Re: [time-nuts] 5060A question

2008-01-08 Thread Magnus Danielson
From: Bruce Griffiths [EMAIL PROTECTED]
Subject: Re: [time-nuts] 5060A question
Date: Tue, 08 Jan 2008 20:56:14 +1300
Message-ID: [EMAIL PROTECTED]

 Dave Carlson wrote:
  Hello,
 
  Since the source of the noise is the vibration of the Hot Wire Ionizer 
  ribbon inside the CBT it will require using something other than pressure 
  on 
  the outer container to snuff it out. The outer steel of the CBT is rigid.
 
  Dave
 

 Which is why it transmits the sound so well.
 Its not the pressure but the acoustic damping.
 However material which has high acoustic damping also tends to be a good
 thermal insulator, so care must be taken to avoid overheating components.

Agree.

 The idea isnt so much to stop the ribbon vibrating but attenuate the
 transmission of the vibration to the surrounding structures and
 ultimately the air.

Indeed, this was my point.

 A correctly sized and located acoustic resonator with appropriate
 internal damping can help by absorbing the sound caused by the vibration.

The fixed frequency suits our purpose here.

In the land of EMC there are there ways to reduce interference:

1) Make the transmitter less efficient
2) Make the transmission path less efficient
3) Make the receiver less efficient

For this case, 3 is clearly out of the question, but there are certainly alot
of options in rule 2 (which can be seens as rule 1 depending on where you draw
the boarder between transmitter and transmission path).

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] How to measure regulator noise?

2008-01-08 Thread Don Collie
I`d just hang an AC millivoltmeter[or microvoltmeter] across the regulator`s 
output.
I use my H/P 400H, which will give readings down to about 50uV. If your 
regulator produces less noise than this [say a 723, with
2uV], then you`ll need a more sensitive meter.
It might be wise to short the input wires of the meter, to check the 
inherent noise in the measurement setup - for meaningful results this should 
be, perhaps, no more than a tenth
of the expected reading [depending on how riggerus you want to be]. Bear in 
mind that most analogue meters are average responding, and calibrated for a 
sinewave - so to get the RMS value for white noise [or whatever] a 
correction factor should be applied [depending also on what sort of accuracy 
you need/want].
 If you want to *look* at the noise, I understand that one of the Tek 
7000 series plug-ins is able to display very small amplitudes of this order. 
Again, because you may be dealing with comparitively small voltages 
[compared with those which might be induced due to hum fields etc],
it would be wise to check the residual noise by shorting the probe[s] at the 
regulator. This is valid since the output impedance of the regulator is 
nearly zero.
Go to it!,..Don C.


- Original Message - 
From: Bruce Griffiths [EMAIL PROTECTED]
To: Discussion of precise time and frequency measurement 
time-nuts@febo.com
Sent: Tuesday, January 08, 2008 10:03 AM
Subject: Re: [time-nuts] How to measure regulator noise?


 Poul-Henning Kamp wrote:
 In message [EMAIL PROTECTED], Dave Brown 
 writes:


 John
 You won't need much of a cap for DC blocking as inputs are hi-Ztoo
 big and the chargeup may well trigger the overrange condition.


 You need to put a shunt resistor after the capacitor if the input is hi-Z
 but an even better way is to offset the gnd clip with a couple of fresh
 batteries so that you can avoid the capacitor.

 Batteries have very low, but not zero noise, so what you do is:

 Put six fresh 1.5 volt batteries in series, so that three of them
 is the wrong way and the resultant voltage is zero and then you
 measure their noise.

 Then you put three of them the wrong way on your 5v supply, so that
 the output voltage is only .5V and then you measure the noise.

 That worked fine for me using a HP6885B


 This technique only offsets the voltage to within 3/4V of zero when
 measuring an arbitrary voltage regulator output.
 With a discrete regulator it is usually easy enough for test purposes to
 adjust the regulator output so that it is within a few tens of
 millivolts of the battery stack voltage.

 A low noise preamp with perhaps 40-60dB gain is still required before
 the spectrum analyser.
 Even residual dc voltages of a few tens of millivolts may saturate such
 an amplifier.
 A simple dc servo can be used to remove the residual offset (1V)
 without using electrolytic capacitors with their attendant leakage and
 noise.
 A low frequency cutoff well below 1Hz is possible without adding
 significant noise.

 Alternatively for measurements in the 10Hz to 100kHz range a preamp like
 that used in Linear Technology's AN83:
 http://www.linear.com/pc/downloadDocument.do?id=4172
 may be useful.
 It uses electrolytic coupling capacitors together with low value 
 resistors.

 For ultralow (milliHertz) frequency noise measurement AC coupling is
 best avoided altogether, if possible.

 Bruce
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there. 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] 5060A acoustic noise issue

2008-01-08 Thread corby d dawson
Brian,

I once obtained a 5060A that was used in the time and frequency lab at
HP.

I guess it had sentimental value to them as they kept it going long after
the 5061B had come out.

It had a newer tube in it and more to the point it had one of the later
5061A/B style A11 oven controller modules installed.

This provides DC to the hot wire and no acoustic noise!

I can't remember where they shoe-horned it in as it is a taller module
than the 5060As.

These A11 units are obtainable and the modification would not be
daunting.

Best regards,

Corby 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] How to measure regulator noise?

2008-01-08 Thread Henk ten Pierick

On Jan 8, 2008, at 12:50, Don Collie wrote:

 I`d just hang an AC millivoltmeter[or microvoltmeter] across the  
 regulator`s
 output.
 I use my H/P 400H, which will give readings down to about 50uV. If  
 your
 regulator produces less noise than this [say a 723, with
 2uV], then you`ll need a more sensitive meter.

Electronic voltmeters or microvoltmeters have a noise bandwidth which  
is larger than the bandwidth on the front panel.  The front panel  
bandwidth is related to accuracy and is not the -3dB bandwidth. The  
noise bandwidth is nearly always range setting dependent and can vary  
very much. I have seen a factor of four in noise bandwidth between  
adjacent range settings. The only way to have a good indication on  
noise is in a known and constant bandwidth. The noise bandwidth of a  
filter is not the same as the 3dB bandwidth and dependent of the  
filter order and shape. It is always more.

Henk

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] How to measure regulator noise?

2008-01-08 Thread Bruce Griffiths
Don Collie wrote:
 I`d just hang an AC millivoltmeter[or microvoltmeter] across the regulator`s 
 output.
 I use my H/P 400H, which will give readings down to about 50uV. If your 
 regulator produces less noise than this [say a 723, with
 2uV], then you`ll need a more sensitive meter.
 It might be wise to short the input wires of the meter, to check the 
 inherent noise in the measurement setup - for meaningful results this should 
 be, perhaps, no more than a tenth
 of the expected reading [depending on how riggerus you want to be]. Bear in 
 mind that most analogue meters are average responding, and calibrated for a 
 sinewave - so to get the RMS value for white noise [or whatever] a 
 correction factor should be applied [depending also on what sort of accuracy 
 you need/want].
  If you want to *look* at the noise, I understand that one of the Tek 
 7000 series plug-ins is able to display very small amplitudes of this order. 
 Again, because you may be dealing with comparitively small voltages 
 [compared with those which might be induced due to hum fields etc],
 it would be wise to check the residual noise by shorting the probe[s] at the 
 regulator. This is valid since the output impedance of the regulator is 
 nearly zero.
 Go to it!,..Don C.


   
Don

This method doesnt give the noise spectrum which can be important for
use when powering VCXOs etc.
The better discrete low noise regulators have very low noise so one
still needs a low noise preamp.
For noise measurements at frequencies below 10Hz a preamp with low noise
and a low cutoff frequency can be challenging.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 5060A acoustic noise issue

2008-01-08 Thread Dave Carlson
You might try Frequency Standards and Services in Colorado Springs for any 
parts. However installing the 5061A/B A11 Dc Oven Controller into a 5060A 
may not be simple.

Dave

- Original Message - 
From: corby d dawson [EMAIL PROTECTED]
To: time-nuts@febo.com
Sent: Tuesday, January 08, 2008 8:48 AM
Subject: [time-nuts] 5060A acoustic noise issue


Brian,

I once obtained a 5060A that was used in the time and frequency lab at
HP.

I guess it had sentimental value to them as they kept it going long after
the 5061B had come out.

It had a newer tube in it and more to the point it had one of the later
5061A/B style A11 oven controller modules installed.

This provides DC to the hot wire and no acoustic noise!

I can't remember where they shoe-horned it in as it is a taller module
than the 5060As.

These A11 units are obtainable and the modification would not be
daunting.

Best regards,

Corby

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there. 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] How to measure regulator noise?

2008-01-08 Thread Don Collie
Hi Bruce and Henk,
Your point about Bandwidth is a valid one, and as you say, it depends on 
the application.
The 400H has corner frequencies of about 10Hz, and 4Mhz, so the noise BW is 
a bit wider than this [Y/N?]. This setup can`t measure noise at very low 
Frequencies, or Frequencies much greater than the [nominal]
4Mhz, because the noise frequency components simply can`t get through.
For my purposes the setup I described is adequate - I mainly want to compare 
one regulator with another [although it would be nice to be able to do so 
over the frequencies
0Hz to infinity]. I suggest that if a regulator proves itself quieter over a 
restricted BW such as I described, it is *probably* quieter at other 
frequencies too.
Take that Bruce!  ;-)..Don.

- Original Message - 
From: Bruce Griffiths [EMAIL PROTECTED]
To: Discussion of precise time and frequency measurement 
time-nuts@febo.com
Sent: Wednesday, January 09, 2008 9:26 AM
Subject: Re: [time-nuts] How to measure regulator noise?


 Henk ten Pierick wrote:
 On Jan 8, 2008, at 12:50, Don Collie wrote:


 I`d just hang an AC millivoltmeter[or microvoltmeter] across the
 regulator`s
 output.
 I use my H/P 400H, which will give readings down to about 50uV. If
 your
 regulator produces less noise than this [say a 723, with
 2uV], then you`ll need a more sensitive meter.


 Electronic voltmeters or microvoltmeters have a noise bandwidth which
 is larger than the bandwidth on the front panel.  The front panel
 bandwidth is related to accuracy and is not the -3dB bandwidth. The
 noise bandwidth is nearly always range setting dependent and can vary
 very much. I have seen a factor of four in noise bandwidth between
 adjacent range settings. The only way to have a good indication on
 noise is in a known and constant bandwidth. The noise bandwidth of a
 filter is not the same as the 3dB bandwidth and dependent of the
 filter order and shape. It is always more.

 Henk


 Thus if one is to make useful comparative measurements with an AC
 voltmeter, an external filter which has a significantly narrower
 bandpass than that of the AC voltmeter is useful.
 A spectrum analyser with the capability of averaging cross power spectra
 has the added advantage (over an AC voltmeter) of being able to make
 meaningful measurements of noise below the noise of its input amplifiers
 (or the noise of external preamplifiers). Since its difficult to build a
 preamp with noise much below 10nV/rtHz whilst ensuring that the preamp
 input will survive worst case transients etc when connected to a power
 supply of 10V or more, such a capability is useful for measuring the
 noise of ultra low noise regulators which may have high frequency noise
 of 20nV/rtHz or less.

 For example the preamp used in Linear technologies AN83 is virtually
 guaranteed to be damaged by connecting its input to a powered up 20V 
 supply.
 Off course, connecting it to the power supply before powering up the
 supply will (if the regulator output slew rate is sufficiently low)
 allow it to be used to measure the noise of higher voltage supplies (at
 least if higher voltage 330uF OSCON caps were available). However
 accidents/mistakes do happen (as do faulty regulators) and eventually
 the preamp input opamp will be destroyed.
 Using standard electrolytics for the input coupling capacitor for
 testing higher voltage supplies isnt particularly useful as they have
 increased leakage current and associated noise. Schemes such as using a
 relatively high resistance in series with the preamp input which is
 shorted out after the coupling capacitor has charged are not foolproof
 and eventually a mistake will lead to destruction of the preamp input 
 stage.

 For circuit schematics of preamps with input protection schemes that
 allow power supplies with outputs greater than 5V to be tested see:
 http://www.ko4bb.com/~bruce/LNPS.html
 http://www.ko4bb.com/%7Ebruce/LNPS.html

 Bruce

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there. 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] 5060A acoustic noise issue

2008-01-08 Thread wa1zms
Corby-

That's interesting to know that somebody has performed
the conversion at least once before.

I have a 5060 donor chassis and maybe that would be
the one to try and see how I might fit an A11 module
in place of the A20.

Thanks!

-Brian, WA1ZMS

-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED]
Behalf Of corby d dawson
Sent: Tuesday, January 08, 2008 11:48 AM
To: time-nuts@febo.com
Subject: [time-nuts] 5060A acoustic noise issue


Brian,

I once obtained a 5060A that was used in the time and frequency lab at
HP.

I guess it had sentimental value to them as they kept it going long after
the 5061B had come out.

It had a newer tube in it and more to the point it had one of the later
5061A/B style A11 oven controller modules installed.

This provides DC to the hot wire and no acoustic noise!

I can't remember where they shoe-horned it in as it is a taller module
than the 5060As.

These A11 units are obtainable and the modification would not be
daunting.

Best regards,

Corby

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] How to measure regulator noise?

2008-01-08 Thread Bruce Griffiths
Don Collie wrote:
 Hi Bruce and Henk,
 Your point about Bandwidth is a valid one, and as you say, it depends on 
 the application.
 The 400H has corner frequencies of about 10Hz, and 4Mhz, so the noise BW is 
 a bit wider than this [Y/N?]. This setup can`t measure noise at very low 
 Frequencies, or Frequencies much greater than the [nominal]
 4Mhz, because the noise frequency components simply can`t get through.
 For my purposes the setup I described is adequate - I mainly want to compare 
 one regulator with another [although it would be nice to be able to do so 
 over the frequencies
 0Hz to infinity]. I suggest that if a regulator proves itself quieter over a 
 restricted BW such as I described, it is *probably* quieter at other 
 frequencies too.
 Take that Bruce!  ;-)..Don.

   
Don

If you have a PC sound card then it can be used with a suitable preamp
to measure regulator noise spectral density over the [20Hz, 20kHz ] range.
Except for switching regulators there's usually not much noise beyond
100kHz.
Often linear regulator output noise may peak somewhere within this range.
The location and the amplitude of the noise peak depends on the
regulator output current and the characteristics of the regulator output
capacitor.
The originator of this thread needs to know the regulator noise spectrum
detail in order to properly evaluate various regulators for use in low
phase noise equipment.
The high frequency noise of a regulator is not a reliable guide to its
flicker noise characteristics.
In low phase noise design guessing isnt good enough, actual measurements
are required for verification.
Whilst it is relatively easy to filter out high frequency power supply
noise, filtering out low frequency noise is much more difficult.

Single ended circuits (most RF circuits) have little or no inherent
power supply rejection and they are particularly susceptible to phase
noise generated by power supply noise modulating the RF circuit's phase
shift.

Bruce