Re: [ccp4bb] Negative electron Density for Zinc

2012-08-29 Thread Eleanor Dodson
I wish I could give a sensible explanation, but this feature is very common 
with heavy atoms.

Possible problems. The default wavelength for all atomic scattering functions 
is Cu Ka - you can see the formula used and details in the file 
$CLIBD/atomsf.lib.

The atomic scattering values are much the same for C N O at a range of 
wavelengths, but metals can have significant f' differences.

You can correct this in REFMAC input - see documentation.

The Zn is not fully occupied - this depends on how it is incorporated.

Eleanor
On 28 Aug 2012, at 18:44, Deepthi wrote:

 Hi everybody,
 
 I am working on a mutant protein structure which is 56 aminoacids long and 
 solved the structure using SAD( Single wavelength  anomalous dispersion) with 
 zinc ions. We used SHELX software to locate the Zinc atoms and solved phases. 
 We got the structure as a monomer. The resolution of the structure is 1.4 A0 .
 
 The problem with the structure is, when i calculate the structure factors in 
 COOT it always shows negative electron density around ZINC atoms which have 
 been located by SHELX.
 
 I initially thought it was because my COOT software was very old and it 
 doesn't recognize the Zinc atoms( I am new to crystallography) . I calculated 
 the structure factors using FFT from CCP4 and it shows the same. I don't 
 understand the problem. It refines well in Refmac and B-factors also look 
 very good.
 
 Anyone know what the problem is?
 
 Thank You in advance
 Deepthi  
 
 -- 
 Deepthi


[ccp4bb] off topic: reduced glutathione interfering with protein activity?

2012-08-29 Thread Peter Hsu
Hi all,

I've been purifying my protein off a GST column and have noticed a massive 
difference in activity of my protein between a prep that was freed from the 
column via on column cleavage, and a prep that was eluted (20mM GSH) and then 
cleaved and further purified. I'm suspecting that the glutathione is somehow 
modifying/inhibiting my protein in some way, despite having removed the 
glutathione from the buffer via dialysis/ion exchange. I don't see anything out 
of the ordinary in my electron density that would suggest that glutathione has 
affected my protein in some way, but the huge difference seen in my activity 
assay suggests otherwise.

My question is, has anyone else seen an effect from glutathione affecting their 
protein in some way? My second question is, what's the minimum amount of 
glutathione necessary to elute your protein from a column? 

Sorry for the off topic question and thanks for any responses,

Peter


Re: [ccp4bb] off topic: reduced glutathione interfering with protein activity?

2012-08-29 Thread Antony Oliver
GSH will reduce your protein quite nicely - is your enzyme activity redox
sensitive?  

---
Dr Antony W Oliver

Senior Research Fellow
CR-UK DNA Repair Enzymes Group
Genome Damage and Stability Centre
Science Park Road
University of Sussex
Falmer, Brighton, BN1 9RQ

email: antony.oli...@sussex.ac.uk
tel (office): +44 (0)1273 678349
tel (lab): +44 (0)1273 677512






On 8/29/12 5:56 PM, Peter Hsu hsuu...@u.washington.edu wrote:

Hi all,

I've been purifying my protein off a GST column and have noticed a
massive difference in activity of my protein between a prep that was
freed from the column via on column cleavage, and a prep that was eluted
(20mM GSH) and then cleaved and further purified. I'm suspecting that the
glutathione is somehow modifying/inhibiting my protein in some way,
despite having removed the glutathione from the buffer via dialysis/ion
exchange. I don't see anything out of the ordinary in my electron density
that would suggest that glutathione has affected my protein in some way,
but the huge difference seen in my activity assay suggests otherwise.

My question is, has anyone else seen an effect from glutathione affecting
their protein in some way? My second question is, what's the minimum
amount of glutathione necessary to elute your protein from a column?

Sorry for the off topic question and thanks for any responses,

Peter


Re: [ccp4bb] off topic: reduced glutathione interfering with protein activity?

2012-08-29 Thread Peter Hsu

I don't think so since I purify in the presence of reducing agents (DTT/BME) 
and I got activity out of the prep that was released by on column cleavage. On 
the other hand, I don't usually add those fresh each time I use the buffers so 
it's entirely possible the reducing agents are nowhere near the initial reduced 
form they were in initially.

On Wed, 29 Aug 2012, Antony Oliver wrote:


GSH will reduce your protein quite nicely - is your enzyme activity redox
sensitive?

---
Dr Antony W Oliver

Senior Research Fellow
CR-UK DNA Repair Enzymes Group
Genome Damage and Stability Centre
Science Park Road
University of Sussex
Falmer, Brighton, BN1 9RQ

email: antony.oli...@sussex.ac.uk
tel (office): +44 (0)1273 678349
tel (lab): +44 (0)1273 677512






On 8/29/12 5:56 PM, Peter Hsu hsuu...@u.washington.edu wrote:


Hi all,

I've been purifying my protein off a GST column and have noticed a
massive difference in activity of my protein between a prep that was
freed from the column via on column cleavage, and a prep that was eluted
(20mM GSH) and then cleaved and further purified. I'm suspecting that the
glutathione is somehow modifying/inhibiting my protein in some way,
despite having removed the glutathione from the buffer via dialysis/ion
exchange. I don't see anything out of the ordinary in my electron density
that would suggest that glutathione has affected my protein in some way,
but the huge difference seen in my activity assay suggests otherwise.

My question is, has anyone else seen an effect from glutathione affecting
their protein in some way? My second question is, what's the minimum
amount of glutathione necessary to elute your protein from a column?

Sorry for the off topic question and thanks for any responses,

Peter





Re: [ccp4bb] Jelly body refinement?

2012-08-29 Thread Robert Nicholls
Hi Gunnar,

A couple of comments, to clarify a few of the similarities and dissimilarities 
between DEN and analogous technologies:

According to your very nice paper from 2010, DEN refinement with gamma=0 gives 
a higher weight to external information, whilst gamma=1 ignores external 
information in favour of self-restraints. Thus, unless I am mistaken, isn't it 
gamma=1 that would be more analogous to jelly-body refinement? 

Both jelly-body and DEN with gamma=1 are similar in that they are both 
independent of explicit externally-derived information. Indeed, DEN with gamma 
in [0,1] is analogous, but not equivalent, to a combination of jelly-body (or 
self-restraints) and external reference structure restraints as implemented in 
REFMAC5. In fact, jelly-body is actually quite different to DEN with gamma=1.

Since jelly-body restraints are not applied to the target function (or 1st 
derivative), the restrained atoms are allowed to move easily if there is 
evidence to suggested that they should, e.g. from the electron density, or from 
other (external) restraints. The principal purpose of jelly-body restraints is 
simply to act as a regulariser thus stabilise refinement, not to inhibit 
deformation of interatomic distances where appropriate.

Jelly-body is only applied to the 2nd derivative simply due to the form of the 
function: X=(d-d_current)^2. Note that d_current is updated at each step, thus 
we always have d=d_current. Thus, X=0, X'=0, but X''!=0. This formulation makes 
sense - in the absence of any external prior knowledge, we shouldn't change the 
likelihood function or the gradient, as we want the minima to remain in the 
same place. However, we can reasonably change the 2nd derivative, and we would 
like to benefit from the decreased effective parameter-to-observation ratio 
from this regulariser. Hopefully, that explains why jelly-body is actually 
quite different to DEN with gamma=1.

Importantly, note that with jelly-body d_current is updated/reset at each step, 
which means that the structure is indeed very deformable. The structure is 
allowed to move away from the start values - in fact, d_current at cycle n is 
not dependent on d_current at cycle 0. I believe this contrasts with DEN, 
unless kappa=1.

In contrast with jelly-body, external restraints and local NCS restraints are 
applied to the target function. In order to allow the inter-atomic distances to 
exhibit large deviations from the prior information, the Geman-McClure robust 
estimator function is used instead of assuming least squares residuals (i.e. 
parameters are estimated using generalised M-estimators instead of the 
traditional maximum likelihood method). Consequently, when using jelly-body and 
external restraints, regions of structure that need to move far should be able 
to do so, whilst the regions that are happy should remain where they are 
(ideally with more stable refinement and less overfitting) .

Hopefully that helps to clarify a few of the similarities and dissimilarities 
between DEN and the analogous technologies implemented in REFMAC5 to anyone who 
may find it useful!

Regards
Rob



On 28 Aug 2012, at 20:23, Gunnar Schroeder wrote:

 Just a quick comment on low resolution refinement: 
 
 The concept of Deformable Elastic Network (DEN) refinement
 is quite similar to jelly-body refinement in the special case of 
 gamma=0, for which the network is not deformable.
 In contrast to jelly-body refinement, the DEN restraints are 
 however actually applied to the target function (and the first 
 derivative).
 
 For gamma0 the minimum of the elastic network potential 
 is allowed to move and, thus, to deform the restraints (which 
 changes their equilibrium distances).  Some individual distances 
 can deform more than others depending on the force they feel 
 from the target function. 
 
 This automatically discriminates between those regions in the 
 structure that need to move far (and are allowed to do so) and 
 those regions that are happy where they are (and remain 
 restrained). 
 
 DEN refinement is implemented in CNS (1.3) and 
 now also in Phenix (=1.7.3).
 
 Cheers,
   Gunnar


Re: [ccp4bb] Plastic and glass plates

2012-08-29 Thread Jim Fairman
Theresa,

The major advantage of the plastic plates is indeed ease of harvest.
 However, the plastic plates also tend to have some evaporation issues and
eventually dry out after a few months, where as the glass plates basically
last forever.  On the other hand, protein crystals in LCP tend to form in
the first several weeks after an experiment is set up so this usually isn't
a problem.  Some people prefer to set up initial screening experiments in
the glass plates and then optimize and set up farm trays using the
plastic plates where it's easier to harvest from (and cheaper).  I've also
set up everything in plastic trays from the start of a project to finish
without any problems.  If you're looking to perform LCP FRAP experiments,
glass is the best way to go.

Cheers, Jim

On Wed, Aug 29, 2012 at 12:24 PM, Theresa Hsu theresah...@live.com wrote:

 Dear all

 Is there any pros and cons of using plastic plates for LCP
 crystallization? The glass is clearer but it is very difficult to open.

 Thank you.




-- 
Jim Fairman, Ph D.
Crystal Core Leader I
Emerald BioStructures http://www.emeraldbiostructures.com/
Tel: 206-780-8914
Cell: 240-479-6575
E-mail: fairman@gmail.com jfair...@embios.com


Re: [ccp4bb] off topic: reduced glutathione interfering with protein activity?

2012-08-29 Thread Vitali Stanevich
Peter,

Yes, in our hands it can eluted completely. Usually 20 ml of elution buffer
per 2 ml of GS4B resin (GE healthcare, fast flow) is enough . We use 30 ml
open-columns for purification - each column has 2 ml of resin.

Vitali

On Wed, Aug 29, 2012 at 3:02 PM, hsuu...@u.washington.edu wrote:

 Hi Vitali,

 I usually prep my GST elution buffers fresh, and make sure to pH it after
 dissolving the powder. So I don't think that's the problem, but good to
 know that you can use as low as 5mM GSH. Do you know if you knock off most
 of the protein from the resin at that concentration of GSH?

 Thanks,
 Peter


 On Wed, 29 Aug 2012, Vitali Stanevich wrote:

  Peter,

 Such high concentration of GSH may change the pH according to our
 experience. We usually use 50 mM Tris pH=8.0, 5 mM GSH, 3 mM DTT - so that
 you can load the sample on ion-exchange column after elution. If your
 protein is not stable without NaCl - you should add it also.

 Vitali

 On Wed, Aug 29, 2012 at 12:42 PM, Peter Hsu hsuu...@u.washington.edu
 wrote:
   I don't think so since I purify in the presence of reducing agents
 (DTT/BME) and I got activity out of the prep that was released by on column
 cleavage. On the other hand, I don't usually add those
   fresh each time I use the buffers so it's entirely possible the
 reducing agents are nowhere near the initial reduced form they were in
 initially.

   On Wed, 29 Aug 2012, Antony Oliver wrote:

 GSH will reduce your protein quite nicely - is your enzyme
 activity redox
 sensitive?

 ---
 Dr Antony W Oliver

 Senior Research Fellow
 CR-UK DNA Repair Enzymes Group
 Genome Damage and Stability Centre
 Science Park Road
 University of Sussex
 Falmer, Brighton, BN1 9RQ

 email: antony.oli...@sussex.ac.uk
 tel (office): +44 (0)1273 678349
 tel (lab): +44 (0)1273 677512






 On 8/29/12 5:56 PM, Peter Hsu hsuu...@u.washington.edu
 wrote:

   Hi all,

   I've been purifying my protein off a GST column and
 have noticed a
   massive difference in activity of my protein between a
 prep that was
   freed from the column via on column cleavage, and a
 prep that was eluted
   (20mM GSH) and then cleaved and further purified. I'm
 suspecting that the
   glutathione is somehow modifying/inhibiting my protein
 in some way,
   despite having removed the glutathione from the buffer
 via dialysis/ion
   exchange. I don't see anything out of the ordinary in
 my electron density
   that would suggest that glutathione has affected my
 protein in some way,
   but the huge difference seen in my activity assay
 suggests otherwise.

   My question is, has anyone else seen an effect from
 glutathione affecting
   their protein in some way? My second question is,
 what's the minimum
   amount of glutathione necessary to elute your protein
 from a column?

   Sorry for the off topic question and thanks for any
 responses,

   Peter










[ccp4bb] Pittsburgh Diffraction Conference 2012

2012-08-29 Thread Smith, Clyde
The 70th Pittsburgh Diffraction Conference will be held at SSRL between 
September 30th and October 2nd 2012 and will feature 2 full days of lectures 
and poster presentations. The conference starts with a welcome reception on 
September 30 and there will be a banquet on the evening of October 1st. The 
conference will precede the 2012 SSRL/LCLS Users' Meeting and Workshops 
http://www-conf.slac.stanford.edu/ssrl-lcls/2012/ October 3rd - 6th. Session 
topic include:

New Science with Femtosecond Diffraction
Exciting Macromolecular Structures
New Ideas in Crystallography, and
Hybrid Methods in Macromolecular Crystallography

There will be a Keynote Address to begin the Conference on Monday October 1st 
by Prof Brian Kobilka (Stanford University) on G protein coupled receptors 
(GPCRs).  All speakers and session topics can be found at this link: 
http://www.pittdifsoc.org/PDC_2012/schedule.htm

Registration is now open (US$100 or US$20 for students and postdocs) and for 
those who register at least 2 weeks in advance you will receive a free ticket 
to the banquet. There will be an award for the best student poster so get your 
abstracts in nice and early. The abstract deadline has been extended to 
September 15. Please visit the website below to register and submit your 
abstract.

http://www.pittdifsoc.org/PDC_2012/index.htm

If you have not made your hotel reservations yet and would like to stay at the 
SLAC Guest House, please reserve your room soon.  The SLAC Guest House will 
only hold the remaining rooms that have been blocked for the PDC and Users' 
Meeting through this week.   Please specify 'Users12' or 'PDS 2012' to take 
advantage of discount rates for SLAC guests.  Visit this link for more 
information about accommodation options: 
http://www.pittdifsoc.org/PDC_2012/accommodations.htm.



-
Clyde A. Smith, Ph.D.
Senior Staff Scientist
Stanford University
Stanford Synchrotron Radiation Lab, MS 99
2575 Sand Hill Rd
Menlo Park, CA 94025 USA

ph   (650)926-8544
fax  (650)926-3292
cell (650)714-6001
smb.slac.stanford.edu