Hi Cindy, I don't use the Dako system or any of the antibodies you mention, but 
this kind of sounds like an antibody diluent problem, where either what you 
dissolve the antibody in doesn't spread out evenly on the slide or the antibody 
doesn't dissolve well in the diluents you use. In an automated system, it just 
could be that the rinse cycle in between reagents is incomplete. I have seen 
something like this happen where oily or waxy blotches contaminated the slide 
before the run . Good luck, Tom T 

-----Original Message-----
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Cindy Pyse
Sent: Friday, August 09, 2013 9:51 AM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] IHC mystery

Happy Friday Histonetters

 

I have a mystery. First a little background. We stain our immuno’s slides on a 
Dako 48 with the PT modules for pretreatment. We have not changed any protocols 
or antibodies since we optimized and validated the instruments. The majority of 
our antibodies are Dako but we do use other vendors, our detection is the Dako 
Flex. The pretreatment used is either high or low Dako pretreatment solution 
with the Her-2 and p16 using there FDA approved pretreatment. The her-2 and p16 
slides are deparaffinized with xylene and rehydrated to water as the FDA 
protocol requires. All other antibodies are placed directly into the preheated 
pretreatment solution. All controls are precut and placed on Dako slides and 
prebaked for 30 minutes in a 58⁰ oven for 30 minutes. We cut our patient tissue 
and place them on the bottom of the slides and bake them in the same oven for 
45 minutes. 

 

Recently we have noticed that some of the ER and PR (breast tissue) is not 
staining completely, with an occasional Her-2 giving us the same staining 
problem. Sometime it is part of the section not staining at all, almost like a 
hydrophobic effect with part of the patient tissue not staining past an almost 
perfect line. Sometimes it seems like the parts of the patient tissue within 
the section is lifted off, but when you look at the slide the tissue is there 
it just does not stain. All of our controls stain perfectly. All other 
antibodies are not effected.

 

Some of the things we have tried. Leaving the slides longer in the oven. Since 
it is breast tissue and the pretreatment has a surfactant in it. We have 
checked the level of the slides on the instruments. Since I do have two 
instruments, we have plotted the slides to see if it is a certain rack or 
instrument. We have checked if the pretreatment modules are running correctly 
since we have three. I have three tech running the slides. I am having an 
engineer coming in to look at both instruments.

 

There is no common denominator I can find. I have talked to my TRS from Dako 
and their Tech Service. No one can come up with the cause of the problem. I 
have been doing histology for over 30 years and immuno’s for over 20 years and 
I have never had a problem that I couldn’t figure out the cause. This time I 
need expert help. Does anyone have ideas? I am willing to trying anything to 
solve this problem. 

 

Please help!

Cindy

 

Cindy Pyse CLT, HT(ASCP)

Laboratory Manager

X-Cell Laboratories of WNY

20 Northpointe Parkway Ste 100

Amherst, NY 14228

716-250-9235 Ext. 232

cp...@x-celllab.com

 

_______________________________________________
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
_______________________________________________
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

Reply via email to