Hi Pushkar,

 

We recently discussed a bit measuring particle size by interpreting 
coordination numbers obtained from fitting EXAFS data, also referring to the 
paper you mention:

 

http://millenia.cars.aps.anl.gov/pipermail/ifeffit/2015-August/thread.html#start

 

The basic idea is that surface atoms are not coordinated to other atoms on the 
outside (i.e., for your Pd foil the surface Pd atoms would have say 6-8 
neighbors instead of 12 for the atoms on the inside) so when the particle size 
becomes less than a few nm the number of surface atoms become a significant 
proportion of the total number of atoms and you can now observe this 
undercoordination as a smaller amplitude (assumed to be due to average 
coordination number) of some coordination peaks in the FT. How you relate this 
average coordination decrease to particle size depends a bit on your system. 
Some approaches have been laid out in the literature, including the paper you 
mention. See the thread above for other. Some time back I used a “brute force” 
approach to calculate the average coordination vs size for a small particle of 
a certain crystal structure (see fig EA-7 in the supplementary material of 
Geochimica et Cosmochimica Acta 71 (2007) 1898–1912). I took the 
crystallographic structure and created the list of atom coordinates for a 10 nm 
spherical particle using the ATOMS program. I then wrote a simple program 
(which I can’t find now :-) to go through the list and calculate the average 
coordination for a certain shell by going to all atoms within a particle radius 
and counting the neighbors in each coordination shell of interest. The result 
is a graph of average coordination for that coordination shell vs radius, which 
you can then compare to the coordination numbers obtained in your fits of the 
EXAFS data to “read off” a particle size. This brute force approach takes into 
account the crystal structure of your material of interest and the slightly 
different surface density of atoms when you cut along different planes to 
create the particle. However, the improved accuracy from that will probably 
give you a difference in coordination number that is within the uncertainty of 
the coordination numbers in an EXAFS fit. You can use a spherical particle 
which is probably OK for your case with Pd, but you can also create say 
rod-shaped particles that emphasize certain crystallographic planes, depending 
on your case. 

 

Be aware of some of the issues raised by Scott in his email and in the thread 
in the link above when comparing EXAFS and TEM determined sizes. Another source 
of discrepancy may be the assumption that the EXAFS peak amplitude reduction is 
due only to a coordination number decrease. It is also conceivable that the 
strain in a 2nm particle gives a contribution to the Debye-Waller factor, 
potentially affecting the determination of the average coordination number by 
EXAFS. 

 

Overall, use particle size determinations as an estimate, regardless of the 
experimental or interpretation method used :-). More importantly, be aware of 
the pitfalls and their effect on the certainty of the conclusions you present. 
Unfortunately there doesn’t seem to be a recipe-like approach to the problem 
yet…

 

Best,

Max. 

 

 

 

 

 

From: Ifeffit [mailto:[email protected]] On Behalf Of 
pushkar shejwalkar
Sent: Thursday, January 21, 2016 3:02 AM
To: XAFS Analysis using Ifeffit <[email protected]>
Subject: [Ifeffit] Regarding calculations for scattering amplitude

 

Dear All,

   I sent a related question in the month of Dec but it seems that the question 
did not appeared on list because of mailbox being full, as we all have received 
an email. Since mailing list is started again I am sending this question once 
more.

My question is actually regarding finding out the nanoparticle size. In this 
respect, I am referring to Harris's work in which by using a simple equation we 
can roughly estimate the nanoparticle size

J. Appl. Phys., Vol. 94, No. 1, 1 July 2003

 Now the question is that in this paper and another paper I found online as 
well, it refers to Nnano and Nbulk. I am guessing that this is basically 
coordination number of nanoparticle and bulk. But my first question is how do I 
calculate it?? e.g. I am working on Pd K-edge. So should I fit Pd foil with Pd 
crystal data and get the CN? but that would be 12 anyways. So would Nbulk be 12 
in that case? and what will be Nnano then?

In another context, CN is proportional to amplitude, so in that case is it 
directly proportional? as in, can I simply replaced in formula, Nbulk/Nnano 
with Amplitudebluk/amplitudenano? 

Third thing is that, if above is true then how should/can I calculated the 
amplitude? isnt that amplitude is always calculated by Feff? if so could 
someone help me with the procedure to do so? How will I know amplitude for 
nanoparticles, because I cannot have FEFF for nanoparticles, right?

I went through lot of material, however, so far my calculations give me a 
solution that is vaguely correlated to TEM images. The TEM value is about 2nm 
(rouhgly 20 angstrom). my current values I am getting by solving the equation 
by my current understanding is only 2Ao. about 10 times less. What I am doing 
is simply plotting the crude data in athena to get x(R) Vs radial distance 
graph, calculate A-3 for nanoparticles and bulk and using this as amplitudes. 
It is possible that I may be doing something totally irrelevant, so please help 
me in this aspect.

Thank you

Sincerely 

Pushkar

 

-- 

Best Regards,
Pushkar Shejwalkar.
Post-doctoral -Researcher,

Tokyo Engineering University,

Tokyo-to

Japan

 

_______________________________________________
Ifeffit mailing list
[email protected]
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit

Reply via email to