Re: [Histonet] effect of power outage on tissue processor

2018-04-20 Thread Allan Wang via Histonet
A generator is probably the cheapest option if you can go manually start it
after a few hours.

I purchased a UPS for a DNA sequencer which shouldn't lose power when in
use. You may also want one to add a few hours of leeway before the
generator is needed.
You should measure the tissue processor's power consumption during usage
with something like a Kill-A-Watt or cheaper ones to help you choose the
right size of UPS.
The E300 manual says it draws 10.5 A at 110 V which is 1200 W, but actual
usage could be significantly less if you aren't processing 300 samples.

I use this UPS and external battery:
https://www.amazon.com/TRIPP-SU2200XLCD-2200VA-1800W-Online/dp/B00CBQNBYI
https://www.amazon.com/BP48V27-2US-External-Battery-Select-
Online/dp/B00I3RROT2

This battery is also an option:
http://www.provantage.com/tripp-lite-bp48v60rt3u~7TRPL1CE.htm

It has a chart for runtimes with external batteries:
https://assets.tripplite.com/ups-runtime-chart/su2200xlcd-ru
ntime-chart-en.pdf

Allan Wang
Lab Manager
US Biomax

On Fri, Apr 20, 2018 at 9:02 AM, Matthew Fleming via Histonet <
histonet@lists.utsouthwestern.edu> wrote:

> Folks,
>
> I'm just wondering about the effects of a power outage on the tissue in a
> tissue processor. I have a small dermatopathology lab, which moved about a
> year ago to a location more prone to power outages. Last weekend the power
> was out for about 7 hours, which meant that the tissue was in 100% alcohol
> for about 5 hours, when it is programmed for 2, I believe. The tech who cut
> the tissue said it seemed a little dehydrated, but the slides looked fine.
>
> After that, I got a quote for an automatic backup generator for the
> building, but it came in at $20,000, which was much more than I was
> expecting and an expense I'd certainly like to avoid if possible. I spoke
> to the guy who maintains my equipment, who said our tissue processor (a
> Sakura E300) should not be harmed by a power outage and would pick up where
> it left off once the power returns. He also said that it can sense when the
> paraffin in the supply bins is melted and wouldn't try to use any wax that
> isn't.
>
> But still, a power outage could certainly mean that the tissue remains in a
> solution longer than programmed, as happened last weekend. My question is,
> how much of a risk does this pose? Could it damage the tissue so much as to
> make the ultimate sections uninterpretable?
>
> I should mention that I know when the power goes out, because the building
> has a fire alarm connected to a monitoring service that calls when the
> power goes out. So, as an alternative to spending $20K, I could get a
> manual generator and just go in and plug it in when the power goes out, or
> if it is out for more than an hour or two.
>
> Thanks in advance for your advice.
>
> Matthew Fleming, MD
> Fleming Dermatopathology
> Brown Deer, WI
> ___
> Histonet mailing list
> Histonet@lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
>
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] effect of power outage on tissue processor

2018-04-20 Thread Matthew Fleming via Histonet
Folks,

I'm just wondering about the effects of a power outage on the tissue in a
tissue processor. I have a small dermatopathology lab, which moved about a
year ago to a location more prone to power outages. Last weekend the power
was out for about 7 hours, which meant that the tissue was in 100% alcohol
for about 5 hours, when it is programmed for 2, I believe. The tech who cut
the tissue said it seemed a little dehydrated, but the slides looked fine.

After that, I got a quote for an automatic backup generator for the
building, but it came in at $20,000, which was much more than I was
expecting and an expense I'd certainly like to avoid if possible. I spoke
to the guy who maintains my equipment, who said our tissue processor (a
Sakura E300) should not be harmed by a power outage and would pick up where
it left off once the power returns. He also said that it can sense when the
paraffin in the supply bins is melted and wouldn't try to use any wax that
isn't.

But still, a power outage could certainly mean that the tissue remains in a
solution longer than programmed, as happened last weekend. My question is,
how much of a risk does this pose? Could it damage the tissue so much as to
make the ultimate sections uninterpretable?

I should mention that I know when the power goes out, because the building
has a fire alarm connected to a monitoring service that calls when the
power goes out. So, as an alternative to spending $20K, I could get a
manual generator and just go in and plug it in when the power goes out, or
if it is out for more than an hour or two.

Thanks in advance for your advice.

Matthew Fleming, MD
Fleming Dermatopathology
Brown Deer, WI
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet