Re: [time-nuts] OXCO Spurious Output at Line Frequencies
Hi Martyn, For such measurement you may shield your DUT and instruments. Ever using a shielded room which might not be easy for everyone or using shielded tents. You can find them made custom for 600-1000€ depending on the size. I've just ordered two of them from a company in netherlands For main line frequencies, you may differentiate what is conducted from what is radiated. For EMI, I'm using now active low pass filter at very low frequency to follow my low noise regulators and I get very good result, but honestly not tested as low as -130dBc... But on low noise PLL or oscillators I can get rid of any spurious from PSU but at 50Hz, my noise floor has never reached -130dBc... I'm working on a very low noise generator (20fs jitter 10Hz-1Mhz) at the moment and at 50 Hz, the phase noise is -80dBc/Hz which is already not bat at 6 GHz I've measured last week a Wentzel VCXO giving -150dBc/Hz @ 50 Hz and I could measure the 50 Hz at -130dBc/Hz but this had not the active filter on it. I test it again using the shielded tent exactly to know what is radiated from conducted. If still there under the tent I will try using my active filter to see if this makes a difference. Cheers Stephane -- Message d'origine -- De : "Martyn Smith"À : time-nuts@febo.com Envoyé 12/07/2016 12:44:31 Objet : [time-nuts] OXCO Spurious Output at Line Frequencies Hello, I have a customer who is measuring the phase noise of my 10 MHz ultra-low phase noise frequency standard. He is seeing spurious signals at line frequencies (50 and 100 Hz as we are in Europe) at a level around -130 dBc. My opinion is that it's impossible to get much better than that. Even running on batteries make little difference, since the equipment is in a test rack with AC signals everywhere. Even the £50k R test set he is using only quotes a spurious spec of -90 dBc. What experience does anyone have here? Best Regards Martyn ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
[time-nuts] DMTD downmixer question
Hi there, I will receive my SR620 soon and want of course to use it as well for stability measurement using the DMTD method. I've read many things on how to design the downmixer. There will be a DDS or low noise generator as LO, the two mixers, and the squarer. There are apparently many ways to do the squarer. Some of the ways I've seen are using fast comparator, logic gate, fast amplifiers... Finally is there a way which looks better than the others ? I was hesitating between a fast comparator and an ECL logic gate for instance. Thanks & cheers Stephane ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] buying a time interval counter
Ji Jerry, The Fluke exhibit a 100ps resolution whereas the SR620 has 25ps. This is the main difference I see from datasheets. cheers Stephane -- Message d'origine -- De : "Jerry" <jster...@att.net> À : "'Stéphane Rey'" <steph@wanadoo.fr>; "'Discussion of precise time and frequency measurement'" <time-nuts@febo.com>; "'Brooke Clarke'" <bro...@pacific.net> Envoyé 30/06/2016 17:01:50 Objet : RE: [time-nuts] buying a time interval counter How does the Fluke PM6690 (same as Pendulum CNT-90) compare to the SR620? A neighbor is selling one in perfect condition (per him) for $900 Jerry NY2KW -Original Message- From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Stéphane Rey Sent: Thursday, June 30, 2016 10:07 AM To: Brooke Clarke; Discussion of precise time and frequency measurement Subject: Re: [time-nuts] buying a time interval counter Hi, I've ordered a SR620 with the option 01 (higher stability standard) Should be there in August Thanks Stephane -- Message d'origine -- De : "Brooke Clarke" <bro...@pacific.net> À : "Stéphane Rey" <steph@wanadoo.fr>; "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Envoyé 29/06/2016 19:20:22 Objet : Re: [time-nuts] buying a time interval counter Hi Stephane: I traded my HP 53132A counter for an SR 620. The 53132 has what I'd call a user hostile interface, so if you are manually controlling the counter the SR 620 has a huge advantage. I also like the long display on the 620 which can be read from across the room. PS Stanford Research is a company founded by physicists and makes some really high quality stuff. In fact some of the products HP/Agilent/Keysight sells are repackaged SR instruments. http://www.prc68.com/I/TandFTE.shtml#SR620 The claim to fame for the HP 53132A is that it can make a frequency (not time interval) measurement to 1E12 in a second. Here's how to get that same result with the SR620: http://www.prc68.com/I/FTS4060.shtml#SR620Fast On the down side the printing functions on the 620 require an Epsom printer. Does anyone have a solution for that? PS SR also makes a 10 MHz crystal oscillator that has options trading stability for aging as well as the EFC tuning polarity and range so as to match other OCXOs. http://prc68.com/I/TandFTE.shtml#SC10 At one point they were looking into making a GPS time receiver where the cable length calibration would be built-in. -- Have Fun, Brooke Clarke http://www.PRC68.com http://www.end2partygovernment.com/2012Issues.html The lesser of evils is still evil. Original Message Hello there, I'm planning to buy a such instrument in order to do some frequency stability measurement at work. The SR620 seems to be discontinued. What model still distributed would you think is good for that at the moment ? Thanks & cheers Stephane ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] buying a time interval counter
Hi, I've ordered a SR620 with the option 01 (higher stability standard) Should be there in August Thanks Stephane -- Message d'origine -- De : "Brooke Clarke" <bro...@pacific.net> À : "Stéphane Rey" <steph@wanadoo.fr>; "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Envoyé 29/06/2016 19:20:22 Objet : Re: [time-nuts] buying a time interval counter Hi Stephane: I traded my HP 53132A counter for an SR 620. The 53132 has what I'd call a user hostile interface, so if you are manually controlling the counter the SR 620 has a huge advantage. I also like the long display on the 620 which can be read from across the room. PS Stanford Research is a company founded by physicists and makes some really high quality stuff. In fact some of the products HP/Agilent/Keysight sells are repackaged SR instruments. http://www.prc68.com/I/TandFTE.shtml#SR620 The claim to fame for the HP 53132A is that it can make a frequency (not time interval) measurement to 1E12 in a second. Here's how to get that same result with the SR620: http://www.prc68.com/I/FTS4060.shtml#SR620Fast On the down side the printing functions on the 620 require an Epsom printer. Does anyone have a solution for that? PS SR also makes a 10 MHz crystal oscillator that has options trading stability for aging as well as the EFC tuning polarity and range so as to match other OCXOs. http://prc68.com/I/TandFTE.shtml#SC10 At one point they were looking into making a GPS time receiver where the cable length calibration would be built-in. -- Have Fun, Brooke Clarke http://www.PRC68.com http://www.end2partygovernment.com/2012Issues.html The lesser of evils is still evil. Original Message Hello there, I'm planning to buy a such instrument in order to do some frequency stability measurement at work. The SR620 seems to be discontinued. What model still distributed would you think is good for that at the moment ? Thanks & cheers Stephane ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] buying a time interval counter
I will try to buy one of these SR620. I've some Standford & Research products like the 535 or the 30 MHz DDS generator. I do admit I'm not fan of the front panel interface as well but this is ok and usable. The SR620 would probably be PC controlled anyway to automate some measurements. The 10 MHz reference will come from a GPSDO which is broadcasted over optical fibers. However the optical SFPs have been tested to have 500fs RMS jitter which mich be pretty high for that. I've a Thunderbolt GPSDO sleeping in a box that will do the job otherwise. I plan to buy a Rb oscillator for reference for DTMD method and design a small circuit for the downmixing. To be continued. Thanks for the comments Stephane -- Message d'origine -- De : "Brooke Clarke" <bro...@pacific.net> À : "Stéphane Rey" <steph@wanadoo.fr>; "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Envoyé 29/06/2016 19:20:22 Objet : Re: [time-nuts] buying a time interval counter Hi Stephane: I traded my HP 53132A counter for an SR 620. The 53132 has what I'd call a user hostile interface, so if you are manually controlling the counter the SR 620 has a huge advantage. I also like the long display on the 620 which can be read from across the room. PS Stanford Research is a company founded by physicists and makes some really high quality stuff. In fact some of the products HP/Agilent/Keysight sells are repackaged SR instruments. http://www.prc68.com/I/TandFTE.shtml#SR620 The claim to fame for the HP 53132A is that it can make a frequency (not time interval) measurement to 1E12 in a second. Here's how to get that same result with the SR620: http://www.prc68.com/I/FTS4060.shtml#SR620Fast On the down side the printing functions on the 620 require an Epsom printer. Does anyone have a solution for that? PS SR also makes a 10 MHz crystal oscillator that has options trading stability for aging as well as the EFC tuning polarity and range so as to match other OCXOs. http://prc68.com/I/TandFTE.shtml#SC10 At one point they were looking into making a GPS time receiver where the cable length calibration would be built-in. -- Have Fun, Brooke Clarke http://www.PRC68.com http://www.end2partygovernment.com/2012Issues.html The lesser of evils is still evil. Original Message Hello there, I'm planning to buy a such instrument in order to do some frequency stability measurement at work. The SR620 seems to be discontinued. What model still distributed would you think is good for that at the moment ? Thanks & cheers Stephane ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. --- L'absence de virus dans ce courrier électronique a été vérifiée par le logiciel antivirus Avast. https://www.avast.com/antivirus ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] buying a time interval counter
Hi Dave, Yep it looks like it's still available. I've seen discontinued on a distributor website which made me thinking the product wasn't available anymore. I'll look what is the price for the HP53230A for comparison. Cheers Stephane -- Message d'origine -- De : "Dr. David Kirkby (Kirkby Microwave Ltd)" <drkir...@kirkbymicrowave.co.uk> À : "Discussion of precise time and frequency measurement" <time-nuts@febo.com>; "Stéphane Rey" <steph@wanadoo.fr> Envoyé 29/06/2016 18:22:50 Objet : Re: [time-nuts] buying a time interval counter On 29 Jun 2016 17:02, "Stéphane Rey" <steph@wanadoo.fr> wrote: > > Hello there, > > I'm planning to buy a such instrument in order to do some frequency stability measurement at work. The SR620 seems to be discontinued. It still looks available to me http://www.thinksrs.com/products/SR620.htm There is also a Keysight 53230A with a slightly better (20 ps vs 25 ps) single shot resolution. It looks more modern, but I don't know how well they compare. Dave. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
[time-nuts] buying a time interval counter
Hello there, I'm planning to buy a such instrument in order to do some frequency stability measurement at work. The SR620 seems to be discontinued. What model still distributed would you think is good for that at the moment ? Thanks & cheers Stephane ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
[time-nuts] Mike Monet, please contact me
Sorry for flooding the list but can't find a working email to contact you Mike. Thanks and cheers Stephane ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] software for HP58503 ?
Well, it works better with correct RS232 wiring :-/ Now I can connect with SatStat ang get my status. I've seen there is GPScon which sounds nice. Any other software suggested ? Cheers Stephane -- Message d'origine -- De : "Stéphane Rey" <steph@wanadoo.fr> À : "Discussion of precise time and frequency measurement" <time-nuts@febo.com> Envoyé 19/03/2016 23:09:00 Objet : [time-nuts] software for HP58503 ? Hi, I'd like to connect my HP58503 to a pc running windosw 7 Enterprise. I do not have com port, I'm using an USB FTDI converter. Which software could I use ? I've tried the Symetricon SatSat but can't connect to my GPSDO any suggestion ? Cheers Stephane --- L'absence de virus dans ce courrier électronique a été vérifiée par le logiciel antivirus Avast. https://www.avast.com/antivirus ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. --- L'absence de virus dans ce courrier électronique a été vérifiée par le logiciel antivirus Avast. https://www.avast.com/antivirus ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
[time-nuts] software for HP58503 ?
Hi, I'd like to connect my HP58503 to a pc running windosw 7 Enterprise. I do not have com port, I'm using an USB FTDI converter. Which software could I use ? I've tried the Symetricon SatSat but can't connect to my GPSDO any suggestion ? Cheers Stephane --- L'absence de virus dans ce courrier électronique a été vérifiée par le logiciel antivirus Avast. https://www.avast.com/antivirus ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] PLL book 3rd edition
Here at CERN in particle accelerators we're not looking at the same properties for PLLs than one would require for radiocom. Two features usually requested are very low short term phase jitters (100fs-1ps) which directly leads into particule beam stability, as well as being able to synchronously reset the dividers for phase sync. which is usually not possible with integrated PLL ICs like NXP/AD/Hittite ones. This is also an unsual application which is not representative of mass usage of PLLs. RF Phase stability and synchronization is a key for accelerators performance and thus the approach is different than for PLLs I'm designing for radiocom... Stephane, F1TJJ -- Message d'origine -- De : "jimlux" <jim...@earthlink.net> À : time-nuts@febo.com Envoyé 08/03/2016 14:30:51 Objet : Re: [time-nuts] PLL book 3rd edition On 3/8/16 12:19 AM, Stéphane Rey wrote: Hi Rick, There are hopefully many applications where monolythics PLL can't achieve the requested functionalitities or performances so that there is still room to build block PLLs. I'm still desiging such things for my job for instance. As do we at JPL. In fact, I'd say that given the advent of more "software driven" radios, with things like PLLs with DDS or NCO in the loop, having a text that covers performance on an analytical basis is useful. However, it's a pretty darn small market. (Considering we do a "new design" every 5-10 years) And, we use the data sheets and ap notes as a much as we'd use the textbook. What is really hard to find is good data on the noise properties of the other components. Everyone makes noise plots for the oscillators and amplifiers and publishes them. For instance, what about the noise added by the Phase Frequency Detector - that's not a spec that shows up on the data sheet. (but does get discussed here on time-nuts) ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] PLL book 3rd edition
Hi Rick, There are hopefully many applications where monolythics PLL can't achieve the requested functionalitities or performances so that there is still room to build block PLLs. I'm still desiging such things for my job for instance. Regards Stephane F1TJJ -- Message d'origine -- De : "Richard (Rick) Karlquist"À : ka2...@aol.com; "Discussion of precise time and frequency measurement" Cc: enrico.rubi...@gmail.com Envoyé 08/03/2016 04:33:24 Objet : Re: [time-nuts] PLL book 3rd edition I know for me, I mainly use the "synthesizer on a chip" IC's from Analog Devices/Hittite and National. Their data sheets and ap notes serve as the "textbook". I'm not sure there will be much call going forward for a book on fundamentals that explains how to design synthesizers from first principles using basic building blocks. Having designed PLL's for over 40 years, I know all about how to do this, yet is now a nearly useless skill with the IC's now available. Only the IC designers themselves need these skills. Occasionally I find myself mentoring these guys in the hope of getting better chips to buy :-) (I have a patent on a phase detector design that was made into a chip, but the chip is built by Keysight's captive foundry which doesn't sell much to the merchant market.) No criticism of the book; it's just a market issue. Rick N6RK On 3/7/2016 4:52 PM, KA2WEU--- via time-nuts wrote: To all : I have published the following book " Microwave and Wireless Synthesizers: Theory and Design, Ulrich L. Rohde, John Wiley & Sons, August 1997, ISBN 0-471-52019-5." and have since kind of drifted into the VCO und high stability oscillators. The first edition "Digital PLL Frequency Synthesizers - Theory and Design, Ulrich L. Rohde, Prentice-Hall, Inc., Englewood Cliffs, NJ, January 1983 " has sold more then 10 000 copies. Is there any of you out there who would like to take over a needed update and take over the resulting revenues and unfortunately also the work and glory and who feels qualified to so ? As I am more or less now in microwave technology and less in PLL IC's, I hate to see this standard textbook disappear Who can help or want to take over? Ulrich In a message dated 3/2/2016 12:04:00 P.M. Eastern Standard Time, time-nuts@febo.com writes: In a message dated 2/16/2016 9:03:59 A.M. Eastern Standard Time, time-nuts@febo.com writes:. http://www.synergymwave.com/articles/2016/calculation-of-fm-and-am.pdf ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Aeroflex IFR 3413
Hi Clint, I do not know this instrument, however let me share my current experience with my IFR2945. I've found a faulty board to be part of the main PLL. The failled component was an old CPLD. I've requested a quotation from both the CPLD and the full board. The CPLD was 90 euros with 9 months of leadtime (!) and the full board 460€ with 1 month. I've bought the full board. The company is actually Cobham Wireless who owns Marconi Aeroflex now. After 3 months nothing yet received. I had to complain. I finally got the board after 3.5 months. Put in the instrument and still not working. New release of the board with FPGA instead of CPLD. The FPGA outputs are not doing anything. Ask to the support again and it appears the board has not been programmed/tested... but they don't want to send me the code for programming (I've the tools at work). I have to send back my board After 3 weeks they tell me I should have ordered the slash T version (which costs 30% more) Well guys.. you made the quotation for that board... What could I do with a non programmed board Meanwhile my support guy tells me the facility he works in France... is going to be closed suddenly. By the way they're all fired. Cobham Wireless is under restructuration and they're closing some sites. My board has been lost ! Finally apparently they've sent me a new board, I hope programmed this time and I'm waiting for it... According to the very nice french support guy I was in contact with, in IFR2945/2965 there are very few components like that which are programmable for which this is no option other than Cobham support but at the moment it looks like an incredible mess there. Good luck Cheers Stephane -- Message d'origine -- De : "Clint Jay"À : time-nuts@febo.com Envoyé 09/02/2016 20:58:08 Objet : [time-nuts] Aeroflex IFR 3413 I've acquired from eBay a 'faulty' Aeroflex IFR 3413 signal generator with Option 001 (no attenuator) and an error 509 which I think would indicate it's had power applied to the output port. Does anyone have service information for these generators? I'm also looking for firmware upgrades and any experience of adding 'options' I.E. is it possible to add the mechanical or electronic attenuator options ? -- Clint. *No trees were harmed in the sending of this mail. However, a large number of electrons were greatly inconvenienced.* ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. --- L'absence de virus dans ce courrier électronique a été vérifiée par le logiciel antivirus Avast. https://www.avast.com/antivirus ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
[time-nuts] looking for rakable GPSDO
Hello, I'm looking for a 10 MHz output GPSDO with external antenna which would be rackable. Symmetricon doesn't seem to propose some neither Keysight. Found some stuff in Oscilloquartz. Any other brand to suggest ? Thanks & cheers Stephane ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] question Alan deviation measured with Timelab and counters
I do understand. Has anyone already compared the performances of squaring the 10 MHz vs squaring the IF ? Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Bob Camp Envoyé : dimanche 25 janvier 2015 19:01 À : Discussion of precise time and frequency measurement Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Hi The approach in the original NIST paper below was sort of a “best guess” about how to do the limiting and filtering. When the paper was presented, a number of us questioned how that part of the circuit was arrived at. The conversation more or less ended up with “that’s something we can investigate further”. The Collins paper (and Bruce’s work based on it) is a much better way to look at the 10 Hz squaring process. At 10 MHz, that stuff is not needed. Bob On Jan 25, 2015, at 10:44 AM, Stéphane Rey steph@wanadoo.fr wrote: Hi everyone. Many thanks for your very useful comments. I had already seen most of the documents you were pointing but not on the collins and Bruce discussion around the multistage filter. However I've already seen this approach in the document from Allan (http://tf.nist.gov/timefreq/general/pdf/84.pdf) At first I had in mind to square the 10 MHz but this is the aim of the evaluation board to evaluate various architectures. So I will implement several squarers including the Collins Approach both at 10 MHz and 100 Hz and all the blocks will have input and output connectors so that I will be able to test several layouts. I will show you the final design. Cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Charles Steinmetz Envoyé : dimanche 25 janvier 2015 08:08 À : Discussion of precise time and frequency measurement Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Stephane wrote: I'm now trying to evaluate various architectures of 2-channels squarers and a DMDT. For that I'm designing a PCB with 4 squarers : simple 74ac04 gate biased at VCC/2, a LT1016 comparator, the transistor based differential amplifier from Winzel and the one from Charles. Note that squaring a 10MHz sine wave and squaring a 10 or 100Hz mixer output are two very different tasks. If you start at baseband, a Collins-style multi-stage limiting amp is a great benefit. That is generally not necessary if you start at 10MHz (or if you do use a Collins-style limiter it needs far fewer stages). All of the squarers you mention work well at 10MHz, but not as well at baseband. The LT1719 is easier to apply and faster than the LT1016. You may want to use that instead of the 1016. The LT1719 and LT1715 datasheets show the simplest possible implementation (see below). The MPSH81 devices in my version are available in surface-mount (MMBTH81) if that is more convenient. Other fast transistors will also work (BFT92, BFT93, BFG31). Best regards, Charles --- L'absence de virus dans ce courrier électronique a été vérifiée par le logiciel antivirus Avast. http://www.avast.com ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. --- L'absence de virus dans ce courrier électronique a été vérifiée par le logiciel antivirus Avast. http://www.avast.com ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] question Alan deviation measured with Timelab and counters
Hi everyone. Many thanks for your very useful comments. I had already seen most of the documents you were pointing but not on the collins and Bruce discussion around the multistage filter. However I've already seen this approach in the document from Allan (http://tf.nist.gov/timefreq/general/pdf/84.pdf) At first I had in mind to square the 10 MHz but this is the aim of the evaluation board to evaluate various architectures. So I will implement several squarers including the Collins Approach both at 10 MHz and 100 Hz and all the blocks will have input and output connectors so that I will be able to test several layouts. I will show you the final design. Cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Charles Steinmetz Envoyé : dimanche 25 janvier 2015 08:08 À : Discussion of precise time and frequency measurement Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Stephane wrote: I'm now trying to evaluate various architectures of 2-channels squarers and a DMDT. For that I'm designing a PCB with 4 squarers : simple 74ac04 gate biased at VCC/2, a LT1016 comparator, the transistor based differential amplifier from Winzel and the one from Charles. Note that squaring a 10MHz sine wave and squaring a 10 or 100Hz mixer output are two very different tasks. If you start at baseband, a Collins-style multi-stage limiting amp is a great benefit. That is generally not necessary if you start at 10MHz (or if you do use a Collins-style limiter it needs far fewer stages). All of the squarers you mention work well at 10MHz, but not as well at baseband. The LT1719 is easier to apply and faster than the LT1016. You may want to use that instead of the 1016. The LT1719 and LT1715 datasheets show the simplest possible implementation (see below). The MPSH81 devices in my version are available in surface-mount (MMBTH81) if that is more convenient. Other fast transistors will also work (BFT92, BFT93, BFG31). Best regards, Charles --- L'absence de virus dans ce courrier électronique a été vérifiée par le logiciel antivirus Avast. http://www.avast.com ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] question Alan deviation measured with Timelab and counters
Hi guys. After several experiments I could discover that the bad ADEV from the two GPSDO DUT are due to GPS lock losses. This is probably because the antenna is outside the windows but half the sky is hidden. We can see the on the frequency plot the sharp change of 0.5Hz and the locking. Good point. I'm now trying to evaluate various architectures of 2-channels squarers and a DMDT. For that I'm designing a PCB with 4 squarers : simple 74ac04 gate biased at VCC/2, a LT1016 comparator, the transistor based differential amplifier from Winzel and the one from Charles. I will add two balanced mixers (minicircuits), IF filters and amplifiers. Does anyone has an idea of what I could add for this evaluation ? Cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Stéphane Rey Envoyé : mardi 20 janvier 2015 23:15 À : 'Discussion of precise time and frequency measurement' Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Hi, Following the tests results in the previous email, today I've performed additional measurements showing that the repeatability of the GPSDO DUT is not great but is coming from the design. I've tested several over sources and repeatability is correct. I can already make some measurement. Good ! Now I'd like to improve. First I'm going to implement a squarer and then I will work on the DMTD... I'm thinking to make a setup on the table, and possibly make a small PCB then. Any comment for the tests results of yesterday here under ? Cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Stéphane Rey Envoyé : lundi 19 janvier 2015 22:32 À : 'Discussion of precise time and frequency measurement' Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Hi Here are the results of today's experiments. plots and TIM files attached to this email. Setup #1 : dark blue I've done again the floor measurement with same conditions : HP58503 for 10 MHz Standard, 1PPS for the EXT gating and the Rb on channel A. Same result (hopefully) Setup #2 : Pink Then I've made what Magnus has suggested, i.e. using the 1 PPS on Channel A, the Rb on channel B and internal gating The ADEV has increased by more than 1 order of magnitude. I guess this confirms the 1PPS stability is lower than the 10 MHz Setup #3-6 : Dark Green, Red, Light blue and Dark yellow. I've measured several times the GPSDO DUT with SEParate inputs. 1PPS on EXT, Rb on channel A and DUT on channel B. This gives 4 different plots... When starting the measurement the plots starts directly at different values... Mmmm very strange. Is it coming for the setup of the GPSDO ? To be investigated further with other sources. This is the plan for tomorrow. However the overal shape of the plot sounds relevant to me. Setup #7-8 + #9 not showed here I've tested the suggested splitted same signal on both inputs with 1m coax for channel B. I've discovered that when swaping the GPSDO on the standard input and the Rb on the channel A I have a slight difference. In order to confirm I've made two time each measurement and this confirms that having the Rb on channel A and GPSDO on the standard input gives the lowest ADEV. The setup #9 which is the same than the light green gives the superimposed plot on that one... So what does it mean ? One of the two sources is better than the other, but which one ? Some other comments : - Swaping signals between channel A and B gives the same ADEV (setup #4 and 5, light blue and red) - On some measurement on the GPSDO DUT, (not displayed here), I could see during the measurement suddenly an increase of one order of magnitude. The HP5370A do not show any difference (the time interval value continues to move with a beat but visually impossible to quantify if the value between two values has increased. No explanation for that. I'll redo the test with some other sources to check if it comes for the measurement system or the GPSDO DUT In conclusion, 1. swaping the Rb and HP58503 doesn't give the same result. The GPSDO has standard seems the best (or the Rb measured) 2. the measurement on the GPSDO DUT gives different results with nearly one order of magnitude difference but shape is still the same. 3. the 1PPS must be connected on the EXT gating input What do you think ? Cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de steph.rey Envoyé : lundi 19 janvier 2015 16:44 À : Discussion of precise time and frequency measurement Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Actually I'm working in the RF department of a big lab, designing RF electronics mainly in microwaves range. I'm luckilly having some tools around to play with and a lot of components like mixers/amplifiers/couplers/splitters/attenuators
Re: [time-nuts] question Alan deviation measured with Timelab and counters
Hi, Just a stupid question on Timelab. Why do I have the plot with 1/4 for the time actually used for the measurement ? I can see that the plot is updated every 4 samples but the scale is not relevant. The sample interval is correctly set (1s) Cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Stéphane Rey Envoyé : mardi 20 janvier 2015 23:15 À : 'Discussion of precise time and frequency measurement' Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Hi, Following the tests results in the previous email, today I've performed additional measurements showing that the repeatability of the GPSDO DUT is not great but is coming from the design. I've tested several over sources and repeatability is correct. I can already make some measurement. Good ! Now I'd like to improve. First I'm going to implement a squarer and then I will work on the DMTD... I'm thinking to make a setup on the table, and possibly make a small PCB then. Any comment for the tests results of yesterday here under ? Cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Stéphane Rey Envoyé : lundi 19 janvier 2015 22:32 À : 'Discussion of precise time and frequency measurement' Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Hi Here are the results of today's experiments. plots and TIM files attached to this email. Setup #1 : dark blue I've done again the floor measurement with same conditions : HP58503 for 10 MHz Standard, 1PPS for the EXT gating and the Rb on channel A. Same result (hopefully) Setup #2 : Pink Then I've made what Magnus has suggested, i.e. using the 1 PPS on Channel A, the Rb on channel B and internal gating The ADEV has increased by more than 1 order of magnitude. I guess this confirms the 1PPS stability is lower than the 10 MHz Setup #3-6 : Dark Green, Red, Light blue and Dark yellow. I've measured several times the GPSDO DUT with SEParate inputs. 1PPS on EXT, Rb on channel A and DUT on channel B. This gives 4 different plots... When starting the measurement the plots starts directly at different values... Mmmm very strange. Is it coming for the setup of the GPSDO ? To be investigated further with other sources. This is the plan for tomorrow. However the overal shape of the plot sounds relevant to me. Setup #7-8 + #9 not showed here I've tested the suggested splitted same signal on both inputs with 1m coax for channel B. I've discovered that when swaping the GPSDO on the standard input and the Rb on the channel A I have a slight difference. In order to confirm I've made two time each measurement and this confirms that having the Rb on channel A and GPSDO on the standard input gives the lowest ADEV. The setup #9 which is the same than the light green gives the superimposed plot on that one... So what does it mean ? One of the two sources is better than the other, but which one ? Some other comments : - Swaping signals between channel A and B gives the same ADEV (setup #4 and 5, light blue and red) - On some measurement on the GPSDO DUT, (not displayed here), I could see during the measurement suddenly an increase of one order of magnitude. The HP5370A do not show any difference (the time interval value continues to move with a beat but visually impossible to quantify if the value between two values has increased. No explanation for that. I'll redo the test with some other sources to check if it comes for the measurement system or the GPSDO DUT In conclusion, 1. swaping the Rb and HP58503 doesn't give the same result. The GPSDO has standard seems the best (or the Rb measured) 2. the measurement on the GPSDO DUT gives different results with nearly one order of magnitude difference but shape is still the same. 3. the 1PPS must be connected on the EXT gating input What do you think ? Cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de steph.rey Envoyé : lundi 19 janvier 2015 16:44 À : Discussion of precise time and frequency measurement Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Actually I'm working in the RF department of a big lab, designing RF electronics mainly in microwaves range. I'm luckilly having some tools around to play with and a lot of components like mixers/amplifiers/couplers/splitters/attenuators, ... almost whatever the frequency is up to several tens of GHz. At home since the last 20 years I could as well get nice instruments. The next two measuring tools really missing and for which I'm limited are the phase noise and stability measurement and possibly a good standard. My Effratom FRK Rb is old and probably not the best from a phase noise and stability point of view but until now has never been characterized. Otherwise I've almost everything I
Re: [time-nuts] question Alan deviation measured with Timelab and counters
Hi, Following the tests results in the previous email, today I've performed additional measurements showing that the repeatability of the GPSDO DUT is not great but is coming from the design. I've tested several over sources and repeatability is correct. I can already make some measurement. Good ! Now I'd like to improve. First I'm going to implement a squarer and then I will work on the DMTD... I'm thinking to make a setup on the table, and possibly make a small PCB then. Any comment for the tests results of yesterday here under ? Cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Stéphane Rey Envoyé : lundi 19 janvier 2015 22:32 À : 'Discussion of precise time and frequency measurement' Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Hi Here are the results of today's experiments. plots and TIM files attached to this email. Setup #1 : dark blue I've done again the floor measurement with same conditions : HP58503 for 10 MHz Standard, 1PPS for the EXT gating and the Rb on channel A. Same result (hopefully) Setup #2 : Pink Then I've made what Magnus has suggested, i.e. using the 1 PPS on Channel A, the Rb on channel B and internal gating The ADEV has increased by more than 1 order of magnitude. I guess this confirms the 1PPS stability is lower than the 10 MHz Setup #3-6 : Dark Green, Red, Light blue and Dark yellow. I've measured several times the GPSDO DUT with SEParate inputs. 1PPS on EXT, Rb on channel A and DUT on channel B. This gives 4 different plots... When starting the measurement the plots starts directly at different values... Mmmm very strange. Is it coming for the setup of the GPSDO ? To be investigated further with other sources. This is the plan for tomorrow. However the overal shape of the plot sounds relevant to me. Setup #7-8 + #9 not showed here I've tested the suggested splitted same signal on both inputs with 1m coax for channel B. I've discovered that when swaping the GPSDO on the standard input and the Rb on the channel A I have a slight difference. In order to confirm I've made two time each measurement and this confirms that having the Rb on channel A and GPSDO on the standard input gives the lowest ADEV. The setup #9 which is the same than the light green gives the superimposed plot on that one... So what does it mean ? One of the two sources is better than the other, but which one ? Some other comments : - Swaping signals between channel A and B gives the same ADEV (setup #4 and 5, light blue and red) - On some measurement on the GPSDO DUT, (not displayed here), I could see during the measurement suddenly an increase of one order of magnitude. The HP5370A do not show any difference (the time interval value continues to move with a beat but visually impossible to quantify if the value between two values has increased. No explanation for that. I'll redo the test with some other sources to check if it comes for the measurement system or the GPSDO DUT In conclusion, 1. swaping the Rb and HP58503 doesn't give the same result. The GPSDO has standard seems the best (or the Rb measured) 2. the measurement on the GPSDO DUT gives different results with nearly one order of magnitude difference but shape is still the same. 3. the 1PPS must be connected on the EXT gating input What do you think ? Cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de steph.rey Envoyé : lundi 19 janvier 2015 16:44 À : Discussion of precise time and frequency measurement Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Actually I'm working in the RF department of a big lab, designing RF electronics mainly in microwaves range. I'm luckilly having some tools around to play with and a lot of components like mixers/amplifiers/couplers/splitters/attenuators, ... almost whatever the frequency is up to several tens of GHz. At home since the last 20 years I could as well get nice instruments. The next two measuring tools really missing and for which I'm limited are the phase noise and stability measurement and possibly a good standard. My Effratom FRK Rb is old and probably not the best from a phase noise and stability point of view but until now has never been characterized. Otherwise I've almost everything I need up to 40 GHz I guess. I'm doing further measurement right now which sounds much much more consistent. I will share tonight. Cheers Stephane On Mon, 19 Jan 2015 08:59:58 -0500, Bob Camp kb...@n1k.org wrote: Hi On Jan 18, 2015, at 5:12 PM, Stéphane Rey steph@wanadoo.fr wrote: Bonsoir Magnus (Are you in Sweeden ?) Being able to measure high stability and low phase noise is definitely a need for me as I'm trying to design low noise synthesizers and I'm already reaching the limits of my current tools for phase noise and I
Re: [time-nuts] question Alan deviation measured with Timelab and counters
Bonsoir Magnus (Are you in Sweeden ?) Being able to measure high stability and low phase noise is definitely a need for me as I'm trying to design low noise synthesizers and I'm already reaching the limits of my current tools for phase noise and I can't afford an E5052 for my own. At work I've one but I will probably not stay after august. And anyway I need such tools in my lab at home... As low-noise and stable synthetizers depends on the standard used, I need as well to measure them as well... Let's start with this simple experiments and once I will understand the ins and outs I will try to improve. I know techniques of cross-correlations and you've already talked about DMTD that for sure I will have to come to... Good night Stephane -Message d'origine- De : Magnus Danielson [mailto:mag...@rubidium.se] Envoyé : dimanche 18 janvier 2015 22:46 À : Stéphane Rey; 'Discussion of precise time and frequency measurement' Cc : mag...@rubidium.se Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Bonsoir Stéphane, On 01/18/2015 10:34 PM, Stéphane Rey wrote: Thanks a lot Bob and Magnus for your very helpful comments. The HP5370a was indeed in TI mode. By the way what is the difference with +/-TI, the button just aside... But I guess I understand where I've missed something : I've tried to put the Rb on channel A and the DUT on channel B but result was always the same but I do understand now that there is indeed a switch to change from COMmon to SEParate and it was always on COM meaning I believe that channel B wasn't used. This explains a lot of things I did not understand. I'm sorry for these so basic issues that might have been solved if I had read carefully the HP5370a manual first. Good. This confirmation makes sense to be and Bob, now we can relax as the mystery is solved. So possible conclusions until now are that I have actually measured the ADEV floor of the system rather than my DUT... which is already nice. The second conclusion from these oscillations seen with the GPSDO under test is that there is very likely in this GPSDO design a systemic noise added to the 10 MHz output (power supply, PCB coupling, ... I'll make further investigations on it later on). It's a great opportunity to learn the tools, and once you have the tools, you can see if you can't improve things. I will experiment all the suggestions you made and will come back. For information the 1PPS from the HP58503b has a positive pulse width that is only few us length. This only makes it hard to view on a scope, but long enough to reliably trigger your counter and scope. Now, when considering that the method is to compare the DUT to an other source, I assume then that the other source shall be at least 1 order of magnitude better than the DUT. Otherwise this will be impossible to distinguish who is the instability contributor between the source and DUT, right ? For a simple setup, yes. But then we are the time-nuts, we have ways of handling these things. :) Let's get you started with the basic measurement, it will be a good start. Then the second question is what kind of very stable source can be used to measure DUT which could be Rb or GPSDO which are already in the range of 10E-10 to 10E-12 100s ? Time-nuts tend to spend their time and money getting even more stable clocks and tools. If you have the right tool, you can measure near and *under* the noise-level of your reference, but not without running into issues. One such trick is called cross-correlation, while another is to use three-corner hat techniques. Cheers, Magnus --- L'absence de virus dans ce courrier électronique a été vérifiée par le logiciel antivirus Avast. http://www.avast.com ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] question Alan deviation measured with Timelab and counters
Hello, First, please do apologize for the confusion answering in the bad email. That's things I'm absolutely able to do when replying at 3 am ! Again, sorry for that and thanks Magnus for having corrected this. Back to my setup : There is indeed nothing on the STOP input of the HP5370a. The standard 10 MHz comes from the GPSDO HP-58503B and feeds the HP5370a Standard input. Its ADEV is given on page 240 of that document : http://www.leapsecond.com/museum/hp58503a/097-58503-13-iss-1.pdf We see that the shape is starting at about 2E-12 at 1s, increase to 2E-11 at 100s before decreasing again down to E-13 for above 10E3s... The setup #1 was using the Racal DANA Rb connected on the START input which is specified at E-9 / E-10, given on page 16 of the manual : http://bee.mif.pg.gda.pl/ciasteczkowypotwor/Racal/9470-9479.pdf The EXT input receives the 1PPS from the HP58503b. It apparently drives both the START and the STOP of the acquisition (the two lights are blinking and the time between two measurement is no longer adjustable from the front panel RATE potentiometer and the period between two samples is 1.0s (detected by Timelab). But yes, the ADEV plot sounds really strange as it goes incredibly low after few seconds which is not consistent with the stability of the sources I'm using which is why I felt something was wrong On Setup #2 I've only replaced the Racal Dana Rb with the GPSDO to test. I've not made this design and not checked yet anything on it. Could these oscillations be from power supply noise ? To be checked. But how can it follow the ADEV plot of the Racal Dana Rb ? mmm Coincidence is not something I like too much and I believe something is clearly wrong in my measurement But what ??? On the Timelab setup screen before launching the acquisition I've left all the parameters as it without touching them. I've just seize 10E6 in the frequency field. Ah, by thay Magnus, for the downmixed test I've forgotten to change this value, I will check on monday when back at the office. Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Bob Camp Envoyé : dimanche 18 janvier 2015 14:44 À : Discussion of precise time and frequency measurement Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Hi Ok, I didn’t think I’d seen the plots before. I agree that the plots look like “counter limited” data. That’s a fine explanation at the shorter Tau’s. I also agree that some sort of periodic “stuff” is getting into one of the signals and creating the ripple. What I’m wondering about (and what makes me question the setup) is the fact that the data is still “counter limited” at the mid to low parts in 10^-13 level at just a bit over 100 seconds. A telecom Rb is doing pretty well to be at 1x10^-12 at 100 seconds. Most GPSDO’s are doing well to be mid parts in 10^-12 at that tau. Simply put, the data continues to be counter limited to a pretty low point. Bob On Jan 18, 2015, at 7:13 AM, Magnus Danielson mag...@rubidium.dyndns.org wrote: Hi Bob, On 01/18/2015 04:25 AM, Bob Camp wrote: Hi I’m a little concerned about the measurement setup here. Based on the quoted text, there have been a few messages in the thread that have not showed up here. The messages got accidentally posted in the wrong thread. = The “start” input on the counter is defined below. The “stop” input is not defined. Is the counter running in time interval mode or in frequency mode? IF it’s in time mode - what is the stop hooked to? IF it’s in frequent mode - what is the gate time set to ? Is the “standard input” the missing front panel input or is it the external reference input on the back panel? I was also considering the setup strange in this regard. Common switched in? = I’m looking at the data on the link: http://www.ptp-images.com/affiche-directement-l-image-kccsz71c9a.html In both cases the slope is roughly 1/tau. Both plots end up in the 3 to 4 x 10^-13 range at 100 to 300 seconds. That’s suspiciously good performance for a rubidium or a GPSDO. Which is what makes me wonder about the setup. The Blue plot (1 pps?) ran for 18 minutes and has 1,114 points in it. The Pink plot ran for about 9 minutes and has a bit over 500 points in it. Both seem reasonable for a 1 pps to 10 MHz sort of setup. That may explain part of my confusion above. Again - I apologize if this all got explained in a post that went missing here. Not really. The plots looks to me like measurement setup baseline plots, with some sine noise in them. Cheers, Magnus ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list --
Re: [time-nuts] question Alan deviation measured with Timelab and counters
Thanks a lot Bob and Magnus for your very helpful comments. The HP5370a was indeed in TI mode. By the way what is the difference with +/-TI, the button just aside... But I guess I understand where I've missed something : I've tried to put the Rb on channel A and the DUT on channel B but result was always the same but I do understand now that there is indeed a switch to change from COMmon to SEParate and it was always on COM meaning I believe that channel B wasn't used. This explains a lot of things I did not understand. I'm sorry for these so basic issues that might have been solved if I had read carefully the HP5370a manual first. So possible conclusions until now are that I have actually measured the ADEV floor of the system rather than my DUT... which is already nice. The second conclusion from these oscillations seen with the GPSDO under test is that there is very likely in this GPSDO design a systemic noise added to the 10 MHz output (power supply, PCB coupling, ... I'll make further investigations on it later on). I will experiment all the suggestions you made and will come back. For information the 1PPS from the HP58503b has a positive pulse width that is only few us length. Now, when considering that the method is to compare the DUT to an other source, I assume then that the other source shall be at least 1 order of magnitude better than the DUT. Otherwise this will be impossible to distinguish who is the instability contributor between the source and DUT, right ? Then the second question is what kind of very stable source can be used to measure DUT which could be Rb or GPSDO which are already in the range of 10E-10 to 10E-12 100s ? Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Magnus Danielson Envoyé : dimanche 18 janvier 2015 16:47 À : time-nuts@febo.com Cc : mag...@rubidium.se Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Bonjour Stéphane, On 01/18/2015 03:37 PM, Stéphane Rey wrote: Hello, First, please do apologize for the confusion answering in the bad email. That's things I'm absolutely able to do when replying at 3 am ! Again, sorry for that and thanks Magnus for having corrected this. Ah well, that's water under the bridge now. I only mentioned it for Bob's reference. Back to my setup : There is indeed nothing on the STOP input of the HP5370a. The standard 10 MHz comes from the GPSDO HP-58503B and feeds the HP5370a Standard input. Its ADEV is given on page 240 of that document : http://www.leapsecond.com/museum/hp58503a/097-58503-13-iss-1.pdf We see that the shape is starting at about 2E-12 at 1s, increase to 2E-11 at 100s before decreasing again down to E-13 for above 10E3s... The setup #1 was using the Racal DANA Rb connected on the START input which is specified at E-9 / E-10, given on page 16 of the manual : http://bee.mif.pg.gda.pl/ciasteczkowypotwor/Racal/9470-9479.pdf The EXT input receives the 1PPS from the HP58503b. It apparently drives both the START and the STOP of the acquisition (the two lights are blinking and the time between two measurement is no longer adjustable from the front panel RATE potentiometer and the period between two samples is 1.0s (detected by Timelab). If you run the counter in frequency or period mode, you normally use the STOP input, which is then internally split to the START and STOP channels. If you run the counter in TI mode, then they are usually separate, but you can force them the same using the START COMMON switch. We tend to use the TI mode, with two basic setup: Stoopid simple: PPS to START and measured clock to STOP. This setup has the down-side that the jitter of the PPS (which can be much higher than that of the clock) can dominate, if so, the next setup is relevant: Standard setup: PPS to ARM/EXT input to trigger measurement. DUT to START channel and reference clock to STOP channel. Sometime the clocks is interchanged, sometimes it is important, somtimes not. Record the TI data. But yes, the ADEV plot sounds really strange as it goes incredibly low after few seconds which is not consistent with the stability of the sources I'm using which is why I felt something was wrong OK, you made what we call a instrument noise limit measurement. Then you do the same thing as a normal measure, but you have start and stop channels see the same signal split. It may be good to let the stop channel has a meter or two of additional coax to de-correlate the rising edges. This setup will let you measure the effect of white noise, slew-rate and counter resolution. It can be good for fault analysis and see if the setup gives reasonable noise or if you can improve it. Adjustment of the trigger points will select a point of optimal slew-rate (and sometimes avoid false-trigger noise) and thus finding the optimum trigger noise. Squaring up the signal may be a nice way
Re: [time-nuts] Current state of optical clocks and the definition of the second
Hi, I've took the time to read carefully your long and detailed message Magnus and this was very interesting. I've learned many things that have enabled me to investigate further. Ah yes, you're right saying that the more you fall into these things, the more you discover that you have to learn. Recently I've worked a lot on PLLs and I've actually learned a lot on special care to ensure low noise Very interesting. By the way I'm still working on this topic to improve again the noise (currently on a 3 GHz LO) Here are some experiment results : http://www.ptp-images.com/affiche-directement-l-image-kccsz71c9a.html 1. Setup #1 (blue plot) HP5370A standard input from HP GPSDO EXT input not connected, internal Arming 0.4s rate START input from 10 MHz distribution unit RacalDana 9478 Rubidium 2. Setup #2 (pink plot) HP5370A standard input from HP GPSDO EXT input not connected, internal Arming 0.4s rate START input from DUT (10 MHz homemade GPSDO) I'm not sure this is the proper way to connect everything... but this is the setup providing the lowest ADEV... which is between 1E-10 and 1E-13. But is the truth ? I feel strange the two plots having the same decreasing path along a linear slope (I mean linear on the log-log plot) ... I'm not sure of what I'm measuring ? Could this be the system measurement floor ? By the way how to measure the ADEV floor of a system other than having a source greater than the measurement system ? What could be these oscillations on the homemade (not by me) GPSDO ? I've tried to downmix the DUT 10 MHz to few kHz using a SR DDS generator and a double balanced mixer from minicircuit via a low pass filter tuned at 100 kHz, but the level wasn't high enough for the counter (which I found strange as it was already nearly 200mV). I hadn't anything in hands to make a squarer quickly so I've just added a Minicircuits RF amplifier. The level was good but the ADEV has jumped to 1E-6. The signal was noisy already on the oscilloscope which I know is for sure the cause. I need to make a squarer. I was hesitating between several methods : using a CMOS gate, but this will increase the flicker noise from what I've read, using an amplifier and clamping diodes or a fast comparator which might create some noise around the trigger point... Any recommendation there ? I'll try to make this squarer next week to continue my investigations Cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Magnus Danielson Envoyé : mercredi 14 janvier 2015 06:05 À : time-nuts@febo.com Cc : mag...@rubidium.se Objet : Re: [time-nuts] Current state of optical clocks and the definition of the second On 01/13/2015 11:41 AM, Attila Kinali wrote: On Mon, 12 Jan 2015 20:09:45 + Gregory Maxwell mailto:gmaxw...@gmail.com gmaxw...@gmail.com wrote: On Mon, Jan 12, 2015 at 12:34 PM, Attila Kinali mailto:att...@kinali.ch att...@kinali.ch wrote: Seems that the state of the art in stabilized lasers has improved a lot lately, e.g. there are commercial available 1550nm devices which have a =3Hz line-width: http://stablelasers.com/products.html http://stablelasers.com/products.html (well on a short term basis, the medium term performance is not so impressive) Laser stabilization, especially for quantum metrology is still an actively researched field. Current state of the art is IIRC 0.3Hz linewidth (sorry, cannot find the reference at the moment). Mid- and long term stability depends highly on the reference used. Current research is fucused mainly on special, low vibration structures made out of low expansion glass or silicon. And these cavities are usually put into a temperature controlled chamber in vacuum. Well, guess what I found standing around in a lab with an optical comb? :) With optical line-widths in sub-Hz range and optical combs you have a nice way of comparing the frequency of that free-running and un-steerable but stable oscillator. However, as you mix it down the noise of the optical comb will dominate, but you can know which multiple of the optical comb and offset it is. Considering the rarity and extreme cost of H-masers, or just really exceptional quarts oscillators; might it be the case that optical LOs start looking interesting for applications which just need stability (or being steered by other sources; e.g. GPSDL)? Well, an 8607 costs more than a Rb-standard. Yes, the 8607 has lower close in phase noise and up to several 1000s it rivals the Rb, but handling it is much more difficult than handling an Rb. Also, if you want to buy one of those exceptionally low noise/high stable 8607's (those that go down into the 10^-14 range) you'd have to sell your car. But, if you buy a H-maser from SpectraTime, you get a 8607 for free ;-) That is also the only
Re: [time-nuts] question Alan deviation measured with Timelab and counters
Bonjour Magnus, Many thanks for your very long and detailed answer. I've read quickly bu will go deeper tonight. Here are the results of today experiments... which are not giving anything valuable... I still don't understand the results I get :-/ With the PM6654C, I've put the HP GPSDO on the standard input, the 1 PPS on channel A and the 10 MHz from the DUT (GPSDO as well) on channel B. This gives something in the range of 2E-9 which looks like the counter resolution, right ? The gating takes 4s and the Time A-B displays a value like 64 E-6 Now if I downmix the channel B to 5 kHz (LO is a DDS Standford Reseach generator), I have a sinus with lower amplitude and no squarer in my hand at the moment to shape the signal. Anyway, I do the same operation and I get on the display two more digits like xx.xx E-6 but the ADEC is in the range of E-7 I do not understand at all this fact. Even if the slew rate is not great, I was expecting an improvement. Note that the values displayed are always changing quite a lot between two samples. For instance with the 5 KHz channel B signal, I can read first sample at 27.11E-6, in the next one is 31.22E-6... which sounds huge, right ? I've then found an HP5370A and tried the same operation. Unfortunately the 5 kHz output is too low for the HP5370A sensitivity. I need an amplifier or sqauerer here but had no time to build on today. Si I could not get anything valuable with the HP5370A at the moment... Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Magnus Danielson Envoyé : mercredi 14 janvier 2015 08:04 À : time-nuts@febo.com Cc : mag...@rubidium.se Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Bonjour Stéphane, On 01/14/2015 02:16 AM, Stéphane Rey wrote: Hi Magnus, For some reason I've missed this message and the one from Jim until now ! This answers many of the questions I had. For my defense, I've 3000 messages since the last 3 months on the list !!! ah, yes, I'd like to get even better than 1E-12. 1E-14 would be perfect but my best standards for now are a HP GPSDO and an Effratrom FRK Rb which both are around 1E-12 'only'. I may have to invest in something better if prices are acceptable. I guess I won't be able to measure beyond the standard itself. The method you describes gives tau=2E-9 ? This is more or less what I could get with the frequency measurement (even a bit lower). So what is the benefit of the time interval measurement here against the frequency measurement ? I've been sloppy with the scaling factor, so there is a fixed scaling factor for the noise that the single-shot resolution produces, and that would be a measurement limit that if everything else is ideal would dominate. This quantization noise is sqrt(1/12) or about 0.289 if I remember correctly, so that is the scale-factor. It will also have a 1/tau slope. So that is how you can expect this noise to behave, it will look like white phase noise, but isn't, it is highly systematic noise, and if you play nicely with it, you can measure below it. However, doing so is non-trivial. I have one counter that does that. The good old HP5328A with the Option 040-series of boards will introduce noise to the counting 100 MHz oscillator such that averaging gets you down towards 10 ps rather than 10 ns resolution in TI mode. However, it does not help you to get nice frequency or stability measures. I've not taken the time to detail-analyse the ADEV scaling factor thought, I should do that, but it follows the general formula of ADEV(tau) = k*t_res/tau where t_res is the single-shot resolution and k is a constant. There is more to this, as counters can show up non-linearities of several sorts, and that the trigger conditions of the input has been optimized, which can be slew-rate limited for many counters and conditions. So, anyway, there is a bit of hand-waving in there, but I thought it was better to get you to get the basic trend there first, and then we can discuss the detailed numbers, as theory is one thing and achieved number can be quite a different one. As for frequency and time-interval measurements, if properly done, they can be used interchangeably without much impact. Realize that frequency and time-interval measurements will both be based on time-interval measurements as the core observation inside the counter, so the single-shot resolution limit applies to them both. However, subtle details lies in how the counter works and there is ways that the frequency precision can be lost. A good counter is the SR620, but the way it does the frequency measure, you need to calibrate the internal delay to make it on the mark measure. Using it in time-interval mode and you can eliminate that offset, because the start and stop measure of your signal under test is done with the same channel, with essentially the same delay both trigger-times
Re: [time-nuts] question Alan deviation measured with Timelab and counters
Hi John, I hadn't noticed before you were here as well ;-) Thanks for answering. So I do understand I can use Timelab in frequency difference even if my counter sends data in TI in nanoseconds. Great. Ah and thanks for the manual link. I didn't remember this was in the manual of the Timepod Will investigate further today -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de John Miles Envoyé : mercredi 14 janvier 2015 07:26 À : 'Discussion of precise time and frequency measurement' Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters - Can I use Frequency difference mode from Timelab to monitor time intervals ? If no is there a way to use the time interval measurement from the counter with Timelab to plot ADEV ? If you feed in frequency samples, it will convert them to phase-difference samples internally, so the program itself doesn't really care. The use of frequency data has a few drawbacks such as less accurate ADEV plots due to the counter's dead time between readings, but it's the easiest way to get started and is perfectly usable for many purposes. In general you should avoid letting the counter do any averaging. Except in very specific circumstances, any apparent improvement in ADEV measurement floor will be illusory. There are exceptions, but this isn't something you want to mess with until you're very comfortable with the rest of the measurement process. Your counter's true ADEV measurement floor at t=1s should be assumed to be close to its single-shot resolution specification (e.g., 100 ps = about 1E-10). - In case the principle of plotting ADEV from Time Interval, what is the interpretation of the result ? The ADEV shows the relative stability between the two GPSDO... So, practically what does it bring ? And how to use this method if I want to characterize a device ? An ADEV graph shows frequency stability statistics at different intervals, ranging from the rate at which the readings are returned from the counter (tau zero, at the left end of the plot) to a maximum interval that's related to how long you let the measurement run. It's much too deep a subject to go into in an email; see http://www.ke5fx.com/stability.htm for more pointers. Again, TimeLab always plots ADEV from time interval/phase data, even if you give it frequency readings. ADEV is fundamentally a frequency stability metric, but it can be computed identically from either TI or frequency samples (assuming zero dead time). - stupid question on Timelab. If I let Timelab in Auto to select the period between two samples (correctly detected), the time scale of the graph is wrong. For instance, a 3h plot stops at 2000s (0.5h)... Here again, I miss something but what ? The TimeLab manual, for one thing. :) Hit the books (specifically http://www.miles.io/TimePod_5330A_user_manual.pdf , page 31). -- john, KE5FX Miles Design LLC ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. --- L'absence de virus dans ce courrier électronique a été vérifiée par le logiciel antivirus Avast. http://www.avast.com ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] question Alan deviation measured with Timelab and counters
Hi there, I'm still working on the ADEV measurement and here is what I've experimented today and few questions which are arising, that for sure have probably an easy answer... After having measured my 10 MHz GPSDOs with 1Hz and 0.01Hz resolution with HP8342A and Phillips PM6654C, I've seen the effect of resolution on the ADEV. With the 0.01 Hz resolution I could only achieve 1E-9 at 10 MHz which is very likely the floor limitation from the counter resolution. Actually I hope my GPSDO is better than that (~1E-12) The PM6654C can measure Time Intervals with a resolution of 0.01ps and an averaging time of 96s. I've then launch a new acquisition from Timelab, selecting Frequency difference instead of frequency only. I've not seen a time interval mode so I think this measurement gives nothing usable. The counter runs upon its own OCXO and I've connected two different GPSDO, one on each channel. I've tried both the internal gating as well as external gating with the 1PPS from the HP GPSDO with same result. The time interval is about 450ps with a variation of about 50ps. I've got an ADEV plot which is now in the range of 1E-12 / 1E-13. However here are my questions : - Can I use Frequency difference mode from Timelab to monitor time intervals ? If no is there a way to use the time interval measurement from the counter with Timelab to plot ADEV ? - In case the principle of plotting ADEV from Time Interval, what is the interpretation of the result ? The ADEV shows the relative stability between the two GPSDO... So, practically what does it bring ? And how to use this method if I want to characterize a device ? - I've googled for the DMTD and discovered the method. The principle seems clear and easy even if I know there might be several pitfails but here again, I don't know how to use the method or perform the result interpretation. In that method there is the LO and at leadt two DUTs inputs. I do understand that the method gives a relative stability between the two DUT... And what ? IF I've a device to characterize how can I get something useful by comparing with an other device that I do not know ? Where am I wrong here ? - stupid question on Timelab. If I let Timelab in Auto to select the period between two samples (correctly detected), the time scale of the graph is wrong. For instance, a 3h plot stops at 2000s (0.5h)... Here again, I miss something but what ? I'm sorry as it's probably weird questions already discussed many times but any comment or URL to point to would be a great help. Many thanks Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Bob Camp Envoyé : vendredi 9 janvier 2015 23:57 À : Discussion of precise time and frequency measurement Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Hi If your only instrument is a counter. — and — You never measure past 1x10^-10 with that counter — and — Measurements that bounce around with a standard deviation of the difference between readings of 1x10^-10 are ok. — then — No, you don’t need anything better than a 1x10^-10 ADEV. Most people would be bothered by a counter that has an typical jump of 1x10^-10 between every reading, so most would want a standard that’s a bit better than that. In addition, if you want to guarantee accuracy of a reading, you probably want something that’s 5X to 10X better than the level that stops the reading jitter. Simply put - ADEV is not standard deviation of frequency. Your frequency counter measures frequency. Going from one to the other means you want to have better ADEV than you might think. Bob On Jan 9, 2015, at 10:42 AM, steph.rey steph@wanadoo.fr wrote: Hi Bob, Many thanks for your prompt and detailled answer. My question on applications wasn't on good ADEV where I perfetcly understand the need, but actually what could be the applications of measuring BAD ADEV (10e-7). That was my point asking what king of application can we cover by measuring such high ADEV when you have counters with resolution not greater than 0.01Hz However you bring to me part of the answer when you talk about the reference and the way to get something cheap and better than 10e-12. I will investigate on DMTD. However, even if you have a beautiful Maser source, will you improve anything above the resolution of your counter. In other words, with my 0.01Hz counter, will I improve my measurement if I replace my GPSDO source with something much better ? I feel the resolution of the counter will anyway limit the ADEV floor, right ? If the last digit of the counter do not move how could we measure something smaller ? The counters I'm using are not running on their own reference (OCXO or TCXO) but with the HP58503b which is a GPS disciplined OCXO but with stability in the range of 10e-11 or 10e-12 at best. I'm working for a big lab where possibly I could have
Re: [time-nuts] question Alan deviation measured with Timelab and counters
Hi Magnus, For some reason I've missed this message and the one from Jim until now ! This answers many of the questions I had. For my defense, I've 3000 messages since the last 3 months on the list !!! ah, yes, I'd like to get even better than 1E-12. 1E-14 would be perfect but my best standards for now are a HP GPSDO and an Effratrom FRK Rb which both are around 1E-12 'only'. I may have to invest in something better if prices are acceptable. I guess I won't be able to measure beyond the standard itself. The method you describes gives tau=2E-9 ? This is more or less what I could get with the frequency measurement (even a bit lower). So what is the benefit of the time interval measurement here against the frequency measurement ? However if I hear what you says, the GPSDO provides the 10 MHz standard reference for the counter, the GPSDO PPS on channel A and channel B receives for instance a 10 MHz signal I want to measure. So what will be the result of Time A-B then ? I do not understand why you put the PPS on channel A instead of something of the same frequency than the DUT ? How the time A-B will behave with these two different frequencies... By letting TimeLab know the frequency, it can adjust for any slipped cycles on the fly. I guess this is what I've not understood. Now if I mix down the 10 MHz DUT with a 10.005 reference to increase the resolution, I'll get 5 kHz on channel B and still PPS on channel A ? Again I do not understand what will happen with these two signals on the time A-B. If I push your method a bit more, I could even get a beat frequency of 1 Hz and with 10-digits I would have increased my resolution by 10E6. Then I will be limited by the standard stability but on the principle would it work as well ? On that document http://www2.nict.go.jp/aeri/sts/2009TrainingProgram/Time%20Keeping/091017_DMTD.pdf it says (page 6) the accuracy of measurement is improved by a factor v/vb (the DUT and offset LO 1/2.PI.f). So it sounds to me that there is a compromise between resolution increase and accuracy. If I chose a beat frequency of 1 Hz the accuracy will not be improved but the resolution will be, right ? What is the transfer clock you're talking about ? and by the way should the offset LO be as stable as the standard reference meaning greater than the DUT ? Well, it's far too late here to let my brain working anymore. I will perform further experiments tomorrow at the office. Thanks cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Magnus Danielson Envoyé : samedi 10 janvier 2015 02:05 À : time-nuts@febo.com Cc : mag...@rubidium.se Objet : Re: [time-nuts] question Alan deviation measured with Timelab and counters Stephane, On 01/09/2015 12:53 PM, steph.rey wrote: Dear all, I'm trying to measure Alan Deviations using Timelab and some frequency counters. The device under test is a GPSDO using a TCXO as référence I've an HP58503B GPSDO which feeds my counters. I've tried an HP5342A, 0-18 GHz, 1 Hz resolution and a Philipps PM6654C, 0.01Hz resolution. In Timelab, the plot with the HP5342A is around 10e-7 which correspond to 1Hz and with the PM6654C, the plot is around 10e-10. I would suspect that this is still the counter which limits the actual response of my device under test. My question are : - how to measure Alan Deviations with levels below 10e-12/10e-13 ? What can be the application of measurement Alan deviation 10e-10 ? I guess most of the low frequency - The HP53503 GPS is given to be 10e-11 / 10e-12. I guess this will limit anyway the measurement floor. I've a Rb source, but it's stability is within the same range. What kind of reference would be more suitable for such measurements ? - With the PM6654C on 15h measurement, I can see some frequency jumps of 800 Hz which are not relevants with the GPSDO undertest. I suspect error in data transmission. This makes the overall measurement totally wrong (10e-5). The counter is in talk only mode. I'd like to get rid of these points maybe 40-50 points out of 1. Is there a way to do that from Timelab or the only option is to export the file and process manually the data ? I've use the PM6654C with TimeLab. I wire the 10 MHz from the GPSDO and then the PPS to Channel A. Channel B has whatever signal I want to measure. By letting TimeLab know the frequency, it can adjust for any slipped cycles on the fly. This works well. The PM6654C has a single-shot resolution of 2 ns, which comes from the internal 500 MHz counting clock. This gives ~ 2E-9/tau (very coarse level) measurement limit. If you want to reach the 1E-12 resolution mark you need another 2000 of resolution gain, which is what you get if you mix your 10 MHz signal with a 10,005 MHz clock or lower. The Dual Mixer Time Difference (DMTD) is more likely to work well, as it provides some cancellation of the transfer clock. Slew-rates
Re: [time-nuts] GPS-disciplining an ordinary VCXO?
With that VCXO you want to have a 5s to 10s or more loop time constant (0.1Hz BW) which typically can only be done in the digital domain.. Hi Said, Could you point us on something describing that ? What kind of digital processing do you think about ? Cheers Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Said Jackson via time-nuts Envoyé : dimanche 28 septembre 2014 07:50 À : Discussion of precise time and frequency measurement Cc : time-nuts@febo.com Objet : Re: [time-nuts] GPS-disciplining an ordinary VCXO? Mark, In the analog domain you can probably do a PLL with a 1Hz loop BW. Using a PLL chip like ADF 4002 or similar. This means all the nasty noise from the NEO will taint your PN up to 20Hz or more, very significantly close-in. If you don't care about noise (jitter) below 100Hz then this is fine. If you do as it will dominate your ADC jitter then you can't use an analog PLL. With that VCXO you want to have a 5s to 10s or more loop time constant (0.1Hz BW) which typically can only be done in the digital domain.. This allows you to use the excellent 1Hz to 100Hz PN of that VCXO without tainting it by the noisy NEO. An even better setup would be to lock a very low noise 5 or 10MHz ocxo to the GPS with 100s time constant, then use the analog PLL with wider bandwidth (say 30Hz) to reduce the VCXO PN close-in even further by using the ocxo to supress the vcxo PN. Welcome to our world, if you look at the archives there are 10++ years of discussions about exactly doing this... Bye, Said Sent from my iPad On Sep 27, 2014, at 21:01, Mark A. Haun hau...@keteu.org wrote: In my quest to learn Verilog and get my hands dirty with software-defined radio, I'm currently designing a direct-sampling shortwave receiver. This uses an 80-MSPS ADC, which requires a low-phase-noise oscillator, e.g. Crystek CVHD-950 or Abracon ABLNO. It would be nice to have some provision for locking this oscillator to an external reference, hence my question: All of the amateur GPSDO designs I've seen are disciplining an OCXO. I understand this is easier because the excellent short-term accuracy of the OCXO means the feedback can run slower, so even a 1 PPS signal can be used. I am wondering what sort of performance could be achieved by disciplining my VCXO directly with a good GPS module. I have a NEO-7N (Ublox) with configurable timepulse up to 10 MHz. Someone mentioned that this is derived from 48 MHz, so jitter is reduced if you pick an integer divisor. That is fine, but I don't have a feel for what other irregularities may be present in the timepulse output, and how they would affect the performance. I also don't know how to go about designing a PLL loop filter. I understand the goal is to marry the long-term GPS stability with the short-term VCXO stability but all I have is a phase-noise plot for the VCXO. How do you know where to split the difference? It is not essential to the larger project, but what I am ideally going for is 1 ppb frequency match between two ends of a radio link, and 1 ppb stability over data symbol times. That is, carrier stability of ~ 1/10 cycle at 10 MHz over one-second symbols. (Channel coherence imposes this limit.) I know the experts here can tell me whether this is impossible, totally doable, or somewhere in between! Thanks, Mark ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. --- Ce courrier électronique ne contient aucun virus ou logiciel malveillant parce que la protection avast! Antivirus est active. http://www.avast.com ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
[time-nuts] various question on stability, jitter, PN, ...
Hi guys, I told you ! Some questions were to arise... ;-) At work I'm working on 1.5, 3 and 12 GHz pulsed systems with pulses length between 0.1 and 5 us. We are especially interested in phase stability pulse to pulse (repetition rate) and possibly with minor priority on the length of the phase pulses, pulse to pulse. 1. When plotting the phase noise response of a CW signal, one can determine the RMS jitter in ps or fs. I'm wondering what is corresponding to this value. As it's RMS I would expect this is the square root of the maximum of the Gaussian distribution of the frequency jitter. Is it right ? If so this correspond roughly to 1 sigma deviation, right ? 2. Is there any link between this frequency jitter and the phase jitter ? I assume no, but... 3. What does bring the Allan deviation plot ? This gives stability vs integration time I know, but how to make an interpretation of this ? Is it a way to plot the frequency jitter in a more detailed way than just giving the rms jitter ? In practical use, for a pulsed system does it mean that only the very short term jitter is of interest ? 4. Is the Allan deviation plot a representation of the jitter vs integration time, meaning there is a direct relation between the RMS jitter computed at various offsets from the carrier in the PN plot ? 5. Is there a practical way to plot phase noise for pulsed signals ? That's all for now. If anyone has clues or can point me into good articles related this would be kind. Thanks Stephane --- Ce courrier électronique ne contient aucun virus ou logiciel malveillant parce que la protection avast! Antivirus est active. http://www.avast.com ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] newcomer
Thanks for the details. This miniaturized device is nice. There is no information regarding stability. Anything there ? What's the technology inside ? crystal, TCXO ? What is the approx price for a such device ? The VCXO I was targetting is this one : http://www.abracon.com/Precisiontiming/ABLNO.pdf This should bring more or less the same level of PN once multiplied than your ULN_1G. However the frequency accuracy is poor and it needs to be disciplined in most applications. Cheers Stephane De : saidj...@aol.com [mailto:saidj...@aol.com] Envoyé : lundi 15 septembre 2014 18:51 À : steph@wanadoo.fr; time-nuts@febo.com Objet : Re: [time-nuts] newcomer Hi Stephanie, I have a similar issue, I can never tell if my messages post or not, some I get back instantly others never show up in my inbox. I think the mail server was just updated.. To answer your questions: 1) attached is the PN plot of the 1.0GHz version. 2) Here is a datasheet for the part: http://www.jackson-labs.com/index.php/products/uln_1g The tricky part in your proposed setup will be getting a 100MHz crystal with low enough phase noise so that the 20log(N/M) added phase noise won't be a problem - +20dB added after all. On our part some of the tricky design issues were size and power as well as the high output power of +22dBm, and a requirement to maintain the output power to within 0.5dB over the entire -40C to +85C temperature range. bye, Said In a message dated 9/15/2014 09:40:37 Pacific Daylight Time, mailto:steph@wanadoo.fr steph@wanadoo.fr writes: Hi, It sounds like all my messages need moderator approbation. Is it the rule on the list or a technical problem at my side ? Cheers Stephane On Mon, 15 Sep 2014 09:17:01 -0700, Said Jackson via time-nuts mailto:time-nuts@febo.com time-nuts@febo.com wrote: Hi Stephanie, Welcome to the list! We designed a 1GHz crystal LO for PLLs (the ULN-1G) using an off the shelf miniature 500MHz crystal oscillator which is run at 3 rd overtone internally then using a diode doubler and a steep bandpass filter using several Mini Circuits ceramic filters and a 20dBm amp. Works like a charm and has phase noise very close to theoretical.. Bye, Said Sent From iPhone On Sep 15, 2014, at 5:50, steph.rey mailto:steph@wanadoo.fr steph@wanadoo.fr wrote: Hi the list, Just wanted to introduce myself for my 1st message. I'm Stephane, 40, living in France, at the moment working in RF electronics for a particles accelerator lab. I'm hamradio as well, and I do enjoy especially weak and accurate signals. I'm desiging various RF circuits. Current design is a universal PLL able to operate from 0.5 to 6 GHz depending on the VCO and supposed to be low-jitter (1ps) regarding the application. I'm also starting a new design of low noise PLL and there will be probably a lot of question arising... I'm starting with the 1 GHz LO made upon a 100 MHz VCXO + multipliers/filters/MMICs. I want to focus deeper on low phase noise/jitter, synchronization and low-noise PLL techniques. I believe this is a good place for most of these topics. Cheers Stephane ___ time-nuts mailing list -- mailto:time-nuts@febo.com time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- mailto:time-nuts@febo.com time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. --- Ce courrier électronique ne contient aucun virus ou logiciel malveillant parce que la protection avast! Antivirus est active. http://www.avast.com ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] newcomer
Thanks for your feedback. This is indeed a bit expensive for that application. The x5 multiplier is indeed just the harmonic 5 capture with the helical filter. I'll let you know the outcome. This probably won't beat any expensive ULN source but might be a good starting point for low cost. Worth to try Stephane -Message d'origine- De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de S. Jackson via time-nuts Envoyé : mardi 16 septembre 2014 02:11 À : time-nuts@febo.com Objet : Re: [time-nuts] newcomer Hi Stephane, our customer just needed +/-50KHz free-running over all conditions including aging, so stability was not of major concern, and the unit easily performs significantly better than that. We did make a VCO version for PLL disciplining so you can lock it to a GPSDO or in a correlated phase noise test system etc. The stability is thus not really good, but sufficient for Radar applications. The technology is a Crystal oscillator multiplied by 5x internally I think (maybe it was 3x, can't remember), then by 2x externally. Since it is a MilSpec compliant part it is not that low-cost, but it is lower-cost and much higher-performance than the legacy part it replaces. More than $1K for sure though. It also has some additional tricks up its sleeve such as a built-in DC-DC switcher and power supply filter allowing operation from +6V to +15V without affecting output power, and allowing noisy external power supplies to be used, and harmonics lower than -32dBc typically by using the steep ceramic low-pass filters. That Abracom part looks pretty nice too, and your approach should work well too if you can get the 5x multiplier working well. Get a good 10MHz OCXO and use an AD PLL chip such as ADF4002 etc to lock that Abracom part with a pretty small loop BW (100Hz), and you will have the best of both worlds. Bye, Said In a message dated 9/15/2014 16:52:18 Pacific Daylight Time, steph@wanadoo.fr writes: Thanks for the details. This miniaturized device is nice. There is no information regarding stability. Anything there ? What's the technology inside ? crystal, TCXO ? What is the approx price for a such device ? The VCXO I was targetting is this one : http://www.abracon.com/Precisiontiming/ABLNO.pdf This should bring more or less the same level of PN once multiplied than your ULN_1G. However the frequency accuracy is poor and it needs to be disciplined in most applications. Cheers Stephane ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. --- Ce courrier électronique ne contient aucun virus ou logiciel malveillant parce que la protection avast! Antivirus est active. http://www.avast.com ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.