As to your issue of tissue not adhering to the slides, you could try to check 
the expiration date of your (+) slides. Perhaps it is just an issue with the 
slide.
As to controlling the "concentration" of an antibody by changing the incubation 
time, that is somewhat unorthodox to say the least. You modify a concentration 
with dilution, not with time. Perhaps you could modify the HIER step, or try to 
dilute the antibody.
René J.

--- On Fri, 5/18/12, Cloughley-Gray, Nancy <cloughl...@rvh.on.ca> wrote:


From: Cloughley-Gray, Nancy <cloughl...@rvh.on.ca>
Subject: [Histonet] Help
To: "'histonet@lists.utsouthwestern.edu'" <histonet@lists.utsouthwestern.edu>
Cc: "Callan, Lisa" <call...@rvh.on.ca>
Date: Friday, May 18, 2012, 4:02 PM


I'm a Histotechnologist working in the Regional Hospital in Barrie, ON Canada. 
We are using the Ventana Ultra for our Immunohistochemistry (IHC). Since the 
end of February, we have been having issues with some tissues lifting off our 
positive (marked with +) charged slides. It seems to be mostly with the fatty 
and/or larger sections. We now dry our slides for one hour at room temperature 
(R.T.) and an additional hour at 60 degrees C. We cut our IHC sections at 4 um. 
Since we have tried 2 different types of + slides and will be trying another 
type of charged slide (from Newcomer this time) I was wondering if anyone has 
any other suggestions?
I also have another question regarding a QC (quality control) issue. We use a 
multi-tissue control that is applied to the top of all our test slides for IHC. 
One of our paths commented that there is some positive staining in the smooth 
muscle nuclei of the normal bowel when we are testing for Progesterone (PR). We 
are using a Heat Induce Epitope Retrieval (HEIR) of 36 minutes with CC1 
(Ventana's proprietary buffer @ pH of 8.0-8.5) and a primary antibody 
incubation time of 16 minutes with PR clone 1E2. (Ventana instrumentation 
provides pre-diluted antibodies and the user adjusts the concentration of the 
antibody by adjusting the time the primary antibody is incubated with the 
tissue).
I am concerned about the implications of this staining and I have not been able 
to find a reference to this kind of unusual staining pattern. The bowel tissue 
that we are using as QC is from a 62 year old female patient. I was wondering 
if anyone has had any experience with this kind of staining and /or any 
references that I could use.

Thanking you in advance,
I look forward to your input,
Nancy Cloughley-Gray MLT

_______________________________________________
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
_______________________________________________
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

Reply via email to