[time-nuts] Lucent GPS "RFTGm-II-Rb"

2005-09-20 Thread Charles S. Osborne
Wonder if anyone on the list has information about a Lucent Rubidium locked
to GPS unit I have? Its relatively new, maybe five years old. Purportedly to
be from a failed cell site venture that caused the unit to be sold surplus
having never been installed. The nomenclature on it is:

Lucent: KS-24019 L109  which I assume may be an in house part number or
contract number.  The model number is:  RFTGm-II-Rb .

I've posted some pictures at:
http://campus.pari.edu/gps-freq/Lucent560F.JPG
http://campus.pari.edu/gps-freq/Lucent561F.JPG
http://campus.pari.edu/gps-freq/Lucent562F.JPG

The unit comes up and locks an OCXO to GPS. Then after some more warm up
time it switches over to the locked Rubidium side and the OCXO switches to
become the standby.

My problem with it is that when I come back the next day the Rb side has
faulted out and pushed the system back onto the OCXO as the main online 10
MHz source. Its unlikely the Rubidium lamp is old enough to be a problem, or
that the initial frequency adjustment has drifted out of range.. though I
suppose its remotely possible.

Have been running the unit on 24vdc but would try more if I thought the
normal -48 ~ 56vdc telco battery strings were used with this. It gets warm
enough on 24vdc that I didn't want to chance it until I asked. I know if its
a switcher that it might actually run cooler on higher voltage due to a
proportionate drop in current. Just haven't had a chance to reverse engineer
it all.

I see RS422 labeling on some  of the ports. But without more info I'm
proceeding with caution and somewhat at a standstill as to what's going on.
Anyone have any experience with these units?

Thanks,

Charles S. Osborne, K4CSO
Technical Director

Pisgah Astronomical Research Institute
1 PARI Drive, HC 73, Box 638
Rosman, NC 28772-9614
http://www.pari.edu
828-862-5813 direct
828-862-5554 main
828-862-5877 FAX
[EMAIL PROTECTED]


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] Unknown Efratom 10MHz Assy, information asked

2005-11-02 Thread Charles S. Osborne
Richard, and Arnold,

Very timely comments. I may be in the middle of working on one of those
Lucent units. Or maybe mine is a slightly different version. The front panel
on my unit says RFTGm-II-Rb (for the Efratom LPRO Rubidium locked side) and
RFTGm-II-XO on the GPS locked OCXO side. I placed the unit's pictures at:

http://campus.pari.edu/k4cso/gps/Lucent/Lucent560F.JPG
http://campus.pari.edu/k4cso/gps/Lucent/Lucent561F.JPG
http://campus.pari.edu/k4cso/gps/Lucent/Lucent562F.JPG

Mine has a string of failures in the 24vdc lines feeding both the Rb and
OCXO sides. Tracked down a blown 3A fuse just prior to two tantalums causing
high current. Removing them powered up a 28vdc to 5vdc DCDC converter but
didn't bring everything back alive.

On the OCXO board the same 3A 24vdc line fuse was blown and diode prior
to it. The OCXO so far isn't getting any DC on its pins. Haven't tracked
down the source of its power yet, but suspect its related to the other
problems. Probably all related to lightning damage that also took out the
GPS receiver in my Z3801A. The Lucent unit did work, but not fully, at one
time as the pictures show. But after it warmed up the Rb side would fail
back over to the OCXO side.

Does anyone have a pinout for the RS422/1PPS or Interface dB15
connectors on this type unit? And is there any software out there which will
look into the lock status/control similar to GPSCon does with the
Z3801A/Z3816As?

Thanks,
Charles S. Osborne, K4CSO
Technical Director

Pisgah Astronomical Research Institute
1 PARI Drive, HC 73, Box 638
Rosman, NC 28772-9614
http://www.pari.edu
828-862-5813 direct
828-862-5554 main
828-862-5877 FAX
[EMAIL PROTECTED]

- Original Message - 
From: "Richard H McCorkle" <[EMAIL PROTECTED]>
To: "Discussion of precise time and frequency measurement"

Sent: Monday, October 31, 2005 4:56 PM
Subject: Re: [time-nuts] Unknown Efratom 10MHz Assy, information asked


> The board you are describing is used in the Lucent RFG-M-XO reference
> frequency generator. It takes in 10 MHz on J2 from a Rubidium oscillator
and
> disciplines the Efratom SC cut OCXO on the main board from the 10 MHz
input.
> The output from the OCXO is converted to 15 MHz and supplied as the
> reference output on J4.
> The RFG-M-RB and RFG-M-XO normally mount together in a chassis (which I
didn
> 't get) and the modules are connected to the chassis by a harness with a
> DB15 on the chassis end. The pin out for the harness is as follows:
>
> DB15M TO/Pin#  Description
> Pin #
> 1RFG0 - P1-1+24v 1.3A/0.6A
> 2RFG0 - P1-2Common
> 3RFG1 - P1-1+24v 0.6A/0.4A
> 4RFG1 - P1-2Common
> 10  RFG0 - J3-1 Alarm
> 11  RFG0 - J3-2 Alarm
> 12  RFG1 - J3-1 Alarm
> 13  RFG1 - J3-2 Alarm
>
> A second interface cable ties the RB to the XO as follows:
>
> RBJ5   XOJ5
> Pin #Pin#
> 15
> 33
>
> The male DB9 is P1 and pin 1 = +24v @ 400ma, Pin 2 = Common. This will
power
> up the board so you can verify operation of the OCXO. Unfortunately that
is
> all the information I have on the unit.
> Hope this helps.
>
> - Original Message - 
> From: "Arnold Tibus" <[EMAIL PROTECTED]>
> To: "Discussion of precise time and frequency measurement"
> 
> Sent: Monday, October 31, 2005 10:02 AM
> Subject: [time-nuts] Unknown Efratom 10MHz Assy, information asked
>
>
> > Hello to the group,
> > I have got in my hands a very nice looking board
> > from EFRATOM equipped with a 10 MHz OCXO.
> > The designation on the top of the board is
> > 55761ASSY104920-001
> > the label underneath shows
> > 55761-104919-001
> > The Ocxo P/N is 105243-001, 10,00MHz
> > The board is complete with 3 x 9pin D-Type connectors
> > 2x fem. 1x male, which seem to be the connector
> > for the powering of the assy.
> > There are 2x SMA conns. fem., J2 and J3, J1 is not
> > assembled.
> >
> > For memory backup there is a mignon battery holder on.
> > The small board monted on top of the main board
> > does have a PLL system with a 74HC4046N and a lot
> > of other typical chips assembled (CD4060, 74HC08,
> > 54HC573F, AD75388Q and others.).
> > I guess this was a part of a Rb-reference device,
> > too nice to disassemble.
> >
> > Therefore I ask into the round table all the experts, if there
> > is somebody who could help me with more details, circuit diagram
> > or I/F informations etc.
> >
> > As I am anyway trying to set up a GPS disciplining system
> > for a 10 MHz OCX

[time-nuts] GPS receiver replacements for UT+

2006-01-04 Thread Charles S. Osborne
I'm looking for replacement GPS receiver daughter boards for the Z3816A or
Z3801A or CNS Clock.

Have been to Synergy's website and asked their support group where to find
Motorola UT+ form/fit/function compatible units, but they kept pointing me
at the newer smaller Motorola FS.

If this has come up before, I must have missed it. Frequent lightning storms
in summer keeps me in the hunt for these boards. Even with lightning
protection on the coaxes the EMP still gets a few every year.

Thanks,
Charles S. Osborne
Technical Director

Pisgah Astronomical Research Institute
1 PARI Drive, HC 73, Box 638
Rosman, NC 28772-9614
http://www.pari.edu
828-862-5813 direct
828-862-5554 main
828-862-5877 FAX
[EMAIL PROTECTED]


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] GPS receiver replacements for UT+

2006-01-05 Thread Charles S. Osborne
Reradiating the GPS antenna signal for isolation is an interesting idea. I
had not considered it. I had a Trimble Scout handheld GPS that used that
method to couple in a magnet mounted car rooftop GPS antenna signal into the
built in antenna inside the handheld. It used a little clip on reradiating
adapter.

Wish the Trimble Scout was still alive. It had an excellent well thought out
feature set in software. It's display gradually quit working and Trimble
considered it obsolete and unreparable when I asked six or seven years ago.
I saw one of those units in a GPS historical display in the Smithsonian
museum in Washington. That's when you know your technology is becoming
dated. Hi!

Thanks,
Charles S. Osborne
Technical Director

Pisgah Astronomical Research Institute
1 PARI Drive, HC 73, Box 638
Rosman, NC 28772-9614
http://www.pari.edu
828-862-5813 direct
828-862-5554 main
828-862-5877 FAX
[EMAIL PROTECTED]
- Original Message - 
From: "Bjorn Gabrielsson" <[EMAIL PROTECTED]>
To: "Charles S. Osborne" <[EMAIL PROTECTED]>
Cc: "Discussion of precise time and frequency measurement"

Sent: Thursday, January 05, 2006 8:53 AM
Subject: Re: [time-nuts] GPS receiver replacements for UT+


> "Charles S. Osborne" <[EMAIL PROTECTED]> writes:
>
> > If this has come up before, I must have missed it. Frequent lightning
storms
> > in summer keeps me in the hunt for these boards. Even with lightning
> > protection on the coaxes the EMP still gets a few every year.
>
> Ever tried a reradiating antenna to further separate the GPS from the
> dangerous outside? Does such a solution have any technical drawbacks
> wrt timing accuracy?
>
> It might not be cheaper than to buy a new lowcost oem receiver board,
> but if you want to protect hard to replace units it could be a
> solution.
>
> I have only used the "professional" type of kits like found here,
>
> http://www.navtechgps.com/supply/rerad.asp
>
> I know there is also a lower end market for use inside cars and the
> like,
>
> http://www.prairie.mb.ca/tri_m.htm
>
>
> -- 
>
> Björn
>


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


[time-nuts] PC clock comparison software?

2006-12-19 Thread Charles S. Osborne
I have a query. Does anyone have a favorite network time sync test software?

Here's my situation. Being at a radio and optical observatory, most
everything we do has varying degrees of time stamp criticality on the stored
data. While I could get obsessive compulsive and try to get everything
sync'd every few seconds with the Z3816A/GPSCon server, much of the data is
quite usable at 100ms or less time accuracy. I'd like better of course.

But for the moment let's just say I need to evaluate where we are, and
improve from that. Most PCs sync to our time server frequently. But PC
clocks being the cheap crap that they are, I need to have some reasonable
way of measuring how quickly they drift. ... particularly if Windows is
experiencing Attention Deficit Syndrome and forgetting to run the core
applications while they obsessively go speak to Father Bill Gates for words
of wisdom or check in with the Mother Ship or whatever it is that they seem
to need an Internet connection to do.

For test purposes I've occassionally set the time server polling routine to
once a day and seen 20 minute errors on delinquent PCs. Other PCs might stay
within a minute. If I try to run the Net-time routine every minute then I
increase the chance that the time server will be busy and not respond. A
goodly handful of these PCs are chugging along quite happily running small
applications on Pentium-166s /Win98se with limited memory. So going to XP or
Linux is not an option in half the PCs. No time, No $$. All the normal
educational, not for profit, zero budget reasons.

Is there a small application that could log the time errors when the system
clock gets set? This would help me decide how often the time needs to update
and what my time error penalty would be if I can't run it that often. I know
temperature will even enter into it as well since several PCs are in
unheated buildings.

It also would be nice if it could log when the time server does not respond.
Some applications respond very ingratiously when the time server won't
answer their query. They hang the program or never resync with the network
even when it comes back. But I can't prove that this is happening. Naturally
our IT folks say the network is running flawlessly. But I'm not that
gulible.

One other idea is to have a central PC poll the outlying PCs for their time
and compare to network time. That reverse net-time sync approach would be
even more useful since the PCs are spread all over many buildings, some a km
of fiber away, and checking/downloading/evaluating their individual logs
would be a chore in itself.

Our network is about 30 PCs on fiber OC12 ( 622Mbps) ATM with 100mbps
ethernet network branches. In the near future some video conferencing and
remote telescope users will add to the chunkiness of the network variable
load. Thoughts on my test software delemma?

Thanks, 73,

Charles S. Osborne, K4CSO
Technical Director

Pisgah Astronomical Research Institute
1 PARI Drive, HC 73, Box 638
Rosman, NC 28772-9614
http://www.pari.edu
828-862-5813 direct
828-862-5554 main
828-862-5877 FAX
[EMAIL PROTECTED]



___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] PC clock comparison software?

2006-12-20 Thread Charles S. Osborne
Lots of good info as always from this list.

I probably should mention what some of these PCs are doing.

Someone asked about the telescope control PCs. The 26m dishes have their own
dedicated GPS receivers for their local time standards. Those do the
celestial navigation and tracking via their control PCs.

Our optical telescope control PCs are almost all on XP or Win2000. So NTP
should have them covered reasonably well. Test software would confirm how
well.

The PCs that tend to misbehave the most are Win98se Pentium 100 ~ 400 MHz.
They are doing a variety of menial tasks like: seismic data, solar flare
monitoring, meteor burst recording, VLF signal recording for Sudden
Ionospheric Disturbances, and cosmic ray montoring. They churn along at 4 ~
40 samples per second typically of constant data flow to the hard drive.
Every minute or so they may push a screen capture to our website.

Others with a somewhat higher workload are doing webcam or weather station
image pushes to our website once a minute.

For time/frequency we have two Z3816As and a Ball Efratom MRT free running
Rubidium frequency standard. I tweak it using an HP53131time interval
counter compared to the Z3816A. GPSCon V1.038 is the software running on a
Win98se box. I have a newer XP box I've been trying to move to for better
serving reasons. But it keeps turning itself off every night even though
we've tried to turn off power saving and automatic updates. Its on a UPS
with four 100AH batteries, plus four 7AH gel cells on float on the Z3816A
48vdc power side. The slightest power glitch still takes the XP computer
down even though its on the UPS. But the Win98se P166 plugged into the same
outlet strip on the same UPS weathers it all for months at a time without
rebooting.

The main network servers are Linux of one variety or another. Not sure what
the latest flavors are, but mostly Red Hat Linux Fedora Core I think.

The network glitches are probably power glitch related even though most
things are on UPSs. We're pretty far up in the mountains, so lightning and
AC supply side glitches are common. And the sheer size of the facility
induces surges from EMP transferred cable to cable on site too. The primary
fiber links are via 3Com ATM Core Builders with redundant fiber links. Its
mostly once things get out to the localized 8~24 port switches that latch up
occurs I think. Diagnostic code would help troubleshoot that since I could
see which branches failed when, and if they recover. Network loading is
minimal except from the main website server to the Internet via T1 ( soon
OC3, then that will be minimally loaded too except during distance learning
and video conferencing).

tnx,
Charles S. Osborne, K4CSO
Technical Director

Pisgah Astronomical Research Institute
1 PARI Drive, HC 73, Box 638
Rosman, NC 28772-9614
http://www.pari.edu
828-862-5813 direct
828-862-5554 main
828-862-5877 FAX
[EMAIL PROTECTED]


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] LUCENT RFTG-m-RB

2006-12-22 Thread Charles S. Osborne
I'll try to get the pinout off mine when I get home and post it.

Charles S. Osborne
Technical Director

Pisgah Astronomical Research Institute
1 PARI Drive, HC 73, Box 638
Rosman, NC 28772-9614
http://www.pari.edu
828-862-5813 direct
828-862-5554 main
828-862-5877 FAX
[EMAIL PROTECTED]

- Original Message - 
From: "Norman J McSweyn" <[EMAIL PROTECTED]>
To: "Discussion of precise time and frequency measurement"

Sent: Friday, December 22, 2006 7:09 PM
Subject: Re: [time-nuts] LUCENT RFTG-m-RB


> Did anyone get the pinout for the "p1 24v" on the rfg-rb??
> Mine has an LPRO.
> Norm n3ykf
>
>
> ___
> time-nuts mailing list
> time-nuts@febo.com
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] Wanted to buy or borrow -- RFTG-m-XO L106A

2008-03-16 Thread Charles S. Osborne
John,

I have my RFTGm-II-XO (L109) open on the workbench (I'm changing the 15MHz
output to 10 MHz) and tried the sequence you mentioned below. Mine reacts
similarly. However, if you reapply a 10 MHz external source to the "10 MHz
REF In" sma it locks after maybe 15 minutes and XO frequency stabilizes.
Initially the Fault lite is on, then clears, and a minute later the "No GPS"
which was on solid, clears.

Note, if you are using the original 15 MHz output frequency, there is muting
applied when things are not stabilized and fault LED is lit.

My RFTGm-II-Rb (L108) unit isn't working. Haven't had time to continue
troubleshooting it since I moved and changed jobs last year, and am just
getting my home lab back in working shape. The Rb unit draws excess current
and blows an on board fuse. I found a few shorted tantalum chip caps two
years ago when I had it opened up. But apparently that's not all of the
problems. It took collateral damage after a lightning hit took out most GPS
connected receivers like my Z3801A in spite of a surge suppressor.

I was sort of waiting for someone on time-nuts, to answer the question of
whether the Rb is disciplined or just used to extend the holdover time.
Seems like I remember it going thru a sequence where the Rb switched over
and acted as primary for a few hours during initial power up. And once the
GPS-XO stabilized it became the ONLINE unit and the Rb went to STANDBY. But
its been almost five years since all that was working, so my memory of that
sequence could be wrong. It would make sense if the GPS-XO was primary and
the Rb acted during holdover wouldn't it?

Wish someone had a manual (or the RS422 magic decoder software). I've sort
of been patiently waiting and watching for someone on time-nuts to uncover a
copy of the Lucent cellsite documentation or something to explain the LED
bahavior on these units. It's more than just ON and OFF indicators, since
they flash as a third status indication format also.

73,
Charles S.Osborne, K4CSO
Duluth, GA

- Original Message - 
From: "John Ackermann N8UR" <[EMAIL PROTECTED]>
To: "Tom Van Baak" <[EMAIL PROTECTED]>; "Discussion of precise time and
frequency measurement" 
Sent: Saturday, March 15, 2008 3:21 PM
Subject: Re: [time-nuts] Wanted to buy or borrow -- RFTG-m-XO L106A


> I'm not sure what's going on.  I ran my XO/RB pair for 30 days, using
> the "official" interconnect cable with the monitoring software
> indicating no problems.  I took phase data from the RB output via the
> TSC and fed it into Stable32 (by the way, with a 20 second timeout, the
> 30 day TSC data capture worked fine).
>
> The results show a reasonably constant +1.5 microsecond slope over 30
> days, about 6x10e-13 offset.  There's no characteristic cliff in the
> ADEV where the GPS kicks in, but yet the offset is better than I would
> expect from the LPRO running in standalone mode.  Plots are at
> http://www.febo.com/pages/oscillators/rftg/
>
> I then pulled power on the RB unit and instead of going into proper XO
> operation, I got "NO GPS" and "FAULT" on the XO.  In separate
> experiments, pulling the interface cable or the 10 MHz cable result in
> the same faults -- the XO doesn't seem to take over as it should.
>
> And, when I measure the XO output, it is just horrible -- ADEV in the 9s
> from 0.1 second out past 10,000 seconds.
>
> I wonder -- has ANYONE gotten the -XO module to work in GPS-disciplined
> mode???
>
> John
> 
> Tom Van Baak said the following on 03/15/2008 03:25 PM:
> >> I suspect the XO module in my RFTG setup is defective, so I'd like to
> >> find another one.  It needs to be the "L106A" (not "B") version to mate
> >> with my -RB (L105A) unit.
> >>
> >> Anyone have one they'd like to sell?  Or, failing that, one I could
> >> borrow to test in my system?
> >
> > John,
> >
> > Interesting, my RFTG pair won't lock right either; I'm thinking
> > my XO module has a problem too. Do you think we're making
> > the same mistake or are both of our XO modules faulty?
> >
> > /tvb
> >
> >
> > ___
> > time-nuts mailing list -- time-nuts@febo.com
> > To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
> >
> >
>
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Racal-Dana 1992 switches

2008-03-27 Thread Charles S. Osborne
Matt,

Not seeing a better alternative posted for the Racal 1992, and having dozens
of similar bad Toko switches on some Pacific Measurements 1038-N10 Scalar
Analyzers, I decided to buy a bunch of the Jameco ones and give it a try.

Definitely not a form fit replacement. But if you're not too asthetically
picky they do fit. The leads can be manually flattened with a pair of needle
nose pliers and reformed to fit under and down to match the closer hole
spacing. Also clip off the small plastic alignment pins to allow the
switches to sit flush on the board. The process only takes a couple of
seconds per switch. A lot quicker than removal of the original switches.

The real catch is the hole size in the 1992 original switch layout is a
couple of sizes smaller on one pin than the other. Using a #54 drill bit (no
bigger) you can clean the hole out and enlarge it without taking the plated
thru barrel out of the holes on that one side.

All my switches are in and working. I didn't even put the keycaps back on as
all the manual lead forming didn't necessarily center within the original
panel cutouts and the plastic posts are adequate for use as switches. The
numerical keys in center keypad may require Dan's epoxy trick. Haven't
gotten to that yet.

Now the real question is... is there a clever way to make the Racal 1992
readout the difference in µHz between two GPS disciplined oscillators? My
only other counter, an HP5384A, is offscale at 10.000 000.000 MHz . I'm
referencing the counter with one Lucent RFTG-m-XO and clocking a Lucent
RFTGm-II-XO. Things are working well enough to be beyond my counter's
ability to see any jitter. The Racal says nanosecond time interval counter,
so I bet there's a way to subtract and increase the resolution similar to an
HP53131?

tnx,
Charles
K4CSO
Duluth, GA
- Original Message - 
From: "Matt Ettus" <[EMAIL PROTECTED]>
To: <[EMAIL PROTECTED]>; "Discussion of precise time and frequency
measurement" 
Sent: Monday, March 24, 2008 1:46 PM
Subject: Re: [time-nuts] Racal-Dana 1992 switches


> On Mon, Mar 10, 2008 at 8:47 AM, Dan Rae <[EMAIL PROTECTED]> wrote:
> > David McGaw wrote:
> >
> >  >Does anyone know of a source of push-button switches for the
> >  >Racal-Dana 1992?  I have one that over half are bad.  No luck from
> >  >Racal-Dana or their service house.
> >  >
> >  >
> >  The original makers of that switch (Omron?) stopped making them years
> >  ago.  I wonder why?
> >
> >  The last time I did this I found a switch at Jameco that was a near
> >  exact replacement.  The only difference was that the button was a
little
> >  too loose a fit and needed a dab of epoxy inside to hold it firmly in
> >  place.  These have been working in mine now for at least five years.
> >  The type I used was KIE22 (29 cents each in 100s), however the ones I
> >  got then with  a cross section shape blue plunger do not look exactly
> >  the same as the ones pictured in the current on line catalog.  Some
> >  further research is indicated obviously...
> >
> >  Once they start to go they will all need to be replaced, I'm afraid.
> >
>
> I bought the ones in the catalog, but they don't fit.  The footprint
> is bigger, for one thing.
>
> Matt
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Racal-Dana 1992 switches

2008-03-29 Thread Charles S. Osborne
Bill,

I guess my reason for wanting to see beyond 0.001 Hz is the usual, I have
reached a limit and wonder if the equipment I have is capable of more, if I
just understood the more esoteric functions I seldom use. The Racal seems
like my scientific calculator.. lots of shift functions that might be hiding
some useful capabilities with a little coaching and study.

Just saw Pete's answer, and I do have a PTS160 synthesizer. So the locked
offset idea is do'able. Have mixers, phase detectors, but may need more info
on the zero crossing detector opamp portion if you could point me in the
right direction.

>From a GPS equipment standpoint more resolution might tell me if my GPS
receivers are working good enough or marginally. I passive split the antenna
line four ways. The Lucent units lock to it with about 34~45 C/N readings.
But my Z3801A has some other issue now that I've replaced the UT+. It hasn't
locked and gives recurring "UTC receiver timeout" messages on GPScon. The
previous UT+ receiver was somewhat lightning damaged and would only lock to
one or two satellites for a few hours a day, but at least it was
communicating.

It may just be a comm issue between the Z3801A motherboard and my present
UT+, since the Z3801A motherboard does answer back with serial number etc to
GPScon. So I know the baud rate at that level is communicating between PC
and Z3801A. The UT+ is a used one from one of the Lucent RFTGm units. So
maybe there's a difference in the way they were setup. The UT+'s are 2000
vintage Synergy R5122U1112 to Lucent with V2.2 firmware and no battery.

I actually have collected enough damaged UT+ boards (from the lightning
prone place I previously worked) that I should build up an NMEA test fixture
to verify they work outside the systems they normally reside in, and that
all the parameters are set correctly. Then I can try to resurrect a few. The
M1501 LNA chips are usually what goes. And I have some ideas I want to try,
to dead bug in a MMIC fix for that obsoleted LNA chip.

Charles Osborne
K4CSO, EM74wa
Duluth, GA
- Original Message - 
From: "Bill Hawkins" <[EMAIL PROTECTED]>
To: "'Discussion of precise time and frequency measurement'"

Sent: Friday, March 28, 2008 9:09 PM
Subject: Re: [time-nuts] Racal-Dana 1992 switches


Charles S. Osborne said, in part,

"Now the real question is... is there a clever way to make the Racal
1992 readout the difference in µHz between two GPS disciplined
oscillators? My only other counter, an HP5384A, is offscale at 10.000
000.000 MHz . I'm referencing the counter with one Lucent RFTG-m-XO and
clocking a Lucent RFTGm-II-XO. Things are working well enough to be
beyond my counter's ability to see any jitter. The Racal says nanosecond
time interval counter, so I bet there's a way to subtract and increase
the resolution similar to an HP53131?"

Haven't seen the answer on this list, so perhaps it occurred privately.

The Racal 1992 is able to read the phase error between two 10 MHz
signals (A and B) in degrees. I have done this with the outputs of two
3801s, which are the only pair of frequency sources that I have. This
would be sub-nanosecond accuracy except that the display shows 3-10
degrees of jitter (difference between two successive readings). This,
however, is only 10E-9. Most people on this list are investigating areas
at least two orders of magnitude lower.

I find that the phase method gives me comparative drift errors soon
enough. An hour gets you near 10E-12. Others require measurement
intervals much shorter than that, but the phase angle method is more
than adequate for time errors that humans will notice. A drift of one
second per year is on the order of 10E-8.

It all depends on your reason for pursuing accurate time/frequency
measurement.

Bill Hawkins





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Lucent RFTG-m-XO 15MHz -> 10MHz

2008-04-13 Thread Charles S. Osborne
I've done three 15 MHz to 10 MHz conversions now, an RFTGm-II-Rb, RFTG-m-XO,
and an RFTGm-II-XO. So I'll mention the quick and dirty mod that worked. The
circuit boards are virtually identical between all three.

I pulled the 15 MHz metal cased filter out. Backtrack from the +23dBm 15 MHz
output, thru the buffer amplifier, and you'll run into the metal bandpass
filter.

I used an oscilloscope to find the 10 MHz output on the OCXO. Using a 470pf
ceramic monolithic capacitor I ac coupled off a sample of the RF via a small
coaxial cable. Then route it over to a resistive power divider into the
missing filter's output socket.

The resistive divider is a very small 1K ohm 10turn pot in series with the
coax center conductor. I used a 51 ohm from the filter output to ground to
form the other part of the divider and help terminate the buffer amp input.

This allows all the level alarm circuitry or whatever is under the daughter
board to continue functioning. The potentiometer is adjusted to drive the
output buffer amp to +23 dBm at 10 MHz output just like it was with 15 MHz
originally.

I also have the original 15 MHz passive external distribution divider for
the system. So the +23 dbm drives the splitter to give six 10 MHz outputs at
+9 dBm each. It's an interesting divider. I'll have to take it apart
sometime. It has two inputs to allow the Rubidium RFTGm-II-Rb unit's output
to become the backup source. When the XO output fails it's muted, and the Rb
takes over as online source.

By the way I modified the RFTGm-II-Rb the same way to use the buffer amp to
output 10 MHz from the LPRO. The associated boards are identical between the
Rb and the XO, just stuffed with different parts. In fact there's a spot for
the GPS receiver, its just missing parts. Maybe when I learn a lot more
about the loop constants and how they interact I'll get ambitious and try
grafting the LPRO into the XO system to see if that results in a GPS locked
Rb.

My RFTGm-II-Rb however still has problems. Thanks to the help of several on
the time-nuts list, I did get both units functioning redundantly again after
the lightning damage several years ago. But after about six or eight hours
the Rb side turns itself off. I'm thinking a DC-DC converter is probably
overheating inside and shutting down. Over the next month I hope to open it
up and confirm that. I may just use an external power supply and remove the
suspect 5v dc-dc converter as a test of the theory. Always hard to test a
thermal problem when the only way to do it is to open the box confining the
heat.

73,
Charles Osborne, K4CSO

- Original Message - 
From: "Bo Granlund" <[EMAIL PROTECTED]>
To: 
Sent: Sunday, April 13, 2008 4:08 PM
Subject: [time-nuts] Lucent RFTG-m-XO 15MHz -> 10MHz


> Hi everybody,
>
> Sorry for asking this kind of stupid question, but I've tried to do my
best
> with google and it hasn't revealed a lot of decisive results.
>
> The thing is, I have a Lucent RFTG-m-XO that outputs 1PPS and 15MHz
signal.
> I was wondering, how do you go ahead and convert the 15MHz output to a
> 10MHz output? What kind of equipment do I need in between there to make
> this happen?
>
> I'm really sorry, because I do realize this is a really stupid question,
but
> I'm kindof out of ideas, and I'm more of a software guy than an
electronics
> guy. :)
>
> Oh, there is a 10MHz input to the RFTG-m-XO. What is the purpose of this
> input?
>
> kind regards,
> Bo Granlund
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Efratom LPRO Rubidium

2008-07-06 Thread Charles S. Osborne
I have one of the Lucent RFTGm-II-Rb Rubidium cell site standards. It's
misbehaving. Knowing the amount of experience on this list, I figured I'd
ask before I dive in much deeper where I have no schematics.

Initially I believed one of the DC-DC converters was not working after
warmup. But after remoting the voltages to test points, they seem to be OK.

The system did suffer a lightning strike some years ago, which took out the
UT+ GPS receiver in the RFTGm-II-XO. But that's working after the UT+ was
replaced. And I doubt the interconnected circuits would have been affected.
Power supply for instance was floated across 24VDC of gell cell batteries,
which usually can absorb a lightning spike pretty well.

The unit powers up about +200 Hz  above 10 MHz (mine is modified to no
longer put out 15 MHz). After 30 minutes it has dropped slowly to +150 Hz.
It briefly shows an online indication then. But a few minutes later that
goes off and the fault light comes on. The frequency drops as low as +129 Hz
as things continue to heat up. A couple of days later its still in the +129
~+149 Hz range. I can heat and cool it and move the frequency ( +157 Hz at
90°F inside the Lucent box, +129Hz at 110°F), so the loop definitely isn't
in control.

I also have its companion RFTGm-II-XO connected to it with the factory
wiring harness. The XO is GPS locked and working fine exceeding my ability
to read its error accurately but sub-milliHz. Counter is clocked by Z3801A
10 MHz as reference.

Lucent diagnostic monitor software says "unit failed", "oscillator
maintenance required". My question then, is there anything one can access
that might be adjusted to bring the OCXO back within capture range? +140 Hz
seems way beyond where it should be. I thought the OCXO tuning range on
something like this was normally +/- a few Hz.

The unit is circa 1999 vintage. Low mileage, probably has two years online
time to date (so likley not an Rb lamp problem). When I bought it, it was
new in the box and still in warranty, victim of a cell site buildup deal
gone bust and surplused out. At one time the two units would back each other
up, switching from one to the other once as they warmed up.

Anything user servicable in the LPRO? I assume a trip to the real service
center would cost as much as a good used car. So if its something I can't
correct, it'll become a dissection learning experience.

One final test I haven't run, is removing the LPRO and powering it up stand
alone, just in case the C field input is the problem. I assume it should be
capable of locking with those inputs floating if the LPRO is indeed OK?
Anyone know what: BITE (Built in Test?), Xtal Mon, Lamp volt, and C-Field
should be normally?

Thanks,
Charles Osborne, K4CSO
Duluth, GA