Rene

I'm sorry but I'm going to have to disagree with you on this one.  But I'm 
coming from a GLP lab.  Different stainers stain differently.  We have two 
different brands of IHC stainers and even with most of the same reagents (rinse 
buffers are different) the staining comes out different on each one with 
basically the same protocol. Plus in a GLP lab we have to validate the stainer 
itself.  All of our equipment has been validated and when we moved to our new 
location we had to re validate everything again. Our validation protocols are 
about 50 to 70 pages long for each of the IHC stainers we have, along with 
validation of the PT links, with data tracers for temp, etc.  Not a simple 
task, but it needs to be done.  Once I have a GLP IHC protocol in place I can't 
change any equipment or reagents without revalidating the procedure.  It might 
be different in clinical but in the GLP world I would revalidate the special 
stains also. It should not be that difficult to do, just stain a couple control 
slides and determine if the staining is correct and document it.  A one page 
sheet could be created for each stain (document vendor, lot numbers and exp 
date of reagents, serial number of stainer, date performed, person performing 
the staining, what control slides were used and then the slides should be 
reviewed by the tech that stained it and then the pathologist, record date 
reviewed with initials, comments, etc.  I don't think much more would be needed 
for specials.  IHC would be different, more slides for validation, additional 
paperwork is required. There is a paper out there on recommendations on how to 
validate IHC staining.  I have a pdf of it if anyone is interested.  That's 
just my two cents, but I have the tendency to do more than to do less, since 
the last thing I need is a 483 from the FDA when they come and visit.


Liz 

Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC
Manager
Premier Laboratory, LLC
PO Box 18592
Boulder, Colorado 80308
office (303) 682-3949 
fax (303) 682-9060
www.premierlab.com
 
 
Ship to Address:
1567 Skyway Drive, Unit E
Longmont, Colorado 80504

-----Original Message-----
From: [email protected] 
[mailto:[email protected]] On Behalf Of Rene J Buesa
Sent: Tuesday, January 05, 2010 10:38 AM
To: '[email protected]'; Dorothy LWebb
Subject: Re: [Histonet] Special stains

Dorothy:
I would NOT do it.
If you do not change the procedure it does not really matter which auto stainer 
you use. Your "validation" is the reaction of your control that should react as 
before.
As a matter of fact, I would be more inclined to made a validation if you 
change your control rather than if you change the auto stainer.
If, for example, you are doing IHC and your dilution has been "fine tuned" for 
a given tissue control and you change the tissue, you will need to "fine tune" 
the dilution to have the same reaction strength of the epitope in the new 
control, but you will not have to change your procedure. That is why I always 
used IHC internal controls, and by using "normal strength epitopes" I don't 
have to keep validating.
That is also why I always shied away from using pathological cases to use as 
epitope controls.
So, I would not advise you to go through validation because of changing the 
auto stainer, as long as the procedure is the same.
René J.

--- On Tue, 1/5/10, Webb, Dorothy L <[email protected]> wrote:


From: Webb, Dorothy L <[email protected]>
Subject: [Histonet] Special stains
To: "'[email protected]'" <[email protected]>
Date: Tuesday, January 5, 2010, 12:23 PM


We are switching automated special stain equipment to a different vendor and am 
wondering if we have to validate all of the special stains we do.  I am from 
the "old school" and the validation process, other than IHC, is not my forte!!  
Any advice would be appreciated, so, thanks ahead of time!!

Dorothy Webb, Ht (ASCP)
Regions Histology Technical Supervisor
651-254-2962



  ________________________________
This e-mail and any files transmitted with it are confidential and are intended 
solely for the use of the individual or entity to whom they are addressed. If 
you are not the intended recipient or the individual responsible for delivering 
the e-mail to the intended recipient, please be advised that you have received 
this e-mail in error and that any use, dissemination, forwarding, printing, or 
copying of this e-mail is strictly prohibited.

If you have received this e-mail in error, please immediately notify the 
HealthPartners Support Center by telephone at (952) 967-6600. You will be 
reimbursed for reasonable costs incurred in notifying us. HealthPartners R001.0
_______________________________________________
Histonet mailing list
[email protected]
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



      
_______________________________________________
Histonet mailing list
[email protected]
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

_______________________________________________
Histonet mailing list
[email protected]
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

Reply via email to