Hi 

In order to adapt sequences on our 3T Prisma, I'm looking for details
both in the Appendix I of the reference manual,
"HCP_Q3_Release_Appendix_I.pdf", and articles published in the dedicated
volume of Neuroimage (n° 80). 

As I'm a new user of Siemens scanner, I may have misinterpreted
something, but I'm confused by different informations. 

In the top of the page 7 of appendix I, it's pointed out that
reconstruction of diffusion weighted images acquired with multi-channel
head coils shouldn't be done with Sum-of-squares, but instead with the
SENSE (R=1) method.
At the end of the appendix I, all Siemens sequence parameters for
diffusion weighted imaging are listed. In the System section, one can
see : 

Coil Combine Mode Sum of Squares 

Does this sequence parameter is the same described page 7 and if yes, is
there a mistake ? 

In the Sotiropoulos et al's paper (NI n° 80, pages 125-143), at the end
of page 139, the point 4 talked about a "root-sum-of-squares"
reconstruction method that should be avoided. Is it the same parameter
as 'Coil Combine Mode' ? 

Thanks in advance 

Sincerely
Franck 