As a student I was told to calibrate the pH meter every time it was used (which 
was about twice a day at the time), using standard buffers at pH4.0 and 7.0. 
Nearly 50 years later this is still good advice, and modern meters are much 
less trouble to calibrate than older ones. If you have to make frequent 
measurements with closely similar solutions, you might get away with 
calibrating twice a day.
 
It's the electrode, not the meter, that changes with time. All pH electrodes 
deteriorate with time, use and (especially) neglect. A slow response indicates 
that a new electrode is needed.
 
Wikipedia has a nice article about pH meters, and there's a web site 
http://www.ph-meter.info/pH-electrode-calibration that gives advice about 
calibrating older and newer instruments.
 
John Kiernan
Anatomy, UWO
London, Canada
= = =
----- Original Message -----
From: Akemi Allison <[email protected]>
Date: Tuesday, February 1, 2011 8:56
Subject: [Histonet] calibration of pH meter
To: histonet <[email protected]>

> Good morning everyone,
> 
> I am curious how frequently you all calibrate your pH meter, 
> particularly when you make up solutions for muscle bx's?  
> Do you do it daily, weekly, or monthly?  I am trying to 
> come up with an acceptable standard in a clinical histology lab, 
> verses a biotech lab, which makes-up large quantities of 
> reagents on a daily basis.
> 
> Thank you in advance for your information,
> Akemi
> 
> 
> Akemi Allison BS, HT (ASCP) HTL
> Director
> Phoenix Lab Consulting
> Tele: 408.335.9994
> E-Mail: [email protected]
> 
> _______________________________________________
> Histonet mailing list
> [email protected]
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
_______________________________________________
Histonet mailing list
[email protected]
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

Reply via email to