Re: [Histonet] Leica bond

2015-07-08 Thread Paula Lucas
Wendy, the Leica Bond is a great machine. We've had it for several years now
and it provides excellent and consistent results. 
Paula 
Bio-Path Medical Group
Fountain Valley, CA



-Original Message-
From: Miller, Wendy [mailto:miller.we...@mhsil.com] 
Sent: Wednesday, July 08, 2015 6:09 AM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Leica bond

Hi!  Does anyone out there use the Leica IHC Bond stainer?  We are looking
into purchasing one and I would like to input.  Do you like it?  Pros and
cons?

Thanks


This message (including any attachments) contains confidential information
intended for a specific individual and purpose, and is protected by law. If
you are not the intended recipient, you should delete this message. Any
disclosure, copying, or distribution of this message, or the taking of any
action based on it, is strictly prohibited.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] GMS Question

2015-05-01 Thread Paula Lucas
Hello,

 

I think I already know the answer but I'm not sure why so if someone can
help me understand the theory behind it, I would greatly appreciate it.

 

Currently, we use the Richard Allen kit for the GMS stain and it uses
Periodic Acid as the 1st step.  

We use a control tissue from a case we had that was positive for fungus and
it's a fungus ball from the Rt Maxillary. 

We ran a test for fungus on a different and current case of the same tissue
(different patient): Rt Maxillary sinus.

 

The control tissue did work, but the patient's tissue did not, so the doctor
ordered a PAS for fungus and this clearly showed the fungal elements nicely.


 

My question is why would the control and patient tissue have different
results when they are both fungus balls from the same specimen source?

 

Thanks in advance,

Paula

Lab Manager

Bio-Path Medical Group

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Leica ASP6025 Processor

2014-11-05 Thread Paula Lucas
Hello,

We're considering the Leica ASP6025 processor for our laboratory and I was
hoping I can get pros and cons or any comments such as in performance,
reagent money savings, ease of use, any comment from you.

 

Thank you so much in advance for taking the time to give me your thoughts,

Paula Lucas

Lab Manager

Bio-Path Medical Group

California

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Study using Resin Embedding

2014-09-26 Thread Paula Lucas
Hello,

We are doing a study on molar roots for a company in Orange County,
California. The study is working with tissues embedded in paraffin blocks
and they would like a study using resin embedded histology.  We do not have
the capability to do this study, so if your company is able to help them,
and would like to contact them to find out more information, please let me
know and I will connect you to the right person.  

Thank you,

Paula Lucas

Lab Manager

Bio-Path Medical Group

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Histotech Job Opportunity/Fountain Valley, Orange County, CA

2014-09-05 Thread Paula Lucas
Please no hiring staff services calling.

 

We are in need of a part-time histotechnician who can work on Tuesdays and
Saturdays, and then also to fill in during the week when needed.

 

Hours start at 5 am. We work Tuesdays through Saturdays.

 

Only experienced histotechs apply, which means previous experience in a lab
work environment (not as an intern), and being able to cut (microtomy), and
embed all types of tissues.  

 

We are a busy lab who supports hospitals, surgery centers and physician
practice offices

Please send resume to:

Paula Lucas

714-755-2984.

Thank you

 

 

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Grossing/PA needed, orange county ca

2014-08-28 Thread Paula Lucas
We need assistance with our volume. Please call if interested and if you
have any questions.  Please ask for Russell Blayney.

Thank you,

Paula Lucas

Lab Manager

Biopath Medical Group

714-433-1330

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] New code 88343

2013-12-17 Thread Paula Lucas
Hello,

 

Can someone shed some light for me on this new code for IHC?  I think the
code refers to double stain antibodies?  

Thanks in advance

: )
Paula Lucas

Lab Manager

Bio-Path Medical Group

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Full Time Histotech Position/Orange County California

2013-08-07 Thread Paula Lucas
A full-time histotech position is opening at the end of August. The opening
is a shift from 4:00am to 12:30pm, Tuesday through Saturdays. On-call duty 1
or 2 Saturdays per month is also required. Certification is required. Duties
include: microtomy, embedding, manual special stains, IHC. 

Please email a resume if you're interested.

Thank you,

 

Paula Lucas

Lab Manager

Bio-Path Medical Group

Fountain Valley, CA 92708

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] recycled chemical testing

2013-05-01 Thread Paula Lucas
Hello,

Could anyone give me a contact where I can send my recycled alcohol and
xylene for quality assessment?

 

Thank you,

Paula Lucas

Lab Manager

Bio-Path Medical Group

Fountain Valley, CA 92708

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Job Opening orange county

2012-03-21 Thread Paula Lucas
We have a part-time histotech position available and anyone who is
interested, please send me your resume by fax or email.  The work days are
Tuesday through Saturday starting at 5 am.  

Thank you,

Paula Lucas

Lab Manager

Bio-Path Medical Group

Fountain Valley, CA 92708

Fax: 714-755-2984

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Job Opening in Orange county California

2012-02-21 Thread Paula Lucas
A part-time histotech position is open in our histology laboratory.  

Tuesday through Saturday starting at 5 am.

If anyone is interested, please either fax or email me your resume.

Thank you,

Paula Lucas

Bio-Path Medical Group

Fountain Valley, CA 92708

Fax: 714-755-2984

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Histotech opening Orange County California

2012-01-27 Thread Paula Lucas
We have a part time histotech position available working from 5 am on
Tuesday through Saturdays.  We are located in Fountain Valley, California.

Please provide me a summary of your histotech experience and please send me
your resume.

Thank you,

Paula Lucas

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Saturday Coverage Orange County California

2011-12-29 Thread Paula Lucas
We are in need of a per diem histotech who can work on Saturday mornings.
We are located in Fountain Valley, California. Please send me an email if
interested or fax resume to:

714 755-2984

Thank you,

Paula Lucas

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Histotech Job Opening in Southern California

2011-10-24 Thread Paula Lucas
A part-time entry level histotech position is available in our laboratory.
Please email or fax resume to:

 

Paula Lucas

Lab Manager

Bio-Path Medical Group

Fountain Valley, CA 92708

 

plu...@biopath.org

714-755-2984 (fax)

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] service on the Sukura Prisma

2011-07-29 Thread Paula Lucas
Hello,

 

Does anyone know of an alternate vendor besides Sakura who can do the PM's
and any necessary adjustments/fixes that come up on the Prisma?  Sakura
bills an arm and a leg for someone to come out.  $175 dollars just for
travel expenses is gouging to me, so if we can find someone else who can do
the work at a cheaper rate, that would be great.

 

Thanks for any info,

Paula Lucas

Lab Manager

BP Labs

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Embedding Media

2011-07-08 Thread Paula Lucas
Hello,

 

We are considering a switch to a different brand of paraffin and this is
because I feel we are having too many compressions in some of our tissue
sections.  Currently, we use Tissue Path Paraplast, regular.

 

I would like to get feedback from you as to what you prefer.  Looking on
line, Richard Allan has a product called Signature Series Paraffin that
comes in a type L that offers compression-free sections and I was also
hoping to get any feedback on that product.

 

I would greatly appreciate any suggestions and thoughts.

Thanks in advance,

Paula

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] IHC platforms/Sales Representatives

2011-04-13 Thread Paula Lucas
Hello,

 

We are inquiring about IHC staining platforms and we've already contacted
the Leica sales rep for the Bond.  I would also like to contact Ventana to
get a quote on their system, and I've tried to get in contact with a sales
rep in our area but I haven't had anyone returning my calls.  If there is a
Vantana rep reading this, would you please contact the sales representative
for my area and let him/her know we are interested in a quote?  We are
located in Orange County (Fountain Valley), California.  

 

For those who have experience with these two platforms, would you mind
taking the time to share with me your views and experiences about the two
systems?

 

Thanks so much,

Paula Lucas

Lab Manager

Bio-Path Medical Group

Fountain Valley, CA

714.433.1330

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Histotech Needed

2011-02-19 Thread Paula Lucas
Histotech needed for vacation coverages and Saturday coverage.  We are an 
independent laboratory located in Fountain Valley, California.  If interested, 
I will give you more details.
Thank you,
Paula Lucas
Bio-Path Medical Group
Fountain Valley, CA

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] TTF-1/background staining

2011-02-04 Thread Paula Lucas
 Siena HT(ASCP)QIHC
 Technical Manager | StatLab Medical Products
 Direct: 972-436-1010  x229 
 
 
 -Original Message-
 From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Paula Lucas
 Sent: Thursday, February 03, 2011 2:38 PM
 To: histonet@lists.utsouthwestern.edu
 Subject: [Histonet] TTF-1/background staining
 
 How can I eliminate the background or as you say, non-specific staining?
 
 
 
 I'm no expert here, I admit, and so I'm asking for your help and
 suggestions. I have searched the Histonet archives, and a lot of what I'm
 seeing deals with animal or rodent tissues, and a lot of it was confusing
to
 me.
 
 
 
 To give you a little background info:
 
 We use the Lab Vision stainer and the TTF-1 antibody we use is from Cell
 Marque. We were having issues with this marker from Lab Vision, so we
 switched to TTF-1 a while ago. We use the UltraVision LP detection kit
from
 Lab Vision. It's a polymer driven detection kit. 
 
 
 
 The antibody is a ready to use, and we have the time set at 30 minutes. 
 
 
 
 We were getting the background staining on a liver specimen last week, and
 we also had this problem today on a lung case. My doctor is getting
 frustrated, and wants me to do something about this, and to repeat the
TTF-1
 stain on the lung, so if someone can give me some suggestions to try, I'm
 ready to try them.
 
 
 
 Thanks in advance,
 
 Paula
 
 Lab Manager
 
 Bio-Path Medical Group
 
 Fountain Valley, CA
 
 ___
 Histonet mailing list
 Histonet@lists.utsouthwestern.edu
 http://lists.utsouthwestern.edu/mailman/listinfo/histonet
 
 ___
 Histonet mailing list
 Histonet@lists.utsouthwestern.edu
 http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Her2 Fixation Requirement Responses/Thanks

2011-02-03 Thread Paula Lucas
Your information and suggestions helped me formulate a plan, so thank you to
those who took the time and responded to my inquiry yesterday.

Paula Lucas 

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] TTF-1/background staining

2011-02-03 Thread Paula Lucas
How can I eliminate the background or as you say, non-specific staining?

 

I'm no expert here, I admit, and so I'm asking for your help and
suggestions.  I have searched the Histonet archives, and a lot of what I'm
seeing deals with animal or rodent tissues, and a lot of it was confusing to
me.

 

To give you a little background info:

We use the Lab Vision stainer and the TTF-1 antibody we use is from Cell
Marque.  We were having issues with this marker from Lab Vision, so we
switched to TTF-1 a while ago.  We use the UltraVision LP detection kit from
Lab Vision. It's a polymer driven detection kit. 

 

The antibody is a ready to use, and we have the time set at 30 minutes.  

 

We were getting the background staining on a liver specimen last week, and
we also had this problem today on a lung case.  My doctor is getting
frustrated, and wants me to do something about this, and to repeat the TTF-1
stain on the lung, so if someone can give me some suggestions to try, I'm
ready to try them.

 

Thanks in advance,

Paula

Lab Manager

Bio-Path Medical Group

Fountain Valley, CA

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Her2 Fixation Requirement

2011-02-02 Thread Paula Lucas
Hello histoland
 
I was just given the task to find a solution that is easy and will also
comply with the CAP guideline for formalin fixation documentation, and so I
started my research on the Histonet archives.  I found some good
information, but was hoping to get more feedback. 
 
Would you mind sharing with me the actions you are taking to comply with the
guideline?  
 
We are a private lab and we provide histology/pathology service for 2
hospitals and a few surgery centers.  We send our blocks to Genzyme for
Her2, and we must document on their test order sheet how many hours the
tissues have been fixed in formalin.
 
I'm assuming I will need to start keeping a log here, with documentation
that shows what time the tissue was excised and placed in formalin from the
OR, and then documentation that shows the time it was dissected and then
placed in the tissue processor.  
 
The problem that I may come across is getting the OR nurse to document the
time for us.  I don't know...maybe we need to put another sections on our
requisition form, or maybe something on the formalin container itself for
the nurse to write on.  It'll be a hassle at first but if I can get the
hospitals lab director involved, I'm sure it will work itself out.
 
Anyway, if you wouldn't mind sharing some of your ideas, I would really
appreciate it.
 
Paula Lucas
Lab Manager
Bio-Path Medical Group
Fountain Valley, CA
 

 

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Cryostat Question Please

2010-11-24 Thread Paula Lucas
Hello all
 
Has anyone used a Richard-Allan/Thermo-Fisher HM550 MP cryostat?  If so,
what is your opinion/pros and cons?
 
Thanks so much in advance,
Paula 
Lab Manager
BP Medical Group
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Coverslip Film

2010-10-07 Thread Paula Lucas
Hello 

Does anyone use a non-Sakura coverslip film on your Sakura coverslipper?  My
boss handed me a flyer from a company that sells film and it's a lot less
money than the Sakura brand.

 

Thanks,

Paula

Lab Manager

Path Lab

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Sakura Film Removal

2010-07-29 Thread Paula Lucas
Hello all - 

 

Is there a more effective and faster way to remove the film from the slide?
I had to decolorize a couple of HE slides to do special stains, and it took
5 days for the film to remove. The slides were only a few days old and I
soaked it in fresh xylene.  I almost removed it too soon because the section
was starting to lift off with the film.

 

I appreciate the help

Paula Lucas

Lab Manager

Bio-Path Medical Group

FV, CA

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Per Diem Histotech Needed (Orange County, California)

2010-07-21 Thread Paula Lucas
Hello-

 

We are located in Fountain Valley, California and we are seeking a per diem
histotech to cover during vacation times or other times when the techs are
scheduled off.  We are seeking a technician who has experience in embedding
and cutting surgical tissue cases.  Coverage START time would be early
morning hours starting between 4:30 am to 7 am, preferably 5 am start time
but I can be somewhat flexible with that.  The total hours in a day would be
between 4 and 6 hours.  

 

If interested, please give me a summary of your work history.

 

Please no recruiters/head hunters respond to my email, only qualified techs
who are interested.

Thank you,

Paula Lucas

Lab Manager

Bio-Path Medical Group

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] HE stainer/coverslippers

2010-03-03 Thread Paula Lucas
I second the Prisma with attached coverslipper.  We have been using it now
for the past 2 months or so.
It's really a non-hassle, state of the art piece of equipment.
Paula

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Liz Chlipala
Sent: Tuesday, March 02, 2010 12:50 PM
To: histonet@lists.utsouthwestern.edu; VickiGauch
Subject: RE: [Histonet] HE stainer/coverslippers

We just purchased the sakura prisma and attached glass coverslipper.  We
just love them.  We tried a lot of glass coverslippers and this by far is
the best one that we have tried.  We run both HE and specials on the prisma
and have not had a problem with the glass coverslipper.  We have been using
them for about a little over 2 months now.

Liz

Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC
Manager
Premier Laboratory, LLC
PO Box 18592
Boulder, Colorado 80308
office (303) 682-3949 
fax (303) 682-9060
www.premierlab.com
 
 
Ship to Address:
1567 Skyway Drive, Unit E
Longmont, Colorado 80504

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Rene J Buesa
Sent: Tuesday, March 02, 2010 1:39 PM
To: histonet@lists.utsouthwestern.edu; VickiGauch
Subject: Re: [Histonet] HE stainer/coverslippers

Sakura
René J.

--- On Tue, 3/2/10, Gauch, Vicki gau...@mail.amc.edu wrote:


From: Gauch, Vicki gau...@mail.amc.edu
Subject: [Histonet] HE stainer/coverslippers
To: histonet@lists.utsouthwestern.edu histonet@lists.utsouthwestern.edu
Date: Tuesday, March 2, 2010, 2:20 PM


Hi,
I was wondering what type of HE stainer/coverslipper units labs are using 
and the pros and cons of each.  We found out we have funding to replace our
old faithful GLX stainer(though it will break my heart to do so)  and are
looking for a stainer/ coverslipper that is one unit- one that we do not
have to unload from the stainer to load onto the coverslipper.  Any help
would be greatly appreciated.

Thanks so much,
Vicki Gauch
AMCH
Albany,NY



-
CONFIDENTIALITY NOTICE: This email and any attachments may contain
confidential information that is protected by law and is for the
sole use of the individuals or entities to which it is addressed.
If you are not the intended recipient, please notify the sender by
replying to this email and destroying all copies of the
communication and attachments. Further use, disclosure, copying,
distribution of, or reliance upon the contents of this email and
attachments is strictly prohibited. To contact Albany Medical
Center, or for a copy of our privacy practices, please visit us on
the Internet at www.amc.edu.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



  
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Saturday Tech Needed Orange County California

2010-02-26 Thread Paula Lucas
Hello histotechs- please email me if you're interested in working a few
hours on Saturday mornings.  We are flexible, and have another Saturday tech
that can fill in if you can not work on certain dates.  I'll give you more
info privately.  I'm only seeking histotechs who have at least 2 years
experience embedding and cutting surgical/biopsy cases in a hospital or
private lab setting.

Thanks,

Paula 

Lab Manager

Bio-Path Medical Group

Fountain Valley, CA 

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] GI, Uro, or Derm Path Lab Set Up

2010-02-05 Thread Paula Lucas
And just remember to those techs out there accepting jobs like these because
they are offering salaries that are way above the salary range, that the
investors will want their money back at some point.  And, if and when the
pod lab is suffering financially, there go the investors, then the business
and there goes your job. 

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Thomas
Jasper
Sent: Thursday, February 04, 2010 4:02 PM
To: Nails, Felton
Cc: histonet@lists.utsouthwestern.edu
Subject: RE: [Histonet] GI, Uro, or Derm Path Lab Set Up

I am in total agreement with both Richard and Felton here.  Sorry
Timothy if you are reading this but to me the bottom line is patient
care.  And in the end set-ups like this hurt patients.  These labs and
I'm using the term loosely are put in place to line the pockets of a
small number of people at the top of a pyramid.

I guess someone will always work to put things like this together if
there is a demand.  But please realize, if this is what you're involved
in, the reality is...

~ Questionable facilities - Cramming professional, OSHA legal and
ergonomically correct lab operations into small office spaces is a poor
idea at best.  Proper plumbing, ventilation and lab space are serious
considerations and fall by the wayside when these operations are put
together.

~ Qualified personnel - All of us posting on this list understand the
fact that qualified, competent Histology Lab personnel are in demand and
difficult to find.  The wage for qualified, competent technical
expertise does not fit into the plans of the originators of these
operations.

~ Proper equipment and instruments - While the aim of these operations
is limited in scope, basic functional equipment and instruments are
required.  In the overall scheme of things, pathology is still a
bargain.  However, money must be spent to properly equip a competent
working lab with modern instruments.

I could go on but that makes the basic point.  And Timothy, lest you
think that I'm shooting from the hip I assure I am not.  Like most
people on this list, I am a serious, diligent professional.  I have
interviewed people from pod labs that have worked under horrible
conditions.  I am also aware of a pod lab trying to get off the ground
that shouldn't have an ice cube's chance in hell.  The plans for this
lab do not have proper plumbing, ventilation or space and there is no
equipment.  While a position has been advertised I cannot think of a
single tech, worth his or her salt, that would consider associating
themselves with such an ill-conceived venture.

One last point - don't forget Timothy, that one day you may be that
patient.  Or it may be your mother, sister, daughter, wife, son,
friend...I'll stop there.  Every block, every slide, every piece of
tissue that I or any of my staff encounter is a precious patient...a
human life.  So trying to dumb everything down, cut corners and make a
small number of people wealthy in the process is an irresponsible and
utterly ridiculous risk.  The professionals in our field have worked
hard for a good number of years to come out of the basement spaces and
after-thought little nooks and crannies and other poor facilities to do
the highly skilled work this profession demands.  I for one will never
abide going backwards and I don't believe people of good conscience and
humanity will either.

Sincerely yours,
Tom Jasper

Thomas Jasper HT (ASCP) BAS
Histology Supervisor
Central Oregon Regional Pathology Services
Bend, Oregon 97701
541/693-2677
tjas...@copc.net 

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Nails,
Felton
Sent: Thursday, February 04, 2010 12:57 PM
To: 'Richard Cartun'; histonet@lists.utsouthwestern.edu; Timothy Jay
Subject: RE: [Histonet] GI, Uro, or Derm Path Lab Set Up

Especially when these physician hire unqualified people to run these
labs and flood the histonet with their uneducated questions.  

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Richard
Cartun
Sent: Thursday, February 04, 2010 2:29 PM
To: histonet@lists.utsouthwestern.edu; Timothy Jay
Subject: Re: [Histonet] GI, Uro, or Derm Path Lab Set Up

We don't need anymore pathology laboratories.  What we need is support
of existing laboratories, especially hospital-based labs.  GI and GU
physicians are Cherry-picking the technical revenue that should be
going to hospital labs.  Let's reform health care; make it more
efficient and less expensive.  We don't need to be putting more money in
clinicians' pockets.

Richard

Richard W. Cartun, Ph.D.
Director, Histology  Immunopathology
Director, Biospecimen Collection Programs Assistant Director, Anatomic
Pathology Hartford Hospital 80 Seymour Street Hartford, CT  06102
(860) 545-1596 Office
(860) 

[Histonet] Saturday Tech Needed- Orange County California

2009-10-16 Thread Paula Lucas
Hello,

Private lab seeking a histoch to work on Saturdays.  Will be rotating with the 
other Saturday techs. Shift starts at 5 or 6 am.

If interested, please send me an email.

Thank you,
Paula Lucas
Lab Manager
Bio-Path Medical Group
Fountain Valley, CA 92708

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Leica Bond/Novocastra Reagents

2009-10-05 Thread Paula Lucas
Hello,

We are considering the Leica Bond and their reagent rental or acquisition 
option, and I have a question regarding back orders.  

If anyone places an order through Leica for their Bond reagents (novocastra), 
are there any problems with back orders on antibodies, detection system and 
other reagents needed to run the Bond?  Or, do you usually get the items right 
away?

Thank you,
Paula Lucas
Lab Manager
Bio-Path Medical Group
Fountain Valley, CA

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Auto-IHC Staining Issues

2009-01-21 Thread Paula Lucas
Hello,

I've been having a problem with the staining quality from an auto-IHC stainer 
and I was wondering if any of you experience the same thing.

I notice sometimes that the staining in the tissue isn't complete, like there 
are areas where the fluid (one or more of the steps involved) didn't get in 
contact with the tissue, thus resulting in patchy, skipped staining.  

I'm thinking that this must be a dispensing problem.

Is this a general thing with all auto stainers - ones like DAKO or Lab Vision's 
stainer or Biocare Medical's stainer?  These stainers dispense fluid onto the 
slide. 

I hope this explains things and I really hope that I can get some feedback.  
It's getting very frustrating.

Thanks in advance,
Paula Lucas
Lab Manager
Bio-Path Medical Group
Fountain Valley, CA

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Cellient Cell Block System

2008-12-06 Thread Paula Lucas
Hello,

Can anyone tell me their experiences with the Hologic's Cellient Cell Block 
System?
I appreciate any thoughts and comments about the system you have.

Paula Lucas
Lab Manager
Bio-Path Medical Group
Fountain Valley, CA


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu