Thanks, Nicolae, for the didactic comment, but I must add that the expression

pol = SIN(PSI)**2 + COS(PSI)**2*COS(2*TET1)**2*COS(2*TET2)**2 
*.....*COS(2*TETm)**2*COS(2*TETb)**2

is an idealization and the real polarization factor may deviate notably from 
the idealized one depending on the crystal type (perfection/mosaicity) and 
other factors especially for multi-bounce systems. So, regarding the 
polarization factor of a real system in use, it is better to either consult the 
manufacturer or try determining it experimentally, for example, by measuring 
the same standard sample with and without monochromator.

Leonid

*******************************************************
 Leonid A. Solovyov
 Institute of Chemistry and Chemical Technology
 660049, K. Marx 42, Krasnoyarsk , Russia
 www.icct.ru/eng/content/persons/Sol_LA
 www.geocities.com/l_solovyov
*******************************************************

--- On Mon, 7/27/09, Nicolae Popa <nicp...@infim.ro> wrote:

> From: Nicolae Popa <nicp...@infim.ro>
> Subject: Re: LP factor in the Rietveld refinement
> To: "Leonid Solovyov" <l_solov...@yahoo.com>, rietveld_l@ill.fr
> Date: Monday, July 27, 2009, 11:04 AM
> Right, but specially for students-
> beginners we must  be much, let say, didactic
> 
> LP means (Lorentz) * (Polarisation)
> What is important in Rietveld refinement when a lot of
> mirrors & monochromators are present is how they change
> (Polarization)
> because (Lorentz) is changed by adding factors independent
> on hkl, then entering in the scaling factor
> 
> Presuming the same scattering plane for all "scatterers"
> the polarization factor is:
> 
> 
> pol = SIN(PSI)**2 +
> COS(PSI)**2*COS(2*TET1)**2*COS(2*TET2)**2
> *.....*COS(2*TETm)**2*COS(2*TETb)**2
> 
> 
> where   TET1, TET2, .....,
> TETm   are the Bragg angles at monochromator
> 1, 2, ....,m
> 
> and where  TETb  is the Bragg angle at 
> sample  (depending on hkl)
> 
> and where PSI  is the angle between polarization 
> vector of the incident beam - IF it is TOTALLY POLARIZED!!!
> - and the scattering plane;
> 
> If the incident beam is NOT POLARIZED the averages of both
> SIN(PSI)**2 and COS(PSI)**2  result in 1/2.
> 
> If the incident beam is partially polarized one replace for
> example SIN(PSI)**2  by  P0 , consequently
> COS(PSI)**2 = 1 - P0  and one refine P0
> 
> If the geometry is much complicated (different scattering
> planes for different monochromators) "pol" should be
> calculated for the given
> 
> geometry  by applying successively the known formula
> (see a book of electrodynamics, e.g.. Landau)
> 
> Ej+1 = (Ej X u)Xu  and taking at the END: 
> |E(last)|**2 / |E0|**2     (X means
> vectorial product)
> 
> where Ej is the electric field vector in the beam scattered
> j times and  u is the unit vector along the scattered
> beam j+1
> 
> Best wishes,
> 
> Nicolae Popa
> 
> 



      