Hi Tim, 

From a user perspective: The slides you’re using may be a bad lot causing 
hydrophobic staining or you’re not cleaning your covertiles correctly. The 
wells on the assembly might also not be pulling adequately. Have you cleaned 
the assembly recently, particularly the wells? You could use a clean kit offer 
by Novocastra or make your own. Performing a decontamination on your instrument 
would be helpful too to include your bulk reagent compartments and reagent 
syringe.  Have you also been performing the suggested manufacturer’s guidance 
on maintaining your Bond-Max IHC/ISH analyzer?
Or...
Simply the instrument needs to be recalibrated by a Leica engineer. Do not PM 
with a third party biomed group as the Leica engineer will need to recalibrate 
the instrument with a PM kit to check and if necessary replace the physical 
components, electronics, and fluidic system of your analyzer along with 
updating the coordinate system to match software build according to their 
latest recommendation from their home office. 

Apart from typical lab reagents or processes only performed by an engineer,  do 
you have electrical or magnetic surges in your lab? The backup battery that 
comes with your Bond is typically good for two years, and would need to be 
replaced every few years, especially as the Bond-Max requires a step down from 
the 120V outlet to 110V. And this consideration can also deeply impact your 
overall staining. 


The application specialist should have more politely said that they didn’t know 
but will try to get a response back to you. I don’t comprehend how the tray 
would be the issue unless it was severely cracked and would prevent the slide 
assembly from functioning properly. 

Very respectfully,
Eddie Martin, HT, HTL, QIHC
Hematology Team Leader
The National Institutes of Health
10 Center Drive
Bethesda, MD 20892


Sent from my iPhone

> On Aug 13, 2020, at 1:00 PM, histonet-requ...@lists.utsouthwestern.edu wrote:
> 
> Send Histonet mailing list submissions to
>    histonet@lists.utsouthwestern.edu
> 
> To subscribe or unsubscribe via the World Wide Web, visit
>    http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> or, via email, send a message with subject or body 'help' to
>    histonet-requ...@lists.utsouthwestern.edu
> 
> You can reach the person managing the list at
>    histonet-ow...@lists.utsouthwestern.edu
> 
> When replying, please edit your Subject line so it is more specific
> than "Re: Contents of Histonet digest..."
> Today's Topics:
> 
>   1. Re: End of shift (Charles Riley)
>   2. Blank spots on IHC tissues - bubbles on Bond (Morken, Timothy)
>   3. Re: Blank spots on IHC tissues - bubbles on Bond (Morken, Timothy)
>   4. Re: Blank spots on IHC tissues - bubbles on Bond
>      (Paula Keene Pierce)
>   5. Protocol or SOP of how to make Tissue microarray
>      (Chakib Boussahmain)
> <mime-attachment>
> <mime-attachment>
> <mime-attachment>
> <mime-attachment>
> <mime-attachment>
> _______________________________________________
> Histonet mailing list
> Histonet@lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet

_______________________________________________
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

Reply via email to