Re: [ccp4bb] How to reduce no. of overlaps

2012-03-07 Thread Edward A. Berry

Jim Pflugrath wrote:

In addition to reducing the beam divergence, you may wish to use a
smaller beam size by using a smaller collimator or making the slits
smaller. A smaller crystal can also help to spatially separate the Bragg
spots as can moving the detector closer to the crystal. Yes, closer to
the crystal. This is not intuitive, but arises since modern homelab
beams are not parallel but are diverging from a focal point near the
sample position. It is just something else you may wish to try.


But the pattern is also diverging from a point at the sample?

I'm guessing the focus point is somewhere between the crystal
and the detector, so by moving the detector closer you are better
approximating "focus on the detector" rather than "focus on the crystal"?

With a home source one probably has room for a Huber goniometer with
arcs, or better yet one of those goniometers that allows rotation
up to 90* about a point at the crystal, so the crystal doesn't move
out of the cold stream as you rotate.

One can also cheat on the mosaicity during integration by fixing
it at a small fraction of the true mosaicity. This is called
"cutting off the wings" or more euphemistically "peak height
sampling". The accuracy will suffer, but not as much as you
might expect- probably because if spot profiles are pretty
similar, the the height at  peak is a good measure of peak
volume.

eab


Re: [ccp4bb] How to reduce no. of overlaps

2012-03-07 Thread Bosch, Juergen
In addition to all the excellent suggestions if you can you can also move your 
detector away from the center of the beam aka increase the detector size in one 
dimension. Not sure if you can do that at your home source though. By moving 
the center of the beam say to the lower 9/10 of the detector you will still 
have the beam position on the detector but you will also have increased the 
total area. Needless to say you will need to collect over a larger wedge to get 
a complete dataset.

Regarding your overlap issues, it does help to spend some time thinking about 
the experiment before collecting the data, I would like to point toward the 
direction of Mosflm & strategy in conjunction with the Testgen option.

Depending how badly your overlapped data is, you might want to run it through 
XDS and see how much it can rescue, assuming you used Mosflm or HKL2000. You 
can also give the Separation Close command a try in Mosflm with Postref Width 
and Profile Optimize (RTFM for more details).

Jürgen
..
Jürgen Bosch
Johns Hopkins University
Bloomberg School of Public Health
Department of Biochemistry & Molecular Biology
Johns Hopkins Malaria Research Institute
615 North Wolfe Street, W8708
Baltimore, MD 21205
Office: +1-410-614-4742
Lab:  +1-410-614-4894
Fax:  +1-410-955-2926
http://web.mac.com/bosch_lab/






Re: [ccp4bb] How to reduce no. of overlaps

2012-03-07 Thread Mark J van Raaij
my experiences:
C2 with a long axis parallel to the shortest crystal dimension, crystals were 
plates. Used prebent loops to fish the crystals. Personally I haven't tried to 
bend loops in mounted crystals as Frank does, but it sounds very useful.
bar-shaped P321 crystals with hexagonal cross-sections, these naturally "fell" 
into loops in the "right" orientations. Too "right" usually, because you also 
don't want the long axis too parallel to the rotation axis and have a large 
blind cone compromising completeness, 10-20º from parallel is great, here a 
minikappa or the bendable loop mounts are very useful.

Mark J van Raaij
Laboratorio M-4
Dpto de Estructura de Macromoleculas
Centro Nacional de Biotecnologia - CSIC
c/Darwin 3
E-28049 Madrid, Spain
tel. (+34) 91 585 4616
http://www.cnb.csic.es/~mjvanraaij



On 7 Mar 2012, at 08:44, Frank von Delft wrote:

> Yes... quite.  Thanks!
> 
> (Beam, rotation axis... how can you expect me to keep all these new-fangled 
> inventions apart?!)
> 
> 
> 
> On 07/03/2012 07:33, VAN RAAIJ , MARK JOHAN wrote:
>> typo correction, you'll want the long axis parallel to the rotation
>> axis, not to the beam.
>> Mark
>> 
>> Quoting Frank von Delft:
>> 
>>> You probably have to tilt your crystal, so that the long axis is
>>> parallel to the beam.  We do this routinely:  cut a plastic pipette
>>> tip to have a sharp point, then push the loop where it attaches to
>>> the pin, to bend the crystal itself.
>>> 
>>> You have to identify from your diffraction whether the long axis is
>>> pointing through the face or the edge of the loop.  As it's P6,
>>> chances are it's through the face, because long-axis P6 tends to
>>> make flat hexagons which lie flush with the face.  So you have to
>>> bend so the face of the loop upwards.
>>> 
>>> You'll have to practice this first, though, so put up an empty loop.
>>>  Top tips:
>>> * Don't breathe!  You'll blow the cryostream away.
>>> * Bend the loop towards (rather than away from) the rim edge of
>>> the pin to which it's glued.
>>> * Don't breathe!
>>> * Practise practise practise.
>>> 
>>> 
>>> Another thing:  most in-house sources allow you to reduce divergence
>>> of the beam.  You lose intensity, but no matter, just expose longer.
>>>  That also improves overlap.
>>> 
>>> Cheers
>>> phx
>>> 
>>> 
>>> 
>>> On 07/03/2012 04:56, Dipankar Manna wrote:
 Dear Crystallographers,
 
 I am working on a protein having SG P6, the cell parameters are a=
 79, b= 79, c= 325. The crystals are forming in big size and with
 very good shape. It also diffracting very well in Home source
 facility both in terms of resolution and intensity. But the only
 problem is the number of overlaps. Its showing much more than the
 good spots. As a result the completeness is showing maximum up to
 65% even after collecting 180 degrees. I am unable to get a
 complete data. I tried with reducing the oscillation angel to 0.3
 degree/0.5 degree but it did not improve that much. Please give me
 some suggestions.
 
 Regards,
 
 Dipankar
 
 /Dipankar Manna/
 
 */Aurigene Discovery Technologies Limited,/*
 
 /#39-40 KIADB Industrial Area, Electronic City, Phase II,/
 
 /Hosur Road, Bangalore- 560 100, India/
 
 /Cell: +91-9538631469 // | Office Ph  : +91 80-66204422 (Extn: 398)
  | Email ID: dipanka...@aurigene.com/
 
 
 
 
 This e-mail and any files transmitted with it are for the sole use
 of the intended recipient(s) and may contain confidential and
 privileged information.If you are not the intended recipient,
 please contact the sender by reply e-mail and destroy all copies of
 the original message.Any unauthorized review, use, disclosure,
 dissemination, forwarding,printing or copying of this email or any
 action taken in reliance on this e-mail is strictly prohibited and
 may be unlawful.
 
 Visit us at http://www.aurigene.com
>> 
>> 
>> Mark J van Raaij
>> Laboratorio M-4
>> Dpto de Estructura de Macromoléculas
>> Centro Nacional de Biotecnología - CSIC
>> c/Darwin 3, Campus Cantoblanco
>> 28049 Madrid
>> tel. 91 585 4616
>> email: mjvanra...@cnb.csic.es
>> 


Re: [ccp4bb] How to reduce no. of overlaps

2012-03-07 Thread Zhijie Li
Hi,

Besides aligning the long axis with the rotation axis, which is the most 
important, there are a few more things that may help:

1) Try optimizing the freezing to reduce the mosaicity (if not ideal), or shoot 
at RT if possible. With higher mosaicity, the shape of the reflections are 
elongated in the reciprocal space, increasing the chance of overlapping. 
2) Collect datesets in two runs, one with detector at far position to get more 
spacing between dots, another one at a closer position to get high res. 
Sometimes you may have to sacrifice some resolution for the completeness.
3) Shoot at a synchrotron with micro beam and a large detector.
4) Try integrating the images with both HKL2000 and mosflm. My past experience 
seems to suggest that mosflm is more tolerant to closely positioned reflections.

Zhijie


From: Dipankar Manna 
Sent: Tuesday, March 06, 2012 11:56 PM
To: CCP4BB@JISCMAIL.AC.UK 
Subject: [ccp4bb] How to reduce no. of overlaps


Dear Crystallographers,



I am working on a protein having SG P6, the cell parameters are a= 79, b= 79, 
c= 325. The crystals are forming in big size and with very good shape. It also 
diffracting very well in Home source facility both in terms of resolution and 
intensity. But the only problem is the number of overlaps. Its showing much 
more than the good spots. As a result the completeness is showing maximum up to 
65% even after collecting 180 degrees. I am unable to get a complete data. I 
tried with reducing the oscillation angel to 0.3 degree/0.5 degree but it did 
not improve that much. Please give me some suggestions.



Regards,



Dipankar





Dipankar Manna

Aurigene Discovery Technologies Limited,

#39-40 KIADB Industrial Area, Electronic City, Phase II,

Hosur Road, Bangalore- 560 100, India

Cell: +91-9538631469  | Office Ph  : +91 80-66204422 (Extn: 398)  | Email ID: 
dipanka...@aurigene.com









This e-mail and any files transmitted with it are for the sole use of the 
intended recipient(s) and may contain confidential and privileged 
information.If you are not the intended recipient, please contact the sender by 
reply e-mail and destroy all copies of the original message.Any unauthorized 
review, use, disclosure, dissemination, forwarding,printing or copying of this 
email or any action taken in reliance on this e-mail is strictly prohibited and 
may be unlawful.

Visit us at http://www.aurigene.com


Re: [ccp4bb] How to reduce no. of overlaps

2012-03-07 Thread Frank von Delft
More accurately:  "adjustable pin".  Their *loops* are all adjustable 
(as opposed to Mitegen or MDL).


Those pins are not very useful if your crystal is already mounted.  And 
once you've bent that pin, it can't be stored again, because robots and 
tongs won't close around them.


They do have the advantage of not sometimes bending back, the way bent 
loops do.



On 07/03/2012 07:56, Zhijie Li wrote:

Hampton sells an adjustable mounted loop for this purpose.
http://hamptonresearch.com/product_detail.aspx?cid=24&sid=136&pid=385 
http://hamptonresearch.com/product_detail.aspx?cid=24&sid=136&pid=385>


*From:* Frank von Delft <mailto:frank.vonde...@sgc.ox.ac.uk>
*Sent:* Wednesday, March 07, 2012 1:26 AM
*To:* CCP4BB@JISCMAIL.AC.UK <mailto:CCP4BB@JISCMAIL.AC.UK>
*Subject:* Re: [ccp4bb] How to reduce no. of overlaps

You probably have to tilt your crystal, so that the long axis is 
parallel to the beam.  We do this routinely:  cut a plastic pipette 
tip to have a sharp point, then push the loop where it attaches to the 
pin, to bend the crystal itself.


You have to identify from your diffraction whether the long axis is 
pointing through the face or the edge of the loop.  As it's P6, 
chances are it's through the face, because long-axis P6 tends to make 
flat hexagons which lie flush with the face.  So you have to bend so 
the face of the loop upwards.


You'll have to practice this first, though, so put up an empty loop.  
Top tips:

* Don't breathe!  You'll blow the cryostream away.
* Bend the loop towards (rather than away from) the rim edge of 
the pin to which it's glued.

* Don't breathe!
* Practise practise practise.


Another thing:  most in-house sources allow you to reduce divergence 
of the beam.  You lose intensity, but no matter, just expose longer.  
That also improves overlap.


Cheers
phx



On 07/03/2012 04:56, Dipankar Manna wrote:


Dear Crystallographers,

I am working on a protein having SG P6, the cell parameters are a= 
79, b= 79, c= 325. The crystals are forming in big size and with very 
good shape. It also diffracting very well in Home source facility 
both in terms of resolution and intensity. But the only problem is 
the number of overlaps. Its showing much more than the good spots. As 
a result the completeness is showing maximum up to 65% even after 
collecting 180 degrees. I am unable to get a complete data. I tried 
with reducing the oscillation angel to 0.3 degree/0.5 degree but it 
did not improve that much. Please give me some suggestions.


Regards,

Dipankar

/Dipankar Manna/

*/Aurigene Discovery Technologies Limited,/*

/#39-40 KIADB Industrial Area, Electronic City, Phase II,/

/Hosur Road, Bangalore- 560 100, India/

/Cell: +91-9538631469 // | Office Ph  : +91 80-66204422 (Extn: 398)  
| Email ID: dipanka...@aurigene.com/





This e-mail and any files transmitted with it are for the sole use of 
the intended recipient(s) and may contain confidential and privileged 
information.If you are not the intended recipient, please contact the 
sender by reply e-mail and destroy all copies of the original 
message.Any unauthorized review, use, disclosure, dissemination, 
forwarding,printing or copying of this email or any action taken in 
reliance on this e-mail is strictly prohibited and may be unlawful.


Visit us at http://www.aurigene.com


Re: [ccp4bb] How to reduce no. of overlaps

2012-03-07 Thread Jim Pflugrath
In addition to reducing the beam divergence, you may wish to use a smaller beam 
size by using a smaller collimator or making the slits smaller.  A smaller 
crystal can also help to spatially separate the Bragg spots as can moving the 
detector closer to the crystal. Yes, closer to the crystal.  This is not 
intuitive, but arises since modern homelab beams are not parallel but are 
diverging from a focal point near the sample position.  It is just something 
else you may wish to try.

How you flashcool your sample will also have a large effect on the spot 
sizes/shapes.

Jim

>.
>Another thing:  most in-house sources allow you to reduce divergence of the 
>beam.  You lose intensity, >but no matter, just expose longer.  That also 
>improves overlap.
>
>Cheers
>phx



Re: [ccp4bb] How to reduce no. of overlaps

2012-03-06 Thread Zhijie Li
Hampton sells an adjustable mounted loop for this purpose.

 http://hamptonresearch.com/product_detail.aspx?cid=24&sid=136&pid=385


From: Frank von Delft 
Sent: Wednesday, March 07, 2012 1:26 AM
To: CCP4BB@JISCMAIL.AC.UK 
Subject: Re: [ccp4bb] How to reduce no. of overlaps


You probably have to tilt your crystal, so that the long axis is parallel to 
the beam.  We do this routinely:  cut a plastic pipette tip to have a sharp 
point, then push the loop where it attaches to the pin, to bend the crystal 
itself.

You have to identify from your diffraction whether the long axis is pointing 
through the face or the edge of the loop.  As it's P6, chances are it's through 
the face, because long-axis P6 tends to make flat hexagons which lie flush with 
the face.  So you have to bend so the face of the loop upwards.

You'll have to practice this first, though, so put up an empty loop.  Top tips: 
* Don't breathe!  You'll blow the cryostream away.
* Bend the loop towards (rather than away from) the rim edge of the pin to 
which it's glued.
* Don't breathe!
* Practise practise practise.


Another thing:  most in-house sources allow you to reduce divergence of the 
beam.  You lose intensity, but no matter, just expose longer.  That also 
improves overlap.

Cheers
phx



On 07/03/2012 04:56, Dipankar Manna wrote: 
  Dear Crystallographers,



  I am working on a protein having SG P6, the cell parameters are a= 79, b= 79, 
c= 325. The crystals are forming in big size and with very good shape. It also 
diffracting very well in Home source facility both in terms of resolution and 
intensity. But the only problem is the number of overlaps. Its showing much 
more than the good spots. As a result the completeness is showing maximum up to 
65% even after collecting 180 degrees. I am unable to get a complete data. I 
tried with reducing the oscillation angel to 0.3 degree/0.5 degree but it did 
not improve that much. Please give me some suggestions.



  Regards,



  Dipankar





  Dipankar Manna

  Aurigene Discovery Technologies Limited,

  #39-40 KIADB Industrial Area, Electronic City, Phase II,

  Hosur Road, Bangalore- 560 100, India

  Cell: +91-9538631469  | Office Ph  : +91 80-66204422 (Extn: 398)  | Email ID: 
dipanka...@aurigene.com







--

  This e-mail and any files transmitted with it are for the sole use of the 
intended recipient(s) and may contain confidential and privileged 
information.If you are not the intended recipient, please contact the sender by 
reply e-mail and destroy all copies of the original message.Any unauthorized 
review, use, disclosure, dissemination, forwarding,printing or copying of this 
email or any action taken in reliance on this e-mail is strictly prohibited and 
may be unlawful.

  Visit us at http://www.aurigene.com


Re: [ccp4bb] How to reduce no. of overlaps

2012-03-06 Thread Frank von Delft

Yes... quite.  Thanks!

(Beam, rotation axis... how can you expect me to keep all these 
new-fangled inventions apart?!)




On 07/03/2012 07:33, VAN RAAIJ , MARK JOHAN wrote:

typo correction, you'll want the long axis parallel to the rotation
axis, not to the beam.
Mark

Quoting Frank von Delft:


You probably have to tilt your crystal, so that the long axis is
parallel to the beam.  We do this routinely:  cut a plastic pipette
tip to have a sharp point, then push the loop where it attaches to
the pin, to bend the crystal itself.

You have to identify from your diffraction whether the long axis is
pointing through the face or the edge of the loop.  As it's P6,
chances are it's through the face, because long-axis P6 tends to
make flat hexagons which lie flush with the face.  So you have to
bend so the face of the loop upwards.

You'll have to practice this first, though, so put up an empty loop.
  Top tips:
 * Don't breathe!  You'll blow the cryostream away.
 * Bend the loop towards (rather than away from) the rim edge of
the pin to which it's glued.
 * Don't breathe!
 * Practise practise practise.


Another thing:  most in-house sources allow you to reduce divergence
of the beam.  You lose intensity, but no matter, just expose longer.
  That also improves overlap.

Cheers
phx



On 07/03/2012 04:56, Dipankar Manna wrote:

Dear Crystallographers,

I am working on a protein having SG P6, the cell parameters are a=
79, b= 79, c= 325. The crystals are forming in big size and with
very good shape. It also diffracting very well in Home source
facility both in terms of resolution and intensity. But the only
problem is the number of overlaps. Its showing much more than the
good spots. As a result the completeness is showing maximum up to
65% even after collecting 180 degrees. I am unable to get a
complete data. I tried with reducing the oscillation angel to 0.3
degree/0.5 degree but it did not improve that much. Please give me
some suggestions.

Regards,

Dipankar

/Dipankar Manna/

*/Aurigene Discovery Technologies Limited,/*

/#39-40 KIADB Industrial Area, Electronic City, Phase II,/

/Hosur Road, Bangalore- 560 100, India/

/Cell: +91-9538631469 // | Office Ph  : +91 80-66204422 (Extn: 398)
  | Email ID: dipanka...@aurigene.com/




This e-mail and any files transmitted with it are for the sole use
of the intended recipient(s) and may contain confidential and
privileged information.If you are not the intended recipient,
please contact the sender by reply e-mail and destroy all copies of
the original message.Any unauthorized review, use, disclosure,
dissemination, forwarding,printing or copying of this email or any
action taken in reliance on this e-mail is strictly prohibited and
may be unlawful.

Visit us at http://www.aurigene.com



Mark J van Raaij
Laboratorio M-4
Dpto de Estructura de Macromoléculas
Centro Nacional de Biotecnología - CSIC
c/Darwin 3, Campus Cantoblanco
28049 Madrid
tel. 91 585 4616
email: mjvanra...@cnb.csic.es



Re: [ccp4bb] How to reduce no. of overlaps

2012-03-06 Thread VAN RAAIJ , MARK JOHAN
typo correction, you'll want the long axis parallel to the rotation  
axis, not to the beam.

Mark

Quoting Frank von Delft:

You probably have to tilt your crystal, so that the long axis is  
parallel to the beam.  We do this routinely:  cut a plastic pipette  
tip to have a sharp point, then push the loop where it attaches to  
the pin, to bend the crystal itself.


You have to identify from your diffraction whether the long axis is  
pointing through the face or the edge of the loop.  As it's P6,  
chances are it's through the face, because long-axis P6 tends to  
make flat hexagons which lie flush with the face.  So you have to  
bend so the face of the loop upwards.


You'll have to practice this first, though, so put up an empty loop.  
 Top tips:

* Don't breathe!  You'll blow the cryostream away.
* Bend the loop towards (rather than away from) the rim edge of  
the pin to which it's glued.

* Don't breathe!
* Practise practise practise.


Another thing:  most in-house sources allow you to reduce divergence  
of the beam.  You lose intensity, but no matter, just expose longer.  
 That also improves overlap.


Cheers
phx



On 07/03/2012 04:56, Dipankar Manna wrote:


Dear Crystallographers,

I am working on a protein having SG P6, the cell parameters are a=  
79, b= 79, c= 325. The crystals are forming in big size and with  
very good shape. It also diffracting very well in Home source  
facility both in terms of resolution and intensity. But the only  
problem is the number of overlaps. Its showing much more than the  
good spots. As a result the completeness is showing maximum up to  
65% even after collecting 180 degrees. I am unable to get a  
complete data. I tried with reducing the oscillation angel to 0.3  
degree/0.5 degree but it did not improve that much. Please give me  
some suggestions.


Regards,

Dipankar

/Dipankar Manna/

*/Aurigene Discovery Technologies Limited,/*

/#39-40 KIADB Industrial Area, Electronic City, Phase II,/

/Hosur Road, Bangalore- 560 100, India/

/Cell: +91-9538631469 // | Office Ph  : +91 80-66204422 (Extn: 398)  
 | Email ID: dipanka...@aurigene.com/





This e-mail and any files transmitted with it are for the sole use  
of the intended recipient(s) and may contain confidential and  
privileged information.If you are not the intended recipient,  
please contact the sender by reply e-mail and destroy all copies of  
the original message.Any unauthorized review, use, disclosure,  
dissemination, forwarding,printing or copying of this email or any  
action taken in reliance on this e-mail is strictly prohibited and  
may be unlawful.


Visit us at http://www.aurigene.com






Mark J van Raaij
Laboratorio M-4
Dpto de Estructura de Macromoléculas
Centro Nacional de Biotecnología - CSIC
c/Darwin 3, Campus Cantoblanco
28049 Madrid
tel. 91 585 4616
email: mjvanra...@cnb.csic.es


Re: [ccp4bb] How to reduce no. of overlaps

2012-03-06 Thread Frank von Delft
You probably have to tilt your crystal, so that the long axis is 
parallel to the beam.  We do this routinely:  cut a plastic pipette tip 
to have a sharp point, then push the loop where it attaches to the pin, 
to bend the crystal itself.


You have to identify from your diffraction whether the long axis is 
pointing through the face or the edge of the loop.  As it's P6, chances 
are it's through the face, because long-axis P6 tends to make flat 
hexagons which lie flush with the face.  So you have to bend so the face 
of the loop upwards.


You'll have to practice this first, though, so put up an empty loop.  
Top tips:

* Don't breathe!  You'll blow the cryostream away.
* Bend the loop towards (rather than away from) the rim edge of the 
pin to which it's glued.

* Don't breathe!
* Practise practise practise.


Another thing:  most in-house sources allow you to reduce divergence of 
the beam.  You lose intensity, but no matter, just expose longer.  That 
also improves overlap.


Cheers
phx



On 07/03/2012 04:56, Dipankar Manna wrote:


Dear Crystallographers,

I am working on a protein having SG P6, the cell parameters are a= 79, 
b= 79, c= 325. The crystals are forming in big size and with very good 
shape. It also diffracting very well in Home source facility both in 
terms of resolution and intensity. But the only problem is the number 
of overlaps. Its showing much more than the good spots. As a result 
the completeness is showing maximum up to 65% even after collecting 
180 degrees. I am unable to get a complete data. I tried with reducing 
the oscillation angel to 0.3 degree/0.5 degree but it did not improve 
that much. Please give me some suggestions.


Regards,

Dipankar

/Dipankar Manna/

*/Aurigene Discovery Technologies Limited,/*

/#39-40 KIADB Industrial Area, Electronic City, Phase II,/

/Hosur Road, Bangalore- 560 100, India/

/Cell: +91-9538631469 // | Office Ph  : +91 80-66204422 (Extn: 398)  | 
Email ID: dipanka...@aurigene.com/





This e-mail and any files transmitted with it are for the sole use of 
the intended recipient(s) and may contain confidential and privileged 
information.If you are not the intended recipient, please contact the 
sender by reply e-mail and destroy all copies of the original 
message.Any unauthorized review, use, disclosure, dissemination, 
forwarding,printing or copying of this email or any action taken in 
reliance on this e-mail is strictly prohibited and may be unlawful.


Visit us at http://www.aurigene.com


[ccp4bb] How to reduce no. of overlaps

2012-03-06 Thread Dipankar Manna
Dear Crystallographers,

I am working on a protein having SG P6, the cell parameters are a= 79, b= 79, 
c= 325. The crystals are forming in big size and with very good shape. It also 
diffracting very well in Home source facility both in terms of resolution and 
intensity. But the only problem is the number of overlaps. Its showing much 
more than the good spots. As a result the completeness is showing maximum up to 
65% even after collecting 180 degrees. I am unable to get a complete data. I 
tried with reducing the oscillation angel to 0.3 degree/0.5 degree but it did 
not improve that much. Please give me some suggestions.

Regards,

Dipankar


Dipankar Manna
Aurigene Discovery Technologies Limited,
#39-40 KIADB Industrial Area, Electronic City, Phase II,
Hosur Road, Bangalore- 560 100, India
Cell: +91-9538631469  | Office Ph  : +91 80-66204422 (Extn: 398)  | Email ID: 
dipanka...@aurigene.com





This e-mail and any files transmitted with it are for the sole use of the 
intended recipient(s) and may contain confidential and privileged 
information.If you are not the intended recipient, please contact the sender by 
reply e-mail and destroy all copies of the original message.Any unauthorized 
review, use, disclosure, dissemination, forwarding,printing or copying of this 
email or any action taken in reliance on this e-mail is strictly prohibited and 
may be unlawful.

Visit us at http://www.aurigene.com