HPLC TROUBLESHOOTING


There are some information for your troubleshooting in conection with the 
practical HPLC instrument. Althougt there is no substitute for experience 
or consulting with an"expert" here are some tips and tricks, some answer 
to the most commonly asked technical inquiris, and some key questions to 
ask to help diagnose our problem. 

But the most important we are following the WI  to minimize the 
troubleshooting in our HPLC .

1.      poor Peak shape, loss of resolution or peak splitting.

�       Probable cause.
Mechanical fouling. Coloumn void. Sample overloud. Sample mobile phase 
incompatibility.dead volume from improperly assembled fittings and 
conectors.
�       Key question
Are all peaks equally affected? Did the peak shape start out fine but 
degrade over time? Has the back pressure increased?
�       Becertain that;
The mobile phase is the proper ph for your analytes. Your analytes are 
soluble in the mobile phase. You are not overloading the coloumn with to 
much sample. Ensure that the sample solvent is not to strong for the 
mobile phase. All connecting fittings are well seated.
�       First pass solution
Make sure your coloumn is properly connected with appropriate fittings for 
perfect, zero dead volume conections.
If the coloumn looks fine, then the problem is specifik to yur analyte or 
system.

2.      high pressure

�       probable casuse;
restricted flow-path
�       be certain that
all components in the mobile phase and spl are compatible with one 
another, filtered, and that precipitation cannot occur.
�       First pass solution
Replace column with a union and measured total system pressure. If 
presurre without coloumn is high, disconnect all conenections starting at 
the end of the flow path until you isolated the section giving high 
pressure. If pressure without coloumn is acceptable, attempt to remove 
particulate matter from the inlet frit by running the coloumn in reverse 
with the outlet disconnected from detector. If pressure is still high, 
refer to cleaning/regenaring coloumns procedure. Use inline filters or 
improve you spl preparation procedure to remove particulates and 
contaminating components.

3.      Retention time or selectivity variation
�       Probable cause
Variable pump flow, temperatur variation, or solvent mixing. Mobile phase 
preparation technique.

�       Be certain that
Comparison are being made under identical system, mobile phase, and spl 
condition. The coloumn are properly equilibrate. Coloumn have not been 
exposed to conditions that could change them. Eqilibrium of coloumn using 
reagents has to be wel controlled.

�       Key question
Is the void time consistent? Using the same spl and mobile phase bottle in 
the same day?

�       First pass solution
Test the coloumn under our QA conditions to make sure they are stil in 
good conditions.

4.      poor sensitivity

�       probable cause
detector lamp energy low. Wrong wavelength.  detector cell contaminated
�       be certain that
HPLC system is clean, calibrated and operating properly. All liquid 
conections are zero dead volume.
�       key questions ; does asimiliar coloumn give better sensitivity?
�       First pass solution
If you can get good sensitivity from another similar coloumn, adsorpsion 
may be occurring. The coloumn must   to be cleaning and regenerations 
coloumn. 

5.      ghost peaks

�       probable causes
contamination of coloumn or system
�       be certain that; the gost peaks is not from electronic noise
�       key questions
what spl have you previously injected to the coloumn? Does the peaks 
diminish with each injection, stay the same, or get larger?
�       First pass solution; make several consecutive 0 uL injections.
�       If peaks decrease with each injection, the contamination is from 
aprevios 
sample injections.
�       If peaks stay the same size with each injection, the contamination is 
from 
mobile phase.
�       If no peaks is seen with 0 uL injections, the gosth peaks is your spl.