Select a series of different tissues (as diverse as possible)  and run half 
with your VIP and the other half with the Peloris. Do not identify which 
processor was used for each block/section and give the sections to as many 
pathologists in unidentified pairs as  a "blind test" They will not know how 
each block was processed.
 
The referees will have only two choices in their selection:
1- one section was better than the other within the pair (V or P), or
2 both were the same = of equal quality (S).
Analyze the results with the chi-square test where "S" includes V and P for a 
theoretical frequency of
(S+P) = (S+V) = 1.00 and for each is the theoreticla frequency is 0.5 
Compare the theoretical distribution with the results obtained from the 
referees.
 
By the way, the Peloris the existing "official" validations were done in 
Australia by the manufacturers of the original instrument (VisionBioSystems) 
and were never externally corroborated.
René J.
 
 

--- On Wed, 9/28/11, Hutton, Allison <ahut...@dh.org> wrote:


From: Hutton, Allison <ahut...@dh.org>
Subject: [Histonet] peloris validation
To: histonet@lists.utsouthwestern.edu
Date: Wednesday, September 28, 2011, 11:05 AM


We are considering purchasing a Peloris II. We are unsure of the validation 
process of a processor, having not had to replace one in 15+ years. We 
currently have 2 VIPs. For those of you using the Peloris, How did you 
validate? how many runs, and what  type of tissue did you use? 
Thank you for your in put.
Allison

_______________________________________________
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
_______________________________________________
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

Reply via email to