>From Callaway

We use the Xradin Shonka Wyckoff chamber, A3, A4 and A6 ionization chamber.  
The readout is a Keithly 35040 electrometer (Not manufactured anymore, but I 
believe Fluke bought the product line and has an electrometer that will 
automatically do the temperature and pressure adjustments).

Standard Imaging in Madison Wisconsin would be the contact for the chambers 
(and probably an electrometer).

The Transfer Calibrations are performed by the University of Wisconsin - 
Madison, Accredited Dosimetry Calibration Laboratory, (John Mickas Lab).  
Although we send the equipment to Standard Imaging for pre checkout and 
interface with ADCL.  ADCL does maintain A2LA accreditation.  You do have to 
specify a 137Cs calibration, or you will get a 60Co calibration, and you do 
need equilibrium caps on the chambers.

We have two Shepherd 89 Box calibrators.  We had custom ion chambers built with 
short stems to provide the geometry (to fit in the box).

We have had a free air (beam port) irradiator in the past, and used the same 
equipment for it, but those were stock stems.

Questions or comments, Call or email.

Dewey
314-225-1061

From: [email protected] [mailto:[email protected]] On Behalf Of 
Treinen, Rick
Sent: Thursday, January 29, 2015 10:37 AM
To: [email protected]
Subject: Powernet: Irradiator Transfer Standards

EXTERNAL SENDER. Do not click on links or open attachments that are not 
expected and do not give out User IDs or passwords.
________________________________

Hello All,

For those with Cs-137 irradiators/calibrators (e.g., Model 81, 78, or 89):

1.       What instruments do you use as transfer standards to verify the source?

a.       Free air (Model 81 or 78)

b.      Box calibrators (Model 89)

2.       Where do you send your instruments for calibration?


Thanks.

---
Rick Treinen
Radiation Protection Engineer
PG&E / DCPP / RP
BLDG 104 / 216 / D
805.545.6481 (O)
805.769.8001 (C)




________________________________
PG&E is committed to protecting our customers' privacy.
To learn more, please visit http://www.pge.com/about/company/privacy/customer/
________________________________
The information contained in this message may be privileged and/or confidential 
and protected from disclosure. If the reader of this message is not the 
intended recipient, or an employee or agent responsible for delivering this 
message to the intended recipient, you are hereby notified that any 
dissemination, distribution or copying of this communication is strictly 
prohibited. Note that any views or opinions presented in this message are 
solely those of the author and do not necessarily represent those of Ameren. 
All e-mails are subject to monitoring and archival. Finally, the recipient 
should check this message and any attachments for the presence of viruses. 
Ameren accepts no liability for any damage caused by any virus transmitted by 
this e-mail. If you have received this in error, please notify the sender 
immediately by replying to the message and deleting the material from any 
computer. Ameren Corporation

Reply via email to