Re: Cell calibration by EN61000-4-3:2002

2004-12-02 Thread owner-emc-p...@listserv.ieee.org
Peter wrote:
 standard requires that a field at each point in a 0.5m matrix be within
6 dB of an empirically identified reference node - except:

CONCESSION 1   On p14, Note 2 says that calibration requirements are
fulfilled if a maximum of 3% of the frequencies does not meet the 6dB
criterion but are at least within the tolerance of -0 to +10 dB.

CONCESSION 2   On p15, i) stop the procedure if at least 12 numbers are
within 6dB.

I have an array of readings of forward power required to create a constant
field at each frequency step at each point in the grid. The answer is
different depending on the order in which the concessions are applied.  Can
someone please translate the requirements into a process that I can apply
to
data in an Excel spreadsheet? 

Hi, Peter.

My understanding is that at least 12 out of the 16 points must normally be
-0 to +6 dB. Thus, the test for that frequency stops (Concession 2) when
this has been met, and one proceeds to the next frequency. However, per
Concession 1, if not all 12 points make it, some being up to +10 dB (but
not -0dB), testing continues and at the end of the test, if 3 percent or
fewer of the frequencies tested have points up to +10 dB, uniformity may
still be claimed.

A simple procedure would be to

1. Store each frequency's 16 point scan on one worksheet page and

2. Evaluate them against a -0 to +10dB limit, discarding any outside that
range. Stop if you can't get 12 inside this limit; you have some work to do
on the chamber.

3. Reevaluate for -0 to +6dB and if 12 points are found, mark the page
compliant and go to the next frequency. If you can't get 12 in the -0 to
+6dB range then flag the worksheet page and continue to the next frequency.

Assuming one percent steps, there are 347 one-percent steps in an 80 to
2500 MHz scan so, if the number of flagged pages is 10 or less, then

4. Continue, stopping only if

5. 11 or more frequencies are flagged,

6. Or the test is done.



Cortland Richmond


This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Ron Pickard:  emc-p...@hypercom.com
 Scott Douglas emcp...@ptcnh.net

For policy questions, send mail to:

 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc



Re: EN61000-4-3:2002

2003-01-28 Thread John Woodgate

I read in !emc-pstc that richwo...@tycoint.com wrote (in 846BF526A205F8
4BA2B6045BBF7E9A6A04675B9C@flbocexu05) about 'EN61000-4-3:2002' on Tue,
28 Jan 2003:
Can anyone verify that EN61000-4-3:2000 does not contain common
modifications to the IEC document - i.e., they are identical?

I can't find an issue dated '2000'. The 2002 edition contains no Common
Modifications. The standard which it supersedes on 2005-04-01 is EN
61000-4-3:1996 + A1:1998 + A2:2001.
-- 
Regards, John Woodgate, OOO - Own Opinions Only. http://www.jmwa.demon.co.uk 
Interested in professional sound reinforcement and distribution? Then go to 
http://www.isce.org.uk
PLEASE do NOT copy news posts to me by E-MAIL!


This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list



EN61000-4-3:2002

2003-01-28 Thread richwo...@tycoint.com

Can anyone verify that EN61000-4-3:2000 does not contain common
modifications to the IEC document - i.e., they are identical?

Richard Woods
Sensormatic Electronics
Tyco International



This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list



RE: SV: Question about EN61000-4-3:2002

2002-05-31 Thread djumbdenstock

Hello Sandy,

I just received a copy and scanned through it.  As you had indicated, the
core of the standard is the same.  The calibration methodology is the same
as EN 61000-4-3 1996.  Annexes have been provided for rationale for the
choice of modulation for tests, a section on the use of anechoic chambers
with a discussion on problems caused by ferrite lining of chambers and
possible solutions, and a discussion of illumination by the independent
windows method.

If you test short range devices, EN 301489 is the product standard that
specifies radiated immunity testing parameters, including the requirement
above 1 GHz.  This standard specifies the level and modulation.  IEC
61000-4-3 2002 provides the guidance for calibration, set up, execution
(with alternatives) and reporting.

Best regards,

Don Umbdenstock
Tyco Safety Products
TEPG -Sensormatic



 --
 From: Sandy Mazzola[SMTP:mazzo...@symbol.com]
 Reply To: Sandy Mazzola
 Sent: Wednesday, May 29, 2002 5:21 PM
 To:   emc-p...@majordomo.ieee.org; am...@westin-emission.no
 Subject:  Re: SV: Question about EN61000-4-3:2002
 
 File: Sandy Mazzola.vcf.txt
 Amund,
  
  Thank you for your response.  Taking a quick look at the 2002 version
 the only changes I could tell were to do with were some information on
 using anechoic chambers above 1 GHz and alternative illumination
 techniques for frequencies above 1 GHz.   
 
 Have a great day
 Sandy Mazzola
 
 Santo Mazzola
 Regulatory Engineer
 Symbol Technologies Inc
 1 Symbol Plaza
 Holtsville, N. Y. 11742-1300
 Phone:  (631) 738-5373
 Fax:  (631) 738-3318
 E-mail: mazzo...@symbol.com
 
 
  am...@westin-emission.no 05/29/02 04:30PM 
 
 Sandy,
 
 I have just seen parts of the 2002 version. It seems to be dealing with
 the
 technique to calibrate the field strength above 1GHz. There are also some
 requirements regarding the measuring chamber.
 
 Best regards
 Amund Westin, Oslo/Norway
 
 
 -Opprinnelig melding-
 Fra: owner-emc-p...@majordomo.ieee.org 
 [mailto:owner-emc-p...@majordomo.ieee.org]Pa vegne av Sandy Mazzola
 Sendt: 28. mai 2002 21:51
 Til: emc-p...@majordomo.ieee.org 
 Emne: Question about EN61000-4-3:2002
 
 
 To all,
 
 
   I noticed that an EN61000-4-3:2002 just came into effectivity.
 Can anyone give me the short concise comparison of EN61000-4-3:2002 versus
 EN61000-4-3:1997.
 
 Thank You
 
 Santo Mazzola
 Regulatory Engineer
 Symbol Technologies Inc
 1 Symbol Plaza
 Holtsville, N. Y. 11742-1300
 Phone:  (631) 738-5373
 Fax:  (631) 738-3318
 E-mail: mazzo...@symbol.com 
 
 
 
 
 
 ---
 This message is from the IEEE EMC Society Product Safety
 Technical Committee emc-pstc discussion list.
 
 Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/ 
 
 To cancel your subscription, send mail to:
  majord...@ieee.org 
 with the single line:
  unsubscribe emc-pstc
 
 For help, send mail to the list administrators:
  Ron Pickard:  emc-p...@hypercom.com 
  Dave Heald:   davehe...@attbi.com 
 
 For policy questions, send mail to:
  Richard Nute:   ri...@ieee.org 
  Jim Bacher: j.bac...@ieee.org 
 
 All emc-pstc postings are archived and searchable on the web at:
 http://ieeepstc.mindcruiser.com/ 
 Click on browse and then emc-pstc mailing list
 
 

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


Re: SV: Question about EN61000-4-3:2002

2002-05-29 Thread Sandy Mazzola
Amund,
 
 Thank you for your response.  Taking a quick look at the 2002 version the 
only changes I could tell were to do with were some information on using 
anechoic chambers above 1 GHz and alternative illumination techniques for 
frequencies above 1 GHz.   

Have a great day
Sandy Mazzola

Santo Mazzola
Regulatory Engineer
Symbol Technologies Inc
1 Symbol Plaza
Holtsville, N. Y. 11742-1300
Phone:  (631) 738-5373
Fax:  (631) 738-3318
E-mail: mazzo...@symbol.com


 am...@westin-emission.no 05/29/02 04:30PM 

Sandy,

I have just seen parts of the 2002 version. It seems to be dealing with the
technique to calibrate the field strength above 1GHz. There are also some
requirements regarding the measuring chamber.

Best regards
Amund Westin, Oslo/Norway


-Opprinnelig melding-
Fra: owner-emc-p...@majordomo.ieee.org 
[mailto:owner-emc-p...@majordomo.ieee.org]Pa vegne av Sandy Mazzola
Sendt: 28. mai 2002 21:51
Til: emc-p...@majordomo.ieee.org 
Emne: Question about EN61000-4-3:2002


To all,


  I noticed that an EN61000-4-3:2002 just came into effectivity.
Can anyone give me the short concise comparison of EN61000-4-3:2002 versus
EN61000-4-3:1997.

Thank You

Santo Mazzola
Regulatory Engineer
Symbol Technologies Inc
1 Symbol Plaza
Holtsville, N. Y. 11742-1300
Phone:  (631) 738-5373
Fax:  (631) 738-3318
E-mail: mazzo...@symbol.com 





---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/ 

To cancel your subscription, send mail to:
 majord...@ieee.org 
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com 
 Dave Heald:   davehe...@attbi.com 

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org 
 Jim Bacher: j.bac...@ieee.org 

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/ 
Click on browse and then emc-pstc mailing list

BEGIN:VCARD
VERSION:2.1
X-GWTYPE:USER
FN:Mazzola, Sandy
TEL;WORK:(631) 738-5373
EMAIL;WORK;PREF;NGW:mazzo...@symbol.com
N:Mazzola;Sandy
END:VCARD

BEGIN:VCARD
VERSION:2.1
X-GWTYPE:USER
FN:Mazzola, Sandy
TEL;WORK:(631) 738-5373
EMAIL;WORK;PREF;NGW:mazzo...@symbol.com
N:Mazzola;Sandy
END:VCARD



SV: Question about EN61000-4-3:2002

2002-05-29 Thread amund

Sandy,

I have just seen parts of the 2002 version. It seems to be dealing with the
technique to calibrate the field strength above 1GHz. There are also some
requirements regarding the measuring chamber.

Best regards
Amund Westin, Oslo/Norway


-Opprinnelig melding-
Fra: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]Pa vegne av Sandy Mazzola
Sendt: 28. mai 2002 21:51
Til: emc-p...@majordomo.ieee.org
Emne: Question about EN61000-4-3:2002


To all,


  I noticed that an EN61000-4-3:2002 just came into effectivity.
Can anyone give me the short concise comparison of EN61000-4-3:2002 versus
EN61000-4-3:1997.

Thank You

Santo Mazzola
Regulatory Engineer
Symbol Technologies Inc
1 Symbol Plaza
Holtsville, N. Y. 11742-1300
Phone:  (631) 738-5373
Fax:  (631) 738-3318
E-mail: mazzo...@symbol.com





---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


Question about EN61000-4-3:2002

2002-05-28 Thread Sandy Mazzola
To all,


  I noticed that an EN61000-4-3:2002 just came into effectivity. 
Can anyone give me the short concise comparison of EN61000-4-3:2002 versus
EN61000-4-3:1997.

Thank You 

Santo Mazzola
Regulatory Engineer
Symbol Technologies Inc
1 Symbol Plaza
Holtsville, N. Y. 11742-1300
Phone:  (631) 738-5373
Fax:  (631) 738-3318
E-mail: mazzo...@symbol.com


BEGIN:VCARD
VERSION:2.1
X-GWTYPE:USER
FN:Mazzola, Sandy
TEL;WORK:(631) 738-5373
EMAIL;WORK;PREF;NGW:mazzo...@symbol.com
N:Mazzola;Sandy
END:VCARD

BEGIN:VCARD
VERSION:2.1
X-GWTYPE:USER
FN:Mazzola, Sandy
TEL;WORK:(631) 738-5373
EMAIL;WORK;PREF;NGW:mazzo...@symbol.com
N:Mazzola;Sandy
END:VCARD