Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-22 Thread TheInfamousFlavio
1.  Same test with a Fluke 6680 yeilds 10MHz mean exactly.

2.Tried different length cables and swapping the cables.  Same result.

3. Autocal was done (a few times)



- Original Message - 
From: Tom Van Baak [EMAIL PROTECTED]
To: Discussion of precise time and frequency measurement 
time-nuts@febo.com
Sent: Thursday, February 22, 2007 00:12
Subject: Re: [time-nuts] SRS SR620 External Source Issue -- Help Request


 I have an SR620 counter that I set up with a Z3801A as an external 
 reference.  If I put a bnc T connector at the output
of the Z3801A and use two equal length bnc cables, one to the ext. ref 
input on the back and the other to channel A
then do a frequency measurement, I get a mean that is about .0015 Hz below 
10,000,000.Hz.

 Does any know why this might be happening? I would expect it to read 
 10,000,000. exactly give or take a couple on
 the last digit.

 Three comments that may help.

 (1)
 0.0015 Hz out of 10 MHz is 1.5e-10 which seems a little
 high but not too bad. But do not expect exactly 10 MHz with
 this sort of test. What you are giving to the channel A
 input is the most highly phase correlated signal you can
 imagine relative to the internal clock and the interpolators.
 This won't happen in real life with real input frequencies.

 (2)
 Try different lengths of the channel A cable and see if the
 number changes. My prediction is it will.

 (3)
 Run the setup menu self-calibration if you haven't in a while.

 /tvb



 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-22 Thread TheInfamousFlavio
I read the same spec on pg 63 but I read it as a fluctuation not an offset.

I'm curious if you try the test Bruce suggested and set the gate time to 
0.1s and then 0.01s and see if the offset increase by a factor of 10 for 
each change?

I'm finding it hard to believe that this top of the line instrument can't 
handle this rudimentary task when my old Fluke 6680 with 500ps one shot 
precision (25 times slower than the SR620)  handles it perfectly.

I'm still not buying this offset is correct.


- Original Message - 
From: Robert Crawford [EMAIL PROTECTED]
To: Discussion of precise time and frequency measurement 
time-nuts@febo.com
Sent: Wednesday, February 21, 2007 23:13
Subject: Re: [time-nuts] SRS SR620 External Source Issue -- Help Request


I ran into this same problem with my first SR620.  I thought it was
 defective, so I sent it to SRS for a cal and refurbishment.  It got
 better, but still had an offset of 0.0004 Hz.  Later I bought a second
 one, and it, too, had an offset similar to yours.

 A careful reading of the manual, page 63, Performance Tests, Accuracy
 (where the 10 MHz output from the back panel is measured at the A input)
 reveals that a +/- 0.0035 Hz offset on a 10 MHz input is acceptable and
 within spec.  I think this also applies to your test configuration.

 I have a photograph of my two SR620s, both using an external clock
 (PRS-10/GPS from an FS710 Distribution Amp), and both measuring the same
 signal on their A inputs.  One reads 10,000,000.00096 Hz and the other
 reads 9.999,999.99827 Hz.  I used a gate time of 1 second averaged for
 100 readings.  Both SR620s had fresh factory calibrations.

 Bob Crawford

 [EMAIL PROTECTED] wrote:

I have an SR620 counter that I set up with a Z3801A as an external 
reference.  If I put a bnc T connector at the output of the Z3801A and use 
two equal length bnc cables, one to the ext. ref input on the back and the 
other to channel A then do a frequency measurement, I get a mean that is 
about .0015 Hz below 10,000,000.Hz.

Does any know why this might be happening? I would expect it to read 
10,000,000. exactly give or take a couple on the last digit.

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts





 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-22 Thread Ulrich Bangert
Dear ???,

I encountered pretty much the same effect when I bought my second hand
SR620. As you, I have been extremly surprised by the fact that the
counter did not display exactly 10.0 MHz under the conditions that you
describe.

However, the answer is in the software: The built in time base has a
coarse gain adjustment screw (electronic adjustment) and the fine gain
adjustment is done by calibration byte number 4 in the software. Please
note that this calibration byte (in fact it is 12 bits) is NOT
influenced by the autocal procedure so no wonder that autocal brought
you no changes. It seems, that even with an external time base connected
this calibration byte stays active and that you have to set it by hand
to its most correct value as explained under point 5 of Calibration
Procedure / Clock Oscillator Calibration on page 77 of the handbook.

If you don't have the handbook available: You may download it (but
missing the schematics) from the SRS web site.

Best regards
Ulrich Bangert, DF6JB 

 -Ursprüngliche Nachricht-
 Von: [EMAIL PROTECTED] 
 [mailto:[EMAIL PROTECTED] Im Auftrag von 
 [EMAIL PROTECTED]
 Gesendet: Mittwoch, 21. Februar 2007 23:57
 An: Discussion of precise time and frequency measurement
 Betreff: [time-nuts] SRS SR620 External Source Issue -- Help Request
 
 
 I have an SR620 counter that I set up with a Z3801A as an 
 external reference.  If I put a bnc T connector at the output 
 of the Z3801A and use two equal length bnc cables, one to the 
 ext. ref input on the back and the other to channel A then do 
 a frequency measurement, I get a mean that is about .0015 Hz 
 below 10,000,000.Hz.
 
 Does any know why this might be happening? I would expect it 
 to read 10,000,000. exactly give or take a couple on the 
 last digit.
 
 ___
 time-nuts mailing list
 time-nuts@febo.com 
 https://www.febo.com/cgi- bin/mailman/listinfo/time-nuts
 


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-22 Thread Robert Crawford
Flavio,

Your experience exactly parallels my surprise and disappointment when I 
bought my first SR620.  In fact, that's why I bought my second one: I 
just couldn't believe the results, even though it had been freshly 
calibrated by the factory.

However, the Specifications on page vii of the manual tells the story.  
The Frequency Error is listed as  +/- ((100 ps typ) [350 ps max])/Gate 
+ Timebase Error) x Frequency.  If you make timebase error zero and use 
a 1 second gate, for a 10 MHz signal you get +/- 0.001 Hz, or +/- 0.0035 
Hz max.  For a 0.1 second gate, the typical error goes up to 0.01 Hz, 
which is exactly what you are seeing.

I haven't played with the CALBYTE 4 value since I sent both of mine to 
SRS for calibration.  If I remember right you need to move a jumper 
inside the unit to enable adjustment of this value, and I didn't want to 
break the factory calibration stickers.  I thought this was simply the 
internal time base frequency adjustment.

Bob Crawford



[EMAIL PROTECTED] wrote:

I read the same spec on pg 63 but I read it as a fluctuation not an offset.

I'm curious if you try the test Bruce suggested and set the gate time to 
0.1s and then 0.01s and see if the offset increase by a factor of 10 for 
each change?

I'm finding it hard to believe that this top of the line instrument can't 
handle this rudimentary task when my old Fluke 6680 with 500ps one shot 
precision (25 times slower than the SR620)  handles it perfectly.

I'm still not buying this offset is correct.


- Original Message - 
From: Robert Crawford [EMAIL PROTECTED]
To: Discussion of precise time and frequency measurement 
time-nuts@febo.com
Sent: Wednesday, February 21, 2007 23:13
Subject: Re: [time-nuts] SRS SR620 External Source Issue -- Help Request


  

I ran into this same problem with my first SR620.  I thought it was
defective, so I sent it to SRS for a cal and refurbishment.  It got
better, but still had an offset of 0.0004 Hz.  Later I bought a second
one, and it, too, had an offset similar to yours.

A careful reading of the manual, page 63, Performance Tests, Accuracy
(where the 10 MHz output from the back panel is measured at the A input)
reveals that a +/- 0.0035 Hz offset on a 10 MHz input is acceptable and
within spec.  I think this also applies to your test configuration.

I have a photograph of my two SR620s, both using an external clock
(PRS-10/GPS from an FS710 Distribution Amp), and both measuring the same
signal on their A inputs.  One reads 10,000,000.00096 Hz and the other
reads 9.999,999.99827 Hz.  I used a gate time of 1 second averaged for
100 readings.  Both SR620s had fresh factory calibrations.

Bob Crawford

[EMAIL PROTECTED] wrote:



I have an SR620 counter that I set up with a Z3801A as an external 
reference.  If I put a bnc T connector at the output of the Z3801A and use 
two equal length bnc cables, one to the ext. ref input on the back and the 
other to channel A then do a frequency measurement, I get a mean that is 
about .0015 Hz below 10,000,000.Hz.

Does any know why this might be happening? I would expect it to read 
10,000,000. exactly give or take a couple on the last digit.

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts




  

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts





___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


  

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-22 Thread TheInfamousFlavio
Ulrich wrote
that even with an external time base connected
this calibration byte stays active and that you have to set it by hand

I tried it, but on my unit calbyte 4 does not have an effect when the 
external timebase is connect, unlike yours.
-thank Flavio

- Original Message - 
From: Ulrich Bangert [EMAIL PROTECTED]
To: 'Discussion of precise time and frequency measurement' 
time-nuts@febo.com
Sent: Thursday, February 22, 2007 11:24
Subject: Re: [time-nuts] SRS SR620 External Source Issue -- Help Request


Dear ???,

I encountered pretty much the same effect when I bought my second hand
SR620. As you, I have been extremly surprised by the fact that the
counter did not display exactly 10.0 MHz under the conditions that you
describe.

However, the answer is in the software: The built in time base has a
coarse gain adjustment screw (electronic adjustment) and the fine gain
adjustment is done by calibration byte number 4 in the software. Please
note that this calibration byte (in fact it is 12 bits) is NOT
influenced by the autocal procedure so no wonder that autocal brought
you no changes. It seems, that even with an external time base connected
this calibration byte stays active and that you have to set it by hand
to its most correct value as explained under point 5 of Calibration
Procedure / Clock Oscillator Calibration on page 77 of the handbook.

If you don't have the handbook available: You may download it (but
missing the schematics) from the SRS web site.

Best regards
Ulrich Bangert, DF6JB

 -Ursprüngliche Nachricht-
 Von: [EMAIL PROTECTED]
 [mailto:[EMAIL PROTECTED] Im Auftrag von
 [EMAIL PROTECTED]
 Gesendet: Mittwoch, 21. Februar 2007 23:57
 An: Discussion of precise time and frequency measurement
 Betreff: [time-nuts] SRS SR620 External Source Issue -- Help Request


 I have an SR620 counter that I set up with a Z3801A as an
 external reference.  If I put a bnc T connector at the output
 of the Z3801A and use two equal length bnc cables, one to the
 ext. ref input on the back and the other to channel A then do
 a frequency measurement, I get a mean that is about .0015 Hz
 below 10,000,000.Hz.

 Does any know why this might be happening? I would expect it
 to read 10,000,000. exactly give or take a couple on the
 last digit.

 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi- bin/mailman/listinfo/time-nuts



___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-22 Thread TheInfamousFlavio
Bob,

I see what your saying about the frequency error spec on pg vii, but I'm 
still reading as a fluctuation not an offset of the mean.  In my mind, I 
don't see why there should be some sort of adjustment for this.  I really 
doesn't make sense that the would make a counter that doesn't display a mean 
that is accurate.
-Flavio


- Original Message - 
From: Robert Crawford [EMAIL PROTECTED]
To: Discussion of precise time and frequency measurement 
time-nuts@febo.com
Sent: Thursday, February 22, 2007 11:58
Subject: Re: [time-nuts] SRS SR620 External Source Issue -- Help Request


 Flavio,

 Your experience exactly parallels my surprise and disappointment when I
 bought my first SR620.  In fact, that's why I bought my second one: I
 just couldn't believe the results, even though it had been freshly
 calibrated by the factory.

 However, the Specifications on page vii of the manual tells the story.
 The Frequency Error is listed as  +/- ((100 ps typ) [350 ps max])/Gate
 + Timebase Error) x Frequency.  If you make timebase error zero and use
 a 1 second gate, for a 10 MHz signal you get +/- 0.001 Hz, or +/- 0.0035
 Hz max.  For a 0.1 second gate, the typical error goes up to 0.01 Hz,
 which is exactly what you are seeing.

 I haven't played with the CALBYTE 4 value since I sent both of mine to
 SRS for calibration.  If I remember right you need to move a jumper
 inside the unit to enable adjustment of this value, and I didn't want to
 break the factory calibration stickers.  I thought this was simply the
 internal time base frequency adjustment.

 Bob Crawford



 [EMAIL PROTECTED] wrote:

I read the same spec on pg 63 but I read it as a fluctuation not an 
offset.

I'm curious if you try the test Bruce suggested and set the gate time to
0.1s and then 0.01s and see if the offset increase by a factor of 10 for
each change?

I'm finding it hard to believe that this top of the line instrument can't
handle this rudimentary task when my old Fluke 6680 with 500ps one shot
precision (25 times slower than the SR620)  handles it perfectly.

I'm still not buying this offset is correct.


- Original Message - 
From: Robert Crawford [EMAIL PROTECTED]
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Sent: Wednesday, February 21, 2007 23:13
Subject: Re: [time-nuts] SRS SR620 External Source Issue -- Help Request




I ran into this same problem with my first SR620.  I thought it was
defective, so I sent it to SRS for a cal and refurbishment.  It got
better, but still had an offset of 0.0004 Hz.  Later I bought a second
one, and it, too, had an offset similar to yours.

A careful reading of the manual, page 63, Performance Tests, Accuracy
(where the 10 MHz output from the back panel is measured at the A input)
reveals that a +/- 0.0035 Hz offset on a 10 MHz input is acceptable and
within spec.  I think this also applies to your test configuration.

I have a photograph of my two SR620s, both using an external clock
(PRS-10/GPS from an FS710 Distribution Amp), and both measuring the same
signal on their A inputs.  One reads 10,000,000.00096 Hz and the other
reads 9.999,999.99827 Hz.  I used a gate time of 1 second averaged for
100 readings.  Both SR620s had fresh factory calibrations.

Bob Crawford

[EMAIL PROTECTED] wrote:



I have an SR620 counter that I set up with a Z3801A as an external
reference.  If I put a bnc T connector at the output of the Z3801A and 
use
two equal length bnc cables, one to the ext. ref input on the back and 
the
other to channel A then do a frequency measurement, I get a mean that is
about .0015 Hz below 10,000,000.Hz.

Does any know why this might be happening? I would expect it to read
10,000,000. exactly give or take a couple on the last digit.

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts






___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts





___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts




 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-22 Thread Hal Murray

 Perhaps you're running into some sort of meta-stability problem since
 your  clock and input are synchronous. Try lengthening one of the
 cables a foot  or so and see if it makes any difference. 

I'd be very suspicious of a metastability problem if the results are even 
somewhat reproducible.

Most of the quirks that get blamed on metastability are just simple 
setup/hold errors such that an input to a state machine has different values 
at different FFs.  The two obvious causes are a longer delay on one path 
(more logic, longer routing) or slightly different setup times on the FFs so 
a signal gets to both at the same time and meets setup one one but doesn't 
the other one.


-- 
These are my opinions, not necessarily my employer's.  I hate spam.




___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-22 Thread Dr Bruce Griffiths
[EMAIL PROTECTED] wrote:
 1.  Same test with a Fluke 6680 yeilds 10MHz mean exactly.

 2.Tried different length cables and swapping the cables.  Same result.

 3. Autocal was done (a few times)



 - Original Message - 
 From: Tom Van Baak [EMAIL PROTECTED]
 To: Discussion of precise time and frequency measurement 
 time-nuts@febo.com
 Sent: Thursday, February 22, 2007 00:12
 Subject: Re: [time-nuts] SRS SR620 External Source Issue -- Help Request


   
 I have an SR620 counter that I set up with a Z3801A as an external 
 reference.  If I put a bnc T connector at the output
 of the Z3801A and use two equal length bnc cables, one to the ext. ref 
 input on the back and the other to channel A
 then do a frequency measurement, I get a mean that is about .0015 Hz below 
 10,000,000.Hz.

 Does any know why this might be happening? I would expect it to read 
 10,000,000. exactly give or take a couple on
 the last digit.
   
 Three comments that may help.

 (1)
 0.0015 Hz out of 10 MHz is 1.5e-10 which seems a little
 high but not too bad. But do not expect exactly 10 MHz with
 this sort of test. What you are giving to the channel A
 input is the most highly phase correlated signal you can
 imagine relative to the internal clock and the interpolators.
 This won't happen in real life with real input frequencies.

 (2)
 Try different lengths of the channel A cable and see if the
 number changes. My prediction is it will.

 (3)
 Run the setup menu self-calibration if you haven't in a while.

 /tvb



 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

 


 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

   
Flavio

The effect that Tom described (differential nonlinearity in the 
interpolators) has a period of about 11.11 ns, so changing the relative 
delay by around 2.7ns (or small odd multiple thereof) is most likely to 
uncover the effect. If your relative delay changes is a multiple of 
11,11ns or close to it you wont see much of an effect.
However the manual indicates that this is relatively small when the time 
interval being measured is longer than a few microseconds or so.

If the differential delay of the 2 synchronisers isn't corrected by the 
calibration procedure then a value of 150 ps would not be unexpected, 
the synchroniser flipflops are MC10H131's which have a clock to output 
delay of up to 1.5ns. As far as I can tell the autocal procedure will 
not correct such an offset.

Bruce

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


[time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-21 Thread TheInfamousFlavio
I have an SR620 counter that I set up with a Z3801A as an external reference.  
If I put a bnc T connector at the output of the Z3801A and use two equal length 
bnc cables, one to the ext. ref input on the back and the other to channel A 
then do a frequency measurement, I get a mean that is about .0015 Hz below 
10,000,000.Hz.

Does any know why this might be happening? I would expect it to read 
10,000,000. exactly give or take a couple on the last digit.

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-21 Thread Magnus Danielson
From: [EMAIL PROTECTED]
Subject: Re: [time-nuts] SRS SR620 External Source Issue -- Help Request
Date: Wed, 21 Feb 2007 18:13:08 -0500
Message-ID: [EMAIL PROTECTED]

 I've done the self calibration. Nothing changes.  I've let everything 
 warm-up and checked both channels A and B, same thing.  I've swapped the 
 cables on the front and the back, same thing.  I've tried a T-connector on 
 the ref input with 50ohm termination, same thing.
 
 The readings are pretty stable, they are just offset by -0.0015 Hz.

Hmm. Interestingly enought I've seen something similar on a HP5372A where the
A channel was slightly low on the frequency but not the B channel. I have not
investigated that in detail to make any conclusions but that 5372A may be in
for a slight trim-up. Since it's interpolators allow for interesting trimming
of each interpolator step it would be interesting to see if I could trim it in
and out of frequency!

Now, if the interpolator trimming is the culprit, I need to wrap my head around
on how it turns out as a frequency error.

I'll check it again tomorrow, as I use it daily right now.

I don't have the manuals (and particular the service manual) for the SR620,
does anybody have them in electronics form?

Cheers,
Magnus

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-21 Thread Dr Bruce Griffiths
Magnus Danielson wrote:
 I don't have the manuals (and particular the service manual) for the 
 SR620,
 does anybody have them in electronics form?

 Cheers,
 Magnus

 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

 
Hej Magnus

http://thinksrs.com/downloads/PDFs/Manuals/SR620m.pdf

Has everything except circuit diagrams - may be able to reconstruct from 
circuit description.
Bruce

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-21 Thread Dr Bruce Griffiths
[EMAIL PROTECTED] wrote:
 I've done the self calibration. Nothing changes.  I've let everything 
 warm-up and checked both channels A and B, same thing.  I've swapped the 
 cables on the front and the back, same thing.  I've tried a T-connector on 
 the ref input with 50ohm termination, same thing.

 The readings are pretty stable, they are just offset by -0.0015 Hz.




 - Original Message - 
 From: Brooke Clarke [EMAIL PROTECTED]
 To: Discussion of precise time and frequency measurement 
 time-nuts@febo.com
 Sent: Wednesday, February 21, 2007 18:04
 Subject: Re: [time-nuts] SRS SR620 External Source Issue -- Help Request


   
 Hi Flavio:

 Have you tried using averaging?  The right most digits are from an
 interpolator and if  you check the manual the displayed resolution is
 much greater than the accuracy, but if you average say 1,000 readings
 then those digits get much better.  Also have you done the self 
 calibration?

 Have Fun,

 Brooke Clarke

 w/Java http://www.PRC68.com
 w/o Java http://www.pacificsites.com/~brooke/PRC68COM.shtml
 http://www.precisionclock.com



 [EMAIL PROTECTED] wrote:

 
 I have an SR620 counter that I set up with a Z3801A as an external 
 reference.  If I put a bnc T connector at the output of the Z3801A and use 
 two equal length bnc cables, one to the ext. ref input on the back and the 
 other to channel A then do a frequency measurement, I get a mean that is 
 about .0015 Hz below 10,000,000.Hz.

 Does any know why this might be happening? I would expect it to read 
 10,000,000. exactly give or take a couple on the last digit.

 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts



   
 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

 


 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

   
Flavio

Your  SR620 is well within its specifications, the tolerance for this 
test is -0.0035 Hz to + 0.0035 Hz.

Bruce

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-21 Thread Mike Fahmie
At 02:57 PM 2/21/2007, you wrote:
I have an SR620 counter that I set up with a Z3801A as an external 
reference.  If I put a bnc T connector at the output of the Z3801A and use 
two equal length bnc cables, one to the ext. ref input on the back and the 
other to channel A then do a frequency measurement, I get a mean that is 
about .0015 Hz below 10,000,000.Hz.

Does any know why this might be happening? I would expect it to read 
10,000,000. exactly give or take a couple on the last digit.

Perhaps you're running into some sort of meta-stability problem since your 
clock and input are synchronous. Try lengthening one of the cables a foot 
or so and see if it makes any difference.

Also, it could be the compound result of noise in both the reference input 
and signal input amplifiers.

-Mike-

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-21 Thread Mike Fahmie
At 03:50 PM 2/21/2007, you wrote:
At 02:57 PM 2/21/2007, you wrote:
 I have an SR620 counter that I set up with a Z3801A as an external
 reference.  If I put a bnc T connector at the output of the Z3801A and use
 two equal length bnc cables, one to the ext. ref input on the back and the
 other to channel A then do a frequency measurement, I get a mean that is
 about .0015 Hz below 10,000,000.Hz.
 
 Does any know why this might be happening? I would expect it to read
 10,000,000. exactly give or take a couple on the last digit.

Perhaps you're running into some sort of meta-stability problem since your
clock and input are synchronous. Try lengthening one of the cables a foot
or so and see if it makes any difference.

Also, it could be the compound result of noise in both the reference input
and signal input amplifiers.

-Mike-

Another thing...It sounds like the Z3801A is being asked to drive 25 ohms 
(two 50 ohm loads in parallel), perhaps the waveform is distorting in such 
a way as to present a slow risetime edge to the counter, thus agravating 
the effects of noise.

-Mike-

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-21 Thread Dr Bruce Griffiths
Mike Fahmie wrote:
 At 02:57 PM 2/21/2007, you wrote:
   
 I have an SR620 counter that I set up with a Z3801A as an external 
 reference.  If I put a bnc T connector at the output of the Z3801A and use 
 two equal length bnc cables, one to the ext. ref input on the back and the 
 other to channel A then do a frequency measurement, I get a mean that is 
 about .0015 Hz below 10,000,000.Hz.

 Does any know why this might be happening? I would expect it to read 
 10,000,000. exactly give or take a couple on the last digit.
 

 Perhaps you're running into some sort of meta-stability problem since your 
 clock and input are synchronous. Try lengthening one of the cables a foot 
 or so and see if it makes any difference.

 Also, it could be the compound result of noise in both the reference input 
 and signal input amplifiers.

 -Mike-

 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

   
Flavio

It could also, depending on the design, be something quite mundane like 
the difference in propagation delay for an ECL gate for rising and 
falling edges.
Or the difference in propagation delay from clock to output of an ECL 
fliflop between 0-1 and 1-0 output transitions.

Bruce

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-21 Thread TheInfamousFlavio

 Your  SR620 is well within its specifications, the tolerance for this
 test is -0.0035 Hz to + 0.0035 Hz.

The readings fluctuate by that amount, but the mean is still off by -0.00150 
Hz.

 Changing cable length...

Tried it, no difference.


 25Ohm load with two terms

Tried all variations of with and without term.  No difference.


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-21 Thread Magnus Danielson
From: Dr Bruce Griffiths [EMAIL PROTECTED]
Subject: Re: [time-nuts] SRS SR620 External Source Issue -- Help Request
Date: Thu, 22 Feb 2007 12:40:57 +1300
Message-ID: [EMAIL PROTECTED]

Hej Bruce,

 http://thinksrs.com/downloads/PDFs/Manuals/SR620m.pdf

Many thanks!

 Has everything except circuit diagrams - may be able to reconstruct from 
 circuit description.

But then there is the patent. :)

Schematics would be good too.

Now, that the measures may be a little off the mark from measure to measure
is not a supprise, but that a frequency measure (of its own timebase) would be
statically off the mark is interesting.

Cheers,
Magnus

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request -- A thought

2007-02-21 Thread TheInfamousFlavio
Here's a thought.  Perhaps there is a problem with the rear reference input; 
A and B channels are fine since they read the same error.  The question is 
Is there a way of adjusting the Reference Input so that it correctly read 
the 10Mhz input?

I tried the same setup with my other counter and it read 10,000,000. or 
thereabouts correctly. So rule out the GPSDO and the cables.



- Original Message - 
From: Dr Bruce Griffiths [EMAIL PROTECTED]
To: Discussion of precise time and frequency measurement 
time-nuts@febo.com
Sent: Wednesday, February 21, 2007 19:47
Subject: Re: [time-nuts] SRS SR620 External Source Issue -- Help Request


 Hej Magnus

 The Frequency measurement logic (before the interpolators) appears to
 use a pair of 2 bit synchonisers, one for the rising edge and one for
 the falling edge of the internally generated frequency gate. The
 synchronisers synchronise the respective transitions of the frequency
 gate signal to the external frequency being measured.
 One synchroniser uses U508A + U509A the other uses U508B + U509B, so
 that any mismatch in propagation delays between U509A And U509B will
 result in a small error in the time interval. Even for ECL flipflops a
 propagation delay difference between U50(A and U509B of 150ps wouldn't
 be unusual. As far as I can tell the calibration procedures do not
 address this issue.  The uncalibrated differential delay through the
 interpolation circuit source selection multiplexer will also contribute
 an offset.
 Perhaps a better assignment of propagation delay differences may be
 U509A and U509B 50ps, multiplexer and wiring delay mismatches 100ps.

 Bruce

 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request -- A thought

2007-02-21 Thread Dr Bruce Griffiths
[EMAIL PROTECTED] wrote:
 Here's a thought.  Perhaps there is a problem with the rear reference input; 
 A and B channels are fine since they read the same error.  The question is 
 Is there a way of adjusting the Reference Input so that it correctly read 
 the 10Mhz input?

 I tried the same setup with my other counter and it read 10,000,000. or 
 thereabouts correctly. So rule out the GPSDO and the cables.



 - Original Message - 
 From: Dr Bruce Griffiths [EMAIL PROTECTED]
 To: Discussion of precise time and frequency measurement 
 time-nuts@febo.com
 Sent: Wednesday, February 21, 2007 19:47
 Subject: Re: [time-nuts] SRS SR620 External Source Issue -- Help Request


   
 Hej Magnus

 The Frequency measurement logic (before the interpolators) appears to
 use a pair of 2 bit synchonisers, one for the rising edge and one for
 the falling edge of the internally generated frequency gate. The
 synchronisers synchronise the respective transitions of the frequency
 gate signal to the external frequency being measured.
 One synchroniser uses U508A + U509A the other uses U508B + U509B, so
 that any mismatch in propagation delays between U509A And U509B will
 result in a small error in the time interval. Even for ECL flipflops a
 propagation delay difference between U50(A and U509B of 150ps wouldn't
 be unusual. As far as I can tell the calibration procedures do not
 address this issue.  The uncalibrated differential delay through the
 interpolation circuit source selection multiplexer will also contribute
 an offset.
 Perhaps a better assignment of propagation delay differences may be
 U509A and U509B 50ps, multiplexer and wiring delay mismatches 100ps.

 Bruce

 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

 


 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

   
Flavio

Not easily.

However if you run a test with say a 0.1 sec gate time and average 
several readings, then if my model is correct the apparent offset in the 
mean will be increased by a factor of 10 over that for a a 1 second gate 
time.

Bruce

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request -- A thought

2007-02-21 Thread TheInfamousFlavio

...if my model is correct the apparent offset in the
 mean will be increased by a factor of 10 over that for a a 1 second gate
 time.

 Bruce


Looks like you're right Bruce.  I did both 1, 0.1 and 0.01 gate times and 
the error increased by a factor of 10.

And it doesn't sound like this is a fixable in any easy sort of manner, 
right? 


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request -- A thought

2007-02-21 Thread Dr Bruce Griffiths
[EMAIL PROTECTED] wrote:
 ...if my model is correct the apparent offset in the
 mean will be increased by a factor of 10 over that for a a 1 second gate
 time.

 Bruce
 


 Looks like you're right Bruce.  I did both 1, 0.1 and 0.01 gate times and 
 the error increased by a factor of 10.

 And it doesn't sound like this is a fixable in any easy sort of manner, 
 right? 


 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

   
Flavio

If the SR620 can do some simple arithmetic before displaying the result, 
then this calibration problem is easily corrected.
Otherwise you will have to do the corrections on a PC (or hand held 
calculator).
The correction will vary with time and temperature so you will need to 
measure it periodically.

Corrected frequency = measured frequency*(nominal gate time)/(nominal 
gate time + differential delay).
In your case the differential delay appears to be about 150 picoseconds.

Bruce

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-21 Thread Robert Crawford
I ran into this same problem with my first SR620.  I thought it was 
defective, so I sent it to SRS for a cal and refurbishment.  It got 
better, but still had an offset of 0.0004 Hz.  Later I bought a second 
one, and it, too, had an offset similar to yours.

A careful reading of the manual, page 63, Performance Tests, Accuracy 
(where the 10 MHz output from the back panel is measured at the A input) 
reveals that a +/- 0.0035 Hz offset on a 10 MHz input is acceptable and 
within spec.  I think this also applies to your test configuration.

I have a photograph of my two SR620s, both using an external clock 
(PRS-10/GPS from an FS710 Distribution Amp), and both measuring the same 
signal on their A inputs.  One reads 10,000,000.00096 Hz and the other 
reads 9.999,999.99827 Hz.  I used a gate time of 1 second averaged for 
100 readings.  Both SR620s had fresh factory calibrations. 

Bob Crawford

[EMAIL PROTECTED] wrote:

I have an SR620 counter that I set up with a Z3801A as an external reference.  
If I put a bnc T connector at the output of the Z3801A and use two equal 
length bnc cables, one to the ext. ref input on the back and the other to 
channel A then do a frequency measurement, I get a mean that is about .0015 Hz 
below 10,000,000.Hz.

Does any know why this might be happening? I would expect it to read 
10,000,000. exactly give or take a couple on the last digit.

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


  


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request -- A thought

2007-02-21 Thread Didier Juges
Dr Bruce Griffiths wrote:
 [EMAIL PROTECTED] wrote:
   
 ...if my model is correct the apparent offset in the
 mean will be increased by a factor of 10 over that for a a 1 second gate
 time.

 Bruce
 
   
 Looks like you're right Bruce.  I did both 1, 0.1 and 0.01 gate times and 
 the error increased by a factor of 10.

 And it doesn't sound like this is a fixable in any easy sort of manner, 
 right? 


 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

   
 
 Flavio

 If the SR620 can do some simple arithmetic before displaying the result, 
 then this calibration problem is easily corrected.
 Otherwise you will have to do the corrections on a PC (or hand held 
 calculator).
 The correction will vary with time and temperature so you will need to 
 measure it periodically.

 Corrected frequency = measured frequency*(nominal gate time)/(nominal 
 gate time + differential delay).
 In your case the differential delay appears to be about 150 picoseconds.

 Bruce

   
Bruce/Flavio,

Could that be mostly corrected (at constant temperature) by inserting 
the proper delay at the right place (in the form of a short piece of 
coax cable)?

It seems that the error could be easily reduced by a factor of 10.

Didier
___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request

2007-02-21 Thread Tom Van Baak
I have an SR620 counter that I set up with a Z3801A as an external reference.  
If I put a bnc T connector at the output 
of the Z3801A and use two equal length bnc cables, one to the ext. ref input 
on the back and the other to channel A 
then do a frequency measurement, I get a mean that is about .0015 Hz below 
10,000,000.Hz.

 Does any know why this might be happening? I would expect it to read 
 10,000,000. exactly give or take a couple on 
 the last digit.

Three comments that may help.

(1)
0.0015 Hz out of 10 MHz is 1.5e-10 which seems a little
high but not too bad. But do not expect exactly 10 MHz with
this sort of test. What you are giving to the channel A
input is the most highly phase correlated signal you can
imagine relative to the internal clock and the interpolators.
This won't happen in real life with real input frequencies.

(2)
Try different lengths of the channel A cable and see if the
number changes. My prediction is it will.

(3)
Run the setup menu self-calibration if you haven't in a while.

/tvb 



___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SRS SR620 External Source Issue -- Help Request -- A thought

2007-02-21 Thread Dr Bruce Griffiths
Didier Juges wrote:
 Dr Bruce Griffiths wrote:
   
 [EMAIL PROTECTED] wrote:
   
 
 ...if my model is correct the apparent offset in the
 mean will be increased by a factor of 10 over that for a a 1 second gate
 time.

 Bruce
 
   
 
 Looks like you're right Bruce.  I did both 1, 0.1 and 0.01 gate times and 
 the error increased by a factor of 10.

 And it doesn't sound like this is a fixable in any easy sort of manner, 
 right? 


 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

   
 
   
 Flavio

 If the SR620 can do some simple arithmetic before displaying the result, 
 then this calibration problem is easily corrected.
 Otherwise you will have to do the corrections on a PC (or hand held 
 calculator).
 The correction will vary with time and temperature so you will need to 
 measure it periodically.

 Corrected frequency = measured frequency*(nominal gate time)/(nominal 
 gate time + differential delay).
 In your case the differential delay appears to be about 150 picoseconds.

 Bruce

   
 
 Bruce/Flavio,

 Could that be mostly corrected (at constant temperature) by inserting 
 the proper delay at the right place (in the form of a short piece of 
 coax cable)?

 It seems that the error could be easily reduced by a factor of 10.

 Didier
 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

   
Didier

No.
Unless one inserts about 2cm or so inside the instrument between a 
synchronisers and the corresponding interpolator.

Better not to do this as it will vary with time and temperature.
Measuring this error sufficiently often an correcting in software is far 
better and more flexible.

Bruce

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts