I like the way that PowerSDR allows you to make relatively precise noise
floor measurements for either the receiver or for the current band
conditions.  The RX meter tends to bounce around a lot on noise, making
it hard to determine a good average.  However, if your receiver has been
calibrated, you can switch the RX meter from "Signal" to "Sig Ave" and
get a much longer term average reading of the noise.  

I like to average over a longer time that used for the defaults, which
can be changed by going to Setup, selecting the Display tab, looking for
the Multimeter section in the lower right corner, and changing Average
Time to something in the 5000 (5 sec) range.  Hit "Apply" when done.

This feature gives an average that settles down gradually (converges?)
to within 0.1 db over 20 to 30 seconds.  In my mind this is an extremely
useful feature.

MDS (Minimum Discernable Signal) is often defined as injecting enough
signal to make the output increase by 3 db.  Increasing the output by 3
db means that the noise power equals the injected signal power and the
two of them add together to double the total power (a 3 db increase).
Thus, at the 3 db point, the injected signal level is at the same level
as the noise floor of the receiver.  Receiver MDS measurements should be
done using a 50 ohm load.  

With PowerSDR and the above averaging, we can measure the noise power
directly, and thus know directly the MDS of the receiver (or the noise
floor of the current band conditions) without the need of injecting an
actual measurement signal.  The receiver bandwidth must be specified for
MDS to be meaningful, and 500 Hz one of the bandwidth standards that is
often used for this specification.

I think it is extremely useful to take note of the noise level of the
band and to see how it changes over the course of the day.

A note of caution: In my home measurements of MDS, I have seen time of
day variations in MDS results.  I have come to realize that because the
shielding of coax is not perfect (30 dB of attenuation?), a noisy band
can impact the measurement of receivers in the 130+ dbm MDS region.  I
don't have an isolation chamber, so I just have to deal with it.  You
know you are in trouble when you can weakly hear signals across the band
when the dummy load is connected.

- Dan, N7VE


-----Original Message-----
From: [EMAIL PROTECTED]
[mailto:[EMAIL PROTECTED] On Behalf Of Dave Haupt
Sent: Friday, June 15, 2007 12:05 PM
To: flexradio@flex-radio.biz
Subject: Re: [Flexradio] Noise floor, specifying

You're getting very close to instrumentation norms.

Noise, by its very nature, varies in apparent strength by the bandwidth
in which you measure it.  When measuring truly random noise, the power
(in watts) is in direct proportion to the bandwidth of the filter used
in the measurement.  Man-made noise often consists of thousands of
semi-correlated carriers, and the noise power can grow faster than
simply being proportional to BW.

To be accurate, an engineer specifies noise either as "dBm/Hz" which
means "dBm measured in a one Hertz bandwidth" or "watts/hz" or something
like that.  It would also be equivalent to say "dBm measured using a
200Hz filter" or something like that.  Audio-frequency guys tend to
measure amplitudes in volts, at which point a square root gets into the
picture, and you will read things like "nanovolts per root Hertz".

Sine waves and noise are described using different mathematical terms
and as an interesting result, to make a truly accurate spectrum
analyzer, instrumentation manufacturers often use an algorithm to
attempt to distinguish a sine wave from a noise
function.   Anybody who's used the digitally-enhanced
analog spectrum analyzers by HP (8566/68, 8590 and 8560 series) has
probably encountered the "noise marker".  The instruments are calibrated
for accuracy with sine wave signals (or sums of sine waves - any
repetitive waveform), and are in error, fundamentally, for noise. When
you turn on the "noise marker", the instrument makes additional
calculations to make the readings accurate for random noise.

When you get down to the nitty-gritty of noise, it's not at all a simple
subject.

73,

Dave W8NF


>>From: "Doug McCormack" <[EMAIL PROTECTED]>
>>Subject: Re: [Flexradio] Noise Floor S meter vs
Panadapter
        
>> Thanks everyone for the quick and detailed
explanation.  I now understand
>> the S meter displays one sum across the entire
width of the green filter,
>> while the panadapter displays hundreds of sums
across the width of each
>> individual pixel.  When I set the filter very
narrow (approaching one pixel)
>> the S metter reading approaches the panadapter
reading.

>> My old Kenwood was similar in that selecting the
narrow CW filter caused a
>> drop in S meter noise floor.  This behavior makes
perfect sense for any radio.

>> I guess if someone asks about the noise floor, they
need to specify at what
>> filter width.

>> 73, Doug, VE3EFC


       
________________________________________________________________________
____________
Yahoo! oneSearch: Finally, mobile search that gives answers, not web
links. 
http://mobile.yahoo.com/mobileweb/onesearch?refer=1ONXIC

_______________________________________________
FlexRadio mailing list
FlexRadio@flex-radio.biz
http://mail.flex-radio.biz/mailman/listinfo/flexradio_flex-radio.biz
Archive Link: http://www.mail-archive.com/flexradio%40flex-radio.biz/
FlexRadio Knowledge Base: http://kb.flex-radio.com/ FlexRadio Homepage:
http://www.flex-radio.com/


_______________________________________________
FlexRadio mailing list
FlexRadio@flex-radio.biz
http://mail.flex-radio.biz/mailman/listinfo/flexradio_flex-radio.biz
Archive Link: http://www.mail-archive.com/flexradio%40flex-radio.biz/
FlexRadio Knowledge Base: http://kb.flex-radio.com/
FlexRadio Homepage: http://www.flex-radio.com/

Reply via email to