Just a little food for thought...Last time I analyzed the 10CFR61 data for FCS, 
I came up with a weighted average alpha energy of about 5.5MeV.  That's right 
about the energy of most of the Pu-238 alphas.  Not as good a choice as Pu-239 
due to the betas and multiple alpha energies, but it is still better than 
Th-230.

I've also noticed that most of the throium sources I've had have more than just 
Th-230 on them -- muddying the results further.  For that matter, an old Pu-239 
source will have Am-241 on it from the Pu-241 that is also there.  A little 
past experimentation with an alpha spectrometer was quite educational.

One good thing about most alpha detectors is a fairly flat energy response 
curve over the energy range we most care about.  However, I have noted a 
fall-off in efficiency as you get down to the Th-230 alpha energy (4.7 MeV), 
particularly with plastic scintillators, that results in some over-reporting of 
alpha results and needless longer count times.  At least it is conservative.

My $0.02 on this.

David Brehm
Fort Calhoun Nuclear Station
Supervisor - Radiological Equipment & Dosimetry
Office: 402.533.7582
Cellular: 402.214.0107
Pager: 402.561.3310

-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of 
GOWDY, GREGORY M
Sent: Thursday, October 28, 2010 4:38 PM
To: [email protected]
Subject: Powernet: RE: Quick benchmarking - Alpha Calibration Sources

Summer uses Am-241 also.  There was an article in Radioactivity and 
Radiochemistry from the 90's that talked about talked about using appropriate 
energy alphas for various uses - Pu-238 and -239, and Am-241 for power plants 
and DOE applications and Th-230 and -232 for environmental labs.  You may want 
to look this reference up to check out the specifics.


Greg Gowdy, PhD, CHP
Sr. Staff Health Physicist
V. C. Summer Nuclear Station
Jenkinsville, SC  29065
(803)-345-4426  (work)
(803)-873-1561  (cell)




-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of 
Barber, Jerry
Sent: Wednesday, October 27, 2010 8:23 PM
To: [email protected]
Subject: Powernet: RE: Quick benchmarking - Alpha Calibration Sources

Robinson uses Am-241

-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of 
BREHM, DAVID M
Sent: Tuesday, October 26, 2010 5:17 PM
To: [email protected]
Subject: Powernet: RE: Quick benchmarking - Alpha Calibration Sources

Fort Calhoun Station currently uses Thorium for calibration and response 
testing gross alpha/beta counters.  I am working towards using Am-241 or Pu-239 
for alpha calibrations as their energy is much closer to what alpha we have.  
We already use the Am or Pu source for setting discriminator voltages as the Th 
betas & gammas make it a less optimal choice.

David Brehm
Supervisor - Radiological Equipment
Fort Calhoun Station
402-533-7582

________________________________
From: [email protected] [[email protected]] On Behalf Of Kinmark, 
Michael L. [[email protected]]
Sent: Tuesday, October 26, 2010 1:29 PM
To: [email protected]
Subject: Powernet: Quick benchmarking - Alpha Calibration Sources

Columbia Generating Station has the following question:

What calibration source do you use for Alpha portable instrument calibrations?  
Pu239, Am241, or other?

Thank you in advance.

Respectfully,

Michael Kinmark
HP Staff Advisor III
Radiological Support
Columbia Generating Station
509-377-2091






This e-mail contains Omaha Public Power District's confidential and proprietary 
information and is for use only by the intended recipient.  Unless explicitly 
stated otherwise, this e-mail is not a contract offer, amendment, nor 
acceptance.  If you are not the intended recipient you are notified that 
disclosing, copying, distributing or taking any action in reliance on the 
contents of this information is strictly prohibited.

-----------------------------------
Powernet - a service of the Health Physics Society Power Reactor Section

Powernet archives at http://www.mail-archive.com/[email protected]/

Reply to: [email protected]
If Questions, contact Mike Russell, CHP at [email protected]
-----------------------------------
Powernet - a service of the Health Physics Society Power Reactor Section

Powernet archives at http://www.mail-archive.com/[email protected]/

Reply to: [email protected]
If Questions, contact Mike Russell, CHP at [email protected]
-----------------------------------
Powernet - a service of the Health Physics Society Power Reactor Section

Powernet archives at http://www.mail-archive.com/[email protected]/

Reply to: [email protected]
If Questions, contact Mike Russell, CHP at [email protected]

This e-mail contains Omaha Public Power District's confidential and proprietary 
information and is for use only by the intended recipient.  Unless explicitly 
stated otherwise, this e-mail is not a contract offer, amendment, nor 
acceptance.  If you are not the intended recipient you are notified that 
disclosing, copying, distributing or taking any action in reliance on the 
contents of this information is strictly prohibited.

-----------------------------------
Powernet - a service of the Health Physics Society Power Reactor Section

Powernet archives at http://www.mail-archive.com/[email protected]/

Reply to: [email protected]
If Questions, contact Mike Russell, CHP at [email protected]

Reply via email to