Re: [ccp4bb] Crystal rescue

2010-01-27 Thread Frederic VELLIEUX
Hi Zhiyi,

There are two ways you can go about this (RT data collection). Either mount the 
crystals in capillaries (not all beam lines at synchrotrons have sufficient 
space on the goniometer setup to allow data collection on capillary-mounted 
crystals) or use a Humidifier (the latter is briefly discussed, for example, on 
web page 
http://www.esrf.eu/UsersAndScience/Experiments/MX/Cryobench/Instruments/microspec
 ).

If nothing is done, the crystals will rapidly dry up and die.

Fred.

 Message du 27/01/10 08:53
 De : Zhiyi Wei 
 A : CCP4BB@JISCMAIL.AC.UK
 Copie à : 
 Objet : Re: [ccp4bb] Crystal rescue
 
 
 Thanks for so many quick responses!
 
 Actually, I have test several different cryo-protectants, including
 glycerol, EG, and PEG400. I did not see much differences between these
 cryo conditions. So, I choose glycerol.
 
 I would like to test my crystals in RT. But I don't know how to do
 this. Just mount crystal to the X-ray machine without cryo stream? Or
 I should use capillaries?
 
 Zhiyi
 
 On Wed, Jan 27, 2010 at 5:45 AM, Kelly Daughtry  wrote:
  Tascimate can be used as the cryo as well. I have had experience with
  crystals in similar condition and moved the crystals to a 20%
  increased Tascimate solution and they froze well.
 
  I agree with Ezra, room temperature mount your crystal before
  freezing. It is the only way to know the true problem.
 
 
  Kelly
  ***
  Kelly Daughtry
  PhD Candidate
  Department of Physiology and Biophysics
  Boston University School of Medicine
  590 Commonwealth Ave
  R 390
  Boston MA, 02215
  (P) 617-358-5548
  ***
 
 
 
  On Tue, Jan 26, 2010 at 1:25 PM, Marcus Winter
   wrote:
  Dear Zhiyi,
 
 
  Ezra is exactly right, of course. The Oxford Diffraction PX Scanner
  system can assess the diffraction qualities of (putative) protein
  crystals in situ - in the crystallisation plate. So, directly, you
  would discover if your 'big and beautiful' crystals actually diffract
  well... in their mother liquor under ambient conditions and before the
  addition of any cryo-protect. Do you have a friend or neighbour with
  a PX Scanner ? If not, please feel most welcome to contact
  Oxford Diffraction: we would be pleased to assist if at all possible.
 
 
  Good Luck and Best Wishes,
 
  Marcus Winter.
 
  www.oxford-diffraction.com
 
 
 
 
  -Original Message-
  From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of
  Ezra Peisach
  Sent: 26 January 2010 16:01
  To: CCP4BB@JISCMAIL.AC.UK
  Subject: Re: [ccp4bb] Crystal rescue
 
  First you need to establish if it is your cryo conditions or the
  crystals. Depending where you are - they might have the equipment to do
 
  a wet mount - without freezing. Yes the crystal will not last - but
  then you know if the problem is in the
  crystal. If it is - you need better crystals. If it is the cryo - you
  need to work on that. Tacsimate is mixture of alot of different
  compounds - but the smears are too close together to be a small salt
  crystal on top...
 
  Good luck,
 
  Ezra
 
  On 1/26/2010 10:42 AM, Zhiyi Wei wrote:
  Dear all,
 
  I got a problem with my crystals. I have two total different proteins
  that both can be crystallized in the condition with PEG3350 and
  Tacsimate
  (although the concentrations are different) with different shapes. The
  crystals look big and beautiful. However, when I test them in
  synchrotron,
  both of these two types of crystals showed poor diffractions. As
  showed in
  the attached diffraction image, the diffraction is up to ~4 A but
  smear in
  one direction while8 A in the other direction. The interesting thing
  is
  that the diffraction pattern is similar for all crystals (from two
  different
  proteins) that I tested without exception although they belong to
  different
  space groups. So, I wonder whether these kind of pattern is caused by
  Tacsimate (I don't know what it is) and how to rescue these crystals.
  Any
  suggestions or comments?
 
  Thanks a lot!
 
  Best,
  Zhiyi
 
 
 
 
 


Re: [ccp4bb] Crystal rescue

2010-01-27 Thread Flip Hoedemaeker

Zhiyi,

You can use a thin cap over your cryo loop, just put a drop of mother 
liquor in the top, place over the loop and make it airtight at the base. 
 Not sure who sells these things though, I guess you can make it from a 
capillary too. Then remove the cryo stream or put it at a temp above 
freezing, say 253K.


Flip

Zhiyi Wei wrote:

Thanks for so many quick responses!

Actually, I have test several different cryo-protectants, including
glycerol, EG, and PEG400. I did not see much differences between these
cryo conditions. So, I choose glycerol.

I would like to test my crystals in RT. But I don't know how to do
this. Just mount crystal to the X-ray machine without cryo stream? Or
I should use capillaries?

Zhiyi

On Wed, Jan 27, 2010 at 5:45 AM, Kelly Daughtry kddau...@bu.edu wrote:

Tascimate can be used as the cryo as well. I have had experience with
crystals in similar condition and moved the crystals to a 20%
increased Tascimate solution and they froze well.

I agree with Ezra, room temperature mount your crystal before
freezing. It is the only way to know the true problem.


Kelly
***
Kelly Daughtry
PhD Candidate
Department of Physiology and Biophysics
Boston University School of Medicine
590 Commonwealth Ave
R 390
Boston MA, 02215
(P) 617-358-5548
***



On Tue, Jan 26, 2010 at 1:25 PM, Marcus Winter
marcus.win...@oxford-diffraction.com wrote:

Dear Zhiyi,


Ezra is exactly right, of course.  The Oxford Diffraction PX Scanner
system can assess the diffraction qualities of (putative) protein
crystals in situ - in the crystallisation plate.  So, directly, you
would discover if your 'big and beautiful' crystals actually diffract
well... in their mother liquor under ambient conditions and before the
addition of any cryo-protect.  Do you have a friend or neighbour with
a PX Scanner ?  If not, please feel most welcome to contact
Oxford Diffraction: we would be pleased to assist if at all possible.


Good Luck and Best Wishes,

Marcus Winter.

www.oxford-diffraction.com




-Original Message-
From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of
Ezra Peisach
Sent: 26 January 2010 16:01
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Crystal rescue

First you need to establish if it is your cryo conditions or the
crystals.  Depending where you are - they might have the equipment to do

a wet mount - without freezing.  Yes the crystal will not last - but
then you know if the problem is in the
crystal.  If it is - you need better crystals.  If it is the cryo - you
need to work on that.  Tacsimate is mixture of alot of different
compounds - but the smears are too close together to be a small salt
crystal on top...

Good luck,

Ezra

On 1/26/2010 10:42 AM, Zhiyi Wei wrote:

Dear all,

I got a problem with my crystals. I have two total different proteins
that both can be crystallized in the condition with PEG3350 and

Tacsimate

(although the concentrations are different) with different shapes. The
crystals look big and beautiful. However, when I test them in

synchrotron,

both of these two types of crystals showed poor diffractions. As

showed in

the attached diffraction image, the diffraction is up to ~4 A but

smear in

one direction while8 A in the other direction. The interesting thing

is

that the diffraction pattern is similar for all crystals (from two

different

proteins) that I tested without exception although they belong to

different

space groups. So, I wonder whether these kind of pattern is caused by
Tacsimate (I don't know what it is) and how to rescue these crystals.

Any

suggestions or comments?

Thanks a lot!

Best,
Zhiyi





Re: [ccp4bb] Crystal rescue

2010-01-27 Thread Matthew BOWLER

Hi Fred and Zhiyi,
the EMBL and ESRF have also developed a dehydration device that can 
also be used for room temperature data collection, see the website here:


http://www.esrf.fr/UsersAndScience/Experiments/MX/About_our_beamlines/ID14-2/HC1b

http://scripts.iucr.org/cgi-bin/paper?S0907444909037822

Cheers, Matt.




Frederic VELLIEUX wrote:

Hi Zhiyi,

There are two ways you can go about this (RT data collection). Either mount the 
crystals in capillaries (not all beam lines at synchrotrons have sufficient 
space on the goniometer setup to allow data collection on capillary-mounted 
crystals) or use a Humidifier (the latter is briefly discussed, for example, on 
web page 
http://www.esrf.eu/UsersAndScience/Experiments/MX/Cryobench/Instruments/microspec
 ).

If nothing is done, the crystals will rapidly dry up and die.

Fred.

  

Message du 27/01/10 08:53
De : Zhiyi Wei 
A : CCP4BB@JISCMAIL.AC.UK
Copie à : 
Objet : Re: [ccp4bb] Crystal rescue



Thanks for so many quick responses!

Actually, I have test several different cryo-protectants, including
glycerol, EG, and PEG400. I did not see much differences between these
cryo conditions. So, I choose glycerol.

I would like to test my crystals in RT. But I don't know how to do
this. Just mount crystal to the X-ray machine without cryo stream? Or
I should use capillaries?

Zhiyi

On Wed, Jan 27, 2010 at 5:45 AM, Kelly Daughtry  wrote:


Tascimate can be used as the cryo as well. I have had experience with
crystals in similar condition and moved the crystals to a 20%
increased Tascimate solution and they froze well.

I agree with Ezra, room temperature mount your crystal before
freezing. It is the only way to know the true problem.


Kelly
***
Kelly Daughtry
PhD Candidate
Department of Physiology and Biophysics
Boston University School of Medicine
590 Commonwealth Ave
R 390
Boston MA, 02215
(P) 617-358-5548
***



On Tue, Jan 26, 2010 at 1:25 PM, Marcus Winter
 wrote:
  

Dear Zhiyi,


Ezra is exactly right, of course. The Oxford Diffraction PX Scanner
system can assess the diffraction qualities of (putative) protein
crystals in situ - in the crystallisation plate. So, directly, you
would discover if your 'big and beautiful' crystals actually diffract
well... in their mother liquor under ambient conditions and before the
addition of any cryo-protect. Do you have a friend or neighbour with
a PX Scanner ? If not, please feel most welcome to contact
Oxford Diffraction: we would be pleased to assist if at all possible.


Good Luck and Best Wishes,

Marcus Winter.

www.oxford-diffraction.com




-Original Message-
From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of
Ezra Peisach
Sent: 26 January 2010 16:01
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Crystal rescue

First you need to establish if it is your cryo conditions or the
crystals. Depending where you are - they might have the equipment to do

a wet mount - without freezing. Yes the crystal will not last - but
then you know if the problem is in the
crystal. If it is - you need better crystals. If it is the cryo - you
need to work on that. Tacsimate is mixture of alot of different
compounds - but the smears are too close together to be a small salt
crystal on top...

Good luck,

Ezra

On 1/26/2010 10:42 AM, Zhiyi Wei wrote:


Dear all,

I got a problem with my crystals. I have two total different proteins
that both can be crystallized in the condition with PEG3350 and
  

Tacsimate


(although the concentrations are different) with different shapes. The
crystals look big and beautiful. However, when I test them in
  

synchrotron,


both of these two types of crystals showed poor diffractions. As
  

showed in


the attached diffraction image, the diffraction is up to ~4 A but
  

smear in


one direction while8 A in the other direction. The interesting thing
  

is


that the diffraction pattern is similar for all crystals (from two
  

different


proteins) that I tested without exception although they belong to
  

different


space groups. So, I wonder whether these kind of pattern is caused by
Tacsimate (I don't know what it is) and how to rescue these crystals.
  

Any


suggestions or comments?

Thanks a lot!

Best,
Zhiyi

  



  


--
Matthew Bowler
Structural Biology Group
European Synchrotron Radiation Facility
B.P. 220, 6 rue Jules Horowitz
F-38043 GRENOBLE CEDEX
FRANCE
===
Tel: +33 (0) 4.76.88.29.28
Fax: +33 (0) 4.76.88.29.04

http://www.esrf.fr/UsersAndScience/Experiments/MX/
=== 



Re: [ccp4bb] Crystal rescue

2010-01-27 Thread Ezra Peisach
MiTeGen (www.mitegen.com) sells a RT device  - plastic capillary to put 
over loop... I do not know how well it would work for a long data 
collection - but people here have used it to evaluate their crystals


Ezra


On 1/27/2010 3:58 AM, Flip Hoedemaeker wrote:

Zhiyi,

You can use a thin cap over your cryo loop, just put a drop of mother 
liquor in the top, place over the loop and make it airtight at the 
base.  Not sure who sells these things though, I guess you can make it 
from a capillary too. Then remove the cryo stream or put it at a temp 
above freezing, say 253K.


Flip

Zhiyi Wei wrote:

Thanks for so many quick responses!

Actually, I have test several different cryo-protectants, including
glycerol, EG, and PEG400. I did not see much differences between these
cryo conditions. So, I choose glycerol.

I would like to test my crystals in RT. But I don't know how to do
this. Just mount crystal to the X-ray machine without cryo stream? Or
I should use capillaries?

Zhiyi

On Wed, Jan 27, 2010 at 5:45 AM, Kelly Daughtry kddau...@bu.edu wrote:

Tascimate can be used as the cryo as well. I have had experience with
crystals in similar condition and moved the crystals to a 20%
increased Tascimate solution and they froze well.

I agree with Ezra, room temperature mount your crystal before
freezing. It is the only way to know the true problem.


Kelly
***
Kelly Daughtry
PhD Candidate
Department of Physiology and Biophysics
Boston University School of Medicine
590 Commonwealth Ave
R 390
Boston MA, 02215
(P) 617-358-5548
***



On Tue, Jan 26, 2010 at 1:25 PM, Marcus Winter
marcus.win...@oxford-diffraction.com wrote:

Dear Zhiyi,


Ezra is exactly right, of course.  The Oxford Diffraction PX Scanner
system can assess the diffraction qualities of (putative) protein
crystals in situ - in the crystallisation plate.  So, directly, you
would discover if your 'big and beautiful' crystals actually diffract
well... in their mother liquor under ambient conditions and before the
addition of any cryo-protect.  Do you have a friend or neighbour with
a PX Scanner ?  If not, please feel most welcome to contact
Oxford Diffraction: we would be pleased to assist if at all possible.


Good Luck and Best Wishes,

Marcus Winter.

www.oxford-diffraction.com




-Original Message-
From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of
Ezra Peisach
Sent: 26 January 2010 16:01
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Crystal rescue

First you need to establish if it is your cryo conditions or the
crystals.  Depending where you are - they might have the equipment 
to do


a wet mount - without freezing.  Yes the crystal will not last - but
then you know if the problem is in the
crystal.  If it is - you need better crystals.  If it is the cryo - 
you

need to work on that.  Tacsimate is mixture of alot of different
compounds - but the smears are too close together to be a small salt
crystal on top...

Good luck,

Ezra

On 1/26/2010 10:42 AM, Zhiyi Wei wrote:

Dear all,

I got a problem with my crystals. I have two total different proteins
that both can be crystallized in the condition with PEG3350 and

Tacsimate
(although the concentrations are different) with different shapes. 
The

crystals look big and beautiful. However, when I test them in

synchrotron,

both of these two types of crystals showed poor diffractions. As

showed in

the attached diffraction image, the diffraction is up to ~4 A but

smear in

one direction while8 A in the other direction. The interesting thing

is

that the diffraction pattern is similar for all crystals (from two

different

proteins) that I tested without exception although they belong to

different

space groups. So, I wonder whether these kind of pattern is caused by
Tacsimate (I don't know what it is) and how to rescue these crystals.

Any

suggestions or comments?

Thanks a lot!

Best,
Zhiyi





Re: [ccp4bb] Crystal rescue

2010-01-27 Thread Clayton, Gina Martyn
 
MiTeGen have instructions on their website for using their system. In
practice I usually put a little silicon grease at the base of the
capillary as sometimes the capillaries fall off plus you need reasonable
steady hand eye coordination to put the capillary over the loop...I have
found MiTeGen RT system very useful.

Good luck!



-Original Message-
From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of
Ezra Peisach
Sent: Wednesday, January 27, 2010 7:41 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Crystal rescue

MiTeGen (www.mitegen.com) sells a RT device  - plastic capillary to put 
over loop... I do not know how well it would work for a long data 
collection - but people here have used it to evaluate their crystals

Ezra


On 1/27/2010 3:58 AM, Flip Hoedemaeker wrote:
 Zhiyi,

 You can use a thin cap over your cryo loop, just put a drop of mother 
 liquor in the top, place over the loop and make it airtight at the 
 base.  Not sure who sells these things though, I guess you can make it

 from a capillary too. Then remove the cryo stream or put it at a temp 
 above freezing, say 253K.

 Flip

 Zhiyi Wei wrote:
 Thanks for so many quick responses!

 Actually, I have test several different cryo-protectants, including
 glycerol, EG, and PEG400. I did not see much differences between
these
 cryo conditions. So, I choose glycerol.

 I would like to test my crystals in RT. But I don't know how to do
 this. Just mount crystal to the X-ray machine without cryo stream? Or
 I should use capillaries?

 Zhiyi

 On Wed, Jan 27, 2010 at 5:45 AM, Kelly Daughtry kddau...@bu.edu
wrote:
 Tascimate can be used as the cryo as well. I have had experience
with
 crystals in similar condition and moved the crystals to a 20%
 increased Tascimate solution and they froze well.

 I agree with Ezra, room temperature mount your crystal before
 freezing. It is the only way to know the true problem.


 Kelly
 ***
 Kelly Daughtry
 PhD Candidate
 Department of Physiology and Biophysics
 Boston University School of Medicine
 590 Commonwealth Ave
 R 390
 Boston MA, 02215
 (P) 617-358-5548
 ***



 On Tue, Jan 26, 2010 at 1:25 PM, Marcus Winter
 marcus.win...@oxford-diffraction.com wrote:
 Dear Zhiyi,


 Ezra is exactly right, of course.  The Oxford Diffraction PX
Scanner
 system can assess the diffraction qualities of (putative) protein
 crystals in situ - in the crystallisation plate.  So, directly, you
 would discover if your 'big and beautiful' crystals actually
diffract
 well... in their mother liquor under ambient conditions and before
the
 addition of any cryo-protect.  Do you have a friend or neighbour
with
 a PX Scanner ?  If not, please feel most welcome to contact
 Oxford Diffraction: we would be pleased to assist if at all
possible.


 Good Luck and Best Wishes,

 Marcus Winter.

 www.oxford-diffraction.com




 -Original Message-
 From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf
Of
 Ezra Peisach
 Sent: 26 January 2010 16:01
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: Re: [ccp4bb] Crystal rescue

 First you need to establish if it is your cryo conditions or the
 crystals.  Depending where you are - they might have the equipment 
 to do

 a wet mount - without freezing.  Yes the crystal will not last -
but
 then you know if the problem is in the
 crystal.  If it is - you need better crystals.  If it is the cryo -

 you
 need to work on that.  Tacsimate is mixture of alot of different
 compounds - but the smears are too close together to be a small
salt
 crystal on top...

 Good luck,

 Ezra

 On 1/26/2010 10:42 AM, Zhiyi Wei wrote:
 Dear all,

 I got a problem with my crystals. I have two total different
proteins
 that both can be crystallized in the condition with PEG3350 and
 Tacsimate
 (although the concentrations are different) with different shapes.

 The
 crystals look big and beautiful. However, when I test them in
 synchrotron,
 both of these two types of crystals showed poor diffractions. As
 showed in
 the attached diffraction image, the diffraction is up to ~4 A but
 smear in
 one direction while8 A in the other direction. The interesting
thing
 is
 that the diffraction pattern is similar for all crystals (from two
 different
 proteins) that I tested without exception although they belong to
 different
 space groups. So, I wonder whether these kind of pattern is caused
by
 Tacsimate (I don't know what it is) and how to rescue these
crystals.
 Any
 suggestions or comments?

 Thanks a lot!

 Best,
 Zhiyi


Notice:  This e-mail message, together with any attachments, contains 
information of Merck  Co., Inc. (One Merck Drive, Whitehouse Station, New 
Jersey, USA 08889), and/or its affiliates Direct contact information for 
affiliates is available at http://www.merck.com/contact/contacts.html) that may 
be confidential, proprietary copyrighted and/or legally

Re: [ccp4bb] Crystal rescue

2010-01-27 Thread Mark J. van Raaij
in my opinion one should always first check RT diffraction and we use  
the Mitegen plastic cap over the loop. We find the solutions slowly  
dry out (6-18 hours), which can be enough to get a useful RT dataset  
on the rotating anode.
If you like, the crystal can also be recovered after RT diffraction  
checking for addition of cryosolution and freezing.

Mark


Quoting Clayton, Gina Martyn gina_clay...@merck.com:



MiTeGen have instructions on their website for using their system. In
practice I usually put a little silicon grease at the base of the
capillary as sometimes the capillaries fall off plus you need reasonable
steady hand eye coordination to put the capillary over the loop...I have
found MiTeGen RT system very useful.

Good luck!



-Original Message-
From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of
Ezra Peisach
Sent: Wednesday, January 27, 2010 7:41 AM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Crystal rescue

MiTeGen (www.mitegen.com) sells a RT device  - plastic capillary to put
over loop... I do not know how well it would work for a long data
collection - but people here have used it to evaluate their crystals

Ezra


On 1/27/2010 3:58 AM, Flip Hoedemaeker wrote:

Zhiyi,

You can use a thin cap over your cryo loop, just put a drop of mother
liquor in the top, place over the loop and make it airtight at the
base.  Not sure who sells these things though, I guess you can make it



from a capillary too. Then remove the cryo stream or put it at a temp
above freezing, say 253K.

Flip

Zhiyi Wei wrote:

Thanks for so many quick responses!

Actually, I have test several different cryo-protectants, including
glycerol, EG, and PEG400. I did not see much differences between

these

cryo conditions. So, I choose glycerol.

I would like to test my crystals in RT. But I don't know how to do
this. Just mount crystal to the X-ray machine without cryo stream? Or
I should use capillaries?

Zhiyi

On Wed, Jan 27, 2010 at 5:45 AM, Kelly Daughtry kddau...@bu.edu

wrote:

Tascimate can be used as the cryo as well. I have had experience

with

crystals in similar condition and moved the crystals to a 20%
increased Tascimate solution and they froze well.

I agree with Ezra, room temperature mount your crystal before
freezing. It is the only way to know the true problem.


Kelly
***
Kelly Daughtry
PhD Candidate
Department of Physiology and Biophysics
Boston University School of Medicine
590 Commonwealth Ave
R 390
Boston MA, 02215
(P) 617-358-5548
***



On Tue, Jan 26, 2010 at 1:25 PM, Marcus Winter
marcus.win...@oxford-diffraction.com wrote:

Dear Zhiyi,


Ezra is exactly right, of course.  The Oxford Diffraction PX

Scanner

system can assess the diffraction qualities of (putative) protein
crystals in situ - in the crystallisation plate.  So, directly, you
would discover if your 'big and beautiful' crystals actually

diffract

well... in their mother liquor under ambient conditions and before

the

addition of any cryo-protect.  Do you have a friend or neighbour

with

a PX Scanner ?  If not, please feel most welcome to contact
Oxford Diffraction: we would be pleased to assist if at all

possible.



Good Luck and Best Wishes,

Marcus Winter.

www.oxford-diffraction.com




-Original Message-
From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf

Of

Ezra Peisach
Sent: 26 January 2010 16:01
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] Crystal rescue

First you need to establish if it is your cryo conditions or the
crystals.  Depending where you are - they might have the equipment
to do

a wet mount - without freezing.  Yes the crystal will not last -

but

then you know if the problem is in the
crystal.  If it is - you need better crystals.  If it is the cryo -



you
need to work on that.  Tacsimate is mixture of alot of different
compounds - but the smears are too close together to be a small

salt

crystal on top...

Good luck,

Ezra

On 1/26/2010 10:42 AM, Zhiyi Wei wrote:

Dear all,

I got a problem with my crystals. I have two total different

proteins

that both can be crystallized in the condition with PEG3350 and

Tacsimate

(although the concentrations are different) with different shapes.



The
crystals look big and beautiful. However, when I test them in

synchrotron,

both of these two types of crystals showed poor diffractions. As

showed in

the attached diffraction image, the diffraction is up to ~4 A but

smear in

one direction while8 A in the other direction. The interesting

thing

is

that the diffraction pattern is similar for all crystals (from two

different

proteins) that I tested without exception although they belong to

different

space groups. So, I wonder whether these kind of pattern is caused

by

Tacsimate (I don't know what it is) and how to rescue these

crystals.

Any

suggestions or comments?

Thanks a lot

Re: [ccp4bb] Crystal rescue

2010-01-27 Thread Ho Leung Ng
Hi Zhiyi,

I think the easiest way is to mount your crystal in an
air-impermeable, viscous oil like Paratone in a loop. Of course, your
crystals may not tolerate oil.


ho


Re: [ccp4bb] Crystal rescue

2010-01-26 Thread Ezra Peisach
First you need to establish if it is your cryo conditions or the 
crystals.  Depending where you are - they might have the equipment to do 
a wet mount - without freezing.  Yes the crystal will not last - but 
then you know if the problem is in the
crystal.  If it is - you need better crystals.  If it is the cryo - you 
need to work on that.  Tacsimate is mixture of alot of different 
compounds - but the smears are too close together to be a small salt 
crystal on top...


Good luck,

Ezra

On 1/26/2010 10:42 AM, Zhiyi Wei wrote:

Dear all,

I got a problem with my crystals. I have two total different proteins
that both can be crystallized in the condition with PEG3350 and Tacsimate
(although the concentrations are different) with different shapes. The
crystals look big and beautiful. However, when I test them in synchrotron,
both of these two types of crystals showed poor diffractions. As showed in
the attached diffraction image, the diffraction is up to ~4 A but smear in
one direction while8 A in the other direction. The interesting thing is
that the diffraction pattern is similar for all crystals (from two different
proteins) that I tested without exception although they belong to different
space groups. So, I wonder whether these kind of pattern is caused by
Tacsimate (I don't know what it is) and how to rescue these crystals. Any
suggestions or comments?

Thanks a lot!

Best,
Zhiyi
   


Re: [ccp4bb] Crystal rescue

2010-01-26 Thread Ed Pozharski
http://hamptonresearch.com/documents/product/hr000175_what_is_tacsimate_new.pdf

turns up once you use google's I'm feeling lucky button.

On Tue, 2010-01-26 at 15:42 +, Zhiyi Wei wrote:
 Dear all,
 
 I got a problem with my crystals. I have two total different proteins
 that both can be crystallized in the condition with PEG3350 and Tacsimate
 (although the concentrations are different) with different shapes. The
 crystals look big and beautiful. However, when I test them in synchrotron,
 both of these two types of crystals showed poor diffractions. As showed in
 the attached diffraction image, the diffraction is up to ~4 A but smear in
 one direction while 8 A in the other direction. The interesting thing is
 that the diffraction pattern is similar for all crystals (from two different
 proteins) that I tested without exception although they belong to different
 space groups. So, I wonder whether these kind of pattern is caused by
 Tacsimate (I don't know what it is) and how to rescue these crystals. Any
 suggestions or comments?
 
 Thanks a lot!
 
 Best,
 Zhiyi


-- 
Edwin Pozharski, PhD, Assistant Professor
University of Maryland, Baltimore
--
When the Way is forgotten duty and justice appear;
Then knowledge and wisdom are born along with hypocrisy.
When harmonious relationships dissolve then respect and devotion arise;
When a nation falls to chaos then loyalty and patriotism are born.
--   / Lao Tse /


Re: [ccp4bb] Crystal rescue

2010-01-26 Thread Zhiyi Wei
I forgot to mention a phenomenon when I did crystal mounting. I found
that if open a coverslip, phase separation will appear in drops in
minutes and damage to the crystals.

And I have tried additive screen. The optimized crystals are really
big (a few hundred microns) with sharp edge but diffraction pattern is
similar.

On Tue, Jan 26, 2010 at 10:42 AM, Zhiyi Wei ustcwebri...@gmail.com wrote:
 Dear all,

 I got a problem with my crystals. I have two total different proteins
 that both can be crystallized in the condition with PEG3350 and Tacsimate
 (although the concentrations are different) with different shapes. The
 crystals look big and beautiful. However, when I test them in synchrotron,
 both of these two types of crystals showed poor diffractions. As showed in
 the attached diffraction image, the diffraction is up to ~4 A but smear in
 one direction while 8 A in the other direction. The interesting thing is
 that the diffraction pattern is similar for all crystals (from two different
 proteins) that I tested without exception although they belong to different
 space groups. So, I wonder whether these kind of pattern is caused by
 Tacsimate (I don't know what it is) and how to rescue these crystals. Any
 suggestions or comments?

 Thanks a lot!

 Best,
 Zhiyi



Re: [ccp4bb] Crystal rescue

2010-01-26 Thread Jürgen Bosch

how much %PEG3350 ?
Try different additives such as glycerol, EG, MPD or smaller PEGs 200  
or 400.

Try flash annealing
Try freezing directly in liquid nitrogen if you have previously frozen  
your crystals in the stream
Try liquid annealing by dipping the frozen crystal into a new cryo  
solution and re-freeze it.

The list can be continued quite a bit.
If you should find out that one of the cryo additives helps then try  
to grow your crystals in the presence of those additives.


Good luck,

Jürgen

On Jan 26, 2010, at 10:42 AM, Zhiyi Wei wrote:


Dear all,

I got a problem with my crystals. I have two total different proteins
that both can be crystallized in the condition with PEG3350 and  
Tacsimate

(although the concentrations are different) with different shapes. The
crystals look big and beautiful. However, when I test them in  
synchrotron,
both of these two types of crystals showed poor diffractions. As  
showed in
the attached diffraction image, the diffraction is up to ~4 A but  
smear in
one direction while 8 A in the other direction. The interesting  
thing is
that the diffraction pattern is similar for all crystals (from two  
different
proteins) that I tested without exception although they belong to  
different

space groups. So, I wonder whether these kind of pattern is caused by
Tacsimate (I don't know what it is) and how to rescue these  
crystals. Any

suggestions or comments?

Thanks a lot!

Best,
Zhiyi
bad.png


-
Jürgen Bosch
Johns Hopkins Bloomberg School of Public Health
Department of Biochemistry  Molecular Biology
Johns Hopkins Malaria Research Institute
615 North Wolfe Street, W8708
Baltimore, MD 21205
Phone: +1-410-614-4742
Lab:  +1-410-614-4894
Fax:  +1-410-955-3655
http://web.mac.com/bosch_lab/



Re: [ccp4bb] Crystal rescue

2010-01-26 Thread Kelly Daughtry
Tascimate can be used as the cryo as well. I have had experience with
crystals in similar condition and moved the crystals to a 20%
increased Tascimate solution and they froze well.

I agree with Ezra, room temperature mount your crystal before
freezing. It is the only way to know the true problem.


Kelly
***
Kelly Daughtry
PhD Candidate
Department of Physiology and Biophysics
Boston University School of Medicine
590 Commonwealth Ave
R 390
Boston MA, 02215
(P) 617-358-5548
***



On Tue, Jan 26, 2010 at 1:25 PM, Marcus Winter
marcus.win...@oxford-diffraction.com wrote:
 Dear Zhiyi,


 Ezra is exactly right, of course.  The Oxford Diffraction PX Scanner
 system can assess the diffraction qualities of (putative) protein
 crystals in situ - in the crystallisation plate.  So, directly, you
 would discover if your 'big and beautiful' crystals actually diffract
 well... in their mother liquor under ambient conditions and before the
 addition of any cryo-protect.  Do you have a friend or neighbour with
 a PX Scanner ?  If not, please feel most welcome to contact
 Oxford Diffraction: we would be pleased to assist if at all possible.


 Good Luck and Best Wishes,

 Marcus Winter.

 www.oxford-diffraction.com




 -Original Message-
 From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of
 Ezra Peisach
 Sent: 26 January 2010 16:01
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: Re: [ccp4bb] Crystal rescue

 First you need to establish if it is your cryo conditions or the
 crystals.  Depending where you are - they might have the equipment to do

 a wet mount - without freezing.  Yes the crystal will not last - but
 then you know if the problem is in the
 crystal.  If it is - you need better crystals.  If it is the cryo - you
 need to work on that.  Tacsimate is mixture of alot of different
 compounds - but the smears are too close together to be a small salt
 crystal on top...

 Good luck,

 Ezra

 On 1/26/2010 10:42 AM, Zhiyi Wei wrote:
 Dear all,

 I got a problem with my crystals. I have two total different proteins
 that both can be crystallized in the condition with PEG3350 and
 Tacsimate
 (although the concentrations are different) with different shapes. The
 crystals look big and beautiful. However, when I test them in
 synchrotron,
 both of these two types of crystals showed poor diffractions. As
 showed in
 the attached diffraction image, the diffraction is up to ~4 A but
 smear in
 one direction while8 A in the other direction. The interesting thing
 is
 that the diffraction pattern is similar for all crystals (from two
 different
 proteins) that I tested without exception although they belong to
 different
 space groups. So, I wonder whether these kind of pattern is caused by
 Tacsimate (I don't know what it is) and how to rescue these crystals.
 Any
 suggestions or comments?

 Thanks a lot!

 Best,
 Zhiyi




Re: [ccp4bb] Crystal rescue

2010-01-26 Thread Joseph Ho
Hi, Zhiyi:
  You can always put a layer of oil on the top of your drop when you
open the coverslip. It will get you more time to mount the crystals.

Good luck

Meng-Chiao Ho

On Tue, 2010-01-26 at 11:48 -0500, Zhiyi Wei wrote:
 I forgot to mention a phenomenon when I did crystal mounting. I found
 that if open a coverslip, phase separation will appear in drops in
 minutes and damage to the crystals.
 
 And I have tried additive screen. The optimized crystals are really
 big (a few hundred microns) with sharp edge but diffraction pattern is
 similar.
 
 On Tue, Jan 26, 2010 at 10:42 AM, Zhiyi Wei ustcwebri...@gmail.com wrote:
  Dear all,
 
  I got a problem with my crystals. I have two total different proteins
  that both can be crystallized in the condition with PEG3350 and Tacsimate
  (although the concentrations are different) with different shapes. The
  crystals look big and beautiful. However, when I test them in synchrotron,
  both of these two types of crystals showed poor diffractions. As showed in
  the attached diffraction image, the diffraction is up to ~4 A but smear in
  one direction while 8 A in the other direction. The interesting thing is
  that the diffraction pattern is similar for all crystals (from two different
  proteins) that I tested without exception although they belong to different
  space groups. So, I wonder whether these kind of pattern is caused by
  Tacsimate (I don't know what it is) and how to rescue these crystals. Any
  suggestions or comments?
 
  Thanks a lot!
 
  Best,
  Zhiyi
 


Re: [ccp4bb] Crystal rescue

2010-01-26 Thread Zhiyi Wei
Thanks for so many quick responses!

Actually, I have test several different cryo-protectants, including
glycerol, EG, and PEG400. I did not see much differences between these
cryo conditions. So, I choose glycerol.

I would like to test my crystals in RT. But I don't know how to do
this. Just mount crystal to the X-ray machine without cryo stream? Or
I should use capillaries?

Zhiyi

On Wed, Jan 27, 2010 at 5:45 AM, Kelly Daughtry kddau...@bu.edu wrote:
 Tascimate can be used as the cryo as well. I have had experience with
 crystals in similar condition and moved the crystals to a 20%
 increased Tascimate solution and they froze well.

 I agree with Ezra, room temperature mount your crystal before
 freezing. It is the only way to know the true problem.


 Kelly
 ***
 Kelly Daughtry
 PhD Candidate
 Department of Physiology and Biophysics
 Boston University School of Medicine
 590 Commonwealth Ave
 R 390
 Boston MA, 02215
 (P) 617-358-5548
 ***



 On Tue, Jan 26, 2010 at 1:25 PM, Marcus Winter
 marcus.win...@oxford-diffraction.com wrote:
 Dear Zhiyi,


 Ezra is exactly right, of course.  The Oxford Diffraction PX Scanner
 system can assess the diffraction qualities of (putative) protein
 crystals in situ - in the crystallisation plate.  So, directly, you
 would discover if your 'big and beautiful' crystals actually diffract
 well... in their mother liquor under ambient conditions and before the
 addition of any cryo-protect.  Do you have a friend or neighbour with
 a PX Scanner ?  If not, please feel most welcome to contact
 Oxford Diffraction: we would be pleased to assist if at all possible.


 Good Luck and Best Wishes,

 Marcus Winter.

 www.oxford-diffraction.com




 -Original Message-
 From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of
 Ezra Peisach
 Sent: 26 January 2010 16:01
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: Re: [ccp4bb] Crystal rescue

 First you need to establish if it is your cryo conditions or the
 crystals.  Depending where you are - they might have the equipment to do

 a wet mount - without freezing.  Yes the crystal will not last - but
 then you know if the problem is in the
 crystal.  If it is - you need better crystals.  If it is the cryo - you
 need to work on that.  Tacsimate is mixture of alot of different
 compounds - but the smears are too close together to be a small salt
 crystal on top...

 Good luck,

 Ezra

 On 1/26/2010 10:42 AM, Zhiyi Wei wrote:
 Dear all,

 I got a problem with my crystals. I have two total different proteins
 that both can be crystallized in the condition with PEG3350 and
 Tacsimate
 (although the concentrations are different) with different shapes. The
 crystals look big and beautiful. However, when I test them in
 synchrotron,
 both of these two types of crystals showed poor diffractions. As
 showed in
 the attached diffraction image, the diffraction is up to ~4 A but
 smear in
 one direction while8 A in the other direction. The interesting thing
 is
 that the diffraction pattern is similar for all crystals (from two
 different
 proteins) that I tested without exception although they belong to
 different
 space groups. So, I wonder whether these kind of pattern is caused by
 Tacsimate (I don't know what it is) and how to rescue these crystals.
 Any
 suggestions or comments?

 Thanks a lot!