Re: [time-nuts] Characterising frequency standards

2009-04-12 Thread Steve Rooke
Ulrich,

2009/4/11 Ulrich Bangert df...@ulrich-bangert.de:
 So why would my counter show any significant differences
 between a 1 sec or 2 sec gate time?

 suppose your source has a 0.5 Hz frequency modulation. Would you see it with
 2 s gate time or a integer multiple of it? Would you notice it with 1 s gate
 time or an odd integer of it?

Agreed, if the source is modulated at exactly 1/2 the input frequency,
the measurement would be blind to it. So the way to account for this
would be to take half the readings, then skip one cycle and take the
other half. Examination of the data would then show the modulation.

 I've just done a Google search for dead time correction
 scheme and I just turn up results relating to particle
 physics where it seems measurements are unable to keep up
 with the flow of data, hence the need to factor in the dead
 time of system.

 Google for the STABLE32 manual. THIS literature will bring you a lot
 further, many well documented source examples in Forth and PL/1, hi. F.e.
 you may look here:

 http://www.wriley.com/

Thanks for the pointers.

Kind regards,
Steve


 Best regards
 Ulrich Bangert

 -Ursprungliche Nachricht-
 Von: time-nuts-boun...@febo.com
 [mailto:time-nuts-boun...@febo.com] Im Auftrag von Steve Rooke
 Gesendet: Freitag, 10. April 2009 12:55
 An: Discussion of precise time and frequency measurement
 Betreff: [!! SPAM] Re: [time-nuts] Characterising frequency standards


 Ulrich,

 2009/4/10 Ulrich Bangert df...@ulrich-bangert.de:
  Steve,
 
  I think the penny has dropped now, thanks. It's
 interesting that the
  ADEV calculation still works even without continuous data
 as all the
  reading I have done has led me to belive this was sacrosanct.
 
  The penny may be falling but it is not completely dropped:
 Of course
  you can feed your ADEV calculation with every second sample removed
  and setting Tau0 = 2. And of course you receive a result
 that now is
  in harmony with your all samples / Tau0 = 1 s
 computation. Had you
  done frequency measurements the reason for this appearant
 harmony is
  that your counter does not show significant different behaviour
  whether set to 1 s gate time or alternate 2 second gate time.

 So why would my counter show any significant differences
 between a 1 sec or 2 sec gate time?

  Nevertheless leaving every second sample out is NOT exactly
 the same
  as continous data with Tau0 = 2 s. Instead it is data with
 Tau0 = 1 s
  and a DEAD TIME of 1s. There are dead time correction schemes
  available in the literature.

 I've just done a Google search for dead time correction
 scheme and I just turn up results relating to particle
 physics where it seems measurements are unable to keep up
 with the flow of data, hence the need to factor in the dead
 time of system. This form of application does not appear to
 correlate with the measurement of plain oscillators. Yes
 there is dead time, per say, but I fail to see how this can
 detract significantly from continuous data given a sufficient
 data set size (as for a total measurement time).

 I guess what we need is a real data set which would show that
 this form of ADEV calculation produces incorrect results, IE.
 the proof of the pudding is in the eating.

 73,
 Steve

  Best regards
  Ulrich Bangert
 
  -Ursprungliche Nachricht-
  Von: time-nuts-boun...@febo.com
 [mailto:time-nuts-boun...@febo.com]
  Im Auftrag von Steve Rooke
  Gesendet: Donnerstag, 9. April 2009 14:00
  An: Tom Van Baak; Discussion of precise time and frequency
  measurement
  Betreff: Re: [time-nuts] Characterising frequency standards
 
 
  Tom,
 
  2009/4/9 Tom Van Baak t...@leapsecond.com:
   The first argument to the adev1 program is the sampling
  interval t0.
   The program doesn't know how far apart the input file
 samples are
   taken so it is your job to specify this. The default is 1 second.
  
   If you have data taken one second apart then t0 = 1.
   If you have data taken two seconds apart then t0 = 2.
   If you have data taken 60 seconds apart then t0 = 60, etc.
  
   If, as in your case, you take raw one second data and
 remove every
   other sample (a perfectly valid thing to do), then t0 = 2.
  
   Make sense now? It's still continuous data in the
 sense that all
   measurements are a fixed interval apart. But in any ADEV
  calculation
   you have to specify the raw data interval.
 
  I think the penny has dropped now, thanks. It's
 interesting that the
  ADEV calculation still works even without continuous data
 as all the
  reading I have done has led me to belive this was sacrosanct.
 
  What I now believe is that it's possible to measure oscillator
  performance with less than optimal test gear. This will
 enable me to
  see the effects of any experiments I make in the future.
 If you can't
  measure it, how can you know that what your doing is good or bad.
 
  73,
  Steve
  --
  Steve Rooke - ZL3TUV  G8KVD  JAKDTTNW
  Omnium finis imminet
 
  

Re: [time-nuts] Characterising frequency standards

2009-04-12 Thread Steve Rooke
Tom,

2009/4/11 Tom Van Baak t...@leapsecond.com:
 Nevertheless leaving every second sample out is NOT exactly the same as
 continous data with Tau0 = 2 s. Instead it is data with Tau0 = 1 s and a
 DEAD TIME of 1s. There are dead time correction schemes available in the
 literature.

 Ulrich, and Steve,

 Wait, are we talking phase measurements here or frequency
 measurements? My assumption with this thread is that Steve
 is simply taking phase (time error) measurements, as in my
 GPS raw data page, in which case there is no such thing as
 dead time.

Yes, phase measurements as in the original GPS.dat data set on your site.

73,
Steve


 /tvb


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




-- 
Steve Rooke - ZL3TUV  G8KVD  JAKDTTNW
Omnium finis imminet

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Characterising frequency standards

2009-04-12 Thread Steve Rooke
2009/4/11 Magnus Danielson mag...@rubidium.dyndns.org:
 Tom Van Baak skrev:
 Nevertheless leaving every second sample out is NOT exactly the same as
 continous data with Tau0 = 2 s. Instead it is data with Tau0 = 1 s and a
 DEAD TIME of 1s. There are dead time correction schemes available in the
 literature.

 Ulrich, and Steve,

 Wait, are we talking phase measurements here or frequency
 measurements? My assumption with this thread is that Steve
 is simply taking phase (time error) measurements, as in my
 GPS raw data page, in which case there is no such thing as
 dead time.

 I agree. I was also considering this earlier but put my mind to rest by
 assuming phase/time samples.

 Dead time is when the counter looses track of time in between two
 consecutive measurements. A zero dead-time counter uses the stop of one
 measure as the start of the next measure.

This becomes very important when the data to be measured has a degree
of randomness and it is therefore important to capture all the data
without any dead time. In the case of measurements of phase error in
an oscillator, it should be possible to miss some data points provided
that the frequency of capture is still known (assuming that accuracy
of drift measurements is required).

 If you have a series of time-error values taken each second and then
 drop every other sample and just recall that the time between the
 samples is now 2 seconds, then the tau0 has become 2s without causing
 dead-time. However, if the original data would have been kept, better
 statistical properties would be given, unless there is a strong
 repetitive disturbance at 2 s period, in which case it would be filtered
 out.

Indeed, there would be a loss of statistical data but this could be
made up by sampling over a period of twice the time. This system is
blind to noise at 1/2 f but ways and means could be taken to account
for that, IE. taking two data sets with a single cycle space between
them or taking another small data set with 2 cycles skipped between
each measurement.

 An example when one does get dead-time, consider a frequency counter
 which measures frequency with a gate-time of say 2 s. However, before it
 re-arms and start the next measures is takes 300 ms. The two samples
 will have 2,3 s between its start and actually spans 4,3 seconds rather
 than 4 seconds. When doing Allan Deviation calculations on such a
 measurement series, it will be biased and the bias may be compensated,
 but these days counters with zero dead-time is readily available or the
 problem can be avoided by careful consideration.

I'm looking at what can be acheieved by a budget strapped amateur who
would have trouble purchasing a later counter capable of measuring
with zero dead time.

 I believe Grenhall made some extensive analysis of the biasing of
 dead-time, so it should be available from NIST FT online library.

I'll see what I can find.

 Before zero dead-time counters was available, a setup of two counters
 was used so that they where interleaved so the dead-time was the measure
 time of the other.

I could look at doing that perhaps.

 I can collect some references to dead-time articles if anyone need them.
 I'd happy to.

73,
Steve


 Cheers,
 Magnus

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




-- 
Steve Rooke - ZL3TUV  G8KVD  JAKDTTNW
Omnium finis imminet

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Characterising frequency standards

2009-04-12 Thread Bruce Griffiths
Steve

Steve Rooke wrote:
 2009/4/11 Magnus Danielson mag...@rubidium.dyndns.org:
   
 Tom Van Baak skrev:
 
 Nevertheless leaving every second sample out is NOT exactly the same as
 continous data with Tau0 = 2 s. Instead it is data with Tau0 = 1 s and a
 DEAD TIME of 1s. There are dead time correction schemes available in the
 literature.
 
 Ulrich, and Steve,

 Wait, are we talking phase measurements here or frequency
 measurements? My assumption with this thread is that Steve
 is simply taking phase (time error) measurements, as in my
 GPS raw data page, in which case there is no such thing as
 dead time.
   
 I agree. I was also considering this earlier but put my mind to rest by
 assuming phase/time samples.

 Dead time is when the counter looses track of time in between two
 consecutive measurements. A zero dead-time counter uses the stop of one
 measure as the start of the next measure.
 

 This becomes very important when the data to be measured has a degree
 of randomness and it is therefore important to capture all the data
 without any dead time. In the case of measurements of phase error in
 an oscillator, it should be possible to miss some data points provided
 that the frequency of capture is still known (assuming that accuracy
 of drift measurements is required).

   
 If you have a series of time-error values taken each second and then
 drop every other sample and just recall that the time between the
 samples is now 2 seconds, then the tau0 has become 2s without causing
 dead-time. However, if the original data would have been kept, better
 statistical properties would be given, unless there is a strong
 repetitive disturbance at 2 s period, in which case it would be filtered
 out.
 

 Indeed, there would be a loss of statistical data but this could be
 made up by sampling over a period of twice the time. This system is
 blind to noise at 1/2 f but ways and means could be taken to account
 for that, IE. taking two data sets with a single cycle space between
 them or taking another small data set with 2 cycles skipped between
 each measurement.

   
 An example when one does get dead-time, consider a frequency counter
 which measures frequency with a gate-time of say 2 s. However, before it
 re-arms and start the next measures is takes 300 ms. The two samples
 will have 2,3 s between its start and actually spans 4,3 seconds rather
 than 4 seconds. When doing Allan Deviation calculations on such a
 measurement series, it will be biased and the bias may be compensated,
 but these days counters with zero dead-time is readily available or the
 problem can be avoided by careful consideration.
 

 I'm looking at what can be acheieved by a budget strapped amateur who
 would have trouble purchasing a later counter capable of measuring
 with zero dead time.

   
You don't need a full featured counter for this application.
One  can easily implement a zero deadtime counter or the equivalent
thereof in an FPGA.
 I believe Grenhall made some extensive analysis of the biasing of
 dead-time, so it should be available from NIST FT online library.
 

 I'll see what I can find.

   
You still need to know the phase noise spectrum of the source being
characterised.
 Before zero dead-time counters was available, a setup of two counters
 was used so that they where interleaved so the dead-time was the measure
 time of the other.
 

 I could look at doing that perhaps.

   
Very easy to do at low cost in an FPGA.
 I can collect some references to dead-time articles if anyone need them.
 I'd happy to.
 

 73,
 Steve

   
 Cheers,
 Magnus

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

 



   

Brice

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Characterising frequency standards

2009-04-12 Thread Bruce Griffiths
Steve

Steve Rooke wrote:
 If I take two sequential phase readings from an input source and place
 this into one data set and aniother two readings from the same source
 but spaced by one cycle and put this in a second data set. From the
 first data set I can calculate ADEV for tau = 1s and can calculate
 ADEV for tau = 2 sec from the second data set. If I now pre-process
 the data in the second set to remove all the effects of drift (given
 that I have already determined this), I now have two 1 sec samples
 which show a statistical difference and can be fed to ADEV with a tau0
 = 1 sec producing a result for tau = 1 sec. The results from this
 second calculation should show equal accuracy as that using the first
 data set (given the limited size of the data set).

   
You need to give far more detail as its unclear exactly what you are
doing with what samples.
Label all the phase samples and then show which samples belong to which
data set.
Also need to show clearly what you mean by skipping a cycle.

 I now collect a large data set but with a single cycle skipped between
 each sample. I feed this into ADEV using tau0 = 2 sec to produce tau
 results = 2 sec. I then pre-process the data to remove any drift and
 feed this to ADEV with a tau0 = 1 sec to produce just the tau = 1 sec
 result. I now have a complete set of results for tau = 1 sec. Agreed,
 there is the issue of modulation at 1/2 input f but ignoring this for
 the moment, this should give a valid result.

   
Again you need to give more detail.
 Now indulge me while I have a flight of fantasy.

 As the effects of jitter and phase noise will produce a statistical
 distribution of measurements, any results from these ADEV calculations
 will be limited on accuracy by the size of the data set. Only if we
 sample for a very long time will we see the very limits of the effects
 of noise. 
What noise from what source?
Noise in such measurements can originate in the measuring instrument or
the source.
For short measurement times quantisation noise and instrumental noise
may mask the noise from the source but they are still present.


 The samples which deviate the most from the median will
 occur very infrequently and it is statistically likely that they will
 not occur adjacent to another highly deviated sample. We could
 pre-process the data to remove all drift and then sort it into an
 array of increasing size. This would give the greatest deviations at
 each end of the array. For 1 sec stability the deviation would be the
 greatest difference from the median of the first and last samples in
 the array. For a 2 sec stability, this same calculation could be made
 taking the first two and last two readings in the array and
 calculating their difference from 2 x the median. This calculation
 could be continued until all the data is used for the final
 calculation. In fact the whole sorted data set could be fed to ADEV to
 produce a result that would show better worse case measurement of the
 input source which still has some statistical probability. In theory,
 if we took an infinite number of samples, there would be a whole
 string of absolutely maximum deviation measurements in a row which
 would show the absolute worse case.

 Is any of this valid or just bad physics, I don't know, but I'm sure
 it will solicit interesting comment.

   
No, not poor physics but poor statistics.

 73,
 Steve

 2009/4/10 Tom Van Baak t...@leapsecond.com:
   
 I think the penny has dropped now, thanks. It's interesting that the
 ADEV calculation still works even without continuous data as all the
 reading I have done has led me to belive this was sacrosanct.
   
 We need to be careful about what you mean by continuous.
 Let me probe a bit further to make sure you or others understand.

 The data that you first mentioned, some GPS and OCXO data at:
http://www.leapsecond.com/pages/gpsdo-sim
 was recorded once per second, for 400,000 samples without any
 interruption; that's over 4 days of continuous data.

 As you see it is very possible to extract every other, or every 10th,
 every 60th, or every Nth point from this large data set to create a
 smaller data set.

 Is it as if you had several counters all connected to the same DUT.
 Perhaps one makes a new phase measurement each second,
 another makes a measurement every 10 seconds; maybe a third
 counter just measures once a minute.

 The key here is not how often they make measurements, but that
 they all keep running at their particular rate.

 The data sets you get from these counters all represent 4 days
 of measurement; what changes is the measurement interval, the
 tau0, or whatever your ADEV tool calls it.

 Now the ADEV plots you get from these counters will all match
 perfectly with the only exception being that the every-60 second
 counter cannot give you any ADEV points for tau less than 60;
 the every-10 second counter cannot give you points for tau less
 than 10 seconds; and for that matter; the every 

Re: [time-nuts] Characterising frequency standards

2009-04-12 Thread Rex
Bruce Griffiths wrote:
 ...

 Brice

   

An impostor? An alias? :-)



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Characterising frequency standards

2009-04-12 Thread Bruce Griffiths
Rex wrote:
 Bruce Griffiths wrote:
   
 ...

 Brice

   
 

 An impostor? An alias? :-)


   
And I thought I was alluding to aliasing of the phase noise spectrum not
the characters of the alphabet.

Bruce
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

   


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Characterising frequency standards

2009-04-12 Thread Magnus Danielson
Steve Rooke skrev:
 2009/4/11 Magnus Danielson mag...@rubidium.dyndns.org:
 Tom Van Baak skrev:
 Nevertheless leaving every second sample out is NOT exactly the same as
 continous data with Tau0 = 2 s. Instead it is data with Tau0 = 1 s and a
 DEAD TIME of 1s. There are dead time correction schemes available in the
 literature.
 Ulrich, and Steve,

 Wait, are we talking phase measurements here or frequency
 measurements? My assumption with this thread is that Steve
 is simply taking phase (time error) measurements, as in my
 GPS raw data page, in which case there is no such thing as
 dead time.
 I agree. I was also considering this earlier but put my mind to rest by
 assuming phase/time samples.

 Dead time is when the counter looses track of time in between two
 consecutive measurements. A zero dead-time counter uses the stop of one
 measure as the start of the next measure.
 
 This becomes very important when the data to be measured has a degree
 of randomness and it is therefore important to capture all the data
 without any dead time. In the case of measurements of phase error in
 an oscillator, it should be possible to miss some data points provided
 that the frequency of capture is still known (assuming that accuracy
 of drift measurements is required).

Depending on the dominant noise type, the ADEV measure will be biased.

 If you have a series of time-error values taken each second and then
 drop every other sample and just recall that the time between the
 samples is now 2 seconds, then the tau0 has become 2s without causing
 dead-time. However, if the original data would have been kept, better
 statistical properties would be given, unless there is a strong
 repetitive disturbance at 2 s period, in which case it would be filtered
 out.
 
 Indeed, there would be a loss of statistical data but this could be
 made up by sampling over a period of twice the time. This system is
 blind to noise at 1/2 f but ways and means could be taken to account
 for that, IE. taking two data sets with a single cycle space between
 them or taking another small data set with 2 cycles skipped between
 each measurement.

Actually, you can take any number of 2 cycle measures and be unable to 
detect the 1/2 f oscillation without detecting it. In order to be able 
to detect it you will need to take 2 measures and be able to make an odd 
number of cycles trigger difference between them to have a chance.

The trouble is that the modulation is at the Nyquist frequency of the 1 
cycle data, so it will fold down to DC on sampling it at half-rate. 
Canceling it from other DC offset errors could be challenging.

Sampling it at 1/3 rate would discover it thought.

 An example when one does get dead-time, consider a frequency counter
 which measures frequency with a gate-time of say 2 s. However, before it
 re-arms and start the next measures is takes 300 ms. The two samples
 will have 2,3 s between its start and actually spans 4,3 seconds rather
 than 4 seconds. When doing Allan Deviation calculations on such a
 measurement series, it will be biased and the bias may be compensated,
 but these days counters with zero dead-time is readily available or the
 problem can be avoided by careful consideration.
 
 I'm looking at what can be acheieved by a budget strapped amateur who
 would have trouble purchasing a later counter capable of measuring
 with zero dead time.

Beleive me, that's where I am too. Patience and saving money for things 
I really want and allowing accumulation over time has allowed me some 
pretty fancy tools in my private lab. Infact I have to lend some of my 
gear to commercial labs as I outperform them...

 I believe Grenhall made some extensive analysis of the biasing of
 dead-time, so it should be available from NIST FT online library.
 
 I'll see what I can find.

I recalled wrong. You should look for Barnes Tables of Bias Functions, 
B1 and B2, for Variance Based on Finite Samples of Processes with Power 
Law Spectral Densities, NBS Technical Note 375, Janurary 1969 as well 
as Barnes and Allan Variance Based on Data with Dead Time Between the 
Mesurements NIST Technical Note 1318, 1990.

A ahort into to the subject is found in NIST Special Publication 1065 by 
W.J. Riley as found on http://www.wriley.com along other excelent 
material. The good thing about that material is that he gives good 
references, as one should.

 Before zero dead-time counters was available, a setup of two counters
 was used so that they where interleaved so the dead-time was the measure
 time of the other.
 
 I could look at doing that perhaps.

You should have two counters of equivalent performance, preferably same 
model. It's a rather expensive approach IMHO.

Have a look at the possibility of picking up a HP 5371A or 5372A. You 
can usually snag one for about 600 USD or 1000 USD respectively on Ebay.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go 

Re: [time-nuts] Characterising frequency standards

2009-04-12 Thread Magnus Danielson
Bruce Griffiths skrev:
 Rex wrote:
 Bruce Griffiths wrote:
   
 ...

 Brice

   
 
 An impostor? An alias? :-)


   
 And I thought I was alluding to aliasing of the phase noise spectrum not
 the characters of the alphabet.

So it is not a case of shot noise of Bruce fingers? :)
I know mine has some, and besides that there are several bugs in the 
language unit...

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Characterising frequency standards

2009-04-12 Thread Bruce Griffiths
Hej Magnus

Magnus Danielson wrote:
 Bruce Griffiths skrev:
   
 Rex wrote:
 
 Bruce Griffiths wrote:
   
   
 ...

 Brice

   
 
 
 An impostor? An alias? :-)


   
   
 And I thought I was alluding to aliasing of the phase noise spectrum not
 the characters of the alphabet.
 

 So it is not a case of shot noise of Bruce fingers? :)
 I know mine has some, and besides that there are several bugs in the 
 language unit...

 Cheers,
 Magnus

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

   
More a case of digital jitter.
Perhaps the control system phase noise was too high.

Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Characterising frequency standards

2009-04-12 Thread Mark Sims

Hello Steve,

Try this...  take Tom's sample data set,  run the numbers.  Then,  using a good 
random number generator,  make another data set by randomly throwing out half 
(or more) of the samples (to simulate a non ZDT counter).  Run the numbers 
again.  See how they change.  This should give you a good idea of how using a 
standard counter would affect your adev numbers.




_
Rediscover HotmailĀ®: Get e-mail storage that grows with you. 
http://windowslive.com/RediscoverHotmail?ocid=TXT_TAGLM_WL_HM_Rediscover_Storage1_042009
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 1938 revisited

2009-04-12 Thread wje
Fluke.l (China) was selling a number of 1938's on Ebay. I snagged one 
just to have a piece of HP history.
It works just fine, but I've noticed something a little strange.

Comparing the 1938 to both my cesium and GPS standards, there's a 
distinct periodic 1ns phase shift every second. Seems to smoothly 
advance  1ns for 500 ms, then retard back to the original phase point 
over the next 500ms. Question: is this to be expected? I'm assuming this 
is from the AFC loop, but I would have expected it to be better damped.

I've searched both this group and the HP group; there really doesn't 
seem to be a great deal of info about these, other than schematics and 
some nice variance plots on leapsecond. Just what is the serial port and 
the PIC data lines useful for, if anything?

-- 
Bill Ezell
--
They said 'Windows or better'
so I used Linux.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 1938 revisited

2009-04-12 Thread Richard (Rick) Karlquist
wje wrote:
 Fluke.l (China) was selling a number of 1938's on Ebay. I snagged one 
 just to have a piece of HP history.
 It works just fine, but I've noticed something a little strange.
 
 Comparing the 1938 to both my cesium and GPS standards, there's a 
 distinct periodic 1ns phase shift every second. Seems to smoothly 
 advance  1ns for 500 ms, then retard back to the original phase point 
 over the next 500ms. Question: is this to be expected? I'm assuming this 
 is from the AFC loop, but I would have expected it to be better damped.

The AFC loop (as opposed to EFC) is a purely analog loop in which there
is no mechanism for a 1 Hz oscillation.

Now if you are talking about the EFC, that is another story.  Depending
on how you are driving the EFC, you can pickup noise from any number
of sources.

I vaguely remember there was an LED that flashed at 1 Hz if everything
was working OK.  You might see if that is the source of what you are
seeing.


 
 I've searched both this group and the HP group; there really doesn't 
 seem to be a great deal of info about these, other than schematics and 
 some nice variance plots on leapsecond. Just what is the serial port and 
 the PIC data lines useful for, if anything?


The serial port can be used to change the oven temperature set point and
IIRC monitor various oven parameters.  Probably nothing you want to
play with if you want a working oscillator.  If you want to experiment,
I have released the software to talk to the serial port.  It should
be archived somewhere.  In production, we ramped the temperature up
and down and found the exact turnover temperature for each individual
oscillator and set the oven to the turnover.  This required that we
make crystals that actually had a turnover.  Many 10811 crystals did
not have turnovers, only an inflection point.

Rick Karlquist N6RK
Designer of E1938A


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Characterising frequency standards

2009-04-12 Thread Steve Rooke
Bruce,

2009/4/12 Bruce Griffiths bruce.griffi...@xtra.co.nz:
 Steve

 Steve Rooke wrote:
 If I take two sequential phase readings from an input source and place
 this into one data set and aniother two readings from the same source
 but spaced by one cycle and put this in a second data set. From the
 first data set I can calculate ADEV for tau = 1s and can calculate
 ADEV for tau = 2 sec from the second data set. If I now pre-process
 the data in the second set to remove all the effects of drift (given
 that I have already determined this), I now have two 1 sec samples
 which show a statistical difference and can be fed to ADEV with a tau0
 = 1 sec producing a result for tau = 1 sec. The results from this
 second calculation should show equal accuracy as that using the first
 data set (given the limited size of the data set).


 You need to give far more detail as its unclear exactly what you are
 doing with what samples.
 Label all the phase samples and then show which samples belong to which
 data set.
 Also need to show clearly what you mean by skipping a cycle.

Say I have a 1Hz input source and my counter measures the period of
the first cycle and assigns this to A1. At the end of the first cycle
the counter is able to be rest and re-triggered to capture the second
cycle and assign this to A2. So far 2 sec have passed and I have two
readings in data set A.

I now repeat the experiment and assign the measurement of the first
period to B1. The counter I am using this time is unable to stop at
the end of the first measurement and retrigger immediately so I'm
unable to measure the second cycle but is left in the armed position.
When the third cycle starts, the counter triggers and completes the
measurement of the third cycle which is now assigned to B2.

For the purposes of my original text, the first data set refers to A1
 A2. Similarly the second data set refers to B1  B2. Reference to
pre-processing of the second data set refers to mathematically
removing the effects of drift from B1  B2 to produce a third data set
which is used as the data input for an ADEV calculation where tau0 = 1
sec with output of tau = 1 sec.


 I now collect a large data set but with a single cycle skipped between
 each sample. I feed this into ADEV using tau0 = 2 sec to produce tau
 results = 2 sec. I then pre-process the data to remove any drift and
 feed this to ADEV with a tau0 = 1 sec to produce just the tau = 1 sec
 result. I now have a complete set of results for tau = 1 sec. Agreed,
 there is the issue of modulation at 1/2 input f but ignoring this for
 the moment, this should give a valid result.


 Again you need to give more detail.

In this case the data set is constructed from the measurement of the
cycle periods of a 1Hz input source where even cycles are skipped,
hence each data point is a measurement of the period of each odd (1,
3, 5, 7...) cycle of the incoming waveform. In this case the time
between each measurement is 2 sec so ADEV is calculated with tau = 2
sec for tau = 2 sec. This data set is then mathematically processed
to remove the effects of drift, bearing in mind the 2 sec spacing of
each data point, and ADEV is then calculated with tau0 = 1 sec for tau
= 1 sec.


 Now indulge me while I have a flight of fantasy.

 As the effects of jitter and phase noise will produce a statistical
 distribution of measurements, any results from these ADEV calculations
 will be limited on accuracy by the size of the data set. Only if we
 sample for a very long time will we see the very limits of the effects
 of noise.

 What noise from what source?

PN - White noise phase WPM, Flicker noise phase FPM, White noise
frequency WFM, Flicker noise frequency FFM and Random walk frequency
RWFM.

 Noise in such measurements can originate in the measuring instrument or
 the source.

Indeed, and this is an important aspect to consider as we have been
discussing the effects of induced jitter/PN to a frequency standard
when it is buffered and divided down. Ideally measurements of ADEV
would be made on the raw frequency standard source (eg. 10MHz) rather
than, say, a divided 1Hz signal.

 For short measurement times quantisation noise and instrumental noise
 may mask the noise from the source but they are still present.

Well, these form the noise floor of our measurement system.



 The samples which deviate the most from the median will
 occur very infrequently and it is statistically likely that they will
 not occur adjacent to another highly deviated sample. We could
 pre-process the data to remove all drift and then sort it into an
 array of increasing size. This would give the greatest deviations at
 each end of the array. For 1 sec stability the deviation would be the
 greatest difference from the median of the first and last samples in
 the array. For a 2 sec stability, this same calculation could be made
 taking the first two and last two readings in the array and
 calculating their difference from 2 x the median. This 

Re: [time-nuts] Characterising frequency standards

2009-04-12 Thread Steve Rooke
2009/4/13 Magnus Danielson mag...@rubidium.dyndns.org:
 Dead time is when the counter looses track of time in between two
 consecutive measurements. A zero dead-time counter uses the stop of one
 measure as the start of the next measure.

 This becomes very important when the data to be measured has a degree
 of randomness and it is therefore important to capture all the data
 without any dead time. In the case of measurements of phase error in
 an oscillator, it should be possible to miss some data points provided
 that the frequency of capture is still known (assuming that accuracy
 of drift measurements is required).

 Depending on the dominant noise type, the ADEV measure will be biased.

If the noise has a component related to the measurement frequency,
agreed, but I have already commented on that before.

 Indeed, there would be a loss of statistical data but this could be
 made up by sampling over a period of twice the time. This system is
 blind to noise at 1/2 f but ways and means could be taken to account
 for that, IE. taking two data sets with a single cycle space between
 them or taking another small data set with 2 cycles skipped between
 each measurement.

 Actually, you can take any number of 2 cycle measures and be unable to
 detect the 1/2 f oscillation without detecting it. In order to be able
 to detect it you will need to take 2 measures and be able to make an odd
 number of cycles trigger difference between them to have a chance.

Agreed.

 The trouble is that the modulation is at the Nyquist frequency of the 1
 cycle data, so it will fold down to DC on sampling it at half-rate.
 Canceling it from other DC offset errors could be challenging.

Comparing the frequency calculated from the data would show a 2Hz
offset with the fundamental frequency of the source.

 Sampling it at 1/3 rate would discover it thought.

Agreed.

 I'm looking at what can be acheieved by a budget strapped amateur who
 would have trouble purchasing a later counter capable of measuring
 with zero dead time.

 Beleive me, that's where I am too. Patience and saving money for things
 I really want and allowing accumulation over time has allowed me some
 pretty fancy tools in my private lab. Infact I have to lend some of my
 gear to commercial labs as I outperform them...

Well, that's a goal for me but I'm looking at what is achievable in
the short term instead of sitting on my hands.

 I recalled wrong. You should look for Barnes Tables of Bias Functions,
 B1 and B2, for Variance Based on Finite Samples of Processes with Power
 Law Spectral Densities, NBS Technical Note 375, Janurary 1969 as well
 as Barnes and Allan Variance Based on Data with Dead Time Between the
 Mesurements NIST Technical Note 1318, 1990.

 A ahort into to the subject is found in NIST Special Publication 1065 by
 W.J. Riley as found on http://www.wriley.com along other excelent
 material. The good thing about that material is that he gives good
 references, as one should.

Thanks for the pointer.

 I could look at doing that perhaps.

 You should have two counters of equivalent performance, preferably same
 model. It's a rather expensive approach IMHO.

It may still be cheaper than the purchase of a counter capable of
continuous collection, especially if you already have a counter that
is capable at 1/2 f.

 Have a look at the possibility of picking up a HP 5371A or 5372A. You
 can usually snag one for about 600 USD or 1000 USD respectively on Ebay.

I'd have to be a really good boy for Santa to bring me something of
that ilk. Perhaps the lotto will come up one day :-)

73,
Steve

 Cheers,
 Magnus

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




-- 
Steve Rooke - ZL3TUV  G8KVD  JAKDTTNW
Omnium finis imminet

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 1938 revisited

2009-04-12 Thread J. L. Trantham
I, too, snagged one of these since it has the reputation of being the
ultimate achievement of crystal oscillator technology with the goal of
building a GPS controlled reference using a Brooks Shera controller card and
a GPS receiver.

Toward that end, since it takes a few minutes for the 1938 to 'lock', is
there a signal that can be used to 'turn on' disciplining?  It doesn't seem
logical to try to discipline the oscillator until it has stabilized and,
therefore, if the device is to be used from time to time, not staying on
continuously, it will need a way to optimize turn on performance.

Thanks in advance and I would appreciate any 'links' to any programs that
would allow communication with the oscillator.

Joe

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Richard (Rick) Karlquist
Sent: Sunday, April 12, 2009 6:33 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 1938 revisited

wje wrote:
 Fluke.l (China) was selling a number of 1938's on Ebay. I snagged one 
 just to have a piece of HP history.
 It works just fine, but I've noticed something a little strange.
 
 Comparing the 1938 to both my cesium and GPS standards, there's a 
 distinct periodic 1ns phase shift every second. Seems to smoothly 
 advance  1ns for 500 ms, then retard back to the original phase point 
 over the next 500ms. Question: is this to be expected? I'm assuming this 
 is from the AFC loop, but I would have expected it to be better damped.

The AFC loop (as opposed to EFC) is a purely analog loop in which there
is no mechanism for a 1 Hz oscillation.

Now if you are talking about the EFC, that is another story.  Depending
on how you are driving the EFC, you can pickup noise from any number
of sources.

I vaguely remember there was an LED that flashed at 1 Hz if everything
was working OK.  You might see if that is the source of what you are
seeing.


 
 I've searched both this group and the HP group; there really doesn't 
 seem to be a great deal of info about these, other than schematics and 
 some nice variance plots on leapsecond. Just what is the serial port and 
 the PIC data lines useful for, if anything?


The serial port can be used to change the oven temperature set point and
IIRC monitor various oven parameters.  Probably nothing you want to
play with if you want a working oscillator.  If you want to experiment,
I have released the software to talk to the serial port.  It should
be archived somewhere.  In production, we ramped the temperature up
and down and found the exact turnover temperature for each individual
oscillator and set the oven to the turnover.  This required that we
make crystals that actually had a turnover.  Many 10811 crystals did
not have turnovers, only an inflection point.

Rick Karlquist N6RK
Designer of E1938A


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 1938 revisited

2009-04-12 Thread Richard (Rick) Karlquist


J. L. Trantham wrote:
 I, too, snagged one of these since it has the reputation of being the
 ultimate achievement of crystal oscillator technology with the goal of

Thanks, we thought it was pretty good :-)

 
 Toward that end, since it takes a few minutes for the 1938 to 'lock', is
 there a signal that can be used to 'turn on' disciplining?  It doesn't seem
 logical to try to discipline the oscillator until it has stabilized and,
 therefore, if the device is to be used from time to time, not staying on
 continuously, it will need a way to optimize turn on performance.

There is nothing in the E1938A that locks.  The closest to lock is the
oven current cutting back.  After that, the oven will settle within
a few minutes to its set point.  However, the crystal will exhibit high
aging rates for hours.  It will not get back to the aging it had when
last turned off for something like a day.  During this recovery time,
you can discipline the oscillator if you like, but it won't give very
good holdover performance if you lose GPS.

No crystal oscillator will give the kind of performance you need for
this application unless it is continuously ovenized.  The SC cut is
a big improvement over the AT cut in terms of a cold start, but is
still not adequate for GPS work right after a cold start.

Hope that helps.

Rick Karlquist N6RK
Designer of E1938A

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] time-nuts Rockwell Jupiter 12-channel GPS receiver OEM module

2009-04-12 Thread Mike Monett

  If nobody  has  already  mentioned   it,  Fluke.1  has  the Rockwell
  Tu00-D205 high  performance 12-channel GPS receiver OEM  modules for
  $9.99 ea with free shipping worldwide, item number: 290306684157

  These appear to have the 10KHz output described at
  http://www.jrmiller.demon.co.uk/projects/ministd/frqstd.htm

  tvb measured the performance at
  http://www.leapsecond.com/pages/gpsdo/

  More documentation is at
  http://www.gpskit.nl/gps-readme.html

  Mike

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 1938 revisited

2009-04-12 Thread J. L. Trantham
Thanks for the info.

My plan is to develop a stable GPS disciplined reference suitable for use as
a reference for Microwave work in the 10 GHz range that can be used in
portable locations with relatively quick start up. 

Perhaps the 1938 would be better in the shop where it could be left on for
weeks/months at a time and an LPRO 101 or a 10811 could be used for the
portable application.  Reasonably high drift rates could be accommodated if
the GPS signal is reliable and the time constants (disciplining rate, drift
rate, etc.) are appropriate.

Thanks,

Joe



-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Richard (Rick) Karlquist
Sent: Sunday, April 12, 2009 7:57 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 1938 revisited



J. L. Trantham wrote:
 I, too, snagged one of these since it has the reputation of being the
 ultimate achievement of crystal oscillator technology with the goal of

Thanks, we thought it was pretty good :-)

 
 Toward that end, since it takes a few minutes for the 1938 to 'lock', is
 there a signal that can be used to 'turn on' disciplining?  It doesn't
seem
 logical to try to discipline the oscillator until it has stabilized and,
 therefore, if the device is to be used from time to time, not staying on
 continuously, it will need a way to optimize turn on performance.

There is nothing in the E1938A that locks.  The closest to lock is the
oven current cutting back.  After that, the oven will settle within
a few minutes to its set point.  However, the crystal will exhibit high
aging rates for hours.  It will not get back to the aging it had when
last turned off for something like a day.  During this recovery time,
you can discipline the oscillator if you like, but it won't give very
good holdover performance if you lose GPS.

No crystal oscillator will give the kind of performance you need for
this application unless it is continuously ovenized.  The SC cut is
a big improvement over the AT cut in terms of a cold start, but is
still not adequate for GPS work right after a cold start.

Hope that helps.

Rick Karlquist N6RK
Designer of E1938A

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] femtosecond jitter anyone?

2009-04-12 Thread Mike Monett
   Chris

   The biggest problem with the OCXO is probably that it has a square
   wave output.

   With careful  design it is possible to achieve a jitter  of  a few
   tens of femtosec for a logic level output from a limiter,  but the
   OCXO designers are unlikely to have used such a limiter.

  [...]

   Bruce

  Bruce

  This would  be  an  excellent subject for  a  tutorial  on precision
  system design. Do you have any links to support your claim of a tens
  of femtosec for a logic level from a limiter?

  I am  not  aware of any logic family that  can  support  that jitter
  performance.

  When you  post items that stretch the state of the art, it  would be
  nice if you would show us all how to do the same.

  Mike

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts