Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-18 Thread Tom Knox
I have asked Agilent 
if stock versions of the 53230A and 53132A switched the internal oscillator out 
of circuit with an Ext Ref signal 
applied. I thought 
Agilent's engineer was intentionally vague but said the oscillators were
indeed switched out of circuit on the counter with Ext Ref signal applied. 
These questions were related to several 53132A's I have seen configured with a 
small board back near the Ext Ref input (OPT H01 I think) that appeared to 
Switch the internal reference out of circuit. Agilent would not share 
information on the option. My question to Agilent is why sell an option and be 
unwilling to say what it does or how your stock unit functions?
Thomas Knox



 From: t...@leapsecond.com
 Date: Mon, 17 Feb 2014 09:38:28 -1000
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,   
 but nobody in?
 
 Bob,
 
 I'm wondering if you (or any else) has measured the PLL performance of the 
 53230-series?
 
 I agree it will clean up the crud but this assumes the ext ref is dirtier 
 than the internal osc.
 
 What I found instead was that if you use a good external ref the PLL actually 
 makes it worse. This was very disappointing. The XO version of the counter 
 performed worse than the OCXO version even with a maser as the ext reference. 
 Did your reading of the schematic show a way to directly use the ext ref, 
 bypassing the noisy PLL?
 
 The other thing I found was that the ref out signal was a very polluted copy 
 of the ref in.
 
 /tvb (i5s)
 
  On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
  
  Hi
  
  If you dig into the schematics (when they supplied them … ):
  
  The external reference goes into a phase detector. It’s one of those 
  digital ones that can lock up to many inputs. You could feed 3. MHz 
  in as a standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The internal 
  oscillator (or an internal oscillator) is phase locked to the external 
  input through a fairly narrow analog loop. The idea is to clean up the crud 
  on the standard line. 
  
  With no external reference, the PLL drops out and you go back to what ever 
  the local reference is. 
  
  Yes there’s a little more to it than that and no the circuit is not exactly 
  the same on every counter HP ever made. 
  
  Bob
  
  On Feb 17, 2014, at 7:55 AM, wb6bnq wb6...@cox.net wrote:
  
  Hi Mike,
  
  The most likely answer is when you select external time base for an input, 
  it disables the connection for the internal oscillator.  The external 
  input signal is probably also routed straight to the reference output jack.
  
  However, it would be good to read the manual, as they usually cover how 
  those connections work.  Otherwise, perhaps someone that owns one could 
  provide further insight.
  
  BillWB6BNQ
  
  mike cook wrote:
  
  Something that must be simple to explain, but that I can't get my head 
  round.
  
  I got a new 53230A.
  When first using it, I measured my T-Bolt 10MHz using the internal 10MHz 
  timebase and it came up short of 10MHz, 9.999 998 5xx. I wasn't worried 
  about it as the counter only has a TCXO internal oscillator. So I fired 
  up my PRS10 and after leaving that on for some time, connected it to  Ext 
  Ref. , changed to the ext time base and measured again. This time 
  10.000.000.00x. Then I switched the two references, measuring the PRS10 
  against the T-Bolt. Again I got 10MHz down to the 11th digit.
  All that looked good so I have been using it with either the PRS10 locked 
  to GPS, or the T-Bolt as the external time base.
  
  After leaving it on (but not inactive) for a month, I did an Autocal. No 
  problem.
  I was wondering if that would have changed the internal time base 
  frequency, but no, using that still gave similar figures to the above.
  
  So at that point I decided to measure the Internal TB against my 
  reference. So I connected the Int. Ref. Out to channel 1, connected my 
  PRS10 ref to Ext. Ref In, selected the EXT time base and found that the 
  count was 10MHz dead on?  I don't get that at all.
  
  in summary:
  DUT against internal TB counts  10MHz.To me that means that the 
  internal timebase is a bit fast. Is that assumption correct?
  DUT against Ext.Ref counts 10MHz
  Internal TB against Ext.Ref counts 10MHz.   If my assumption above is 
  correct, the count should be greater than 10MHz, no?
  
  Can anyone shed any light on that?
  
  
  
  
  
  
  
  
  
  
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to 
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
  
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to 
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
  
  ___
  

[time-nuts] HP 10811-60165 Double Oven Crystal Oscillator Pin-Outs

2014-02-18 Thread Chris Smith
I've seen a pin-out for the outer-oven 6-position connector (2 heater wires,
2 thermistor wires), but I've not found anything on the pin-out of the other
6-position connector. Has anyone come across the pin-outs for the
10811-60165 connectors?

CS
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-18 Thread Tom Van Baak (lab)
TomK,

If anyone has technical contacts deep within Agilent, let's see if this issue 
can be resolved. I would have bought a 53230A when it came out a few years ago 
but my eval units showed this clock noise problem. That plus the poor quality 
of the ref out made me think the designers were cutting corners, or had little 
experience in metrology, or maybe they thought this was ok for a bench 
instrument.

Otherwise it's a really nice counter; the first one from Agilent than can 
actually do ADEV properly (since it is a time stamping counter).

I should dig out my old data and send it to you. Maybe as group we can help 
them fix the problem.

/tvb

 On Feb 18, 2014, at 12:10 AM, Tom Knox act...@hotmail.com wrote:
 
 I have asked Agilent 
 if stock versions of the 53230A and 53132A switched the internal oscillator 
 out of circuit with an Ext Ref signal 
 applied. I thought 
 Agilent's engineer was intentionally vague but said the oscillators were
 indeed switched out of circuit on the counter with Ext Ref signal applied. 
 These questions were related to several 53132A's I have seen configured with 
 a small board back near the Ext Ref input (OPT H01 I think) that appeared to 
 Switch the internal reference out of circuit. Agilent would not share 
 information on the option. My question to Agilent is why sell an option and 
 be unwilling to say what it does or how your stock unit functions?
 Thomas Knox
 
 
 
 From: t...@leapsecond.com
 Date: Mon, 17 Feb 2014 09:38:28 -1000
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,
 but nobody in?
 
 Bob,
 
 I'm wondering if you (or any else) has measured the PLL performance of the 
 53230-series?
 
 I agree it will clean up the crud but this assumes the ext ref is dirtier 
 than the internal osc.
 
 What I found instead was that if you use a good external ref the PLL 
 actually makes it worse. This was very disappointing. The XO version of the 
 counter performed worse than the OCXO version even with a maser as the ext 
 reference. Did your reading of the schematic show a way to directly use the 
 ext ref, bypassing the noisy PLL?
 
 The other thing I found was that the ref out signal was a very polluted copy 
 of the ref in.
 
 /tvb (i5s)
 
 On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 If you dig into the schematics (when they supplied them … ):
 
 The external reference goes into a phase detector. It’s one of those 
 digital ones that can lock up to many inputs. You could feed 3. MHz 
 in as a standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The internal 
 oscillator (or an internal oscillator) is phase locked to the external 
 input through a fairly narrow analog loop. The idea is to clean up the crud 
 on the standard line. 
 
 With no external reference, the PLL drops out and you go back to what ever 
 the local reference is. 
 
 Yes there’s a little more to it than that and no the circuit is not exactly 
 the same on every counter HP ever made. 
 
 Bob
 
 On Feb 17, 2014, at 7:55 AM, wb6bnq wb6...@cox.net wrote:
 
 Hi Mike,
 
 The most likely answer is when you select external time base for an input, 
 it disables the connection for the internal oscillator.  The external 
 input signal is probably also routed straight to the reference output jack.
 
 However, it would be good to read the manual, as they usually cover how 
 those connections work.  Otherwise, perhaps someone that owns one could 
 provide further insight.
 
 BillWB6BNQ
 
 mike cook wrote:
 
 Something that must be simple to explain, but that I can't get my head 
 round.
 
 I got a new 53230A.
 When first using it, I measured my T-Bolt 10MHz using the internal 10MHz 
 timebase and it came up short of 10MHz, 9.999 998 5xx. I wasn't worried 
 about it as the counter only has a TCXO internal oscillator. So I fired 
 up my PRS10 and after leaving that on for some time, connected it to  Ext 
 Ref. , changed to the ext time base and measured again. This time 
 10.000.000.00x. Then I switched the two references, measuring the PRS10 
 against the T-Bolt. Again I got 10MHz down to the 11th digit.
 All that looked good so I have been using it with either the PRS10 locked 
 to GPS, or the T-Bolt as the external time base.
 
 After leaving it on (but not inactive) for a month, I did an Autocal. No 
 problem.
 I was wondering if that would have changed the internal time base 
 frequency, but no, using that still gave similar figures to the above.
 
 So at that point I decided to measure the Internal TB against my 
 reference. So I connected the Int. Ref. Out to channel 1, connected my 
 PRS10 ref to Ext. Ref In, selected the EXT time base and found that the 
 count was 10MHz dead on?  I don't get that at all.
 
 in summary:
 DUT against internal TB counts  10MHz.To me that means that the 
 internal timebase is a bit fast. Is that assumption correct?
 DUT against Ext.Ref counts 10MHz
 Internal TB against 

Re: [time-nuts] HP 10811-60165 Double Oven Crystal Oscillator Pin-Outs

2014-02-18 Thread Richard H McCorkle
Chris,

The 10811-60165 may be similar to the HP 10811-60158
that uses using the following pin-out:

1 - BRN Oscillator Return (Com)
2 - RED Oscillator Power (+12V)
3 - ORG Oven Monitor Return (Com)
4 - YEL Oven Monitor Output
5 - GRN Oven Power (+18-24V)
6 - BLU Oven Return (Com)

The following description is from the 10811 A/B Manual
where the recommended oven monitor circuit is shown:

The Oven Monitor Output is an indicator of oven warm-up.
At initial turn-on (warmup) the oven monitor will go to
approximately 1.5 volts below the oven power supply
voltage. After the oven cuts back, the output will drop
to approximately 3.5 volts (at 25°C). The output
impedance of this circuit is 10,000 ohms.

Richard

 I've seen a pin-out for the outer-oven 6-position connector (2 heater wires,
 2 thermistor wires), but I've not found anything on the pin-out of the other
 6-position connector. Has anyone come across the pin-outs for the
 10811-60165 connectors?

 CS
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE-5680A sweep range setting

2014-02-18 Thread Paul Berger

Hi,

I have a couple that look like the one in these pictures minus the the 
little frequency control board. 
http://www.flickr.com/photos/14336723@N08/sets/72157632394339366/
One of mine would not lock so what I did I looked at my other 5680A 
which is the type referred to in the tip about C217 and looked at how it 
is connected in relation to the crystal  for the VCXO, which is very 
close to C217.  In the ones that I have with the stacked cards this 
crystal is on the middle card with a little block of foam over it.  Near 
this crystal is a trim cap C245 that seems to be connected the same way 
relative to the crystal and by adjusting it I was able to get mine to 
lock, but did find the adjustment to be a bit twitchy and since it is on 
the middle card, you need to remove the top one every time you want to 
tweak it a bit.


The one that I have are marked with option 57 and instead of having a 
flange around the edge like many of the telecom surplus ones it is 
mounted on a piece of 1/4 aluminum plate.The top card has a PIC on 
it and there is a RS232 level converter chip there but it looks like the 
connection only go to the 5 pin connector next to the SP232ACT RS232 
chip.  This one does not require external +5V and in fact bring out the 
10MHz on pin 4 where others seem to connect +5V. there does not appear 
to be any PPS output either.  These also have a cutout to expose the 15 
pin connector that is on the base board in front of the physics package.


Paul.

On 2/17/14 11:03 PM, Simon Lyons wrote:

Hello everyone,

I have a 5680 which is failing to lock. My DDS frequency is 8388608Hz, 
but it's sweeping between about 8388638 and 8388740. My unit has 3 
levels of PCBs in the DDS/VCO corner and there is no trimcap labeled 
C217. Does anyone know how to adjust the sweep center frequency on 
this type of 'triple decker' unit?


Thanks,

=Simon=
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 10811-60165 Double Oven Crystal Oscillator Pin-Outs

2014-02-18 Thread Chris Smith
That pin-out sounds promising as this unit has
BLK-RED-BLK-ORG-YEL-GRN-BLU, which seems consistent with the one that
you provided. Did you source that from one of HP's various guides or
somewhere else?

CS

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com]On
Behalf Of Richard H McCorkle
Sent: Tuesday, February 18, 2014 11:07 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 10811-60165 Double Oven Crystal Oscillator
Pin-Outs


Chris,

The 10811-60165 may be similar to the HP 10811-60158
that uses using the following pin-out:

1 - BRN Oscillator Return (Com)
2 - RED Oscillator Power (+12V)
3 - ORG Oven Monitor Return (Com)
4 - YEL Oven Monitor Output
5 - GRN Oven Power (+18-24V)
6 - BLU Oven Return (Com)

The following description is from the 10811 A/B Manual
where the recommended oven monitor circuit is shown:

The Oven Monitor Output is an indicator of oven warm-up.
At initial turn-on (warmup) the oven monitor will go to
approximately 1.5 volts below the oven power supply
voltage. After the oven cuts back, the output will drop
to approximately 3.5 volts (at 25°C). The output
impedance of this circuit is 10,000 ohms.

Richard

 I've seen a pin-out for the outer-oven 6-position connector (2 heater
wires,
 2 thermistor wires), but I've not found anything on the pin-out of the
other
 6-position connector. Has anyone come across the pin-outs for the
 10811-60165 connectors?

 CS
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-18 Thread Bob Camp
Hi

Well at least this got me digging a little. 

If you grab a copy of the 53230A spec sheet and look under the external 
reference input, it’s pretty well described. It will accept 1, 5,10 MHz as an 
external reference. It will lock over a 1 ppm range with the XO option and 0.1 
ppm with the OCXO option. Based on that I’d guess they are still using the same 
basic PLL approach as on the older counters (5335 era). 

The “Microsoft Windows inside” sticker on the back of the counter was a bit of 
a surprise ….

Bob

On Feb 18, 2014, at 11:51 AM, Tom Van Baak (lab) t...@leapsecond.com wrote:

 TomK,
 
 If anyone has technical contacts deep within Agilent, let's see if this issue 
 can be resolved. I would have bought a 53230A when it came out a few years 
 ago but my eval units showed this clock noise problem. That plus the poor 
 quality of the ref out made me think the designers were cutting corners, or 
 had little experience in metrology, or maybe they thought this was ok for a 
 bench instrument.
 
 Otherwise it's a really nice counter; the first one from Agilent than can 
 actually do ADEV properly (since it is a time stamping counter).
 
 I should dig out my old data and send it to you. Maybe as group we can help 
 them fix the problem.
 
 /tvb
 
 On Feb 18, 2014, at 12:10 AM, Tom Knox act...@hotmail.com wrote:
 
 I have asked Agilent 
 if stock versions of the 53230A and 53132A switched the internal oscillator 
 out of circuit with an Ext Ref signal 
 applied. I thought 
 Agilent's engineer was intentionally vague but said the oscillators were
 indeed switched out of circuit on the counter with Ext Ref signal applied. 
 These questions were related to several 53132A's I have seen configured with 
 a small board back near the Ext Ref input (OPT H01 I think) that appeared to 
 Switch the internal reference out of circuit. Agilent would not share 
 information on the option. My question to Agilent is why sell an option and 
 be unwilling to say what it does or how your stock unit functions?
 Thomas Knox
 
 
 
 From: t...@leapsecond.com
 Date: Mon, 17 Feb 2014 09:38:28 -1000
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,
 but nobody in?
 
 Bob,
 
 I'm wondering if you (or any else) has measured the PLL performance of the 
 53230-series?
 
 I agree it will clean up the crud but this assumes the ext ref is dirtier 
 than the internal osc.
 
 What I found instead was that if you use a good external ref the PLL 
 actually makes it worse. This was very disappointing. The XO version of the 
 counter performed worse than the OCXO version even with a maser as the ext 
 reference. Did your reading of the schematic show a way to directly use the 
 ext ref, bypassing the noisy PLL?
 
 The other thing I found was that the ref out signal was a very polluted 
 copy of the ref in.
 
 /tvb (i5s)
 
 On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 If you dig into the schematics (when they supplied them … ):
 
 The external reference goes into a phase detector. It’s one of those 
 digital ones that can lock up to many inputs. You could feed 3. 
 MHz in as a standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The 
 internal oscillator (or an internal oscillator) is phase locked to the 
 external input through a fairly narrow analog loop. The idea is to clean 
 up the crud on the standard line. 
 
 With no external reference, the PLL drops out and you go back to what ever 
 the local reference is. 
 
 Yes there’s a little more to it than that and no the circuit is not 
 exactly the same on every counter HP ever made. 
 
 Bob
 
 On Feb 17, 2014, at 7:55 AM, wb6bnq wb6...@cox.net wrote:
 
 Hi Mike,
 
 The most likely answer is when you select external time base for an 
 input, it disables the connection for the internal oscillator.  The 
 external input signal is probably also routed straight to the reference 
 output jack.
 
 However, it would be good to read the manual, as they usually cover how 
 those connections work.  Otherwise, perhaps someone that owns one could 
 provide further insight.
 
 BillWB6BNQ
 
 mike cook wrote:
 
 Something that must be simple to explain, but that I can't get my head 
 round.
 
 I got a new 53230A.
 When first using it, I measured my T-Bolt 10MHz using the internal 10MHz 
 timebase and it came up short of 10MHz, 9.999 998 5xx. I wasn't worried 
 about it as the counter only has a TCXO internal oscillator. So I fired 
 up my PRS10 and after leaving that on for some time, connected it to  
 Ext Ref. , changed to the ext time base and measured again. This time 
 10.000.000.00x. Then I switched the two references, measuring the PRS10 
 against the T-Bolt. Again I got 10MHz down to the 11th digit.
 All that looked good so I have been using it with either the PRS10 
 locked to GPS, or the T-Bolt as the external time base.
 
 After leaving it on (but not inactive) for a month, I did an Autocal. No 
 problem.
 I 

Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-18 Thread Tom Knox
Thanks Tom and Bob, I have been thinking of contacting Agilent for some time. I 
think they are a great company with some good products, but there are a few 
real blind spots in some current products. I also have seen in the past a 
genuine interest in listening. I would be willing to approach them if I could 
enlist your help in addressing potential changes to improve the product. 
Thanks;
Thomas Knox



 From: li...@rtty.us
 Date: Tue, 18 Feb 2014 18:00:17 -0500
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,   
 but nobody in?
 
 Hi
 
 Well at least this got me digging a little. 
 
 If you grab a copy of the 53230A spec sheet and look under the external 
 reference input, it’s pretty well described. It will accept 1, 5,10 MHz as an 
 external reference. It will lock over a 1 ppm range with the XO option and 
 0.1 ppm with the OCXO option. Based on that I’d guess they are still using 
 the same basic PLL approach as on the older counters (5335 era). 
 
 The “Microsoft Windows inside” sticker on the back of the counter was a bit 
 of a surprise ….
 
 Bob
 
 On Feb 18, 2014, at 11:51 AM, Tom Van Baak (lab) t...@leapsecond.com wrote:
 
  TomK,
  
  If anyone has technical contacts deep within Agilent, let's see if this 
  issue can be resolved. I would have bought a 53230A when it came out a few 
  years ago but my eval units showed this clock noise problem. That plus the 
  poor quality of the ref out made me think the designers were cutting 
  corners, or had little experience in metrology, or maybe they thought this 
  was ok for a bench instrument.
  
  Otherwise it's a really nice counter; the first one from Agilent than can 
  actually do ADEV properly (since it is a time stamping counter).
  
  I should dig out my old data and send it to you. Maybe as group we can help 
  them fix the problem.
  
  /tvb
  
  On Feb 18, 2014, at 12:10 AM, Tom Knox act...@hotmail.com wrote:
  
  I have asked Agilent 
  if stock versions of the 53230A and 53132A switched the internal 
  oscillator out of circuit with an Ext Ref signal 
  applied. I thought 
  Agilent's engineer was intentionally vague but said the oscillators were
  indeed switched out of circuit on the counter with Ext Ref signal applied. 
  These questions were related to several 53132A's I have seen configured 
  with a small board back near the Ext Ref input (OPT H01 I think) that 
  appeared to Switch the internal reference out of circuit. Agilent would 
  not share information on the option. My question to Agilent is why sell an 
  option and be unwilling to say what it does or how your stock unit 
  functions?
  Thomas Knox
  
  
  
  From: t...@leapsecond.com
  Date: Mon, 17 Feb 2014 09:38:28 -1000
  To: time-nuts@febo.com
  Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,   
   but nobody in?
  
  Bob,
  
  I'm wondering if you (or any else) has measured the PLL performance of 
  the 53230-series?
  
  I agree it will clean up the crud but this assumes the ext ref is 
  dirtier than the internal osc.
  
  What I found instead was that if you use a good external ref the PLL 
  actually makes it worse. This was very disappointing. The XO version of 
  the counter performed worse than the OCXO version even with a maser as 
  the ext reference. Did your reading of the schematic show a way to 
  directly use the ext ref, bypassing the noisy PLL?
  
  The other thing I found was that the ref out signal was a very polluted 
  copy of the ref in.
  
  /tvb (i5s)
  
  On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
  
  Hi
  
  If you dig into the schematics (when they supplied them … ):
  
  The external reference goes into a phase detector. It’s one of those 
  digital ones that can lock up to many inputs. You could feed 3. 
  MHz in as a standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The 
  internal oscillator (or an internal oscillator) is phase locked to the 
  external input through a fairly narrow analog loop. The idea is to clean 
  up the crud on the standard line. 
  
  With no external reference, the PLL drops out and you go back to what 
  ever the local reference is. 
  
  Yes there’s a little more to it than that and no the circuit is not 
  exactly the same on every counter HP ever made. 
  
  Bob
  
  On Feb 17, 2014, at 7:55 AM, wb6bnq wb6...@cox.net wrote:
  
  Hi Mike,
  
  The most likely answer is when you select external time base for an 
  input, it disables the connection for the internal oscillator.  The 
  external input signal is probably also routed straight to the reference 
  output jack.
  
  However, it would be good to read the manual, as they usually cover how 
  those connections work.  Otherwise, perhaps someone that owns one could 
  provide further insight.
  
  BillWB6BNQ
  
  mike cook wrote:
  
  Something that must be simple to explain, but that I can't get my head 
  round.
  
  I got a new 53230A.
  When 

Re: [time-nuts] HP 10811-60165 Double Oven Crystal Oscillator Pin-Outs

2014-02-18 Thread Richard H McCorkle
Chris,

I bought a number of 10811-60158 so I determined the pin-out
from an actual unit.

Richard



 That pin-out sounds promising as this unit has
 BLK-RED-BLK-ORG-YEL-GRN-BLU, which seems consistent with the one that
 you provided. Did you source that from one of HP's various guides or
 somewhere else?

 CS

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com]On
 Behalf Of Richard H McCorkle
 Sent: Tuesday, February 18, 2014 11:07 AM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] HP 10811-60165 Double Oven Crystal Oscillator
 Pin-Outs


 Chris,

 The 10811-60165 may be similar to the HP 10811-60158
 that uses using the following pin-out:

 1 - BRN Oscillator Return (Com)
 2 - RED Oscillator Power (+12V)
 3 - ORG Oven Monitor Return (Com)
 4 - YEL Oven Monitor Output
 5 - GRN Oven Power (+18-24V)
 6 - BLU Oven Return (Com)

 The following description is from the 10811 A/B Manual
 where the recommended oven monitor circuit is shown:

 The Oven Monitor Output is an indicator of oven warm-up.
 At initial turn-on (warmup) the oven monitor will go to
 approximately 1.5 volts below the oven power supply
 voltage. After the oven cuts back, the output will drop
 to approximately 3.5 volts (at 25°C). The output
 impedance of this circuit is 10,000 ohms.

 Richard

 I've seen a pin-out for the outer-oven 6-position connector (2 heater
 wires,
 2 thermistor wires), but I've not found anything on the pin-out of the
 other
 6-position connector. Has anyone come across the pin-outs for the
 10811-60165 connectors?

 CS
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.



 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.