Sure you can, all you need is a bright enough lightsource which emits at the 
correct wavelength. If the spectrum of the light source is too wide, you dump 
in a filter to narrow it down to your desired excitation.
Flow cytometers nowadays even use LED lasers instead of the huge, bulky and 
vulnerable gas lasers of the old days. Which is a good thing.

Jonathan
---
Jonathan Cremer
Laboratory Technician
TARGID - KU Leuven


Van: histonet-boun...@lists.utsouthwestern.edu 
[histonet-boun...@lists.utsouthwestern.edu] namens Emily Sours 
[talulahg...@gmail.com]
Verzonden: woensdag 6 juni 2012 1:31
Aan: histonet@lists.utsouthwestern.edu
Onderwerp: Re: [Histonet] Fluorescence-filters

Can you do fluorescence with LED?
(That may be a stupid question, as I always thought LED was just for
brightfield).

Emily

You see a peanut, day's off to a good start; you witness some soil it's a
jamboree for Vince Noir.
--Howard Moon, in Charlie, The Mighty Boosh
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Fluorescence-filters

2012-06-06 Thread Sue Hunter
We recently installed a LED light source on our fluorescence scope that our 
pathologists use for reading IF on kidneys and skins.  They love it.  No more 
having to warm up the bulb, not turning it off too soon,  or my nightmare - 
leaving it on overnight.  The bulb is supposed to last a really long time and 
gives off a nice bright light.

Sue Hunter, Supervisor
Advanced Diagnostics
Beaumont Health System
Royal Oak MI
248-898-5146
shun...@beaumont.edu



-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Emily Sours
Sent: Tuesday, June 05, 2012 7:32 PM
To: histonet@lists.utsouthwestern.edu
Subject: Re: [Histonet] Fluorescence-filters

Can you do fluorescence with LED?
(That may be a stupid question, as I always thought LED was just for 
brightfield).

Emily

You see a peanut, day's off to a good start; you witness some soil it's a 
jamboree for Vince Noir.
--Howard Moon, in Charlie, The Mighty Boosh 
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
To report this email as SPAM, please forward it to s...@websense.com.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Histotech scheduling

2012-06-06 Thread Duddey, Aimee
I am looking for some creative scheduling ideas.  We have a day shift
crew that starts at 4am and goes til 530pm.  The schedule is below.
Techs would like to rotate to do different jobs within the department
but don't really want to rotate hours.  Can I please get some examples
of how others of you are staffing?  Even if you have a second or third
shift I would like to see some different workflow plans and how they are
staffed.

 

1 HT 400am - Embedder

2 HTs 500am - Cutters

1 HT 730 - Immunos/specials/orders

1 HT 830 - Grossing/pathologist assist

1 lab assistant 800am - cytology prep/gross room assist

 

Aimee M. Duddey, MLT(ASCP)

Moore Regional Hospital Laboratory

Assistant Director of Laboratory - Pathology

910-715-5286

FirstHealth OF THE CAROLINAS

 

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Histotech scheduling

2012-06-06 Thread joelle weaver

http://www.biomedcentral.com/1472-6890/10/2




Joelle Weaver MAOM, HTL (ASCP) QIHC
  Date: Wed, 6 Jun 2012 08:30:47 -0400
 From: adud...@firsthealth.org
 To: histonet@lists.utsouthwestern.edu
 Subject: [Histonet] Histotech scheduling
 
 I am looking for some creative scheduling ideas.  We have a day shift
 crew that starts at 4am and goes til 530pm.  The schedule is below.
 Techs would like to rotate to do different jobs within the department
 but don't really want to rotate hours.  Can I please get some examples
 of how others of you are staffing?  Even if you have a second or third
 shift I would like to see some different workflow plans and how they are
 staffed.
 
  
 
 1 HT 400am - Embedder
 
 2 HTs 500am - Cutters
 
 1 HT 730 - Immunos/specials/orders
 
 1 HT 830 - Grossing/pathologist assist
 
 1 lab assistant 800am - cytology prep/gross room assist
 
  
 
 Aimee M. Duddey, MLT(ASCP)
 
 Moore Regional Hospital Laboratory
 
 Assistant Director of Laboratory - Pathology
 
 910-715-5286
 
 FirstHealth OF THE CAROLINAS
 
  
 
 ___
 Histonet mailing list
 Histonet@lists.utsouthwestern.edu
 http://lists.utsouthwestern.edu/mailman/listinfo/histonet
  
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Histotech scheduling

2012-06-06 Thread Fahy, Claire
Please unsubscribe me.
thanks

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of joelle weaver
Sent: 06 June 2012 15:14
To: adud...@firsthealth.org; histonet@lists.utsouthwestern.edu
Subject: RE: [Histonet] Histotech scheduling


http://www.biomedcentral.com/1472-6890/10/2




Joelle Weaver MAOM, HTL (ASCP) QIHC
  Date: Wed, 6 Jun 2012 08:30:47 -0400
 From: adud...@firsthealth.org
 To: histonet@lists.utsouthwestern.edu
 Subject: [Histonet] Histotech scheduling

 I am looking for some creative scheduling ideas.  We have a day shift
 crew that starts at 4am and goes til 530pm.  The schedule is below.
 Techs would like to rotate to do different jobs within the department
 but don't really want to rotate hours.  Can I please get some examples
 of how others of you are staffing?  Even if you have a second or third
 shift I would like to see some different workflow plans and how they
 are staffed.



 1 HT 400am - Embedder

 2 HTs 500am - Cutters

 1 HT 730 - Immunos/specials/orders

 1 HT 830 - Grossing/pathologist assist

 1 lab assistant 800am - cytology prep/gross room assist



 Aimee M. Duddey, MLT(ASCP)

 Moore Regional Hospital Laboratory

 Assistant Director of Laboratory - Pathology

 910-715-5286

 FirstHealth OF THE CAROLINAS



 ___
 Histonet mailing list
 Histonet@lists.utsouthwestern.edu
 http://lists.utsouthwestern.edu/mailman/listinfo/histonet
  
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

Department of Agriculture, Food and the Marine

The information contained in this email and in any attachments is confidential 
and is designated solely for the attention and use of the intended 
recipient(s). This information may be subject to legal and professional 
privilege.  If you are not an intended recipient of this email, you must not 
use, disclose, copy, distribute or retain this message or any part of it. If 
you have received this email in error, please notify the sender immediately and 
delete all copies of this email from your computer system(s).

An Roinn Talmhaíochta, Bia agus Mara

Tá an t-eolais san ríomhphost seo, agus in aon ceangláin leis, faoi phribhléid 
agus faoi rún agus le h-aghaigh an seolaí amháin. D’fhéadfadh ábhar an seoladh 
seo bheith faoi phribhléid profisiúnta nó dlíthiúil. Mura tusa an seolaí a bhí 
beartaithe leis an ríomhphost seo a fháil, tá cosc air, nó aon chuid de, a 
úsáid, a chóipeál, nó a scaoileadh.  Má tháinig sé chugat de bharr dearmad, 
téigh i dteagmháil leis an seoltóir agus scrios an t-ábhar ó do ríomhaire le do 
thoil.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] staffing questions

2012-06-06 Thread Hutton, Allison
I have been asked to do some research and provide some facts and figures as it 
relates to staffing of histology labs.  If anybody is willing to answer the 
questions listed below, I would appreciate your input.  You may email me at 
privately and all information will be kept confidental.
1. Staffing levels, i.e. # of pathologists, techs, PAs, and Lab aides
2. Approximate workload, either blocks per day or cases per year
3. Salary for HTs and supervisors (I know this can be a sticky area but any 
info would be helpful)
Thank You in advance
Allison Hutton, HTL(ASCP)cm
Lead Tech Histology
Doylestown Hospital
ahut...@dh.org

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] staffing questions

2012-06-06 Thread Rene J Buesa
Staff level (your question 1) depends on the answer to your question 2 and also 
depends on the productivity level, both individual and collective.
The salaries depend on the nature of your lab (non for profit, private for 
profit or governmental) and with the total benefits offered. But also depend on 
the market availability and the willingness to negotiate both from the 
administration and the potential or existing employees.
So you ca understand both issues I am sending you some information under 
separate cover.
René J.



From: Hutton, Allison ahut...@dh.org
To: histonet@lists.utsouthwestern.edu 
Sent: Wednesday, June 6, 2012 11:09 AM
Subject: [Histonet] staffing questions

I have been asked to do some research and provide some facts and figures as it 
relates to staffing of histology labs.  If anybody is willing to answer the 
questions listed below, I would appreciate your input.  You may email me at 
privately and all information will be kept confidental.
1. Staffing levels, i.e. # of pathologists, techs, PAs, and Lab aides
2. Approximate workload, either blocks per day or cases per year
3. Salary for HTs and supervisors (I know this can be a sticky area but any 
info would be helpful)
Thank You in advance
Allison Hutton, HTL(ASCP)cm
Lead Tech Histology
Doylestown Hospital
ahut...@dh.org

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Sequenza units for manual immunohistochemistry

2012-06-06 Thread Keri Colwell
Hello Histonet,

I'm just wondering if anyone out there is using these units for their
manual staining, and if anyone has ever given any consideration to what
happens when you have a large volume of reagent waste in the bottom
from all your additions and washes.  Since the units draw reagents
through with capillary action, what do you think happens when the bottom
of the slide/coverplate combo is immersed in the cocktail that is
created in the bottom.

Has anyone had any issues using these units?  How do people prevent a
large volume from building up in the bottom?  How are people cleaning
their coverplates and the units themselves?

Thanks in advance for your thoughts!


Keri

Keri Colwell, BSc
Laboratory Technologist | Technologiste de laboratoire
TSE and Pathology
Lethbridge Laboratory | Laboratoire de Lethbridge
Canadian Food Inspection Agency | Agence candienne d'inspection des
aliments
Township Road 9-1 | Ch de Canton 9-1
Box 640  | CP 640
Lethbridge, AB T1J 3Z4
E-mail | Courriel: keri.colw...@inspection.gc.ca 
Telephone | Téléphone:  403-382-5500 ext 5613/5511
Facsimile | Télécopieur: 403-382-5585
Government of Canada | Gouvernement du Canada

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] Sequenza units for manual immunohistochemistry

2012-06-06 Thread Beckham, Sharon
I have used them for quite some time.  I was having problems with uneven 
staining and determined that it was from the being too full and was drawing up 
liquid from the bottom.  There is a small opening on the end and I just empty 
the solution from it.  I usually empty it after rinsing the antibody off and 
then usually before putting DAB or DAPI on depending on the number of slides 
stained.  As for washing, I just wash in a solution of soapy water and then 
rinse in distilled.  I use the coverplates for several staining times.  I love 
the sequenza and prefer it to having an automatic stainer.
 


Sharon Beckham, HT (ASCP)
Head, Histotechnology
Stowers Institute for Medical Research
1000 E. 50th St.
Kansas City, Missouri 64110
(816)926-4305
s...@stowers.org

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Keri Colwell
Sent: Wednesday, June 06, 2012 12:23 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Sequenza units for manual immunohistochemistry

Hello Histonet,

I'm just wondering if anyone out there is using these units for their manual 
staining, and if anyone has ever given any consideration to what happens when 
you have a large volume of reagent waste in the bottom from all your 
additions and washes.  Since the units draw reagents through with capillary 
action, what do you think happens when the bottom of the slide/coverplate combo 
is immersed in the cocktail that is created in the bottom.

Has anyone had any issues using these units?  How do people prevent a large 
volume from building up in the bottom?  How are people cleaning their 
coverplates and the units themselves?

Thanks in advance for your thoughts!


Keri

Keri Colwell, BSc
Laboratory Technologist | Technologiste de laboratoire TSE and Pathology 
Lethbridge Laboratory | Laboratoire de Lethbridge Canadian Food Inspection 
Agency | Agence candienne d'inspection des aliments Township Road 9-1 | Ch de 
Canton 9-1 Box 640  | CP 640 Lethbridge, AB T1J 3Z4 E-mail | Courriel: 
keri.colw...@inspection.gc.ca Telephone | Téléphone:  403-382-5500 ext 
5613/5511 Facsimile | Télécopieur: 403-382-5585 Government of Canada | 
Gouvernement du Canada

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] processor solution rotation

2012-06-06 Thread Steve McClain
May I suggest?
Define a standard that works for you, combining both elements,
blocks and runs e.g.,
Every week or 200 blocks then check the first reagent  volume and quality.
Again at 400
And again at 600 blocks.
We change the first reagent whenever it appears dirty.

Consider adding a few ccs. of eosin as a visual indicator in the first 100% 
alcohol to monitor carryover and number of runs.
When the indicator reaches the first xylene, time to rotate or change.
NOTE Eosin fluoresces.

The actual blocks and runs number varies
on specimen size
and fixation, (unfixed tissues use up the alcohols)
and also whether you use sponges (more carryover)
and whether the processor drains/pumps out completely-carryover.
And on how fast the processing cycle is- short cycles demand fresh reagents and 
leave little room for error.

When in doubt do not hesitate to replenish the last solution of each type, 
alcohol, xylene and paraffin or change all solutions.
Steve
Steve A. McClain, MD
McClain Labs, LLC 45 Manor Road, Smithtown, NY 11787 631 361 4000

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Pathologists' Assistant Job in Mid-West

2012-06-06 Thread Melissa Phelan
Hi Everyone,

I have a Pathologists' Assistant Job in the MidWest that is a full time
permanent job. I am looking for an ASCP Certified Pathologists' Assistant,
with a Master's Degree from an NAACLS accredited institution. Please email
me for more details. Thank you@

To view a complete list of Allied Search Partners current openings go to:
http://www.alliedsearchpartners.com/careers.php

Melissa Phelan
LinkedIn: http://www.linkedin.com/in/melissaphelan
President, Laboratory Staffing
Allied Search Partners
P: 888.388.7571
F: 888.388.7572
M: 407.697.1175
www.alliedsearchpartners.com http://www.alliedsearchpartners.com/




___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Approval and review of procedures

2012-06-06 Thread Sproles, Judy A.
Hello,
I am currently helping organize histology SOPs and need some help clarifying 
who can perform SOP review. In cytology, the technical supervisor (MD) reviews 
and approves procedures. CLIA rules for clinical departments allow MT as 
technical supervisors. Can an HT be a technical supervisor and perform annual 
review of SOPs?

Judy Sproles CT(ASCP

Methodist. Leading Medicine. Recognized by U.S.News  World Report as one of 
America's Best Hospitals in 13 specialties. Named to FORTUNE®  Magazine's 
100 Best Companies to Work For® list seven years in a row. Designated as a 
Magnet hospital for excellence in nursing. Visit us at  methodisthealth.com. 
Follow us at  twitter.com/MethodistHosp and  
www.facebook.com/MethodistHospital.  ***CONFIDENTIALITY NOTICE*** This e-mail 
is the property of The Methodist Hospital and/or its relevant affiliates and 
may contain restricted and privileged material for the sole use of the intended 
recipient(s). Any review, use, distribution or disclosure by others is strictly 
prohibited. If you are not the intended recipient (or authorized to receive for 
the recipient), please contact the sender and delete all copies of the message. 
Thank you.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: Approval and review of procedures

2012-06-06 Thread Morken, Timothy
No, it has to be the pathologist who is designated as the technical supervisor 
for the lab.


Tim Morken
Department of Pathology
UC San Francisco Medical Center
505 Parnassus Ave, Box 1656
Room S570
San Francisco, CA 94132

(415) 353-1266 (ph)
(415) 514-3403 (fax)
tim.mor...@ucsfmedctr.org




-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Sproles, Judy A.
Sent: Wednesday, June 06, 2012 2:30 PM
To: 'histonet@lists.utsouthwestern.edu'
Subject: [Histonet] Approval and review of procedures

Hello,
I am currently helping organize histology SOPs and need some help clarifying 
who can perform SOP review. In cytology, the technical supervisor (MD) reviews 
and approves procedures. CLIA rules for clinical departments allow MT as 
technical supervisors. Can an HT be a technical supervisor and perform annual 
review of SOPs?

Judy Sproles CT(ASCP

Methodist. Leading Medicine. Recognized by U.S.News  World Report as one of 
America's Best Hospitals in 13 specialties. Named to FORTUNE®  Magazine's 
100 Best Companies to Work For® list seven years in a row. Designated as a 
Magnet hospital for excellence in nursing. Visit us at  methodisthealth.com. 
Follow us at  twitter.com/MethodistHosp and  
www.facebook.com/MethodistHospital.  ***CONFIDENTIALITY NOTICE*** This e-mail 
is the property of The Methodist Hospital and/or its relevant affiliates and 
may contain restricted and privileged material for the sole use of the intended 
recipient(s). Any review, use, distribution or disclosure by others is strictly 
prohibited. If you are not the intended recipient (or authorized to receive for 
the recipient), please contact the sender and delete all copies of the message. 
Thank you.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: Approval and review of procedures

2012-06-06 Thread Weems, Joyce K.
And not her/his designee?

Joyce Weems
Pathology Manager
678-843-7376 Phone
678-843-7831 Fax
joyce.we...@emoryhealthcare.org



www.saintjosephsatlanta.org
5665 Peachtree Dunwoody Road
Atlanta, GA 30342

This e-mail, including any attachments is the property of Saint Joseph’s 
Hospital and is intended for the sole use of the intended recipient(s).  It may 
contain information that is privileged and confidential.  Any unauthorized 
review, use, disclosure, or distribution is prohibited. If you are not the 
intended recipient, please delete this message, and reply to the sender 
regarding the error in a separate email.


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Morken, Timothy
Sent: Wednesday, June 06, 2012 5:40 PM
To: Sproles, Judy A.; 'histonet@lists.utsouthwestern.edu'
Subject: [Histonet] RE: Approval and review of procedures

No, it has to be the pathologist who is designated as the technical supervisor 
for the lab.


Tim Morken
Department of Pathology
UC San Francisco Medical Center
505 Parnassus Ave, Box 1656
Room S570
San Francisco, CA 94132

(415) 353-1266 (ph)
(415) 514-3403 (fax)
tim.mor...@ucsfmedctr.org




-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Sproles, Judy A.
Sent: Wednesday, June 06, 2012 2:30 PM
To: 'histonet@lists.utsouthwestern.edu'
Subject: [Histonet] Approval and review of procedures

Hello,
I am currently helping organize histology SOPs and need some help clarifying 
who can perform SOP review. In cytology, the technical supervisor (MD) reviews 
and approves procedures. CLIA rules for clinical departments allow MT as 
technical supervisors. Can an HT be a technical supervisor and perform annual 
review of SOPs?

Judy Sproles CT(ASCP

Methodist. Leading Medicine. Recognized by U.S.News  World Report as one of 
America's Best Hospitals in 13 specialties. Named to FORTUNE®  Magazine's 
100 Best Companies to Work For® list seven years in a row. Designated as a 
Magnet hospital for excellence in nursing. Visit us at  methodisthealth.com. 
Follow us at  twitter.com/MethodistHosp and  
www.facebook.com/MethodistHospital.  ***CONFIDENTIALITY NOTICE*** This e-mail 
is the property of The Methodist Hospital and/or its relevant affiliates and 
may contain restricted and privileged material for the sole use of the intended 
recipient(s). Any review, use, distribution or disclosure by others is strictly 
prohibited. If you are not the intended recipient (or authorized to receive for 
the recipient), please contact the sender and delete all copies of the message. 
Thank you.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



This e-mail message (including any attachments) is for the sole use of
the intended recipient(s) and may contain confidential and privileged
information. If the reader of this message is not the intended
recipient, you are hereby notified that any dissemination, distribution
or copying of this message (including any attachments) is strictly
prohibited.

If you have received this message in error, please contact
the sender by reply e-mail message and destroy all copies of the
original message (including attachments).
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: Reagent rotation

2012-06-06 Thread vijaykanthg

Dear Mark

We change the solution depend upon no of cassettes run. In your case the size 
of specimen is small so u can change/ rotate depend upon no of run.i felt 
that the size of the specimen make big impact on reagents in Tissue Processer.

With regards
Dr G vijay
Orchid Pharma
-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of 
histonet-requ...@lists.utsouthwestern.edu
Sent: Wednesday, June 06, 2012 10:30 PM
To: histonet@lists.utsouthwestern.edu
Subject: Histonet Digest, Vol 103, Issue 7

Send Histonet mailing list submissions to
histonet@lists.utsouthwestern.edu

To subscribe or unsubscribe via the World Wide Web, visit
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
or, via email, send a message with subject or body 'help' to
histonet-requ...@lists.utsouthwestern.edu

You can reach the person managing the list at
histonet-ow...@lists.utsouthwestern.edu

When replying, please edit your Subject line so it is more specific
than Re: Contents of Histonet digest...


Today's Topics:

   1. Reagent rotation (Mark Perrin)
   2. RE: Reagent rotation (Goins, Tresa)
   3. Reagent rotation (Stephenson, Sheryl)
   4. Re: Fluorescence-filters (Eric Hoy)
   5. Current Histology Openings (Brannon Owens)
   6. Re: Fluorescence-filters (Mark Tarango)
   7. NEW Position Alert - Quality Assurance Histologist (Matt Ward)
   8. Re: Fluorescence-filters (Emily Sours)
   9. Re: Fluorescence-filters (Emily Sours)
  10. RE: Reagent rotation (Ian R Bernard)
  11. RE: Fluorescence-filters (Jonathan Cremer)
  12. RE: Fluorescence-filters (Sue Hunter)
  13. Histotech scheduling (Duddey, Aimee)
  14. RE: Histotech scheduling (joelle weaver)
  15. RE: Histotech scheduling (Fahy, Claire)
  16. staffing questions (Hutton, Allison)
  17. Re: staffing questions (Rene J Buesa)


--

Message: 1
Date: Tue, 5 Jun 2012 11:04:50 -0700 (PDT)
From: Mark Perrin perrintos...@yahoo.com
Subject: [Histonet] Reagent rotation
To: histonet@lists.utsouthwestern.edu
histonet@lists.utsouthwestern.edu
Message-ID:
1338919490.15637.yahoomail...@web44714.mail.sp1.yahoo.com
Content-Type: text/plain; charset=iso-8859-1

What is the normal procedure for changing the reagents in the tissue processor? 
We have a Sakura VIP6. Should the schedule of changing be based on number of 
runs vs. number of blocks processed? About 80% of our blocks are small GI 
biospies.

Thanks in advance for your input.

?
Toshia Perrin
Medical Practice Coordinator
Southern Pathology

This message contains privileged and confidential information intended only for 
the use of the addressee named above. If you are not the intended recipient of 
this message you must not disseminate, copy, or take any action in reliance on 
it.

--

Message: 2
Date: Tue, 5 Jun 2012 18:35:14 +
From: Goins, Tresa tgo...@mt.gov
Subject: RE: [Histonet] Reagent rotation
To: 'Mark Perrin' perrintos...@yahoo.com,
histonet@lists.utsouthwestern.edu
histonet@lists.utsouthwestern.edu
Message-ID:
ca4df32ed505d94bb55e95487d8e98411f24c...@doaisd5205.state.mt.ads
Content-Type: text/plain; charset=iso-8859-1

We rotate by cassette number - our daily load can vary from 10 to more than 100 
cassettes, so an exchange schedule based on number of runs was not appropriate.

Tresa

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Mark Perrin
Sent: Tuesday, June 05, 2012 12:05 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Reagent rotation

What is the normal procedure for changing the reagents in the tissue processor? 
We have a Sakura VIP6. Should the schedule of changing be based on number of 
runs vs. number of blocks processed? About 80% of our blocks are small GI 
biospies.

Thanks in advance for your input.

?
Toshia Perrin
Medical Practice Coordinator
Southern Pathology

This message contains privileged and confidential information intended only for 
the use of the addressee named above. If you are not the intended recipient of 
this message you must not disseminate, copy, or take any action in reliance on 
it.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet





--

Message: 3
Date: Tue, 5 Jun 2012 14:02:36 -0500
From: Stephenson, Sheryl sstephen...@lifecell.com
Subject: [Histonet] Reagent rotation
To: 'Mark Perrin' perrintos...@yahoo.com,
histonet@lists.utsouthwestern.edu
histonet@lists.utsouthwestern.edu
Message-ID:
c2a8fe4da0ed8c468a3713f8607829d30d3f922...@amwpvex03.kci.com
Content-Type: text/plain; charset=iso-8859-1


We do it by the number of Runs.  About every 6 run for