Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-16 Thread Joseph Gwinn
In article 
[EMAIL PROTECTED],
 jlevine [EMAIL PROTECTED] wrote:

 Hello,
 
  While it's unlikely that I will soon get to build such an instrument, I
  am quite interested in how they are built, if only to understand what
  can happen and why.  Can you suggest some articles and/or books and/or
  patents delving into both the theory and the practicalities of building
  DMTD instruments?
 
We (the time and frequency division of NBS/NIST) designed and built
 a dual-mixer systerm in 1980 (more or less). This same system is the one
 that still runs the atomic clock ensemble in Boulder. You can get the 
 publications
 that describe this instrument from the publications database on our web site.
 Go to tf.nist.gov and click on the publications menu. When the menu appears,
 look for author Glaze. The stuff was published in about 1983 or so.
 There were several papers as I recall with various combinations of the folks 
 who
 built the system and the software drivers for it.

This is precisely the kind of pointer I was hoping for.  Thanks.


The system we built was totally analog, but a modern system would probably
 be fully digital. Our system had a resolution of about 0.2 ps and a
 stability of about 3-4 ps. A digital system could do better, mostly because 
 the
 temperature sensitive stuff could be confined to the analog front end whereas 
 we
 had to worry about temperature pretty much everywhere in the system.

That isn't bad for 1980 analog electronics.  I think that the 5120 is 
the digital realization, as discussed in other postings.  That said, the 
5120 is temperature sensitive, and one had to allow many hours for 
temperatures to stabilize, but then the resolution appeared to be about 
0.01 pS. I assume that the improvement from 0.2 pS was due to the fancy 
matched-mixers trick, combined with use of a very low noise oscillator.


 However, the job is not trivial, since even tiny impedance mismatches can
 cause problems at this sub-picosecond resolution. You should watch especially
 for the connectors and the cables. We typically use SMA connectors and
 rigid coax. The inputs are buffered with distribution amplifiers with
 a reverse isolation that is as good as we can make it. About -165 db, I think,
 although I have not looked at that recently. (Note that the problems are not
 adequate digital computing power but plain old analog electronics.)

As I said, I don't think I will be building such an instrument.  But 
it's just this kind of nitty gritty detail I want to be aware of, for 
interest, and for self-protection in the lab.


Even so, we have a detectable sensitivity to temperature at the
 level of ps. This noise level tends to be too small to affect the
 data from cesium standards, but it could be a problem if you were trying to
 calibrate the long-period performance of a device or a transmission system 
 that
 had a small delay, since the residual diurnal temperature sensitivity could
 come to get you. 

What we were doing was to measure the temperature coefficient of 
electrical length of a temperature-stable 10 MHz distribution amplifier, 
the goal being a tempco not exceeding 1.0 pS per degree centigade.  Some 
of the tested amps achieve ~0.5 pS/degree C, in a total delay of ~4.5 
nanoseconds, or ~111 ppm per degree C, call it 100 ppm.

The test consisted of measuring changes in total delay at three 
temperatures, 17, 24, and 31 degrees C.  The problem is that it took at 
least an hour for the amplifier to stabilize at each temperature, so 
instrument drift is a significant source of error.  The measured RC 
time constant of delay of the amplifier in chamber is 14 minutes.

My solution was to compare the amplifier under test to a mechanical 
variable delay unit (Colby Instruments PDL-100A-625PS-5.0NS), using a 
fast sampling scope (200 femtosecond rms jitter(?), averaged down to ~50 
fS) as the null detector.  

The specific circuit is a low-noise oscillator (Symmetricom 1050A) 
driving the first splitter, one output driving the scope sync input, the 
other driving the input of the second splitter.  One output of the 
second splitter drives the reference path, which contains the variable 
delay unit.  The other output drives the device path, which contains the 
amplifier under test.  Both device and reference path cables pass 
through the environmental chamber, with the heated lengths held equal.  
The cables are low tempco as well (~1.5 ppm per degree C).  Everything 
was 50-ohm, at least nominally, but no attempt at precision matching or 
isolation was made, and the connectors and adapters were a mix of 
whatever could be scrounged up in the lab.

This setup yielded clean data, easily sufficient to the purpose.  The 
main limits to accuracy appear to be hysteresis in the amplifiers under 
test, and the cyclic temperature variation of the environmental chamber 
itself.


 If you are in this business then you need professional help.

Heh.  I've been told this before, but the issue 

Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-15 Thread jlevine
Hello,

 While it's unlikely that I will soon get to build such an instrument, I
 am quite interested in how they are built, if only to understand what
 can happen and why.  Can you suggest some articles and/or books and/or
 patents delving into both the theory and the practicalities of building
 DMTD instruments?

   We (the time and frequency division of NBS/NIST) designed and built
a dual-mixer systerm in 1980 (more or less). This same system is the
one
that still runs the atomic clock ensemble in Boulder. You can get the
publications
that describe this instrument from the publications database on our
web site.
Go to tf.nist.gov and click on the publications menu. When the menu
appears,
look for author Glaze. The stuff was published in about 1983 or so.
There were
several papers as I recall with various combinations of the folks who
built the
system and the software drivers for it.
   The system we built was totally analog, but a modern system would
probably
be fully digital. Our system had a resolution of about 0.2 ps and a
stability of
about 3-4 ps. A digital system could do better, mostly because the
temperature
sensitive stuff could be confined to the analog front end whereas we
had to
worry about temperature pretty much everywhere in the system.
However,
the job is not trivial, since even tiny impedance mismatches can
cause
problems at this sub-picosecond resolution. You should watch
especially
for the connectors and the cables. We typically use SMA connectors and
rigid coax. The inputs are buffered with distribution amplifiers with
a reverse
isolation that is as good as we can make it. About -165 db, I think,
although I
have not looked at that recently. (Note that the problems are not
adequate
digital computing power but plain old analog electronics.)
   Even so, we have a detectable sensitivity to temperature at the
level of ps. This noise level tends to be too small to affect the
data  from
cesium standards, but it could be a problem if you were trying to
calibrate
the long-period performance of a device or a transmission system that
had a
small delay, since the residual diurnal temperature sensitivity could
come to
get you. If you are in this business then you need professional help.

Judah Levine
Time and Frequency Division
NIST Boulder

___
questions mailing list
questions@lists.ntp.org
https://lists.ntp.org/mailman/listinfo/questions


Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-14 Thread Uwe Klein
Joseph Gwinn wrote:

 OK.  It sounds like what the 5120 does.  I be that there are a lot of 
 details to get *exactly* right, though.
Right.

But with having a ten year old Cray in every laptop ...


uwe

___
questions mailing list
questions@lists.ntp.org
https://lists.ntp.org/mailman/listinfo/questions


Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-14 Thread jlevine
 I may need a Dual Mixer Time Difference (DMTD) instrument, to measure
 picosecond changes in electrical length in a coax plus amplifier time
 reference signal distribution system with total delays in the hundreds
 of nanoseconds, currently operating at 10 MHz (sinewave), but with 100
 MHz likely at some future date.

 What DMTD instruments are commercially available?  A google search was
 not successful - all noise no detectable signal, probably because DMTD
 instruments are not that common, and many people build their own.

   We use dual-mixer systems in our primary time scale and also to
calibrate and evaluate oscillators and timing hardware. So far as I
know,
the only units that are commercially available are made by Timing
Solutions, which was recently acquired by Symmetricom. There
are a number of different configurations, depending how how many
devices you want to measure, whether they all run at the same
frequency, etc.
   It is possible to build these devices on your own,  but it is not
trivial to get pico-second resolution and stability. Almost everything
is temperature sensitive at this level of resolution.

Judah Levine
Time and Frequency Division
NIST Boulder

___
questions mailing list
questions@lists.ntp.org
https://lists.ntp.org/mailman/listinfo/questions


Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-14 Thread Joseph Gwinn
In article [EMAIL PROTECTED],
 Uwe Klein [EMAIL PROTECTED] wrote:

 Joseph Gwinn wrote:
 
  OK.  It sounds like what the 5120 does.  I be that there are a lot of 
  details to get *exactly* right, though.
 Right.
 
 But with having a ten year old Cray in every laptop ...

Computational power must be harnessed to be useful.  I'm talking about 
the considerable human effort required for the harnessing.

Joe Gwinn

___
questions mailing list
questions@lists.ntp.org
https://lists.ntp.org/mailman/listinfo/questions


Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-14 Thread Joseph Gwinn
In article 
[EMAIL PROTECTED],
 jlevine [EMAIL PROTECTED] wrote:

  I may need a Dual Mixer Time Difference (DMTD) instrument, to measure
  picosecond changes in electrical length in a coax plus amplifier time
  reference signal distribution system with total delays in the hundreds
  of nanoseconds, currently operating at 10 MHz (sinewave), but with 100
  MHz likely at some future date.
 
  What DMTD instruments are commercially available?  A google search was
  not successful - all noise no detectable signal, probably because DMTD
  instruments are not that common, and many people build their own.
 
We use dual-mixer systems in our primary time scale and also to
 calibrate and evaluate oscillators and timing hardware. So far as I
 know, the only units that are commercially available are made by Timing
 Solutions, which was recently acquired by Symmetricom. There
 are a number of different configurations, depending how how many
 devices you want to measure, whether they all run at the same
 frequency, etc.

That's been what I'm finding, and now this is being confirmed.

I don't know why Symmetricom keeps the 5120 under their hat.  It's 
really a strange story - the only way to find out that the 5120 is a 
DMTD instrument (done up in all-digital DSP form) was by knowing that 
TSC used to make an analog DMTD instrument, and following TSC's (and 
specifically Dr Stein's) trail in the literature.


It is possible to build these devices on your own,  but it is not
 trivial to get pico-second resolution and stability. Almost everything
 is temperature sensitive at this level of resolution.

I think such instruments are also sensitive to user mood.


While it's unlikely that I will soon get to build such an instrument, I 
am quite interested in how they are built, if only to understand what 
can happen and why.  Can you suggest some articles and/or books and/or 
patents delving into both the theory and the practicalities of building 
DMTD instruments?

Thanks,

Joe Gwinn

___
questions mailing list
questions@lists.ntp.org
https://lists.ntp.org/mailman/listinfo/questions

Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-13 Thread Uwe Klein
Joseph Gwinn wrote:
 I may need a Dual Mixer Time Difference (DMTD) instrument, to measure 
 picosecond changes in electrical length in a coax plus amplifier time 
 reference signal distribution system with total delays in the hundreds 
 of nanoseconds, currently operating at 10 MHz (sinewave), but with 100 
 MHz likely at some future date.
 
 What DMTD instruments are commercially available?  A google search was 
 not successful - all noise no detectable signal, probably because DMTD 
 instruments are not that common, and many people build their own.
 
 Thanks,
 
 Joe Gwinn

Take one of the better GS DSO's that have high storage depth.
Read the shots from the DSO and do all further processing in software?

uwe

___
questions mailing list
questions@lists.ntp.org
https://lists.ntp.org/mailman/listinfo/questions


Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-13 Thread John Ackermann N8UR
Joseph Gwinn said the following on 05/12/2008 10:38 PM:

 What DMTD instruments are commercially available?  A google search was 
 not successful - all noise no detectable signal, probably because DMTD 
 instruments are not that common, and many people build their own.
 
 The silence, the silence.  I have not found too many commercial DMTF 
 units, but I have found one, although the maker does not market it a 
 such:
 
 The Symmetricom 5120 
 http://www.symmttm.com/products_pn_adev_test_sets_5120A.asp is at 
 heart a digital DMTD instrument, and will make all the usual DMTD 
 measurements, although it is marketed primarily as a phase noise test 
 set.
 
 What else is available?  

The 5120A is truly a wonderful box, but it's also not cheap (about
$30K).  It's fully DSP based so all the interesting stuff is done in
software.  One huge advantage is that the reference and
device-under-test do not have to be at the same frequency.  There's an
older version, the 5110A, that has been discontinued but should sell
used for less than $10K if you can find one.  It's more of a pure DMTD
box and doesn't do phase noise in a useful way.

I don't know of other commercially marketed products that provide a DMTD
function.  However, there's been quite a bit of discussion about this
over on the time-nuts list, and that's probably a better place for your
question (https://www.febo.com/mailman/listinfo/time-nuts).

The single most critical piece of a DMTD system is the zero crossing
detector.  Unless you have a way to increase the slew rate of the low
frequency beat note by a million or so, trigger jitter in the counter
will eat up almost all the advantages of the down-mix.  Again, there's
been some discussion about this on time-nuts, and there are some folks
there working on designing and building bits of the hardware (at least,
a couple of months ago there was a fair bit of discussion on the point).

John
___
questions mailing list
questions@lists.ntp.org
https://lists.ntp.org/mailman/listinfo/questions


Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-13 Thread Joseph Gwinn
In article [EMAIL PROTECTED],
 David L. Mills [EMAIL PROTECTED] wrote:

 Joseph,
 
 I took a look at the instrument instruction manual to see what is going 
 on. In typical todayspeak, Symmetricom doesn't say how the gadget works. 
 I make it what used to be called a Costas direct-conversion receiver. 
 The test signal is connected to two mixers; the reference oscillator is 
 connected to the other mixer inputs in quadrature. The mixer outputs are 
 digitized and filtered, the Q signal is shifted 90 degrees from the I 
 signal and combinted. The result is a baseband SSB dignal which is then 
 Fourier transformed for display. Is this what you have in mind?

Yes, but not quite the whole story.  Although impossible to discern from 
Symmetricom's 5120 datasheet and users guide, there is more to it than 
that.

I found this instrument by accident while researching the literature for 
DMTD information.  This search led me to Timing Solutions Corp (which 
was bought by Symmetricom in 2006) and  Direct-Digital Phase-Noise 
Measurement,  J. Grove, J. Hein, J. Retta, P. Schweiger, W. Solbrig, 
and S.R. Stein, 2004 IEEE International Ultrasonics, Ferroelectrics, and 
Frequency Control Joint 50th Anniversary Conference, pages 287-291.  But 
if this is an advance in the technology, there could be a patent, and 
there was: Two-Channel Digital Phase Detector, US Patent 7,227,346 to 
Wayne E. Solbrig.

I then approached Symmetricom, which led me to the 5120 (1 MHz to 30 
MHz) and the 5125 (future, 1 MHz to 400 MHz).  A section of the above 
article appears in the 5120 users guide.  

I have no idea why Symmetricom doesn't really mention that the 5120 can 
do these things, but I assume that the market for phase noise test sets 
vastly exceeds all other markets for a 5120-like instrument.  

I borrowed an early demo 5120 instrument, and in my somewhat slapdash 
lab setup, it was easily able to resolve 0.01 picosecond (eyeball rms 
width of the traces) changes in delay at 10 MHz while using a very quiet 
oscillator (a Symmetricom 1050A), after warming up overnight.

Joe Gwinn


 Dave
 
 Joseph Gwinn wrote:
 
  In article [EMAIL PROTECTED],
   Joseph Gwinn [EMAIL PROTECTED] wrote:
  
  
 I may need a Dual Mixer Time Difference (DMTD) instrument, to measure 
 picosecond changes in electrical length in a coax plus amplifier time 
 reference signal distribution system with total delays in the hundreds 
 of nanoseconds, currently operating at 10 MHz (sinewave), but with 100 
 MHz likely at some future date.
 
 What DMTD instruments are commercially available?  A google search was 
 not successful - all noise no detectable signal, probably because DMTD 
 instruments are not that common, and many people build their own.
  
  
  The silence, the silence.  I have not found too many commercial DMTF 
  units, but I have found one, although the maker does not market it a 
  such:
  
  The Symmetricom 5120 
  http://www.symmttm.com/products_pn_adev_test_sets_5120A.asp is at 
  heart a digital DMTD instrument, and will make all the usual DMTD 
  measurements, although it is marketed primarily as a phase noise test 
  set.
  
  What else is available?  
  
  
  Joe Gwinn

___
questions mailing list
questions@lists.ntp.org
https://lists.ntp.org/mailman/listinfo/questions


Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-13 Thread Joseph Gwinn
In article [EMAIL PROTECTED],
 Uwe Klein [EMAIL PROTECTED] wrote:

 Joseph Gwinn wrote:
  I may need a Dual Mixer Time Difference (DMTD) instrument, to measure 
  picosecond changes in electrical length in a coax plus amplifier time 
  reference signal distribution system with total delays in the hundreds 
  of nanoseconds, currently operating at 10 MHz (sinewave), but with 100 
  MHz likely at some future date.
  
  What DMTD instruments are commercially available?  A google search was 
  not successful - all noise no detectable signal, probably because DMTD 
  instruments are not that common, and many people build their own.
  
  Thanks,
  
  Joe Gwinn
 
 Take one of the better GS DSO's that have high storage depth.
 Read the shots from the DSO and do all further processing in software?

I don't understand how this would work.  Could you expand the 
description?   And what is GS?

Thanks,

Joe Gwinn

___
questions mailing list
questions@lists.ntp.org
https://lists.ntp.org/mailman/listinfo/questions


Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-13 Thread Uwe Klein
Joseph Gwinn wrote:
 In article [EMAIL PROTECTED],
  Uwe Klein [EMAIL PROTECTED] wrote:
 
 
Joseph Gwinn wrote:

I may need a Dual Mixer Time Difference (DMTD) instrument, to measure 
picosecond changes in electrical length in a coax plus amplifier time 
reference signal distribution system with total delays in the hundreds 
of nanoseconds, currently operating at 10 MHz (sinewave), but with 100 
MHz likely at some future date.

What DMTD instruments are commercially available?  A google search was 
not successful - all noise no detectable signal, probably because DMTD 
instruments are not that common, and many people build their own.

Thanks,

Joe Gwinn

Take one of the better GS DSO's that have high storage depth.
Read the shots from the DSO and do all further processing in software?
 
 
 I don't understand how this would work.  Could you expand the 
 description?   And what is GS? 
GS as in GigaSample

http://www.unusualresearch.com/AppNotes/TimeNuts/OptDualMixer.pdf
http://www.wriley.com/paper6ht.htm

if my understanding is correct:
take a large syncronous sampling of both signals.
extract the data.
retrace in software the math done in hardware on the aquired data set.
i.e. if you do a soft mixdown to DC you should get two vectors (R/I)
describing the phase relationship between both signals.

uwe

___
questions mailing list
questions@lists.ntp.org
https://lists.ntp.org/mailman/listinfo/questions


Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-13 Thread David L. Mills
Uwe,

A Costas receiver does what I think the 5120 does. You can buy one, 
called a software defined rario, for less than $1000. It consists of two 
double-balanced mixers converting to baseband. The I and Q signals are 
sent to a 24-bit sound card and ordinary PC. The rest is done by a DSP 
program, which does the filtering and combining. What makes this a 
Costas receiver is that the synthesized local oscillator generates the I 
and Q mixer signal 90 degrees out of phase and the I and Q channels 
baseband processing has to by 90 degrees out of phase as well, which is 
the hard part. The sound card of course has a lower frequency limit of a 
few Hz; it really should use direct-coupled ADCs.

The result is exactly equivalent to an SSB receiver, which reveals the 
baseband phase noise and anything else that gets in the way. Do a fast 
Fourier transform and see the dBc characteristic. Not really very novel 
and I would think not patentable due prior art. Do you know the patent 
number or name?

Dave

Uwe Klein wrote:
 Joseph Gwinn wrote:
 
 In article [EMAIL PROTECTED],
  Uwe Klein [EMAIL PROTECTED] wrote:


 Joseph Gwinn wrote:

 I may need a Dual Mixer Time Difference (DMTD) instrument, to 
 measure picosecond changes in electrical length in a coax plus 
 amplifier time reference signal distribution system with total 
 delays in the hundreds of nanoseconds, currently operating at 10 MHz 
 (sinewave), but with 100 MHz likely at some future date.

 What DMTD instruments are commercially available?  A google search 
 was not successful - all noise no detectable signal, probably 
 because DMTD instruments are not that common, and many people build 
 their own.

 Thanks,

 Joe Gwinn


 Take one of the better GS DSO's that have high storage depth.
 Read the shots from the DSO and do all further processing in software?



 I don't understand how this would work.  Could you expand the 
 description?   And what is GS? 
 
 GS as in GigaSample
 

 http://www.unusualresearch.com/AppNotes/TimeNuts/OptDualMixer.pdf
 http://www.wriley.com/paper6ht.htm
 
 if my understanding is correct:
 take a large syncronous sampling of both signals.
 extract the data.
 retrace in software the math done in hardware on the aquired data set.
 i.e. if you do a soft mixdown to DC you should get two vectors (R/I)
 describing the phase relationship between both signals.
 
 uwe

___
questions mailing list
questions@lists.ntp.org
https://lists.ntp.org/mailman/listinfo/questions


Re: [ntp:questions] Dual Mixer Time Difference (DMTD) instruments sought

2008-05-13 Thread Joseph Gwinn
In article [EMAIL PROTECTED],
 [EMAIL PROTECTED] (John Ackermann N8UR) wrote:

 Joseph Gwinn said the following on 05/12/2008 10:38 PM:
 
  What DMTD instruments are commercially available?  A google search was 
  not successful - all noise no detectable signal, probably because DMTD 
  instruments are not that common, and many people build their own.
  
  The silence, the silence.  I have not found too many commercial DMTF 
  units, but I have found one, although the maker does not market it a 
  such:
  
  The Symmetricom 5120 
  http://www.symmttm.com/products_pn_adev_test_sets_5120A.asp is at 
  heart a digital DMTD instrument, and will make all the usual DMTD 
  measurements, although it is marketed primarily as a phase noise test 
  set.
  
  What else is available?  
 
 The 5120A is truly a wonderful box, but it's also not cheap (about
 $30K).  It's fully DSP based so all the interesting stuff is done in
 software.  One huge advantage is that the reference and
 device-under-test do not have to be at the same frequency.  There's an
 older version, the 5110A, that has been discontinued but should sell
 used for less than $10K if you can find one.  It's more of a pure DMTD
 box and doesn't do phase noise in a useful way.

The 5110A is analog, I think, although I never did get a users guide.


 I don't know of other commercially marketed products that provide a DMTD
 function.  However, there's been quite a bit of discussion about this
 over on the time-nuts list, and that's probably a better place for your
 question (https://www.febo.com/mailman/listinfo/time-nuts).

I joined, but will lurk for now.

 
 The single most critical piece of a DMTD system is the zero crossing
 detector.  Unless you have a way to increase the slew rate of the low
 frequency beat note by a million or so, trigger jitter in the counter
 will eat up almost all the advantages of the down-mix.  Again, there's
 been some discussion about this on time-nuts, and there are some folks
 there working on designing and building bits of the hardware (at least,
 a couple of months ago there was a fair bit of discussion on the point).

Yes.  And don't forget ground loops.  Noise at 1 Hz is very difficult to 
shield.

I bet one big advantage of the DSP approach is that math is cleaner than 
practical analog hardware.

Joe Gwinn

___
questions mailing list
questions@lists.ntp.org
https://lists.ntp.org/mailman/listinfo/questions