email to