[time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Brooke Clarke
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi:

While trying to figure out the O-1814 Rb source I found a 1 PPS signal that's 
less than 10 us wide.  But using an HP 54501A DSO it's almost impossible to 
determine the period.  http://www.prc68.com/I/O1814.shtml

If the display is 1,000 dots wide:
at 2 us/sq the trace takes up about 4.5 squares or 450 dots,
at 200 us/sq the trace takes up about 0.045 squares or 4 dots but the second 
pulse is way off screen.  At slower sweep speeds you can't see the pulse.

Are there scopes that allow the period of a narrow pulse to be measured that 
don't cost an arm and leg?

-- 
Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com
http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Alan Melia
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

I presume it has a calibrated delayed TB how about using that ??
Alan G3NYK

- Original Message -
From: "Brooke Clarke" <[EMAIL PROTECTED]>
To: "Discussion of precise time and frequency measurement"

Sent: Tuesday, October 23, 2007 11:05 PM
Subject: [time-nuts] Scopes and 1 PPS signals?


> Hi:
>
> While trying to figure out the O-1814 Rb source I found a 1 PPS signal
that's
> less than 10 us wide.  But using an HP 54501A DSO it's almost impossible
to
> determine the period.  http://www.prc68.com/I/O1814.shtml
>
> If the display is 1,000 dots wide:
> at 2 us/sq the trace takes up about 4.5 squares or 450 dots,
> at 200 us/sq the trace takes up about 0.045 squares or 4 dots but the
second
> pulse is way off screen.  At slower sweep speeds you can't see the pulse.
>
> Are there scopes that allow the period of a narrow pulse to be measured
that
> don't cost an arm and leg?
>
> --
> Have Fun,
>
> Brooke Clarke
> http://www.PRC68.com
> http://www.precisionclock.com
> http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread WB6BNQ
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi Brook,

You could try using a 74123 (or newer variants) retriggerable one shot to extend
the pulse width to a level that would allow you to see it on the scope.

BillWB6BNQ

Brooke Clarke wrote:

> Hi:
>
> While trying to figure out the O-1814 Rb source I found a 1 PPS signal that's
> less than 10 us wide.  But using an HP 54501A DSO it's almost impossible to
> determine the period.  http://www.prc68.com/I/O1814.shtml
>
> If the display is 1,000 dots wide:
> at 2 us/sq the trace takes up about 4.5 squares or 450 dots,
> at 200 us/sq the trace takes up about 0.045 squares or 4 dots but the second
> pulse is way off screen.  At slower sweep speeds you can't see the pulse.
>
> Are there scopes that allow the period of a narrow pulse to be measured that
> don't cost an arm and leg?
>
> --
> Have Fun,
>
> Brooke Clarke
> http://www.PRC68.com
> http://www.precisionclock.com
> http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Daun Yeagley
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

What you need then is the FatPPS from TAPR.  Our own John Ackermann (N8UR)
designed it primarily for use with time servers by stretching the 1PPS pulse out
to several milliseconds so that the computer could see it.  I suspect the same
principle would be useful for a scope.  Check out:
http://www.tapr.org/kits_fatpps.html

Daun 

-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On Behalf
Of WB6BNQ
Sent: Tuesday, October 23, 2007 6:29 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Scopes and 1 PPS signals?

Hi Brook,

You could try using a 74123 (or newer variants) retriggerable one shot to extend
the pulse width to a level that would allow you to see it on the scope.

BillWB6BNQ

Brooke Clarke wrote:

> Hi:
>
> While trying to figure out the O-1814 Rb source I found a 1 PPS signal that's
> less than 10 us wide.  But using an HP 54501A DSO it's almost impossible to
> determine the period.  http://www.prc68.com/I/O1814.shtml
>
> If the display is 1,000 dots wide:
> at 2 us/sq the trace takes up about 4.5 squares or 450 dots,
> at 200 us/sq the trace takes up about 0.045 squares or 4 dots but the second
> pulse is way off screen.  At slower sweep speeds you can't see the pulse.
>
> Are there scopes that allow the period of a narrow pulse to be measured that
> don't cost an arm and leg?
>
> --
> Have Fun,
>
> Brooke Clarke
> http://www.PRC68.com
> http://www.precisionclock.com
> http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread steve heidmann
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi ,
   
  Isn't  5370 ownership a prerequisite  for this group ?

Brooke Clarke <[EMAIL PROTECTED]> wrote:
  Hi:

While trying to figure out the O-1814 Rb source I found a 1 PPS signal that's 
less than 10 us wide. But using an HP 54501A DSO it's almost impossible to 
determine the period. http://www.prc68.com/I/O1814.shtml

If the display is 1,000 dots wide:
at 2 us/sq the trace takes up about 4.5 squares or 450 dots,
at 200 us/sq the trace takes up about 0.045 squares or 4 dots but the second 
pulse is way off screen. At slower sweep speeds you can't see the pulse.

Are there scopes that allow the period of a narrow pulse to be measured that 
don't cost an arm and leg?

-- 
Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com
http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


 __
Do You Yahoo!?
Tired of spam?  Yahoo! Mail has the best spam protection around 
http://mail.yahoo.com 
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Dave Brown
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY


- Original Message - 
From: "Brooke Clarke" <[EMAIL PROTECTED]>
To: "Discussion of precise time and frequency measurement" 

Sent: Wednesday, October 24, 2007 11:05 AM
Subject: [time-nuts] Scopes and 1 PPS signals?


> ); SAEximRunCond expanded to false
> Errors-To: [EMAIL PROTECTED] RETRY
>
> Hi:
>
> While trying to figure out the O-1814 Rb source I found a 1 PPS 
> signal that's
> less than 10 us wide.  But using an HP 54501A DSO it's almost 
> impossible to
> determine the period.  http://www.prc68.com/I/O1814.shtml
>
> If the display is 1,000 dots wide:
> at 2 us/sq the trace takes up about 4.5 squares or 450 dots,
> at 200 us/sq the trace takes up about 0.045 squares or 4 dots but 
> the second
> pulse is way off screen.  At slower sweep speeds you can't see the 
> pulse.
>
> Are there scopes that allow the period of a narrow pulse to be 
> measured that
> don't cost an arm and leg?
>

 If your DSO has external triggering capability then using an external 
digital delay unit will do the trick.  Something as simple as a one 
shot driven from the preceding pulse to do the triggering?

This sort of thing is one of the reasons some analog scopes are still 
useful.  A tek 7K mainframe with a 7D11 delay unit is a good solution. 
That has up to 1 second precise low jitter delay available.

DaveB, NZ 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Hal Murray
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY


> Are there scopes that allow the period of a narrow pulse to be
> measured that  don't cost an arm and leg? 

Can you use delayed sweep to look at the next pulse?

If you have a good PPS source, that's a handy way to calibrate your scope.


-- 
These are my opinions, not necessarily my employer's.  I hate spam.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Brooke Clarke
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi Alan:

I tried.  At 100 ms/sq the second pulse would be just off screen to the right. 
  At 200 ms/sq the second pulse should be at center screen.  Using the window 
function the window time base can be no more than 10X faster than the main 
trace which would be 20 ms/sq, way too slow to see the pulse.

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com
http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam


Alan Melia wrote:
> ); SAEximRunCond expanded to false
> Errors-To: [EMAIL PROTECTED] RETRY
> 
> I presume it has a calibrated delayed TB how about using that ??
> Alan G3NYK
> 
> - Original Message -
> From: "Brooke Clarke" <[EMAIL PROTECTED]>
> To: "Discussion of precise time and frequency measurement"
> 
> Sent: Tuesday, October 23, 2007 11:05 PM
> Subject: [time-nuts] Scopes and 1 PPS signals?
> 
> 
> 
>>Hi:
>>
>>While trying to figure out the O-1814 Rb source I found a 1 PPS signal
> 
> that's
> 
>>less than 10 us wide.  But using an HP 54501A DSO it's almost impossible
> 
> to
> 
>>determine the period.  http://www.prc68.com/I/O1814.shtml
>>
>>If the display is 1,000 dots wide:
>>at 2 us/sq the trace takes up about 4.5 squares or 450 dots,
>>at 200 us/sq the trace takes up about 0.045 squares or 4 dots but the
> 
> second
> 
>>pulse is way off screen.  At slower sweep speeds you can't see the pulse.
>>
>>Are there scopes that allow the period of a narrow pulse to be measured
> 
> that
> 
>>don't cost an arm and leg?
>>
>>--
>>Have Fun,
>>
>>Brooke Clarke
>>http://www.PRC68.com
>>http://www.precisionclock.com
>>http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
>>
>>___
>>time-nuts mailing list -- time-nuts@febo.com
>>To unsubscribe, go to
> 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> 
>>and follow the instructions there.
> 
> 
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Brooke Clarke
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi Bill:

I'm sure that would work, but it's something else that needs to be done 
sometime in the future.

I checked the pulse coming from the M12+T and it's more like 100 ms wide, 
probably for this reason.

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com
http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam


WB6BNQ wrote:
> Hi Brook,
> 
> You could try using a 74123 (or newer variants) retriggerable one shot to 
> extend
> the pulse width to a level that would allow you to see it on the scope.
> 
> BillWB6BNQ
> 
> Brooke Clarke wrote:
> 
> 
>>Hi:
>>
>>While trying to figure out the O-1814 Rb source I found a 1 PPS signal that's
>>less than 10 us wide.  But using an HP 54501A DSO it's almost impossible to
>>determine the period.  http://www.prc68.com/I/O1814.shtml
>>
>>If the display is 1,000 dots wide:
>>at 2 us/sq the trace takes up about 4.5 squares or 450 dots,
>>at 200 us/sq the trace takes up about 0.045 squares or 4 dots but the second
>>pulse is way off screen.  At slower sweep speeds you can't see the pulse.
>>
>>Are there scopes that allow the period of a narrow pulse to be measured that
>>don't cost an arm and leg?
>>
>>--
>>Have Fun,
>>
>>Brooke Clarke
>>http://www.PRC68.com
>>http://www.precisionclock.com
>>http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
>>
>>___
>>time-nuts mailing list -- time-nuts@febo.com
>>To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>and follow the instructions there.
> 
> 
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Didier Juges
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Some DSOs (my lowly Tek 210 does) have a peak detect acquisition mode where
it will analyze the signal at the highest speed of the A/D converter
regardless of sweep speed and report the min and max for each displayed
pixel line. Main advantages are 1) to make aliasing obvious and 2) to be
able to see very narrow pulses at low sweep speed. I can see 20nS wide
pulses at 1 PPS easily. Of course, the precision of the measurement (if you
use the period function) is limited to the display resolution, but at least
you can get in the ballpark. My scope is in that mode all the time, little
reason not to.

If you do not have such mode, put a diode (i.e. 1N4148) in series with the
probe. The diode will charge the input capacitance of the probe, which will
discharge exponentially in the 10 Mohm resistance and give you something you
can see. Can't beat that for cost. Add some capacitance in parallel with the
probe if needed to widen the pulse.

Didier KO4BB

> -Original Message-
> From: [EMAIL PROTECTED] 
> [mailto:[EMAIL PROTECTED] On Behalf Of Brooke Clarke
> Sent: Tuesday, October 23, 2007 6:24 PM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] Scopes and 1 PPS signals?
> 
> Hi Bill:
> 
> I'm sure that would work, but it's something else that needs 
> to be done sometime in the future.
> 
> I checked the pulse coming from the M12+T and it's more like 
> 100 ms wide, probably for this reason.
> 
> Have Fun,
> 
> Brooke Clarke
> http://www.PRC68.com
> http://www.precisionclock.com
> http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
> 
> 
> WB6BNQ wrote:
> > Hi Brook,
> > 
> > You could try using a 74123 (or newer variants) 
> retriggerable one shot 
> > to extend the pulse width to a level that would allow you 
> to see it on the scope.
> > 
> > BillWB6BNQ
> > 
> > Brooke Clarke wrote:
> > 
> > 
> >>Hi:
> >>
> >>While trying to figure out the O-1814 Rb source I found a 1 
> PPS signal 
> >>that's less than 10 us wide.  But using an HP 54501A DSO 
> it's almost 
> >>impossible to determine the period.  
> >>http://www.prc68.com/I/O1814.shtml
> >>
> >>If the display is 1,000 dots wide:
> >>at 2 us/sq the trace takes up about 4.5 squares or 450 dots, at 200 
> >>us/sq the trace takes up about 0.045 squares or 4 dots but 
> the second 
> >>pulse is way off screen.  At slower sweep speeds you can't 
> see the pulse.
> >>
> >>Are there scopes that allow the period of a narrow pulse to be 
> >>measured that don't cost an arm and leg?
> >>
> >>--
> >>Have Fun,
> >>
> >>Brooke Clarke


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Alan Melia
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi Brooke, ah well its probably a case of "not the right instrument for the
job" but that's a bit tough if its all you got, maybe the extra delay
mentioned by others is what's called for. My old analogue scopes will do it
:-)) Im surprised you can only get 10 screens worth delay. Some scopes will
allow delay and sweep on different ranges of the TB.

Alan

- Original Message -
From: "Brooke Clarke" <[EMAIL PROTECTED]>
To: "Discussion of precise time and frequency measurement"

Sent: Wednesday, October 24, 2007 12:20 AM
Subject: Re: [time-nuts] Scopes and 1 PPS signals?


> Hi Alan:
>
> I tried.  At 100 ms/sq the second pulse would be just off screen to the
right.
>   At 200 ms/sq the second pulse should be at center screen.  Using the
window
> function the window time base can be no more than 10X faster than the main
> trace which would be 20 ms/sq, way too slow to see the pulse.
>
> Have Fun,
>
> Brooke Clarke
> http://www.PRC68.com
> http://www.precisionclock.com
> http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
>
>
> Alan Melia wrote:
> > ); SAEximRunCond expanded to false
> > Errors-To: [EMAIL PROTECTED] RETRY
> >
> > I presume it has a calibrated delayed TB how about using that ??
> > Alan G3NYK
> >
> > - Original Message -
> > From: "Brooke Clarke" <[EMAIL PROTECTED]>
> > To: "Discussion of precise time and frequency measurement"
> > 
> > Sent: Tuesday, October 23, 2007 11:05 PM
> > Subject: [time-nuts] Scopes and 1 PPS signals?
> >
> >
> >
> >>Hi:
> >>
> >>While trying to figure out the O-1814 Rb source I found a 1 PPS signal
> >
> > that's
> >
> >>less than 10 us wide.  But using an HP 54501A DSO it's almost impossible
> >
> > to
> >
> >>determine the period.  http://www.prc68.com/I/O1814.shtml
> >>
> >>If the display is 1,000 dots wide:
> >>at 2 us/sq the trace takes up about 4.5 squares or 450 dots,
> >>at 200 us/sq the trace takes up about 0.045 squares or 4 dots but the
> >
> > second
> >
> >>pulse is way off screen.  At slower sweep speeds you can't see the
pulse.
> >>
> >>Are there scopes that allow the period of a narrow pulse to be measured
> >
> > that
> >
> >>don't cost an arm and leg?
> >>
> >>--
> >>Have Fun,
> >>
> >>Brooke Clarke
> >>http://www.PRC68.com
> >>http://www.precisionclock.com
> >>http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
> >>
> >>___
> >>time-nuts mailing list -- time-nuts@febo.com
> >>To unsubscribe, go to
> >
> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> >
> >>and follow the instructions there.
> >
> >
> >
> > ___
> > time-nuts mailing list -- time-nuts@febo.com
> > To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
> >
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Brooke Clarke
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi Daun:

The previous message suggested using a 74123 one shot, and the FatPPS uses a 
74HC123.  But $45 seems a bit high for just this measurement.

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com
http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam


Daun Yeagley wrote:
> ); SAEximRunCond expanded to false
> Errors-To: [EMAIL PROTECTED] RETRY
> 
> What you need then is the FatPPS from TAPR.  Our own John Ackermann (N8UR)
> designed it primarily for use with time servers by stretching the 1PPS pulse 
> out
> to several milliseconds so that the computer could see it.  I suspect the same
> principle would be useful for a scope.  Check out:
> http://www.tapr.org/kits_fatpps.html
> 
> Daun 
> 
> -Original Message-
> From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On Behalf
> Of WB6BNQ
> Sent: Tuesday, October 23, 2007 6:29 PM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] Scopes and 1 PPS signals?
> 
> Hi Brook,
> 
> You could try using a 74123 (or newer variants) retriggerable one shot to 
> extend
> the pulse width to a level that would allow you to see it on the scope.
> 
> BillWB6BNQ
> 
> Brooke Clarke wrote:
> 
> 
>>Hi:
>>
>>While trying to figure out the O-1814 Rb source I found a 1 PPS signal that's
>>less than 10 us wide.  But using an HP 54501A DSO it's almost impossible to
>>determine the period.  http://www.prc68.com/I/O1814.shtml
>>
>>If the display is 1,000 dots wide:
>>at 2 us/sq the trace takes up about 4.5 squares or 450 dots,
>>at 200 us/sq the trace takes up about 0.045 squares or 4 dots but the second
>>pulse is way off screen.  At slower sweep speeds you can't see the pulse.
>>
>>Are there scopes that allow the period of a narrow pulse to be measured that
>>don't cost an arm and leg?
>>
>>--
>>Have Fun,
>>
>>Brooke Clarke
>>http://www.PRC68.com
>>http://www.precisionclock.com
>>http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
>>
>>___
>>time-nuts mailing list -- time-nuts@febo.com
>>To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>and follow the instructions there.
> 
> 
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> 
> 
> 
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Brooke Clarke
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi Alan:

It's not the delay that's limited, it's the sweep speed, or maybe that's saying 
the same thing.

I do have a SR DG535 that can be used to generate the needed delay per Dave's 
suggestion, but I'm already short of 50 Ohm coax cables.  Tried to get them at 
Radio shack but they no longer carry any 50 Ohm cables or bulk cable.

Any recommendations for good value coax that's on line?

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com
http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam


Alan Melia wrote:
> Hi Brooke, ah well its probably a case of "not the right instrument for the
> job" but that's a bit tough if its all you got, maybe the extra delay
> mentioned by others is what's called for. My old analogue scopes will do it
> :-)) Im surprised you can only get 10 screens worth delay. Some scopes will
> allow delay and sweep on different ranges of the TB.
> 
> Alan
> 
> - Original Message -
> From: "Brooke Clarke" <[EMAIL PROTECTED]>
> To: "Discussion of precise time and frequency measurement"
> 
> Sent: Wednesday, October 24, 2007 12:20 AM
> Subject: Re: [time-nuts] Scopes and 1 PPS signals?
> 
> 
> 
>>Hi Alan:
>>
>>I tried.  At 100 ms/sq the second pulse would be just off screen to the
> 
> right.
> 
>>  At 200 ms/sq the second pulse should be at center screen.  Using the
> 
> window
> 
>>function the window time base can be no more than 10X faster than the main
>>trace which would be 20 ms/sq, way too slow to see the pulse.
>>
>>Have Fun,
>>
>>Brooke Clarke
>>http://www.PRC68.com
>>http://www.precisionclock.com
>>http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
>>
>>
>>Alan Melia wrote:
>>
>>>); SAEximRunCond expanded to false
>>>Errors-To: [EMAIL PROTECTED] RETRY
>>>
>>>I presume it has a calibrated delayed TB how about using that ??
>>>Alan G3NYK
>>>
>>>- Original Message -
>>>From: "Brooke Clarke" <[EMAIL PROTECTED]>
>>>To: "Discussion of precise time and frequency measurement"
>>>
>>>Sent: Tuesday, October 23, 2007 11:05 PM
>>>Subject: [time-nuts] Scopes and 1 PPS signals?
>>>
>>>
>>>
>>>
>>>>Hi:
>>>>
>>>>While trying to figure out the O-1814 Rb source I found a 1 PPS signal
>>>
>>>that's
>>>
>>>
>>>>less than 10 us wide.  But using an HP 54501A DSO it's almost impossible
>>>
>>>to
>>>
>>>
>>>>determine the period.  http://www.prc68.com/I/O1814.shtml
>>>>
>>>>If the display is 1,000 dots wide:
>>>>at 2 us/sq the trace takes up about 4.5 squares or 450 dots,
>>>>at 200 us/sq the trace takes up about 0.045 squares or 4 dots but the
>>>
>>>second
>>>
>>>
>>>>pulse is way off screen.  At slower sweep speeds you can't see the
> 
> pulse.
> 
>>>>Are there scopes that allow the period of a narrow pulse to be measured
>>>
>>>that
>>>
>>>
>>>>don't cost an arm and leg?
>>>>
>>>>--
>>>>Have Fun,
>>>>
>>>>Brooke Clarke
>>>>http://www.PRC68.com
>>>>http://www.precisionclock.com
>>>>http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
>>>>
>>>>___
>>>>time-nuts mailing list -- time-nuts@febo.com
>>>>To unsubscribe, go to
>>>
>>>https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>
>>>
>>>>and follow the instructions there.
>>>
>>>
>>>
>>>___
>>>time-nuts mailing list -- time-nuts@febo.com
>>>To unsubscribe, go to
> 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> 
>>>and follow the instructions there.
>>>
>>
>>___
>>time-nuts mailing list -- time-nuts@febo.com
>>To unsubscribe, go to
> 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> 
>>and follow the instructions there.
> 
> 
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Didier Juges
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

I took some pictures of my Thunderbolt's PPS output. While it's pulse is not
quite as narrow as what you report, it's 10uS, way too narrow for the scope
to display it in normal ("Sample") mode at 250mS/div (the fastest setting
that will allow to measure the period of a PPS signal) on my TDS-210 digital
storage scope.

There are at http://www.ko4bb.com/Test_Equipment/Thunderbolt/PulseStretching

The names are self explanatory, but just in case :-)

1_PPS_10xProbe is the PPS measured at the end of a 3 foot cable with a 10x
probe

2_PPS_Sample Mode is the PPS signal in normal (sample) mode, as you probably
would see it on any DSO without any special mode. It's fair to say it would
be hard to make sure what you are seing or measuring. Please note that some
of the time, the scope triggers, but the trace is flat. I had to wait to
make this shot, which was the best I saw in about 30 seconds.

3_PPS_PeakDetect_Mode is the same signal, with the scope in Peak-Detect
mode. It is easy to check the period (if not very acurate). In this mode,
all sweeps look the same.

4_PPS_Diode is with a 1N4148 between the PPS output and the scope probe. See
how the 10uS pulse is now a solid 250uS at 1/3 the height. It would be much
easier to see on any scope, I need to pull the old trusty Tek 2245A and see
what it does on it.

5_PPS_Diode_5.6nF is the same diode with a 5600pF capacitor across the
probe. The pulse is now strecthed to ~10mS.

6_PPS_Diode_5.6nF is the same signal at the 250mS/div sweep rate.

That's a heck of a lot cheaper than $45. Does not even need a power supply
:-)

Didier KO4BB



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Brooke Clarke
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi Dider:

I'll give that a try tomorrow.  What does your camera setup look like?
Digital SLR on tripod?  Does that mean the TDS-210 can not do a screen dump to 
file?

I've been looking at the Rigol DS1102CD mixed signal scope.  The 100 MHz 2 
analog channel & 16 digital channels is very reasonably priced.  But after 3 
emails I still have not been able to see a programming manual for it.  Although 
they do have a software package.

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com
http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam


Didier Juges wrote:
> ); SAEximRunCond expanded to false
> Errors-To: [EMAIL PROTECTED] RETRY
> 
> I took some pictures of my Thunderbolt's PPS output. While it's pulse is not
> quite as narrow as what you report, it's 10uS, way too narrow for the scope
> to display it in normal ("Sample") mode at 250mS/div (the fastest setting
> that will allow to measure the period of a PPS signal) on my TDS-210 digital
> storage scope.
> 
> There are at http://www.ko4bb.com/Test_Equipment/Thunderbolt/PulseStretching
> 
> The names are self explanatory, but just in case :-)
> 
> 1_PPS_10xProbe is the PPS measured at the end of a 3 foot cable with a 10x
> probe
> 
> 2_PPS_Sample Mode is the PPS signal in normal (sample) mode, as you probably
> would see it on any DSO without any special mode. It's fair to say it would
> be hard to make sure what you are seing or measuring. Please note that some
> of the time, the scope triggers, but the trace is flat. I had to wait to
> make this shot, which was the best I saw in about 30 seconds.
> 
> 3_PPS_PeakDetect_Mode is the same signal, with the scope in Peak-Detect
> mode. It is easy to check the period (if not very acurate). In this mode,
> all sweeps look the same.
> 
> 4_PPS_Diode is with a 1N4148 between the PPS output and the scope probe. See
> how the 10uS pulse is now a solid 250uS at 1/3 the height. It would be much
> easier to see on any scope, I need to pull the old trusty Tek 2245A and see
> what it does on it.
> 
> 5_PPS_Diode_5.6nF is the same diode with a 5600pF capacitor across the
> probe. The pulse is now strecthed to ~10mS.
> 
> 6_PPS_Diode_5.6nF is the same signal at the 250mS/div sweep rate.
> 
> That's a heck of a lot cheaper than $45. Does not even need a power supply
> :-)
> 
> Didier KO4BB
> 
> 
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Didier Juges
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Digital camera (Samsung D53, 5Mpx, 3x optical zoom, 2.4" LCD, was $88 at
WalMart last year) on a bipod, my arms :-)
Flash is turned off, scope contrast adjusted for normal use.

I have the Comm module for the TDS-210, and I have the GPIB controller and
John Miles' software, which I use with the spectrum analyzer. But, my main
computer is not very close to the lab bench, and the laptop is in another
room at the moment, so it's just easier to take the digital camera. 

I bought my TDS-210 at a local hamfest a few years ago and I never regretted
it. It's a great little instrument.

I only have a couple of gripes with it: 1) the contrast voltage is not
temperature compensated, so you have to adjust it when you turn the scope
on, then a couple more times during the first few dozen minutes of
operation, then it will stay good forever, not a biggie.

2) the trigger point wants to be at the center of the screen, and if it's
not, it's going to be so many mS or uS from the center of the screen when
you adjust it for instance. From that point, if you change the time base,
the trigger point moves to remain at the same time delay from the center of
the screen. You can very rapidly have the trigger point a mile or so to the
left of the screen, so to speak, and that can be aggravating when you turn
the horizontal control knob to bring it back into the screen :-) I'm still
not used to it after many years. I wish we could set the scope to have the
trigger point at one or nine divisions, and it would not move when you
change the time base, like the old 468.

These are nothing major, the scope is great.

Didier KO4BB

> -Original Message-
> From: [EMAIL PROTECTED] 
> [mailto:[EMAIL PROTECTED] On Behalf Of Brooke Clarke
> Sent: Tuesday, October 23, 2007 10:20 PM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] Scopes and 1 PPS signals?
> 
> Hi Dider:
> 
> I'll give that a try tomorrow.  What does your camera setup look like?
> Digital SLR on tripod?  Does that mean the TDS-210 can not do 
> a screen dump to file?
> 
> I've been looking at the Rigol DS1102CD mixed signal scope.  
> The 100 MHz 2 analog channel & 16 digital channels is very 
> reasonably priced.  But after 3 emails I still have not been 
> able to see a programming manual for it.  Although they do 
> have a software package.
> 
> Have Fun,
> 
> Brooke Clarke
> http://www.PRC68.com
> http://www.precisionclock.com
> http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
> 
> 
> Didier Juges wrote:
> > ); SAEximRunCond expanded to false
> > Errors-To: [EMAIL PROTECTED] RETRY
> > 
> > I took some pictures of my Thunderbolt's PPS output. While 
> it's pulse 
> > is not quite as narrow as what you report, it's 10uS, way 
> too narrow 
> > for the scope to display it in normal ("Sample") mode at 250mS/div 
> > (the fastest setting that will allow to measure the period of a PPS 
> > signal) on my TDS-210 digital storage scope.
> > 
> > There are at 
> > http://www.ko4bb.com/Test_Equipment/Thunderbolt/PulseStretching
> > 
> > The names are self explanatory, but just in case :-)
> > 
> > 1_PPS_10xProbe is the PPS measured at the end of a 3 foot 
> cable with a 
> > 10x probe
> > 
> > 2_PPS_Sample Mode is the PPS signal in normal (sample) mode, as you 
> > probably would see it on any DSO without any special mode. 
> It's fair 
> > to say it would be hard to make sure what you are seing or 
> measuring. 
> > Please note that some of the time, the scope triggers, but 
> the trace 
> > is flat. I had to wait to make this shot, which was the 
> best I saw in about 30 seconds.
> > 
> > 3_PPS_PeakDetect_Mode is the same signal, with the scope in 
> > Peak-Detect mode. It is easy to check the period (if not very 
> > acurate). In this mode, all sweeps look the same.
> > 
> > 4_PPS_Diode is with a 1N4148 between the PPS output and the scope 
> > probe. See how the 10uS pulse is now a solid 250uS at 1/3 
> the height. 
> > It would be much easier to see on any scope, I need to pull the old 
> > trusty Tek 2245A and see what it does on it.
> > 
> > 5_PPS_Diode_5.6nF is the same diode with a 5600pF capacitor 
> across the 
> > probe. The pulse is now strecthed to ~10mS.
> > 
> > 6_PPS_Diode_5.6nF is the same signal at the 250mS/div sweep rate.
> > 
> > That's a heck of a lot cheaper than $45. Does not even need a power 
> > supply
> > :-)
> > 
> > Didier KO4BB
> > 
&

Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Dean Weiten
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi there,

One option that I've used is to put a flip-flop on the signal, turning
the 1 PPS short pulse into a 1/2 Hz square wave.  You can always see
that on the DSO, and it may even be easier to catch the rising and
falling edges on other assorted equipment (depending on the equipment).

Regards,

Dean Weiten.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Dave Brown
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY


- Original Message - 
From: "Didier Juges" <[EMAIL PROTECTED]>
To: "'Discussion of precise time and frequency measurement'" 

Sent: Wednesday, October 24, 2007 3:58 PM
Subject: Re: [time-nuts] Scopes and 1 PPS signals?


> ); SAEximRunCond expanded to false
> Errors-To: [EMAIL PROTECTED] RETRY
>
> I took some pictures of my Thunderbolt's PPS output. While it's 
> pulse is not
> quite as narrow as what you report, it's 10uS, way too narrow for 
> the scope
> to display it in normal ("Sample") mode at 250mS/div (the fastest 
> setting
> that will allow to measure the period of a PPS signal) on my TDS-210 
> digital
> storage scope.
>
> There are at 
> http://www.ko4bb.com/Test_Equipment/Thunderbolt/PulseStretching
>
> The names are self explanatory, but just in case :-)
>
> 1_PPS_10xProbe is the PPS measured at the end of a 3 foot cable with 
> a 10x
> probe
>
> 2_PPS_Sample Mode is the PPS signal in normal (sample) mode, as you 
> probably
> would see it on any DSO without any special mode. It's fair to say 
> it would
> be hard to make sure what you are seing or measuring. Please note 
> that some
> of the time, the scope triggers, but the trace is flat. I had to 
> wait to
> make this shot, which was the best I saw in about 30 seconds.
>
> 3_PPS_PeakDetect_Mode is the same signal, with the scope in 
> Peak-Detect
> mode. It is easy to check the period (if not very acurate). In this 
> mode,
> all sweeps look the same.
>
> 4_PPS_Diode is with a 1N4148 between the PPS output and the scope 
> probe. See
> how the 10uS pulse is now a solid 250uS at 1/3 the height. It would 
> be much
> easier to see on any scope, I need to pull the old trusty Tek 2245A 
> and see
> what it does on it.
>
> 5_PPS_Diode_5.6nF is the same diode with a 5600pF capacitor across 
> the
> probe. The pulse is now strecthed to ~10mS.
>
> 6_PPS_Diode_5.6nF is the same signal at the 250mS/div sweep rate.
>
> That's a heck of a lot cheaper than $45. Does not even need a power 
> supply
> :-)
>
> Didier KO4BB
>


I took Brookes original inquiry to be about measuring the width of the 
nominally 10 uS  wide PPS pulse-in which case anything that alters the 
pulse width is a no-no.  Precise and stable triggering delay to place 
one of the pulses in the on-screen area is all  that is required.
DaveB, NZ

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-23 Thread Didier Juges
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

I read this:

"But using an HP 54501A DSO it's almost impossible to determine the period.
http://www.prc68.com/I/O1814.shtml";

It seems he wanted to measure the period. For once, the thread actually was
about the original question :-)

Didier

> -Original Message-
> From: [EMAIL PROTECTED] 
> [mailto:[EMAIL PROTECTED] On Behalf Of Dave Brown
> Sent: Tuesday, October 23, 2007 11:21 PM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] Scopes and 1 PPS signals?
> 
> I took Brookes original inquiry to be about measuring the 
> width of the 
> nominally 10 uS  wide PPS pulse-in which case anything that 
> alters the 
> pulse width is a no-no.  Precise and stable triggering delay to place 
> one of the pulses in the on-screen area is all  that is required.
> DaveB, NZ


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-24 Thread Dave Brown
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

But he knows the period already to parts in 10e-9 or better.  His 
email sez it's a PPS signal ex an Rb source.
Not much point in trying to measure that with a scope timebase!
So I still think he wanted to measure the  ~10uS pulse width..

DaveB, NZ

- Original Message - 
From: "Didier Juges" <[EMAIL PROTECTED]>
To: "'Discussion of precise time and frequency measurement'" 

Sent: Wednesday, October 24, 2007 5:27 PM
Subject: Re: [time-nuts] Scopes and 1 PPS signals?


> ); SAEximRunCond expanded to false
> Errors-To: [EMAIL PROTECTED] RETRY
>
> I read this:
>
> "But using an HP 54501A DSO it's almost impossible to determine the 
> period.
> http://www.prc68.com/I/O1814.shtml";
>
> It seems he wanted to measure the period. For once, the thread 
> actually was
> about the original question :-)
>
> Didier
>
>> -Original Message-
>> From: [EMAIL PROTECTED]
>> [mailto:[EMAIL PROTECTED] On Behalf Of Dave Brown
>> Sent: Tuesday, October 23, 2007 11:21 PM
>> To: Discussion of precise time and frequency measurement
>> Subject: Re: [time-nuts] Scopes and 1 PPS signals?
>>
>> I took Brookes original inquiry to be about measuring the
>> width of the
>> nominally 10 uS  wide PPS pulse-in which case anything that
>> alters the
>> pulse width is a no-no.  Precise and stable triggering delay to 
>> place
>> one of the pulses in the on-screen area is all  that is required.
>> DaveB, NZ
>
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
> -- 
> No virus found in this incoming message.
> Checked by AVG Free Edition.
> Version: 7.5.503 / Virus Database: 269.15.6/1086 - Release Date: 
> 22/10/2007 19:57
>
> 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-24 Thread Brooke Clarke
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi Dave:

The problem is I don't have any documentation on the O-1814.  The label does 
not say Frequency Standard or anything else about what it does.  There is a 
very nice Rb source (M-100) inside and front panel connectors labeled "5 MHz 
In", "5 Mhz Out" and "10 MHz Out" so this may be a frequency standard.  But 
it's not clear how any of the radios in the GRC-106 Pacer Speak system can take 
advantage of a reference frequency.  http://www.prc68.com/I/PRC104.shtml#206

I'm beginning to think this is a Time Of Day standard.  There are buttons for 
"Rvc TOD" and "Send TOD" as well as a lamp labeled "Set Clock".

I was looking for a serial clock stream that would be used to clock in the TOD 
serial data stream when I found the narrow pulse.

I tried the SR620.  It's  bouncing around a 32 Hz period.

My web host has changed the server that's hosting PRC68.com and I've sort of 
been distracted, but will try the diode pulse stretcher.  If you look at the 
link above just below the GRC-206 system photo let me know if you see
(new server)
  Starting with the top row:
First VRC-83 Aircraft Radio <- bold title
(or the old server)
GRC-206(V)1 can be fitted with

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com
http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam


Dave Brown wrote:
> ); SAEximRunCond expanded to false
> Errors-To: [EMAIL PROTECTED] RETRY
> 
> But he knows the period already to parts in 10e-9 or better.  His 
> email sez it's a PPS signal ex an Rb source.
> Not much point in trying to measure that with a scope timebase!
> So I still think he wanted to measure the  ~10uS pulse width..
> 
> DaveB, NZ
> 
> - Original Message - 
> From: "Didier Juges" <[EMAIL PROTECTED]>
> To: "'Discussion of precise time and frequency measurement'" 
> 
> Sent: Wednesday, October 24, 2007 5:27 PM
> Subject: Re: [time-nuts] Scopes and 1 PPS signals?
> 
> 
> 
>>); SAEximRunCond expanded to false
>>Errors-To: [EMAIL PROTECTED] RETRY
>>
>>I read this:
>>
>>"But using an HP 54501A DSO it's almost impossible to determine the 
>>period.
>>http://www.prc68.com/I/O1814.shtml";
>>
>>It seems he wanted to measure the period. For once, the thread 
>>actually was
>>about the original question :-)
>>
>>Didier
>>
>>
>>>-Original Message-
>>>From: [EMAIL PROTECTED]
>>>[mailto:[EMAIL PROTECTED] On Behalf Of Dave Brown
>>>Sent: Tuesday, October 23, 2007 11:21 PM
>>>To: Discussion of precise time and frequency measurement
>>>Subject: Re: [time-nuts] Scopes and 1 PPS signals?
>>>
>>>I took Brookes original inquiry to be about measuring the
>>>width of the
>>>nominally 10 uS  wide PPS pulse-in which case anything that
>>>alters the
>>>pulse width is a no-no.  Precise and stable triggering delay to 
>>>place
>>>one of the pulses in the on-screen area is all  that is required.
>>>DaveB, NZ
>>
>>
>>___
>>time-nuts mailing list -- time-nuts@febo.com
>>To unsubscribe, go to 
>>https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>and follow the instructions there.
>>
>>
>>-- 
>>No virus found in this incoming message.
>>Checked by AVG Free Edition.
>>Version: 7.5.503 / Virus Database: 269.15.6/1086 - Release Date: 
>>22/10/2007 19:57
>>
>>
> 
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-24 Thread Tom Van Baak
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

> I tried the SR620.  It's  bouncing around a 32 Hz period.

Brooke,

You have a house standard, right? (GPSDO or cesium). Use
your SR620 to take a set of N (10 or 100) TI measurements
between your local reference and the O-1814 *rising edge*.
Then again using the O-1814 *falling edge*. The difference
between these two measurements is the average width of the
O-1814 pulse. Easy to do & very high accuracy.

Check the sdev of each measurement to make sure nothing
funny is going on; a couple of ns is OK. Hint: you get better
data if you 1) sync the O-1814 1PPS to less than, say, 10 uS
of UTC, and 2) if the O-1814 is tuned to within, say, 1e-11, of
your house reference.

/tvb


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-24 Thread Brooke Clarke
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi Tom:

It's the period or rep rate I'm trying to determine.

The SR620 has the PRS10 Rb as it's external time base.  It's the period of the 
pulse I'm trying to determine.  i.e. I'm not at all sure it's a 1 PPS and 
expect it may be a serial data clock.

I've added a photo showing the front of the M100 Rb std.
http://www.prc68.com/I/O1814.shtml#M100

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com
http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam


Tom Van Baak wrote:
> ); SAEximRunCond expanded to false
> Errors-To: [EMAIL PROTECTED] RETRY
> 
> 
>>I tried the SR620.  It's  bouncing around a 32 Hz period.
> 
> 
> Brooke,
> 
> You have a house standard, right? (GPSDO or cesium). Use
> your SR620 to take a set of N (10 or 100) TI measurements
> between your local reference and the O-1814 *rising edge*.
> Then again using the O-1814 *falling edge*. The difference
> between these two measurements is the average width of the
> O-1814 pulse. Easy to do & very high accuracy.
> 
> Check the sdev of each measurement to make sure nothing
> funny is going on; a couple of ns is OK. Hint: you get better
> data if you 1) sync the O-1814 1PPS to less than, say, 10 uS
> of UTC, and 2) if the O-1814 is tuned to within, say, 1e-11, of
> your house reference.
> 
> /tvb
> 
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-25 Thread Rob Kimberley
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

>From the info given, the TOD will be a HaveQuick serial time string. Need to
know what version of HaveQuick.

I have some info on an Excel spreadsheet which may help. Will dig it out.

Rob  

-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
Behalf Of Brooke Clarke
Sent: 25 October 2007 02:54
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Scopes and 1 PPS signals?

); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi Dave:

The problem is I don't have any documentation on the O-1814.  The label does

not say Frequency Standard or anything else about what it does.  There is a 
very nice Rb source (M-100) inside and front panel connectors labeled "5 MHz

In", "5 Mhz Out" and "10 MHz Out" so this may be a frequency standard.  But 
it's not clear how any of the radios in the GRC-106 Pacer Speak system can
take 
advantage of a reference frequency.  http://www.prc68.com/I/PRC104.shtml#206

I'm beginning to think this is a Time Of Day standard.  There are buttons
for 
"Rvc TOD" and "Send TOD" as well as a lamp labeled "Set Clock".

I was looking for a serial clock stream that would be used to clock in the
TOD 
serial data stream when I found the narrow pulse.

I tried the SR620.  It's  bouncing around a 32 Hz period.

My web host has changed the server that's hosting PRC68.com and I've sort of

been distracted, but will try the diode pulse stretcher.  If you look at the

link above just below the GRC-206 system photo let me know if you see
(new server)
  Starting with the top row:
First VRC-83 Aircraft Radio <- bold title
(or the old server)
GRC-206(V)1 can be fitted with

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com
http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam


Dave Brown wrote:
> ); SAEximRunCond expanded to false
> Errors-To: [EMAIL PROTECTED] RETRY
> 
> But he knows the period already to parts in 10e-9 or better.  His 
> email sez it's a PPS signal ex an Rb source.
> Not much point in trying to measure that with a scope timebase!
> So I still think he wanted to measure the  ~10uS pulse width..
> 
> DaveB, NZ
> 
> - Original Message - 
> From: "Didier Juges" <[EMAIL PROTECTED]>
> To: "'Discussion of precise time and frequency measurement'" 
> 
> Sent: Wednesday, October 24, 2007 5:27 PM
> Subject: Re: [time-nuts] Scopes and 1 PPS signals?
> 
> 
> 
>>); SAEximRunCond expanded to false
>>Errors-To: [EMAIL PROTECTED] RETRY
>>
>>I read this:
>>
>>"But using an HP 54501A DSO it's almost impossible to determine the 
>>period.
>>http://www.prc68.com/I/O1814.shtml";
>>
>>It seems he wanted to measure the period. For once, the thread 
>>actually was
>>about the original question :-)
>>
>>Didier
>>
>>
>>>-Original Message-
>>>From: [EMAIL PROTECTED]
>>>[mailto:[EMAIL PROTECTED] On Behalf Of Dave Brown
>>>Sent: Tuesday, October 23, 2007 11:21 PM
>>>To: Discussion of precise time and frequency measurement
>>>Subject: Re: [time-nuts] Scopes and 1 PPS signals?
>>>
>>>I took Brookes original inquiry to be about measuring the
>>>width of the
>>>nominally 10 uS  wide PPS pulse-in which case anything that
>>>alters the
>>>pulse width is a no-no.  Precise and stable triggering delay to 
>>>place
>>>one of the pulses in the on-screen area is all  that is required.
>>>DaveB, NZ
>>
>>
>>___
>>time-nuts mailing list -- time-nuts@febo.com
>>To unsubscribe, go to 
>>https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>and follow the instructions there.
>>
>>
>>-- 
>>No virus found in this incoming message.
>>Checked by AVG Free Edition.
>>Version: 7.5.503 / Virus Database: 269.15.6/1086 - Release Date: 
>>22/10/2007 19:57
>>
>>
> 
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-25 Thread Brooke Clarke
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Hi Didier:

When just a diode is used the stretching is only at the bottom of the trailing 
edge and not enough to help.  But when a little capacitance is added it works.

The period is the same as what the SR620 reported, 1/ 32 Hz.  Seems a strange 
number.

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.precisionclock.com
http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam


Didier Juges wrote:
> Some DSOs (my lowly Tek 210 does) have a peak detect acquisition mode where
> it will analyze the signal at the highest speed of the A/D converter
> regardless of sweep speed and report the min and max for each displayed
> pixel line. Main advantages are 1) to make aliasing obvious and 2) to be
> able to see very narrow pulses at low sweep speed. I can see 20nS wide
> pulses at 1 PPS easily. Of course, the precision of the measurement (if you
> use the period function) is limited to the display resolution, but at least
> you can get in the ballpark. My scope is in that mode all the time, little
> reason not to.
> 
> If you do not have such mode, put a diode (i.e. 1N4148) in series with the
> probe. The diode will charge the input capacitance of the probe, which will
> discharge exponentially in the 10 Mohm resistance and give you something you
> can see. Can't beat that for cost. Add some capacitance in parallel with the
> probe if needed to widen the pulse.
> 
> Didier KO4BB
> 
> 
>>-Original Message-
>>From: [EMAIL PROTECTED] 
>>[mailto:[EMAIL PROTECTED] On Behalf Of Brooke Clarke
>>Sent: Tuesday, October 23, 2007 6:24 PM
>>To: Discussion of precise time and frequency measurement
>>Subject: Re: [time-nuts] Scopes and 1 PPS signals?
>>
>>Hi Bill:
>>
>>I'm sure that would work, but it's something else that needs 
>>to be done sometime in the future.
>>
>>I checked the pulse coming from the M12+T and it's more like 
>>100 ms wide, probably for this reason.
>>
>>Have Fun,
>>
>>Brooke Clarke
>>http://www.PRC68.com
>>http://www.precisionclock.com
>>http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
>>
>>
>>WB6BNQ wrote:
>>
>>>Hi Brook,
>>>
>>>You could try using a 74123 (or newer variants) 
>>
>>retriggerable one shot 
>>
>>>to extend the pulse width to a level that would allow you 
>>
>>to see it on the scope.
>>
>>>BillWB6BNQ
>>>
>>>Brooke Clarke wrote:
>>>
>>>
>>>
>>>>Hi:
>>>>
>>>>While trying to figure out the O-1814 Rb source I found a 1 
>>
>>PPS signal 
>>
>>>>that's less than 10 us wide.  But using an HP 54501A DSO 
>>
>>it's almost 
>>
>>>>impossible to determine the period.  
>>>>http://www.prc68.com/I/O1814.shtml
>>>>
>>>>If the display is 1,000 dots wide:
>>>>at 2 us/sq the trace takes up about 4.5 squares or 450 dots, at 200 
>>>>us/sq the trace takes up about 0.045 squares or 4 dots but 
>>
>>the second 
>>
>>>>pulse is way off screen.  At slower sweep speeds you can't 
>>
>>see the pulse.
>>
>>>>Are there scopes that allow the period of a narrow pulse to be 
>>>>measured that don't cost an arm and leg?
>>>>
>>>>--
>>>>Have Fun,
>>>>
>>>>Brooke Clarke
> 
> 
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Scopes and 1 PPS signals?

2007-10-25 Thread Didier Juges
); SAEximRunCond expanded to false
Errors-To: [EMAIL PROTECTED] RETRY

Glad it works. I found the same here and that's why the better pictures were
made with a 5600pF cap that just happened to lay on the bench...

Didier 

> -Original Message-
> From: [EMAIL PROTECTED] 
> [mailto:[EMAIL PROTECTED] On Behalf Of Brooke Clarke
> Sent: Thursday, October 25, 2007 1:25 PM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] Scopes and 1 PPS signals?
> 
> Hi Didier:
> 
> When just a diode is used the stretching is only at the 
> bottom of the trailing edge and not enough to help.  But when 
> a little capacitance is added it works.
> 
> The period is the same as what the SR620 reported, 1/ 32 Hz.  
> Seems a strange number.
> 
> Have Fun,
> 
> Brooke Clarke
> http://www.PRC68.com
> http://www.precisionclock.com
> http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
> 
> 
> Didier Juges wrote:
> > Some DSOs (my lowly Tek 210 does) have a peak detect 
> acquisition mode 
> > where it will analyze the signal at the highest speed of the A/D 
> > converter regardless of sweep speed and report the min and max for 
> > each displayed pixel line. Main advantages are 1) to make aliasing 
> > obvious and 2) to be able to see very narrow pulses at low sweep 
> > speed. I can see 20nS wide pulses at 1 PPS easily. Of course, the 
> > precision of the measurement (if you use the period function) is 
> > limited to the display resolution, but at least you can get in the 
> > ballpark. My scope is in that mode all the time, little 
> reason not to.
> > 
> > If you do not have such mode, put a diode (i.e. 1N4148) in 
> series with 
> > the probe. The diode will charge the input capacitance of 
> the probe, 
> > which will discharge exponentially in the 10 Mohm 
> resistance and give 
> > you something you can see. Can't beat that for cost. Add some 
> > capacitance in parallel with the probe if needed to widen the pulse.
> > 
> > Didier KO4BB
> > 
> > 
> >>-Original Message-----
> >>From: [EMAIL PROTECTED]
> >>[mailto:[EMAIL PROTECTED] On Behalf Of Brooke Clarke
> >>Sent: Tuesday, October 23, 2007 6:24 PM
> >>To: Discussion of precise time and frequency measurement
> >>Subject: Re: [time-nuts] Scopes and 1 PPS signals?
> >>
> >>Hi Bill:
> >>
> >>I'm sure that would work, but it's something else that needs to be 
> >>done sometime in the future.
> >>
> >>I checked the pulse coming from the M12+T and it's more like 100 ms 
> >>wide, probably for this reason.
> >>
> >>Have Fun,
> >>
> >>Brooke Clarke
> >>http://www.PRC68.com
> >>http://www.precisionclock.com
> >>http://www.prc68.com/I/WebCam2.shtml 24/7 Sky-Weather-Astronomy Cam
> >>
> >>
> >>WB6BNQ wrote:
> >>
> >>>Hi Brook,
> >>>
> >>>You could try using a 74123 (or newer variants)
> >>
> >>retriggerable one shot
> >>
> >>>to extend the pulse width to a level that would allow you
> >>
> >>to see it on the scope.
> >>
> >>>BillWB6BNQ
> >>>
> >>>Brooke Clarke wrote:
> >>>
> >>>
> >>>
> >>>>Hi:
> >>>>
> >>>>While trying to figure out the O-1814 Rb source I found a 1
> >>
> >>PPS signal
> >>
> >>>>that's less than 10 us wide.  But using an HP 54501A DSO
> >>
> >>it's almost
> >>
> >>>>impossible to determine the period.  
> >>>>http://www.prc68.com/I/O1814.shtml
> >>>>
> >>>>If the display is 1,000 dots wide:
> >>>>at 2 us/sq the trace takes up about 4.5 squares or 450 
> dots, at 200 
> >>>>us/sq the trace takes up about 0.045 squares or 4 dots but
> >>
> >>the second
> >>
> >>>>pulse is way off screen.  At slower sweep speeds you can't
> >>
> >>see the pulse.
> >>
> >>>>Are there scopes that allow the period of a narrow pulse to be 
> >>>>measured that don't cost an arm and leg?
> >>>>
> >>>>--
> >>>>Have Fun,
> >>>>
> >>>>Brooke Clarke
> > 
> > 
> > 
> > ___
> > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
> > 
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, 
> go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.