Re: [time-nuts] question Alan deviation measured with Timelab and counters

2015-01-27 Thread steph.rey

Thanks a lot for your comment Bruce,

I need to feel a bit deeper the ins and outs of the methods so I guess 
I will anyway implement both methods on an evaluation PCB and 
characterize each method. This will bring to me some actual data to 
compare. I will share the results of course.
The plan is to have an eval PCB with 4 independant 10 MHz squarers, 
isolation amplifiers, mixers, low pass filters and multistage limiting 
amplifier. Each block will have input/output connectors so that I can 
combine any architecture with these blocks. The PCB will receive a low 
noise PSU as well.
Before I start the design if one thinks about something to add in the 
evaluation, this is very welcome.


Stephane




On Tue, 27 Jan 2015 03:24:44 + (UTC), Bruce Griffiths 
bruce.griffi...@xtra.co.nz wrote:

The performance of the 2 systems should be comparable provided the
similar equivalent noise bandwidths are used.Every 10Mhz edge needs 
to

be timestamped with ps resolution and the resulting phase samples low
pass filtered and decimated to achieve this.The 10MSPS picosecond or
better resolution time stamping with femtosecond integral linearity
will be a bit of a challenge to achieve.
Bruce

 On Tuesday, 27 January 2015 3:26 PM, Stéphane Rey
steph@wanadoo.fr wrote:


 I do understand.
Has anyone already compared the performances of squaring the 10 MHz
vs squaring the IF ?

Stephane

-Message d'origine-
De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de Bob 
Camp

Envoyé : dimanche 25 janvier 2015 19:01
À : Discussion of precise time and frequency measurement
Objet : Re: [time-nuts] question Alan deviation measured with Timelab
and counters

Hi

The approach in the original NIST paper below was sort of a “best
guess” about how to do the limiting and filtering. When the paper was
presented, a number of us questioned how that part of the circuit was
arrived at. The conversation more or less ended up with “that’s
something we can investigate further”. The Collins paper (and Bruce’s
work based on it) is a much better way to look at the 10 Hz squaring
process. At 10 MHz, that stuff is not needed.

Bob

On Jan 25, 2015, at 10:44 AM, Stéphane Rey steph@wanadoo.fr 
wrote:


Hi everyone.

Many thanks for your very useful comments.
I had already seen most of the documents you were pointing but not 
on
the collins and Bruce discussion around the multistage filter. 
However

I've already seen this approach in the document from Allan
(http://tf.nist.gov/timefreq/general/pdf/84.pdf)

At first I had in mind to square the 10 MHz but this is the aim of 
the

evaluation board to evaluate various architectures. So I will
implement several squarers including the Collins Approach both at 10
MHz and 100 Hz and all the blocks will have input and output
connectors so that I will be able to test several layouts.

I will show you the final design.

Cheers
Stephane


-Message d'origine-
De : time-nuts [mailto:time-nuts-boun...@febo.com] De la part de
Charles Steinmetz Envoyé : dimanche 25 janvier 2015 08:08 À :
Discussion of precise time and frequency measurement Objet : Re:
[time-nuts] question Alan deviation measured with Timelab and 
counters


Stephane wrote:


I'm now trying to evaluate various architectures of 2-channels
squarers and a DMDT. For that I'm designing a PCB with 4 squarers :
simple 74ac04 gate biased at VCC/2, a LT1016 comparator, the
transistor based differential amplifier from Winzel and the one 
from Charles.


Note that squaring a 10MHz sine wave and squaring a 10 or 100Hz 
mixer

output are two very different tasks.  If you start at baseband, a
Collins-style multi-stage limiting amp is a great benefit.  That is
generally not necessary if you start at 10MHz (or if you do use a
Collins-style limiter it needs far fewer stages).  All of the 
squarers

you mention work well at 10MHz, but not as well at baseband.

The LT1719 is easier to apply and faster than the LT1016.  You may
want to use that instead of the 1016.  The LT1719 and LT1715
datasheets show the simplest possible implementation (see below).

The MPSH81 devices in my version are available in surface-mount
(MMBTH81) if that is more convenient.  Other fast transistors will
also work (BFT92, BFT93, BFG31).

Best regards,

Charles



---
L'absence de virus dans ce courrier électronique a été vérifiée par 
le logiciel antivirus Avast.

http://www.avast.com

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



---
L'absence de virus dans ce courrier électronique a été vérifiée par
le logiciel antivirus Avast.
http://www.avast.com


Re: [time-nuts] question Alan deviation measured with Timelab and counters

2015-01-19 Thread steph.rey
Actually I'm working in the RF department of a big lab, designing RF 
electronics mainly in microwaves range. I'm luckilly having some tools 
around to play with and a lot of components like 
mixers/amplifiers/couplers/splitters/attenuators, ... almost whatever 
the frequency is up to several tens of GHz.
At home since the last 20 years I could as well get nice instruments. 
The next two measuring tools really missing and for which I'm limited 
are the phase noise and stability measurement and possibly a good 
standard. My Effratom FRK Rb is old and probably not the best from a 
phase noise and stability point of view but until now has never been 
characterized. Otherwise I've almost everything I need up to 40 GHz I 
guess.


I'm doing further measurement right now which sounds much much more 
consistent. I will share tonight.

Cheers
Stephane



On Mon, 19 Jan 2015 08:59:58 -0500, Bob Camp kb...@n1k.org wrote:

Hi

On Jan 18, 2015, at 5:12 PM, Stéphane Rey steph@wanadoo.fr 
wrote:


Bonsoir Magnus (Are you in Sweeden ?)

Being able to measure high stability and low phase noise is 
definitely a need for me as I'm trying to design low noise 
synthesizers and I'm already reaching the limits of my current tools 
for phase noise and I can't afford an E5052 for my own. At work I've 
one but I will probably not stay after august. And anyway I need such 
tools in my lab at home…


If you have tools at work, the best possible thing to do is to get
some oscillators / standards characterized. If you *know* what this 
or

that oscillator is doing in terms of ADEV or phase noise at this Tau
or frequency offset, it’s much easier to figure a lot of this out.

The most basic way to do phase noise in the basement is with a single
mixer setup running into some sort of audio FFT device. A sound card
can be used or an audio spectrum analyzer. Parts are  $100 to get 
one

setup once you can do the audio measurements.

For ADEV, a DMTD or it’s cousin, the single mixer is the easy way to
go. The single mixer does not get a lot of discussion these days. It
is much easier to set up than a DMTD. It does require an offset
oscillator. Once you have a single mixer phase noise setup, you are
about half way to a single mixer ADEV setup. Cost for one is  $100 
in

parts. You already have a counter to collect the data out of it.

In both cases you are running a comparison device. Having a
characterized OCXO to compare to is a really nice thing.

Bob



As low-noise and stable synthetizers depends on the standard used, I 
need as well to measure them as well...


Let's start with this simple experiments and once I will understand 
the ins and outs I will try to improve. I know techniques of 
cross-correlations and you've already talked about DMTD that for sure 
I will have to come to...


Good night
Stephane

-Message d'origine-
De : Magnus Danielson [mailto:mag...@rubidium.se]
Envoyé : dimanche 18 janvier 2015 22:46
À : Stéphane Rey; 'Discussion of precise time and frequency 
measurement'

Cc : mag...@rubidium.se
Objet : Re: [time-nuts] question Alan deviation measured with 
Timelab and counters


Bonsoir Stéphane,

On 01/18/2015 10:34 PM, Stéphane Rey wrote:

Thanks a lot Bob and Magnus for your very helpful comments.

The HP5370a was indeed in TI mode. By the way what is the 
difference with +/-TI, the button just aside...


But I guess I understand where I've missed something : I've tried 
to put the Rb on channel A and the DUT on channel B but result was 
always the same but I do understand now that there is indeed a switch 
to change from COMmon to SEParate and it was always on COM meaning I 
believe that channel B wasn't used. This explains a lot of things I 
did not understand. I'm sorry for these so basic issues that might 
have been solved if I had read carefully the HP5370a manual first.


Good. This confirmation makes sense to be and Bob, now we can relax 
as the mystery is solved.


So possible conclusions until now are that I have actually measured 
the ADEV floor of the system rather than my DUT... which is already 
nice. The second conclusion from these oscillations seen with the 
GPSDO under test is that there is very likely in this GPSDO design a 
systemic noise added to the 10 MHz output (power supply, PCB 
coupling, ... I'll make further investigations on it later on).


It's a great opportunity to learn the tools, and once you have the 
tools, you can see if you can't improve things.


I will experiment all the suggestions you made and will come back. 
For information the 1PPS from the HP58503b has a positive pulse width 
that is only few us length.


This only makes it hard to view on a scope, but long enough to 
reliably trigger your counter and scope.


Now, when considering that the method is to compare the DUT to an 
other source, I assume then that the other source shall be at least 1 
order of magnitude better than the DUT. Otherwise this will be 
impossible to distinguish who is the instability 

Re: [time-nuts] question Alan deviation measured with Timelab and counters

2015-01-09 Thread steph.rey

Hi Bob,

Many thanks for your prompt and detailled answer.

My question on applications wasn't on good ADEV where I perfetcly 
understand the need, but actually what could be the applications of 
measuring BAD ADEV (10e-7). That was my point asking what king of 
application can we cover by measuring such high ADEV when you have 
counters with resolution not greater than 0.01Hz


However you bring to me part of the answer when you talk about the 
reference and the way to get something cheap and better than 10e-12. I 
will investigate on DMTD. However, even if you have a beautiful Maser 
source, will you improve anything above the resolution of your counter. 
In other words, with my 0.01Hz counter, will I improve my measurement if 
I replace my GPSDO source with something much better ? I feel the 
resolution of the counter will anyway limit the ADEV floor, right ? If 
the last digit of the counter do not move how could we measure something 
smaller ?
The counters I'm using are not running on their own reference (OCXO or 
TCXO) but with the HP58503b which is a GPS disciplined OCXO but with 
stability in the range of 10e-11 or 10e-12 at best.


I'm working for a big lab where possibly I could have nice piece of 
equipment but this is always easier to find alternatives solutions at 
lower price. On the application I'm working on we're looking for phase 
stability in the range of fs at several GHz. One of the project I'm 
working will use a femtosecond laser modulated at 88 Mhz that some 
people want to use as RF reference for the 3 GHz source. I'm pretty sure 
this can't achieve the phase stability requirement and I'm trying to 
illustrate this.
However even for my ham activites where I'm trying to design low noise 
LOs, I'd like to have a tool able to measure goog frequency and phase 
stability...


Stephane



On Fri, 9 Jan 2015 07:48:42 -0500, Bob Camp kb...@n1k.org wrote:

Hi

Welcome to the world of trying to measure this stuff …


On Jan 9, 2015, at 6:53 AM, steph.rey steph@wanadoo.fr wrote:

Dear all,

I'm trying to measure Alan Deviations using Timelab and some 
frequency counters.

The device under test is a GPSDO using a TCXO as référence

I've an HP58503B GPSDO which feeds my counters. I've tried an 
HP5342A, 0-18 GHz, 1 Hz resolution and a Philipps PM6654C, 0.01Hz 
resolution.


In Timelab, the plot with the HP5342A is around 10e-7 which 
correspond to 1Hz and with the PM6654C, the plot is around 10e-10.
I would suspect that this is still the counter which limits the 
actual response of my device under test.


Yes, the counters and TCXO are limiting your measurements.



My question are :
- how to measure Alan Deviations with levels below 10e-12/10e-13 ?


How much money do you have to spend? ( There are expensive commercial
ways to do this).

No matter what, you will need a “better than” reference. That’s not
going to be cheap. Most of us simply get a second GPSDO and compare
them. The assumption is that they both are the same and you can
allocate the error equally between them. With three you can more
accurately allocate the error.

A DMTD is the “cheap” way to get the actual measurement done.

What can be the application of measurement Alan deviation  10e-10 ? 
I guess most of the low frequency


There are a number of systems applications that very much need good
ADEV. Getting into why this or that nav or com system needs it would
take a bit of time.

- The HP53503 GPS is given to be 10e-11 / 10e-12. I guess this will 
limit anyway the measurement floor. I've a Rb source, but it's 
stability is within the same range. What kind of reference would be 
more suitable for such measurements ?


If you want to do it directly, a hydrogen maser is a good way to go.
That’s silly expensive. Just compare GPSDO’s, that’s a lot cheaper.

- With the PM6654C on 15h measurement, I can see some frequency 
jumps of 800 Hz which are not relevants with the GPSDO undertest. I 
suspect error in data transmission. This makes the overall measurement 
totally wrong (10e-5). The counter is in talk only mode. I'd like to 
get rid of these points maybe 40-50 points out of 1. Is there a 
way to do that from Timelab or the only option is to export the file 
and process manually the data ?


You can expand the data and zap the offending segments. It’s done on
the phase plot.

Have Fun.

Bob



Thanks  cheers
Stephane
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow

[time-nuts] question Alan deviation measured with Timelab and counters

2015-01-09 Thread steph.rey

Dear all,

I'm trying to measure Alan Deviations using Timelab and some frequency 
counters.

The device under test is a GPSDO using a TCXO as référence

I've an HP58503B GPSDO which feeds my counters. I've tried an HP5342A, 
0-18 GHz, 1 Hz resolution and a Philipps PM6654C, 0.01Hz resolution.


In Timelab, the plot with the HP5342A is around 10e-7 which correspond 
to 1Hz and with the PM6654C, the plot is around 10e-10.
I would suspect that this is still the counter which limits the actual 
response of my device under test.


My question are :
- how to measure Alan Deviations with levels below 10e-12/10e-13 ? What 
can be the application of measurement Alan deviation  10e-10 ? I guess 
most of the low frequency
- The HP53503 GPS is given to be 10e-11 / 10e-12. I guess this will 
limit anyway the measurement floor. I've a Rb source, but it's stability 
is within the same range. What kind of reference would be more suitable 
for such measurements ?
- With the PM6654C on 15h measurement, I can see some frequency jumps 
of 800 Hz which are not relevants with the GPSDO undertest. I suspect 
error in data transmission. This makes the overall measurement totally 
wrong (10e-5). The counter is in talk only mode. I'd like to get rid of 
these points maybe 40-50 points out of 1. Is there a way to do that 
from Timelab or the only option is to export the file and process 
manually the data ?


Thanks  cheers
Stephane
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] newcomer

2014-09-15 Thread steph.rey

Hi the list,

Just wanted to introduce myself for my 1st message.
I'm Stephane, 40, living in France, at the moment working in RF  
electronics for a particles accelerator lab. I'm hamradio as well, and I 
do enjoy especially weak and accurate signals.
I'm desiging various RF circuits. Current design is a universal PLL 
able to operate from 0.5 to 6 GHz depending on the VCO and supposed to 
be low-jitter (1ps) regarding the application.
I'm also starting a new design of low noise PLL and there will be 
probably a lot of question arising... I'm starting with the 1 GHz LO 
made upon a 100 MHz VCXO + multipliers/filters/MMICs.
I want to focus deeper on low phase noise/jitter, synchronization and 
low-noise PLL techniques.

I believe this is a good place for most of these topics.


Cheers
Stephane


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] newcomer

2014-09-15 Thread steph.rey

Hello Said,

Thanks for the answer.

Sounds interresting. Do you have a description of that ? Especially a 
phase noise plot ?


As said, I'm planning to use a VCXO (low cost and low noise) at 100 MHz 
followed by a MMIC (ERA) and a 500 MHz 3-cells helical filter from 
Temwell. Then a doubler from minicircuit, an other MMIC and a 3-cell 
helical filter at 1 GHz...


Cheers
Stephane



On Mon, 15 Sep 2014 09:17:01 -0700, Said Jackson via time-nuts 
time-nuts@febo.com wrote:

Hi Stephanie,

Welcome to the list!

We designed a 1GHz crystal LO for PLLs (the ULN-1G) using an off the
shelf miniature 500MHz crystal oscillator which is run at 3 rd
overtone internally then using a diode doubler and a steep bandpass
filter using several Mini Circuits ceramic filters and a 20dBm amp.

Works like a charm and has phase noise very close to theoretical..

Bye,
Said

Sent From iPhone


On Sep 15, 2014, at 5:50, steph.rey steph@wanadoo.fr wrote:

Hi the list,

Just wanted to introduce myself for my 1st message.
I'm Stephane, 40, living in France, at the moment working in RF  
electronics for a particles accelerator lab. I'm hamradio as well, and 
I do enjoy especially weak and accurate signals.
I'm desiging various RF circuits. Current design is a universal PLL 
able to operate from 0.5 to 6 GHz depending on the VCO and supposed to 
be low-jitter (1ps) regarding the application.
I'm also starting a new design of low noise PLL and there will be 
probably a lot of question arising... I'm starting with the 1 GHz LO 
made upon a 100 MHz VCXO + multipliers/filters/MMICs.
I want to focus deeper on low phase noise/jitter, synchronization 
and low-noise PLL techniques.

I believe this is a good place for most of these topics.


Cheers
Stephane


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] newcomer

2014-09-15 Thread steph.rey

Hi,

It sounds like all my messages need moderator approbation. Is it the 
rule on the list or a technical problem at my side ?

Cheers
Stephane



On Mon, 15 Sep 2014 09:17:01 -0700, Said Jackson via time-nuts 
time-nuts@febo.com wrote:

Hi Stephanie,

Welcome to the list!

We designed a 1GHz crystal LO for PLLs (the ULN-1G) using an off the
shelf miniature 500MHz crystal oscillator which is run at 3 rd
overtone internally then using a diode doubler and a steep bandpass
filter using several Mini Circuits ceramic filters and a 20dBm amp.

Works like a charm and has phase noise very close to theoretical..

Bye,
Said

Sent From iPhone


On Sep 15, 2014, at 5:50, steph.rey steph@wanadoo.fr wrote:

Hi the list,

Just wanted to introduce myself for my 1st message.
I'm Stephane, 40, living in France, at the moment working in RF  
electronics for a particles accelerator lab. I'm hamradio as well, and 
I do enjoy especially weak and accurate signals.
I'm desiging various RF circuits. Current design is a universal PLL 
able to operate from 0.5 to 6 GHz depending on the VCO and supposed to 
be low-jitter (1ps) regarding the application.
I'm also starting a new design of low noise PLL and there will be 
probably a lot of question arising... I'm starting with the 1 GHz LO 
made upon a 100 MHz VCXO + multipliers/filters/MMICs.
I want to focus deeper on low phase noise/jitter, synchronization 
and low-noise PLL techniques.

I believe this is a good place for most of these topics.


Cheers
Stephane


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.

___